Re: Need specs for 015-0221-00 DC block


Craig Sawyers
 

I measure a series capacitance between input and output of 47nF. So into 50
ohms that implies that its 3dB low frequency is about 70kHz.

And since it is specced for use with the 7L12/13 it must go to around 2GHz
with a reasonable VSWR.

But beyond that, nothing firm.

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...]
On Behalf Of Mike Czuhajewski
Sent: 24 September 2012 13:32
To: TekScopes@...
Subject: [TekScopes] Need specs for 015-0221-00 DC block

This is their BNC DC block. Finding any data on it is pretty much
impossible;
tried lots of Google searches, old Tek catalogs, etc.
(It's listed under spectrum analyzer accessories.) About the only thing I
can
find is the voltage limit (50) and nothing more. They do give detailed
info on
the type N block, though.

Does anyone have any actual data on this? I'm looking for the frequency
range, especially the low end. I've seen it mentioned a few times but with
different sets of frequencies; nothing agrees. Does anyone know the actual
Tek spec on these?

Thanks.



------------------------------------