开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育
Manuals for an HP3456B DMM?
Does anybody know where I can find manuals for an HP3456B DMM? The HP site has a manual for the 3456A, how different is the 3456B from the 3456A? Thanks, BobH
By BobH · #151178 ·
Re: Success: Repair HP3455A Multimeter
Thank you for posting your story, some other 3455 owner will benefit someday. Dave Wise ________________________________ Sent: Wednesday, April 2, 2025 11:49 AM To:
By Dave Wise · #151177 ·
Success: Repair HP3455A Multimeter
Hi all, I checked U16 the latch connected to the driver U5 with a scooe. The output pin 15 ( connected to pin 9 of U5) was not stable and so I replaced U16. Now the 3455 is working again. At least no
By Ralf Rosche · #151176 ·
Re: hp-410b, early manual
Thanks Ron, that is the oldest 410B manual I've seen. It's for early non-prefixed Serial Number 401 through 7228. Dave Wise ________________________________ Sent: Wednesday, April 2, 2025 10:11
By Dave Wise · #151175 ·
hp-410b, early manual
In the course of refurbishing an old -410b, I was a bit confused in that my unit does not correspond to a manual stored here in our Files section.? So, I purchased a manual online and discovered that
By Ron na2o · #151174 ·
Re: LCR Meter E4980A Power On Test Fail
Thanks Nigel, I have updated the firmware from A.04.00 to A.05.00. Unfortunately, the failure persisted. Since it was able to passed all other tests, except CPU[2]: 16 tests failed, it seems like a
By KC Chong · #151173 ·
RF Alignment Failure in Agilent CXA signal analyzer N9000A
I have Agilent CXA Signal Analyzer N9000A, I'm getting these errors, please help. I referred manual and it suggested to check 50Mhz with -24dB, But when I check in my equipment its showing 50Mhz with
By ipother6@... · #151172 ·
Re: Request HDD image for Agilent Infiniium DCA 86100C mainframe
I'm still looking..
By Alex T · #151171 ·
Re: HP83712B EPROM content
No, it is not the same as for the 83732A. I tried but it is not working properly.
By tom_iphi · #151170 ·
Re: HP83712B EPROM content
"tom_iphi via groups.io" <iphi@...> writes: ko4bb has https://www.ko4bb.com/getsimple/index.php?id=manuals&dir=HP_Agilent_Keysight/HP_83732A_Synthesized_Signal_Generator Not sure
By Sven Schnelle · #151169 ·
HP83712B EPROM content
Hi, does anyone happen to have the data of EPROMs U7 and U24 of the HP83712B? I seem to have lost data... Tom
By tom_iphi · #151168 ·
Re: HP 85047A test set on 8753A analyzer -- switch control board jumpers to retrofit for opt 009 (mechanical switch)
[email protected]> wrote: I have looked at the circuit diagram of the 85046A, and noticed a capacitor value changed between the mechanical and electronic relays. Could that be the problem? Dave
By Dr. David Kirkby, Kirkby Microwave Ltd · #151167 ·
Re: E5052A Power On Test (3.3V Bus Supply)
Appreciate the suggestion. I reinstalled the A3 and tried cycling all the PSU connectors, including the motherboard one. I also brushed a tiny dab of Deoxit D100 on each male pin and cycled the
By Kevin A. · #151166 ·
Re: Tricks and thoughts about the 3457A battery
Thanks, I will order a few pieces to have them just in case. Best regards, Razvan
By Razvan Popescu · #151165 ·
Re: HP 53310A modulation domain analyzer memory upgrade
After a set from Aliexpress that didn't work, I got some from an eBay vendor I've found reliable (his stocks are all old, before the dodgy cloning started). I don't remember exactly what I bought but
By Adrian Godwin · #151164 ·
Re: HP3325A Trouble, Calibration FAIL 2 & 3
Hi all, I spent some more hours with my 3325A. I made a new effort and studied the service manual from scratch. The main trouble shooting flowchart directed me to group J, so I thoroughly examined
By Ralf Rosche · #151163 ·
Re: E5052A Power On Test (3.3V Bus Supply)
The first thing I recommend that you do is to look at all the PSU cables in the PSU area including the big multiwire one that loops back to the motherboard. As a first attempt, try partially removing
By jmr · #151162 ·
Re: HP6209B component location layout
Forgot to mention, I am thinking of leaving the 50K voltage adjustment pot in place as is, even tough already ordered a 100K replacement Bourns "Helipot". Everything seems to work fine, and I can
By Alex · #151161 ·
Re: HP6209B component location layout
I poked around a bit further and discovered that the +43V reference supply that was reading low at 34.2V actually had 14V of ripple on it. After removing and checking C12 a 450uF 85V cap, it was
By Alex · #151160 ·
Re: HP 85047A test set on 8753A analyzer -- switch control board jumpers to retrofit for opt 009 (mechanical switch)
I believe the control module (rear panel) has one or more abnormalities - I tried setting attenuator (stimulus>power>attenuator 1) and rather than adjusting the attenuator, the test set switched ports
By David Feldman · #151159 ·