Re: Manuals for an HP3456B DMM?
Hi. It's worth checking the DC power rails for the GPIB logic section.? It's possibly isolated from all the "measurement" analogue/digital stuff. My old 3438A had "a lot" of 50Hz ripple on the GPIB
By Dave_G0WBX · #151260 ·
Re: Manuals for an HP3456B DMM?
Regarding the meter that only returns values, could it be that the ‘talk only’ switch is set and therefore the instrument is not accepting commands. As was mentioned earlier, it could simply be a
By nigel adams · #151259 ·
Re: 8591E CAL AMPTD failure
Hi Andrew I redid the alignment of the crystal filters of my 8591A. I used a 10nF capacitor and a 91 Ohm surface mount resistor to build the 3 CRYSTAL SHORT. Here are the results obtained; it
By Yves Tardif · #151258 ·
HP8935 Spectrum Analyzer Markers
Is there a way to take control of the spectrum analyzer marker position on a HP8935 and dial it around like you can on a basic spectrum analyzer? Thanks Jeff KC8VFN
By Jeff Acree · #151257 ·
Re: 8591E CAL AMPTD failure
Hi Andrew, This is the detail of the "short crystal", and yes, it is possible to do it easily. Yves De :
By Yves Tardif · #151256 ·
Re: Manuals for an HP3456B DMM?
And deoxit the address dip switches. They tend to take a set... -Chuck Harris <SteveHx@...> wrote:
By Chuck Harris · #151254 ·
Re: 8594E Cal Signal not Found ?
I should also make a correction that I keep saying "EEPROM" out of habit, when I actually mean "EPROM" in relation to this older equipment.? My usual work is in embedded microcontrollers. -mark
By Mark Litwack · #151253 ·
Re: Manuals for an HP3456B DMM?
Could be, but I'd try some deoxit on the connectors first, especially if it hasn't been used in a while. Steve Hendrix
By Steve Hendrix · #151252 ·
Re: Manuals for an HP3456B DMM?
It turns out that one of my meters is not accepting GPIB commands. It can return the display value, but no other commands are accepted. I am about half way through porting my 3455A tool to the 3456A
By BobH · #151251 ·
Re: 8495G Step ATT OPT 011 5V Cable Pinout
TNX, what pin number is the Minus?
By 4X6MU · #151250 ·
Re: 8495G Step ATT OPT 011 5V Cable Pinout
Minus is used to switch the attenuators Am 06.04.2025 um 18:50 schrieb 4X6MU via groups.io:
By DF6NA Rainer · #151249 ·
Re: 8495G Step ATT OPT 011 5V Cable Pinout
TNX Rainer, Which pin is (-) or GND? ILAN 4X6MU
By 4X6MU · #151248 ·
Looking for crystal test fixture 41901A details
Does anyone have 41901A SMD crystals PI-network test fixture? https://www.keysight.com/gb/en/assets/9018-01432/user-manuals/9018-01432.pdf I would like to have a good close-up photos of what the
By Leo Bodnar · #151247 ·
Re: Manuals for an HP3456B DMM?
Thanks for the inputs on the issues for this meter. I have written tools for several "R2D2" instruments. The command set is similar to the 3455A which I have already written a tool set for. One thing
By BobH · #151246 ·
Re: Manuals for an HP3456B DMM?
I'm not sure how luck plays into programming the 3456A, it is a very rigidly defined state machine that is well documented in the operation manual. The 3456A is programmed using an slightly enhanced
By Chuck Harris · #151245 ·
Re: E5052A Power On Test (3.3V Bus Supply)
Glad you got it fixed! I did look through the program code and found references to your errors but I couldn't find an obvious cause, only ways to patch past the errors. There are some diagnostic
By jmr · #151244 ·
Re: 8495G Step ATT OPT 011 5V Cable Pinout
Am 06.04.2025 um 08:00 schrieb 4X6MU via groups.io:
By DF6NA Rainer · #151243 ·
8495G Step ATT OPT 011 5V Cable Pinout
Hi, If anyone can pls help with the cable pinout for the 8495G Step ATT OPT 011 5V. Service Manual Documentation is not clear or missleading, it doesn't show where the common Ground and +5V
By 4X6MU · #151242 ·
Re: Manuals for an HP3456B DMM?
I love that I can calibrate them with a screwdriver rather than stepping through software menus. Good luck with the GPIB programming. They predate the SCPI programming standards. John??? KK6IL
By John Kolb · #151241 ·
Re: E5052A Power On Test (3.3V Bus Supply)
Figured it out. So the LT1084 arrived from digikey. I swapped U18 and U19 on the A3; no changes in LED state or boot up error. At this point, I had also removed PSU and removed PSU daughterboard,
By Kevin A. · #151240 ·