Re: Tek 7904 random redout and noisy pot
Hakan, After knowing about the socket problem, is it best as PM to do a sweep of the sockets in the scopes and plugins? Is there a point where the sockets were better so replacing would not be needed?
By
Mark Vincent
·
#209638
·
|
Re: Tek 7904 random redout and noisy pot
While you are at it check if they are the bad TI C95 sockets. They are trouble waiting to happen. Some spray might help but probably only for a short time.
By
zenith5106
·
#209637
·
|
Re: Tek 2432 & 2430A Power up fails
That worked, thx. After cold start (with J156 removed), self cal and ext cal the scope returned with 0 Fail messages.
By
Gerdy
·
#209636
·
|
Re: Tek 2432 & 2430A Power up fails
Finally solved! I replaced the battery, that did not clear all the Fail messages but then I forced a cold start (after J156 removed). Then a 'self cal' and an 'ext cal' (the latter needed to apply
By
Gerdy
·
#209635
·
|
Re: Tek 7904 random redout and noisy pot
I'd try taking the knobs off and spraying Faderlube down the shaft before disassembling it, and I'd also try that from the exterior and see if it'll wick through any gaps. For garbage and missing
By
Adam R. Maxwell
·
#209634
·
|
Re: Tek 7904 random redout and noisy pot
Roger, You may have to, unsolder the leads from the diodes from the rear wafer, loosen the nuts on the rear of the switch enough to keep them on, loosen the nut for the pot, move the pot back enough
By
Mark Vincent
·
#209633
·
|
Re: 067-0587-01 amplitude pot wire connections
Thank you very much for having time to check and post this reply, Barry Roger
By
Rogerio O
·
#209632
·
|
Tektronix 2236 Seven-Segment Display 670-7423-00
Hello and need help for this part. I¡¯m looking for the vfd display used in the counter timer of the 2236. Part number is 670-7423-01 for the small board and 150-1110-00 for just the display tube.
By
lbrown8008
·
#209631
·
|
Tek 7904 random redout and noisy pot
dear all, I dropped my 7904 from its cart last week. Up to the moment the identified losses are: 1- broken 7B85 time/div pot and cal/uncal knob and rod 2- Very noisy redout pot 3- Garbage number
By
Rogerio O
·
#209630
·
|
Re: 067-0587-01 amplitude pot wire connections
Looking at that pot from the back, on mine the center pot has the left tab grounded, the center tab has a wire with white insulation and a violet stripe, and the right tab has a wire with red
By
n4buq
·
#209629
·
|
067-0587-01 amplitude pot wire connections
Hi to everybody. I bought a 067-0587-01 on Epay and, for the moment, I have two issues: 1- One red wire going to the amplitude Pot is "floating" as shown in photo
By
Rogerio O
·
#209628
·
|
Added photo album 067-0587-01
#photo-notice
Rogerio O <rodd414@...> added the photo album 067-0587-01 ( /g/TekScopes/album?id=302607 ) : one wire connection is broken
By
Group Notification <noreply@...>
·
#209627
·
|
Re: 7A13 overshoot
Looking again: the offending relay was K490, which is the bandpass limiting relay; I misread the numbers on the board since I wasn't using a flashlight. From the description and schematic, K480
By
Adam R. Maxwell
·
#209626
·
|
Re: 7A13 overshoot
I had opened an older version of the schematic which doesn't have that relay. I see it now on the later versions. Apologies for the interruption. Barry - N4BUQ
By
n4buq
·
#209625
·
|
Re: 7A13 overshoot
I'm looking at the manual on Tekwiki for B200000 up, and K480 is at the very back center of the output amplifier board on the labeled component photo. I need to find it on the schematic and see if
By
Adam R. Maxwell
·
#209624
·
|
2784 Spectrum Analyzer
Hi, my 2784 Spec lost all the data, because Li- Battery has not been changed last fifteen years. Now, when I power the Spectrum Analyzer the display shows me a 2782 . How can I fix this problem? I can
By
J¨¹rgen W
·
#209623
·
|
Re: 7A13 overshoot
Hi Adam, Was that, perhaps, K48 or K490? I don't see a K480 but I might be looking at a different version of the manual. Glad you found/fixed it. Those relays are definitely a pain. Barry - N4BUQ
By
n4buq
·
#209622
·
|
Re: 7A13 overshoot
Well, after replacing C187 and C163 with no improvement, it turns out that I overlooked relay K480. Swapping with my known good relay immediately got rid of the giant hump at the front of the pulse,
By
Adam R. Maxwell
·
#209621
·
|
Re: Tektronix 494P 1st LO and Frequency failure(s)
P.s. test tne above with rbw well over 200kHz. That cuts out the inner yig loop which is 2/3 of the first if hw, making debugging easier.
By
Edward Prest
·
#209620
·
|
Re: Tektronix 494P 1st LO and Frequency failure(s)
no if 1st if osc...I would do in this order the following..Check... Power supplies, good, not close but right on the money. See john miles notes. Check there is any yig output , i.e.1st IF is on front
By
Edward Prest
·
#209619
·
|