Is this normal?
thanks, -Kurt
|
Re: TDS744A adjustment software needed
--- In TekScopes@..., "baltimora86" <acuffe@...> wrote: Does anyone have a copy of the adjustment software for the TDS744A that they could share? I have one that had its input attenuators used to repair another scope. Now I've managed find a set of attenuators in unknown condition, but it's having problems that I think might be related to calibration.
The symptoms are that on three channels the trace goes off the screen on all ranges in the most sensitive decade. Also, they will only trigger if the trigger mode is set to AC. The last channel has a small DC offset on the most sensitive ranges, triggers fine, but reads about 20% low. It also fails SPC and the acquisition self test. I know the scope worked fine with its original attenuators, so I would like to at least try calibration before I part it out.
--------------------
Seems like you have a bunch of bad attenuator hybrids. I doubt that you would be able to run adjustment software at all due to the errors. SPC is part of the sequence and must pass to continue and if I recall correctly the program will not start unless the error log is empty. I think maybe the program was able to clear the error log, but the actual faults will still there. If you still want the software contact me off list for a copy. /H?kan
|
EIA Power Connector; Delete Chassis Tracks; Rack Mount Ears Punched for EIA Standard; +$215.00.
toggle quoted message
Show quoted text
--- In TekScopes@..., DON CRAMER <donlcramer@...> wrote:
Would anyone with the secret code book know what comprises MOD WT on the 5440 mainframe? Would anyone with access to the 'secret code book' know what comprises "MOD WT" on the 5440 mainframe (circa 1986)? I'm guessing one item might be the convenient IEC power entry. Thanks in advance, Don
|
That sounds like caps which might be part of Horizontal Display Logic circuit. If remember correctly, that circuit is in that area of the board.
toggle quoted message
Show quoted text
--- In TekScopes@..., "pdxareaid" <public_email@...> wrote:
trigger view did nothing. but i had it on long enough that i noticed a smoking tantalum. and it was even moaning! ikyn it was making noise when it started to smoke. it was in a cluster of 3 tants on the large rightmost board (viewed from the front) towards the rear...i don't even know what board that is yet but i will as soon as i can get to look at the service manual. phil
--- In TekScopes@..., "pdxareaid" <public_email@> wrote:
arggg...i fogot to poke the trigger view. one last trip out there right now.
--- In TekScopes@..., "Brad King" <cpgixxer@> wrote:
Trigger view as far as I know detaches the vertical pre from the circuit. The service manuals have block diagrams to help segregate the circuit to narrow it down. Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:48:37 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
lol..i have not had time to do much of anything with it yet (only had it a couple of hours) but i can't resist dropping everything and running out to poke it with these tips rolling in...
when the screen is glowing, none of the controls seems to affect it and it generally lights up the whole screen.
using the beamfinder, i get a bright dot that jumps around and in and out of what looks like it might be trying to show me a reasonable beamfinder display, but severely shift off screen to the right. the position controls seem to affect it but not in a consistent way...but there is some effect. all of the beamfinder action appears to be on the right half of the screen. phil
ps. i'm going to be in trouble if i don't finish some other stuff today...i'll leave poking it again til tomorrow but will check in to soak up the tips for when i do get out there. thanks for all the help. keep em coming :-)
--- In TekScopes@..., "Brad King" <cpgixxer@> wrote:
The glow is because the trace isn't centered, could be a bad resistor or transistor in the vertical. Does it move h or v with the adjustment? Did you try pressing the trigger view button? Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:25:55 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
ah, the beamfinder...never was very fond of it. now maybe i see why it is there. thanks for the tip.
sweep rate does not seem to affect anything.
beamfinder: I got the usual bright connected 2 (i assume) point display but severely shifted to the right such that only 1/2 of the expected display is on the screen. curiously, when i tried that, and released the beamfinder, i got a bright point on the display rather than a full screen glow. then it went back to the glow. i tried the beamfinder a few more times and it seemed to start bringing something back from the dead. i even saw it flash what looked like it might have been a reasonable trace...but it disappeared very quickly. the burning smell was making me nervous so i shut it down.
are these useful clues?
--- In TekScopes@..., larrys@ wrote:
"pdxareaid" <public_email@> wrote:
Whole screen glows off and on, no trace to speak of. Does the rate of the on / off vary with the sweep speed setting? Sounds like you may have a trace, just offscreen. See if the beamfinder shows it too you.
-ls-
|
TDS744A adjustment software needed
Does anyone have a copy of the adjustment software for the TDS744A that they could share? I have one that had its input attenuators used to repair another scope. Now I've managed find a set of attenuators in unknown condition, but it's having problems that I think might be related to calibration.
The symptoms are that on three channels the trace goes off the screen on all ranges in the most sensitive decade. Also, they will only trigger if the trigger mode is set to AC. The last channel has a small DC offset on the most sensitive ranges, triggers fine, but reads about 20% low. It also fails SPC and the acquisition self test. I know the scope worked fine with its original attenuators, so I would like to at least try calibration before I part it out.
|
Re: Need specs for 015-0221-00 DC block
That is the value of coupling capacitance I measured on mine as well and the low frequency cutoff was indeed pretty close to 70kHz. On Mon, 24 Sep 2012 14:33:22 +0100, "Craig Sawyers" <c.sawyers@...> wrote: I measure a series capacitance between input and output of 47nF. So into 50 ohms that implies that its 3dB low frequency is about 70kHz.
And since it is specced for use with the 7L12/13 it must go to around 2GHz with a reasonable VSWR.
But beyond that, nothing firm.
-----Original Message----- From: TekScopes@... [mailto:TekScopes@...] On Behalf Of Mike Czuhajewski Sent: 24 September 2012 13:32 To: TekScopes@... Subject: [TekScopes] Need specs for 015-0221-00 DC block
This is their BNC DC block. Finding any data on it is pretty much impossible;
tried lots of Google searches, old Tek catalogs, etc. (It's listed under spectrum analyzer accessories.) About the only thing I can
find is the voltage limit (50) and nothing more. They do give detailed info on
the type N block, though.
Does anyone have any actual data on this? I'm looking for the frequency range, especially the low end. I've seen it mentioned a few times but with different sets of frequencies; nothing agrees. Does anyone know the actual Tek spec on these?
Thanks.
------------------------------------
Yahoo! Groups Links
|
On Mon, 24 Sep 2012 10:43:12 -0000, "tjeenz" <tjeenz@...> wrote: I removed u460. (ca3102 from rca). it is broken (20ohm between 1 and 14 in both directions) also i found a diode connected on the top to pin 12. i couldn't detect pin 12 in the schematics... now start looking for a ca3102 On Mon, 24 Sep 2012 08:50:17 -0000, "tjeenz" <tjeenz@...> wrote: Voltage U460 pin3 and 9 is -8.6V, R456 is ok. pin 1 and 8 are 3.5V not 2.2
That fits with pin 1 being 3.5 volts instead of 2.2 volts. The low resistance between the collector and base is pulling that node up via the high output of U480C. Pin 2 and pin 10 at about 0.0 volts while pin 3 is -8.6 volts suggests that the emitters of U460E and U460B is open as well. Pin 12 and pin 5 are connected internally to the substrate. The CA3102 is built on a junction isolated process so the substrate connects to each transistor via a diode which is reverse biased to provide isolation. u460 is unfortunately not on a socket.
When replacing U460, I would cut the collet pins out of an IC socket and individually install them into the PC board to make a low profile socket for the new IC. To access u460 and area, i see i have to remove some components. I will start on that a.s.a.p. but a question: some loose contact in the timebase switch. sometimes it generates only half screen image when i move with that knob. is that the pot-meter in the middle?
From what I remember of the 2213, the timebase switch uses a different and unreliable design from the other 22xx series oscilloscopes. Someone else may have suggestions for repairing it.
|
Re: Can you display the left plugin using a 5D10?
I wasn't aware of the 7k readout behavior so thanks for the tip. I'll have to try it and get back to you.
I wonder if this solution would apply to the 5440's with the 5a48 plugin?
jamie
toggle quoted message
Show quoted text
--- In TekScopes@..., "Rob" <rgwood@...> wrote: "Now if only I could get the readout to reflect a 100x probe when doing tube work!"
Is the 5000 series readout system different than the 7000?
On the 7000 series all you need to do is have 6.2K +- 5% of resistance between the readout code ring around the BNC input and signal ground. I have successfully made some BNC spring pin coded rings for my probes by obtaining the outer ring part from a forum member here and then placing a 6.2K surface mount resistor on top of the spring for the pin. It works great.
My struggle however has been to get a 1000x resistor value. Actually I can get a 1000X with a 4.3K, 4.7K or so. However it will not add zeros so only forces the readout to correct values for the mv range. So far I have not figured out if either a "proper" resistance value is my problem or something different s done for 1000x probes for the readout.
Anyway, try putting a 6.2K resistor from your scopes readout code ring to ground and see if you get the 100X values showing. If you do than that means the 5000 series readout is similar enough to the 7000 series that you could build your own code collar like I did. Assuming that the 6.2K test will be successful; a search in the old messages for "BNC collar" will lead you to a discussion back around June 7th of this year that will help with the rest.
If the 6.2K test fails..... well at least I tried to help * smile *. Good luck Rob
|
Re: Need specs for 015-0221-00 DC block
I measure a series capacitance between input and output of 47nF. So into 50 ohms that implies that its 3dB low frequency is about 70kHz.
And since it is specced for use with the 7L12/13 it must go to around 2GHz with a reasonable VSWR.
But beyond that, nothing firm.
toggle quoted message
Show quoted text
-----Original Message----- From: TekScopes@... [mailto:TekScopes@...] On Behalf Of Mike Czuhajewski Sent: 24 September 2012 13:32 To: TekScopes@... Subject: [TekScopes] Need specs for 015-0221-00 DC block
This is their BNC DC block. Finding any data on it is pretty much impossible; tried lots of Google searches, old Tek catalogs, etc. (It's listed under spectrum analyzer accessories.) About the only thing I can find is the voltage limit (50) and nothing more. They do give detailed info on the type N block, though.
Does anyone have any actual data on this? I'm looking for the frequency range, especially the low end. I've seen it mentioned a few times but with different sets of frequencies; nothing agrees. Does anyone know the actual Tek spec on these?
Thanks.
------------------------------------
Yahoo! Groups Links
|
Need specs for 015-0221-00 DC block
This is their BNC DC block. Finding any data on it is pretty much impossible; tried lots of Google searches, old Tek catalogs, etc. (It's listed under spectrum analyzer accessories.) About the only thing I can find is the voltage limit (50) and nothing more. They do give detailed info on the type N block, though.
Does anyone have any actual data on this? I'm looking for the frequency range, especially the low end. I've seen it mentioned a few times but with different sets of frequencies; nothing agrees. Does anyone know the actual Tek spec on these?
Thanks.
|
Re: Tek Curve tracer model 477 with 178 test fixture
Louis
...... and I think you mean 577 ( not 477) curve tracer
Dave
On 9/24/2012 5:43 AM, Artekmedia wrote:
?
Louis
If you don find a free copy contact me off list at manuals@...
we have a very nice professional quality scan of the 178
manual at a very affordable price
Cheers
Dave
On 9/24/2012 4:23 AM, vk4bvy wrote:
?
Got one of these recently and need copies of the
connection cards to
test IC's....is there a member of this great group,
who has scanned
these and can make them available????
Thanks in advance
Louis VK4BVY
--
Dave Henderson
Manuals@...
PO Box 175
Welch,MN 55089
651-269-4265
--
Dave Henderson
Manuals@...
PO Box 175
Welch,MN 55089
651-269-4265
|
Re: Tek Curve tracer model 477 with 178 test fixture
Louis
If you don find a free copy contact me off list at
manuals@... we have a very nice professional quality
scan of the 178 manual at a very affordable price
Cheers
Dave
On 9/24/2012 4:23 AM, vk4bvy wrote:
?
Got one of these recently and need copies of the
connection cards to
test IC's....is there a member of this great group, who
has scanned
these and can make them available????
Thanks in advance
Louis VK4BVY
--
Dave Henderson
Manuals@...
PO Box 175
Welch,MN 55089
651-269-4265
|
I removed u460. (ca3102 from rca). it is broken (20ohm between 1 and 14 in both directions) also i found a diode connected on the top to pin 12. i couldn't detect pin 12 in the schematics... now start looking for a ca3102
T.
toggle quoted message
Show quoted text
--- In TekScopes@..., "tjeenz" <tjeenz@...> wrote: Hi, Voltage U460 pin3 and 9 is -8.6V, R456 is ok. pin 1 and 8 are 3.5V not 2.2 pin 4 and 11 check out ok, 2 and 2.4V dependent on the slope setting.
u460 is unfortunately not on a socket.
To access u460 and area, i see i have to remove some components. I will start on that a.s.a.p. but a question: some loose contact in the timebase switch. sometimes it generates only half screen image when i move with that knob. is that the pot-meter in the middle?
T.
--- In TekScopes@..., David <davidwhess@> wrote:
U460 is not necessarily bad but if it is, the CA3102 transistor array is not expensive or difficult to acquire. If U480 is bad, the 10102 ECL Quad 2-input NOR gate is readily available.
4 volts at both pin 10 of U480C and pin 7 of U480B is consistent with a logical high for a 10102 when operated at PECL levels but one pin of that pair should be low.
Check the voltage at pin 3 or 9 of U460. R456 might be open so check its resistance.
Check pin 8 or 1 of U460 which should be 2.2 volts.
Check pin 11 or 4 of U460. It should be either 2.4 volts or 2.0 volts depending on the trigger slope setting.
If U460 is socketted, then you could pull it and check the transistor junctions with the diode test function on your multimeter.
On Mon, 24 Sep 2012 02:58:59 -0000, "tjeenz" <tjeenz@> wrote:
Hi David,
no square wave on the pins 10 and 7, both slopes.... it just measures a +4V dc, at both pins. Does it mean my U460 has died?
T.
--- In TekScopes@..., David <davidwhess@> wrote:
That narrows the problem down to some point at or after transistor array U460. The good news is that the 2213 does not use a custom trigger IC so if the problem can be found, it should be relatively easy to fix.
With the trigger set to LINE and AUTO, adjust the LEVEL to midway and measure pin 10 of U480C and pin 7 of U480B. With both SLOPE settings, there should be at least a 300mV to 400mV square wave on each pin.
If there is, then check pin 2 of U480A. It should have a 300mV to 400mV square wave.
On Sun, 23 Sep 2012 10:51:11 -0000, "tjeenz" <tjeenz@> wrote:
Hi David, still out there?
back from trip... I checked tp444: with trig on "line" i have a sinus with about 0.5V tt
if i connect a 4V top-top sinus at ch1 with trig on "int" i have 150mV sinus on tp444 (very small)
U460B, pin 2: norm: there is indeed a +0.65 - -0.65V which can be controlled by the trig-lev. pot. auto: sinus 50 mV (again with 4V tt at ch1). which seems very small...
in all situations the display does not show a "still" image.
tjeenz
--- In TekScopes@..., "tjeenz" <tjeenz@> wrote:
hi David, thx for the tip! i will check this as soon as possible (couple days out of town). i check tp444 before with ext. trigger and that was very small compared to the input signal. don't know the exact amplitude but much lower. probably back on saturday. tjeenz.
--- In TekScopes@..., David <davidwhess@> wrote:
I would start by looking at TP444 (middle of schematic 4) which should be a duplicate of the trigger input signal as selected by the source and coupling controls. At the very least, there should be a 60Hz sine wave at that point when the trigger source is set to line.
Then I would check pin 2 of U460B which should be a duplicate of the tri.gger level voltage as set by the trigger level control. In Auto mode this voltage will be between the peak levels of the trigger signal but in normal mode, it should be between +0.65V and -0.65V.
It would be nice to have a nicer scan of the service manual. I extracted and cleaned up schematic 4 from the one available online:
On Tue, 18 Sep 2012 11:00:41 -0000, "tjeenz" <tjeenz@> wrote:
Hi, Have a trigger problem with a 2213 tektronix.
In auto and line trigger i have a baseline. In norm not. Also ext. trigger fails (signal from function generator 4v pp and ch1 on 1v/div)
cannot get still image to see frequency, amplitude is ok. with signal on one of the channels, trig in auto i can almost get still image if freq is low enough with var hold-off. trig-level does nothing. Also the trigger led never blinks.
Checked all p.s. voltages. Are ok (no ripples), but according to service manual there should be some specific resistance on the voltages. These are way too low in my scope: 8.6v 114 ohm at tp500= 58 ohm +8.6v 95 ohm at W975 = 16 ohm +5v 330 ohm at W985 = 16 ohm +30v 905 ohm at W965 = 16 ohm +100v 12.5 kohm at W966 = 16 ohm
so i guess something is wrong somewhere. checked the big capacitors in the p.s, for 30V and +8.6V, both were in spec.
Are these resistance values in the manual a good reference for finding the faulty area?
regards, tjeenz
|
Tek Curve tracer model 477 with 178 test fixture
Got one of these recently and need copies of the connection cards to test IC's....is there a member of this great group, who has scanned these and can make them available????
Thanks in advance
Louis VK4BVY
|
Hi, Voltage U460 pin3 and 9 is -8.6V, R456 is ok. pin 1 and 8 are 3.5V not 2.2 pin 4 and 11 check out ok, 2 and 2.4V dependent on the slope setting.
u460 is unfortunately not on a socket.
To access u460 and area, i see i have to remove some components. I will start on that a.s.a.p. but a question: some loose contact in the timebase switch. sometimes it generates only half screen image when i move with that knob. is that the pot-meter in the middle?
T.
toggle quoted message
Show quoted text
--- In TekScopes@..., David <davidwhess@...> wrote: U460 is not necessarily bad but if it is, the CA3102 transistor array is not expensive or difficult to acquire. If U480 is bad, the 10102 ECL Quad 2-input NOR gate is readily available.
4 volts at both pin 10 of U480C and pin 7 of U480B is consistent with a logical high for a 10102 when operated at PECL levels but one pin of that pair should be low.
Check the voltage at pin 3 or 9 of U460. R456 might be open so check its resistance.
Check pin 8 or 1 of U460 which should be 2.2 volts.
Check pin 11 or 4 of U460. It should be either 2.4 volts or 2.0 volts depending on the trigger slope setting.
If U460 is socketted, then you could pull it and check the transistor junctions with the diode test function on your multimeter.
On Mon, 24 Sep 2012 02:58:59 -0000, "tjeenz" <tjeenz@...> wrote:
Hi David,
no square wave on the pins 10 and 7, both slopes.... it just measures a +4V dc, at both pins. Does it mean my U460 has died?
T.
--- In TekScopes@..., David <davidwhess@> wrote:
That narrows the problem down to some point at or after transistor array U460. The good news is that the 2213 does not use a custom trigger IC so if the problem can be found, it should be relatively easy to fix.
With the trigger set to LINE and AUTO, adjust the LEVEL to midway and measure pin 10 of U480C and pin 7 of U480B. With both SLOPE settings, there should be at least a 300mV to 400mV square wave on each pin.
If there is, then check pin 2 of U480A. It should have a 300mV to 400mV square wave.
On Sun, 23 Sep 2012 10:51:11 -0000, "tjeenz" <tjeenz@> wrote:
Hi David, still out there?
back from trip... I checked tp444: with trig on "line" i have a sinus with about 0.5V tt
if i connect a 4V top-top sinus at ch1 with trig on "int" i have 150mV sinus on tp444 (very small)
U460B, pin 2: norm: there is indeed a +0.65 - -0.65V which can be controlled by the trig-lev. pot. auto: sinus 50 mV (again with 4V tt at ch1). which seems very small...
in all situations the display does not show a "still" image.
tjeenz
--- In TekScopes@..., "tjeenz" <tjeenz@> wrote:
hi David, thx for the tip! i will check this as soon as possible (couple days out of town). i check tp444 before with ext. trigger and that was very small compared to the input signal. don't know the exact amplitude but much lower. probably back on saturday. tjeenz.
--- In TekScopes@..., David <davidwhess@> wrote:
I would start by looking at TP444 (middle of schematic 4) which should be a duplicate of the trigger input signal as selected by the source and coupling controls. At the very least, there should be a 60Hz sine wave at that point when the trigger source is set to line.
Then I would check pin 2 of U460B which should be a duplicate of the tri.gger level voltage as set by the trigger level control. In Auto mode this voltage will be between the peak levels of the trigger signal but in normal mode, it should be between +0.65V and -0.65V.
It would be nice to have a nicer scan of the service manual. I extracted and cleaned up schematic 4 from the one available online:
On Tue, 18 Sep 2012 11:00:41 -0000, "tjeenz" <tjeenz@> wrote:
Hi, Have a trigger problem with a 2213 tektronix.
In auto and line trigger i have a baseline. In norm not. Also ext. trigger fails (signal from function generator 4v pp and ch1 on 1v/div)
cannot get still image to see frequency, amplitude is ok. with signal on one of the channels, trig in auto i can almost get still image if freq is low enough with var hold-off. trig-level does nothing. Also the trigger led never blinks.
Checked all p.s. voltages. Are ok (no ripples), but according to service manual there should be some specific resistance on the voltages. These are way too low in my scope: 8.6v 114 ohm at tp500= 58 ohm +8.6v 95 ohm at W975 = 16 ohm +5v 330 ohm at W985 = 16 ohm +30v 905 ohm at W965 = 16 ohm +100v 12.5 kohm at W966 = 16 ohm
so i guess something is wrong somewhere. checked the big capacitors in the p.s, for 30V and +8.6V, both were in spec.
Are these resistance values in the manual a good reference for finding the faulty area?
regards, tjeenz
|
Would anyone with the secret code book know what comprises MOD WT on the 5440 mainframe?
Would anyone with access to the 'secret code book' know what comprises "MOD WT" on the 5440 mainframe (circa 1986)?? I'm guessing one item might be the convenient IEC power entry. Thanks in advance, Don
|
Re: Help I got a Tracking Generator for my Tek 2710 but how do I install it?
Rolynn Thank you so very much for that :) I will be in touch soon.
toggle quoted message
Show quoted text
--- In TekScopes@..., "ROLYNN PRECHTL K7DFW" <k7dfw@...> wrote:
I think my biggest hurdle is going to be getting the proper Semi Rigid cables that are needed I have no clue where to buy these........
|
Re: Interesting account of HP's attempt to compete with the 7000 series
" review would be with two guys notoriously chary about poor
engineering detail execution ". That is probably why hp, unlike
Tektronix, did not have any overloaded/discolored resistors or
electrolytics running on only few volts of voltage margin.
Miroslav Pokorni
On 9/21/2012 5:38 PM, John Miles wrote:
toggle quoted message
Show quoted text
?
Chuck House talks about where the HP
logic-analyzer product line came from -- a classic
solution-in-search-of-a-problem scenario, but one where it turned out
that there actually was a problem to be solved.? (Warning: very long.)
?
"We'd already spent $2 million, with a burn rate that had to
escalate for our next phase. We were, at best, two years from delivery.
The Tek engineers told me that they had spent $34 million dollars and
seven years of effort, to get to this (IEEE) show with an incomplete
line. Whew!
Worse, my team was having trouble debugging our own prototype
scheduled for a division review in four weeks. That review would be
with two guys notoriously chary about poor engineering detail execution
- Packard and Hewlett, headed to Colorad! o Springs in late April. It
was not hard to figure out what to do. I had to cancel Next Gen.
The first problem was that I was in love with it..."
-- john, KE5FX
?
|
Re: Help I got a Tracking Generator for my Tek 2710 but how do I install it?
Hi Herbert
Thanks, I had seen those numbers but I thought they would have a corresponding long part number like the one I see on Ebay and in the SV manual for other parts. I was looking through the PDF for it but never found it, now I know why LOL.
Thanks for the offer Herb I may take you up on it pretty soon, maybe ask you instead for some photos if possible of how yours are routed so I can see where to make the bends. For now I think I can just estimate the amount I will need and buy 25% extra just in case.
BTW any suggestions on a good online place to order the parts? Digikey or is there some place better?
Thank you Robert
toggle quoted message
Show quoted text
--- In TekScopes@..., Measurement <measurement@...> wrote: Hi,
the parts are named W52 and W53 and W55.
- W55 is solder end to SMB - W52 SMA-male to N-female - W53 SMA-male to SMA-male
Would it help, to know the length ??? I can measure in mz 2712 !
Herbert
Am 23.09.2012 21:48, schrieb dreadlk:
For starters I found this on ebay but it looks too short.
<>
I am scouring the 2712 SVC Manual looking for the correct part number but so far no luck, it is diagramed but the part numbers are not listed for any of the cables.
--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com>, "dreadlk" <dreadlk@> wrote:
Thanks for the information, I agree with both of your assessments. It was easy to find the Cables in a flex format so I went out and bought some on day one. Now that I read your posts I know it won't work and my dilemma is that I have no idea where to find the Semi Rigid versions. I am going to do some searching but I have a feeling I am going to need a lot of luck on my side to make this happen. If the unit will error out it makes little sense for me to proceed further until I locate these cables so at this point I am going to mark out the plate for cutting and drilling based on Jims info and then when I know where to get the needed cables I will start the cutting and drilling.
If anybody has any of these cables and is willing to sell them please let me know or if you have any ideas on a source that might sell them the info would be greatly appreciated.
Robert KD8MJR
--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com>, "Rich Miller" <av8torrich@> wrote:
Hello Robert:
With regard to the rigid coax cable used to connect the TG to the
front
panel, I think you are going to want to use this. While at low frequency
other types of cable may work fine, when you get into microwave you may run
into a problem with the TG Gain Flatness Normalization. If the output of the
TG is off by more than +/-3dBm (This is from memory so double check), it
will show up as a fault when you first attempt to normalize.
Rich, AJ3G
From: TekScopes@... <mailto:TekScopes%40yahoogroups.com> [mailto:TekScopes@... <mailto:TekScopes%40yahoogroups.com>] On Behalf
Of dreadlk Sent: Friday, September 21, 2012 2:33 PM To: TekScopes@... <mailto:TekScopes%40yahoogroups.com> Subject: [TekScopes] Re: Help I got a Tracking Generator for my Tek 2710 but
how do I install it?
Hi Jim
Thank you so much for the reply. I do have a VR BW board in that cavity and
I received the TG with no cables at all, I will have to figure out how to
obtain the needed parts and where to source them. I was wondering if flexible cables with SMA and Type N will work or does it
have to be the rigid stuff? I obtained a copy of the 2712 SV Manual and am
going to look through it tonight to try and get a better idea of what I am
dealing with, I want to thank you for your post, it certainly clarified
things for me, I hope I am up to the task of doing this Job as I must admit
I am a bit worried about messing up my 2710 it has been absolutely a pleasure to use over the years and is in fantastic condition.
Let me ask you Jim is there any chance of having my proverbial cake and
eating it too? I mean could I leave in the VR BW board and run cables to the Back of the
unit and put the TG in a separate enclosure, kind of like a DIY 2707? I am
just wondering if anybody has ever done that, I would imagine that it might
not be practical but I just wanted to get your opinion.
Anyway Jim I am going to take out the VR BW board and see what I am dealing
with. If you have any suggestions please post them, and thanks so much for
your help.
Robert KD8MJR
--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com> <mailto:TekScopes%40yahoogroups.com> ,
"Jim" <jflan@> wrote:
Hi Robert - The TG mounts in the cavity on the left side of the analyzer.
This also is
the same cavity which houses the Variable Resolution BW board. You can only
use one or the other, not both. It wasn't clear whether you have the VR BW
board or not.
There are two semi-rigid cables that will need to be made. One
routes the
RF to the front panel and has an SMA on one end and a Type N on the other.
The other routes the 2nd LO (2 GHz) to the TG. Did you receive these cables?
The ribbon cable will plug into the main power supply board to provide the
interface. in fact, I just remembered that one end of the ribbon cable is
crimped to a IDC connector that is *soldered* to the power supply board. The
other end uses at typical IDC connector that will plug into the TG. The 2712
(not 2710) service manual has a block diagram that will aide in making the
necessary connections. This manual is down loadable from the Tek.com site
and will be a better reference for what you are wanting to do. You will have
some work to do to get your TG up and running. Take your time and it will go
smoothly.
Good luck. Jim WB5KYE
--- In TekScopes@...
<mailto:TekScopes%40yahoogroups.com> <mailto:TekScopes%40yahoogroups.com> ,
"dreadlk" <dreadlk@> wrote:
I posted about a month ago about searching for an internal
tracking
generator for my 2710 SA and I finally found one from an old Tektronix dealer.
Now I have this unit with Id No 119-3889-07 it's in a golden
colored all
metal case with 3 female SMA connectors and a 10 pin 2 row IDC connector.
Luckily I found several old Mobo serial connectors that are
the right
size but only use 9 of the 10 pins so I removed the wire and made a 10 wire
straight pin to pin cable from the connector ends.
Now I open up my 2710 to locate grounds and power pins on this IDC
connector and have a WTF moment. Where the Heck does this thing go!! I remember a side section that was free but the unit won't fit and all the
card slots are used up! Plus its just too big, so I have no clue how this
fits into a 2710.
I can find nothing on the net about this Option 04 tracking
generator,
it's not even listed in the 2710 service manual but I have seen several 2710
sold with Tracking Gens built in and the guy who sold it to me told me it
works with the 2710 Arggggh....
Anybody have any idea what I should do next? I can't even
locate the 10
pin connector that it supposed to plug into much less fit the unit into my
2710. Man I hope I did not waste all that money, I am having serious regrets
now. Sorry if this comes of as a Crazy Rant but I should have done my homework before making this purchase, I am upset with myself.
Robert KD8MJR
|
U460 is not necessarily bad but if it is, the CA3102 transistor array is not expensive or difficult to acquire. If U480 is bad, the 10102 ECL Quad 2-input NOR gate is readily available.
4 volts at both pin 10 of U480C and pin 7 of U480B is consistent with a logical high for a 10102 when operated at PECL levels but one pin of that pair should be low.
Check the voltage at pin 3 or 9 of U460. R456 might be open so check its resistance.
Check pin 8 or 1 of U460 which should be 2.2 volts.
Check pin 11 or 4 of U460. It should be either 2.4 volts or 2.0 volts depending on the trigger slope setting.
If U460 is socketted, then you could pull it and check the transistor junctions with the diode test function on your multimeter.
toggle quoted message
Show quoted text
On Mon, 24 Sep 2012 02:58:59 -0000, "tjeenz" <tjeenz@...> wrote: Hi David,
no square wave on the pins 10 and 7, both slopes.... it just measures a +4V dc, at both pins. Does it mean my U460 has died?
T.
--- In TekScopes@..., David <davidwhess@...> wrote:
That narrows the problem down to some point at or after transistor array U460. The good news is that the 2213 does not use a custom trigger IC so if the problem can be found, it should be relatively easy to fix.
With the trigger set to LINE and AUTO, adjust the LEVEL to midway and measure pin 10 of U480C and pin 7 of U480B. With both SLOPE settings, there should be at least a 300mV to 400mV square wave on each pin.
If there is, then check pin 2 of U480A. It should have a 300mV to 400mV square wave.
On Sun, 23 Sep 2012 10:51:11 -0000, "tjeenz" <tjeenz@...> wrote:
Hi David, still out there?
back from trip... I checked tp444: with trig on "line" i have a sinus with about 0.5V tt
if i connect a 4V top-top sinus at ch1 with trig on "int" i have 150mV sinus on tp444 (very small)
U460B, pin 2: norm: there is indeed a +0.65 - -0.65V which can be controlled by the trig-lev. pot. auto: sinus 50 mV (again with 4V tt at ch1). which seems very small...
in all situations the display does not show a "still" image.
tjeenz
--- In TekScopes@..., "tjeenz" <tjeenz@> wrote:
hi David, thx for the tip! i will check this as soon as possible (couple days out of town). i check tp444 before with ext. trigger and that was very small compared to the input signal. don't know the exact amplitude but much lower. probably back on saturday. tjeenz.
--- In TekScopes@..., David <davidwhess@> wrote:
I would start by looking at TP444 (middle of schematic 4) which should be a duplicate of the trigger input signal as selected by the source and coupling controls. At the very least, there should be a 60Hz sine wave at that point when the trigger source is set to line.
Then I would check pin 2 of U460B which should be a duplicate of the tri.gger level voltage as set by the trigger level control. In Auto mode this voltage will be between the peak levels of the trigger signal but in normal mode, it should be between +0.65V and -0.65V.
It would be nice to have a nicer scan of the service manual. I extracted and cleaned up schematic 4 from the one available online:
On Tue, 18 Sep 2012 11:00:41 -0000, "tjeenz" <tjeenz@> wrote:
Hi, Have a trigger problem with a 2213 tektronix.
In auto and line trigger i have a baseline. In norm not. Also ext. trigger fails (signal from function generator 4v pp and ch1 on 1v/div)
cannot get still image to see frequency, amplitude is ok. with signal on one of the channels, trig in auto i can almost get still image if freq is low enough with var hold-off. trig-level does nothing. Also the trigger led never blinks.
Checked all p.s. voltages. Are ok (no ripples), but according to service manual there should be some specific resistance on the voltages. These are way too low in my scope: 8.6v 114 ohm at tp500= 58 ohm +8.6v 95 ohm at W975 = 16 ohm +5v 330 ohm at W985 = 16 ohm +30v 905 ohm at W965 = 16 ohm +100v 12.5 kohm at W966 = 16 ohm
so i guess something is wrong somewhere. checked the big capacitors in the p.s, for 30V and +8.6V, both were in spec.
Are these resistance values in the manual a good reference for finding the faulty area?
regards, tjeenz
|