Re: 2246 Mod A - no trace
After installing the socket, I did an ohmmeter test on all connections to the socket from other chips and they are all fine. By default (no voltage applied), the input selection relay selects GND, so it should still show the trace. Ch2, Ch3 and Ch4 were working before I installed the socket, so it is baffling.
toggle quoted message
Show quoted text
It does show "Self check Volts Cal error" (this is new) and also complains about the battery not working correctly (this error was there before). With the beam finder on, it shows a very bright horizontal line display with some vertical deflection at the center of the screen. Believe that is how it is supposed to work. I will pull back and see if somehow the front panel controls are far off somehow, though that would not result in a Volts Cal error. Something is not good with the motherboard near the repair - the same something that caused the failure to begin with. Priya. --- In TekScopes@..., "Tom Miller" <tmiller11147@...> wrote:
|
|
tektronix 318 logic analyzer probe
|
|
Re: IC programmers slightly OT
I'm partial to Advin myself. Alternatively, a decent used Data I/O
programmer. Happy hunting. * REPLY SEPARATOR * On 19-Feb-13 at 10:46 Alexandre Souza - Listas wrote: forAnybody familiar with the Xeltec line of programmers? Good, bad, seemingly high prices.Good programers, but I'd buy Elnec |
|
Re: 2246 Mod A - no trace
¿ªÔÆÌåÓýWhen you removed the chip did any of the plated through holes loose the
plating? I think the board is up to six layers. You may have lost an internal
connection to the chip.
?
?
Tom
?
|
|
Re: 2246 Mod A - no trace
Priya,
toggle quoted message
Show quoted text
Wow, I didn't know it was all channels! Check your trigger setting, try a single shot too. maybe your focus is so far out you can't see it. Look at it in the dark too. Check the rejection section in the trigger area, set it for DC. Use the menu to set things to default just in case you got into unfamiliar territory. I don't see how putting in that socket would have done anything else barring you accidently damaged something in the process or zapped something that is static sensitive. Pull back for a second. Hopefully you just overlooked something! Nick Nielsen --- In TekScopes@..., "Mover" <priya_jakatdar@...> wrote:
|
|
Re: 2400 series of scopes -- automatic voltage measurement
The 24xx digital scopes lack any such limit because they make
measurements in a completely different matter. What I wonder is why the 2247A series which *do* operate in a similar way do not have such a limit listed. If the peak detection circuit in the 2465B series is good down to the specified 10ns, then I have no idea what the 1 MHz restriction refers to. Those two specifications seem to conflict. On Tue, 19 Feb 2013 03:33:14 -0000, "victor_j_silva" <daejon1@...> wrote: I assume you mean a 2445A/65A/67 and 2445B/65B/67B. |
|
Re: 2400 series of scopes -- automatic voltage measurement
Are you referring to Page 1-11, Volts, Maximum sine-wave frequency?
toggle quoted message
Show quoted text
This is very confusing because here it actually says greater than or equal to 1MHz. --Victor --- In TekScopes@..., Steve <steve65@...> wrote:
|
|
Re: IC programmers slightly OT
Mark Wendt (Contractor)
Magnus/Alexandre,
toggle quoted message
Show quoted text
Thanks for the feedback! Mark On 2/19/2013 9:47 AM, magnustoelle wrote:
|
|
Re: DC504 counter display
HankC <hankc918@...> wrote:
I wasn't clear in describing my symptom.Nope. They're muxed, so the digit & segment drivers are ok. Gotta be LED-specific (or socket, or solder joint...) -ls- |
|
Re: DC504 counter display
Jim
If these are socketed, first try wiggling the individual displays in the socket, then try swapping a known good display for an apparently defective display. I've known these displays to lose a segment. ?I've also known driver lines to go bad. ?Swapping the display will diagnose both problems. 73 Jim N6OTQ
|
|
Re: DC504 counter display
I wasn't clear in describing my symptom.
5 of the 7 digits are OK; they display an "8" when you hold the reset button. One digit has a bad center horiz segment; another digit has a bad top horiz segment. Does that change anyone's diagnosis? HankC ________________________________ From: "TekScopes@..." <TekScopes@...> To: TekScopes@... Sent: Tuesday, February 19, 2013 2:49 AM Subject: [TekScopes] Digest Number 7355 All about classic Tektronix CRT o'scopes All about classic Tektronix CRT o'scopes Group 15 New Messages Digest #7355 1a Re: What use for a 640 Ohm 1x Probe? by "KeepIt SimpleStupid" KeepItSimpleStupid 1b Re: What use for a 640 Ohm 1x Probe? by "Don Black" donblack1au 1c Re: What use for a 640 Ohm 1x Probe? by "Philip" ndpmcintosh 1d Re: What use for a 640 Ohm 1x Probe? by "David" david_william_hess 2a Re: Tektronix 2465 capacitors etc. by "Chris" digitalinuxguy 2b Re: Tektronix 2465 capacitors etc. by "John Snyder" ykochcal 3a DC504 counter display by "HankC" hankc918 3b Re: DC504 counter display by "Chris van Lint" 3c Re: DC504 counter display by "David" david_william_hess 3d Re: DC504 counter display by "taylorvandy" taylorvandy 3e Re: DC504 counter display by "Miroslav Pokorni" mpokorni1 3f Re: DC504 counter display by "Glenn Little WB4UIV" glittle_29445 4a Manual / schematic for Leader 524S counter? by "Dave C" davec2468 4b Manual / schematic for Leader 524S counter? by "Dave C" davec2468 5a Tek 485 by "John Polakowski" johnpolakow Messages 1a Re: What use for a 640 Ohm 1x Probe? Mon Feb?18,?2013 8:19?pm (PST) . Posted by: "KeepIt SimpleStupid" KeepItSimpleStupid I'm wondering if it could be an RF probe.? . com/rfprobe1. htm has a resistance of 640 ohms end to end on the center conductor. To what use Reply to sender . Reply to group . Reply via Web Post . All Messages (7) . Top ^ 1b Re: What use for a 640 Ohm 1x Probe? Mon Feb?18,?2013 8:43?pm (PST) . Posted by: "Don Black" donblack1au I think you are measuring the inner conductor resistance. This is made of resistance wire (often wound in a spiral) to damp reflections back the cable. The cable isn't terminated (the input impedance of the scope is 1 Meg) and like any transmission line energy is reflected where it isn't absorbed by matching termination. This is a compromise to damp such reflections, useful at low frequencies. Tektronix explains their techniques in their concept book series, the one you want is called 'Oscilloscope Probe Circuits" . It's available to download on the Internet and is an excellent guide to probe design. Silicon Chip also has a good article on probes a while back that I think explained it. Don Black. I'm wondering if it could be an RF probe. . com/rfprobe1. htm Reply to sender . Reply to group . Reply via Web Post . All Messages (7) . Top ^ 1c Re: What use for a 640 Ohm 1x Probe? Mon Feb?18,?2013 8:55?pm (PST) . Posted by: "Philip" ndpmcintosh Yeah. It doesn't have any markings on it other than the name "Fairchild&quo t; on the cable, and I am glad I checked it before I actually tried to test it. In looking at an old but classic oscilloscope book it mentions a "direct" probe that has essentially 0 resistance used for checking low impedance, low frequency circuits. And, it also describes the "isolation&quo t; probe which has an R of 4.7 to 10kOhms. It might be something intended to serve as one of these. It looks like it is from the early 60's and perhaps was used on the Fairchild 766H. I doubt there would be much current use for this probe.
Reply to sender . Reply to group . Reply via Web Post . All Messages (7) . Top ^ 1d Re: What use for a 640 Ohm 1x Probe? Mon Feb?18,?2013 9:38?pm (PST) . Posted by: "David" david_william_hess I agree with what Don posted. The 640 ohms is just the resistance of the inner conductor of the coaxial cable for a x1 oscilloscope probe. I measured about 250 ohms on a x1 probe I happen to have in reach. On Tue, 19 Feb 2013 04:55:41 -0000, "Philip" ndpmcintosh@ mac.com> wrote: Yeah. It doesn't have any markings on it other than the name "Fairchild&quo t; on the cable, and I am glad I checked it before I actually tried to test it. Reply to sender . Reply to group . Reply via Web Post . All Messages (7) . Top ^ 2a Re: Tektronix 2465 capacitors etc. Mon Feb?18,?2013 8:36?pm (PST) . Posted by: "Chris" digitalinuxguy John, A HUGE thank you on checking that and giving me your value! I was hoping the short wasn't in the main board somewhere as that doesn't look like fun to pull in the least. I just checked the resistance from the pin 7 (blue 87v wire) on the connector (power supply still out and on the bench) and I got a value of 25.55Kohm, so it would appear I should be OK to go at that (assuming that powering something else up on the main board doesn't cause it to short/overload the supply). I was testing the diodes on the power supply and diode CR1244 is giving me 0.538V forward and 0.630v reverse. I think I might pull it and check it out of the board later tonight after I eat etc. Thanks for the help! Chris --- In TekScopes@yahoogrou ps.com, "John Snyder" wrote:
Reply to sender . Reply to group . Reply via Web Post . All Messages (14) . Top ^ 2b Re: Tektronix 2465 capacitors etc. Mon Feb?18,?2013 11:49?pm (PST) . Posted by: "John Snyder" ykochcal Chris The CR1244 is across the B-E junction of the transistor, so in circuit it would not measure Open like a diode out of the circuit would. I have my board out waiting for the inverter board to dry from it's cleaning, and I measure .706V forward (of the diode) and .581V reverse (instead of open, which is forward for the transistor junction.) You could measure all the voltages Q1220, Q1221, Q1222, and Q1224 E,B and C along with U1281A pin 1, 2 and 3 CR1220 anode to see what is out of wack John -----Original Message----- From: TekScopes@yahoogrou ps.com [mailto:TekScopes@yahoogrou ps.com] On Behalf Of Chris Sent: Monday, February 18, 2013 8:37 PM To: TekScopes@yahoogrou ps.com Subject: [TekScopes] Re: Tektronix 2465 capacitors etc. John, A HUGE thank you on checking that and giving me your value! I was hoping the short wasn't in the main board somewhere as that doesn't look like fun to pull in the least. I just checked the resistance from the pin 7 (blue 87v wire) on the connector (power supply still out and on the bench) and I got a value of 25.55Kohm, so it would appear I should be OK to go at that (assuming that powering something else up on the main board doesn't cause it to short/overload the supply). I was testing the diodes on the power supply and diode CR1244 is giving me 0.538V forward and 0.630v reverse. I think I might pull it and check it out of the board later tonight after I eat etc. Thanks for the help! Chris --- In TekScopes@yahoogrou ps.com, "John Snyder" wrote: the J121 connector unitBehalf Of Chrisget into all this, so maybe my understanding is skewed. :)three groups, and the wrong voltages on the right side, so it would seem that'slook at the three 10 uF caps and all the diodes in that area. Be sure to onlyuse fast diodes for any replacements.capacitors can come from the factory new. Did you measure all these voltages in the------------ --------- --------- ------ Yahoo! Groups Links Reply to sender . Reply to group . Reply via Web Post . All Messages (14) . Top ^ 3a DC504 counter display Mon Feb?18,?2013 8:52?pm (PST) . Posted by: "HankC" hankc918 Anyone familiar with the TM500 counter, model DC504 ? I just picked one up & I noticed a couple of the digits are missing the middle horizontal segment. So, it displays an "8" as a "0" , or a "0" as a "U" . In your experience, is this a display problem or one of the chips driving the display ? HankC Reply to sender . Reply to group . Reply via Web Post . All Messages (6) . Top ^ 3b Re: DC504 counter display Mon Feb?18,?2013 8:55?pm (PST) . Posted by: "Chris van Lint" More than likely the LED read outs. Chris VK4CVL At 03:52 PM 19/02/2013, you wrote: Reply to sender . Reply to group . Reply via Web Post . All Messages (6) . Top ^ 3c Re: DC504 counter display Mon Feb?18,?2013 9:18?pm (PST) . Posted by: "David" david_william_hess Going by the schematic, the display digits are multiplexed which is the common configuration so if the middle horizontal segment is working on at least some digits, then the 7 segment display driver is fine. If some segments of every 7 segment display work, then the column display driver is fine as well. That pretty much narrows any problem to bad display segments or possibly an open connection to the missing display segments. I would normally consider the later less likely than the former except that two of the same segment are missing. On Mon, 18 Feb 2013 20:52:57 -0800 (PST), HankC hankc918@yahoo. com> wrote: Anyone familiar with the TM500 counter, model DC504 ? Reply to sender . Reply to group . Reply via Web Post . All Messages (6) . Top ^ 3d Re: DC504 counter display Mon Feb?18,?2013 9:48?pm (PST) . Posted by: "taylorvandy" taylorvandy --- In TekScopes@yahoogrou ps.com, HankC wrote: Common problem - FND357 displays go bad - fixed mine with two new LED displays. 6 bucks each, do a google search to see who might have them cheap. Cheers, Taylor Reply to sender . Reply to group . Reply via Web Post . All Messages (6) . Top ^ 3e Re: DC504 counter display Mon Feb?18,?2013 10:48?pm (PST) . Posted by: "Miroslav Pokorni" mpokorni1 Tektronix had propensity to use chip IC sockets. Try to remove display package and reseat it. That might re-establish pin connection. Sometimes it takes several re-seatings to clear the socket. Miroslav Pokorni On 2/18/2013 9:18 PM, David wrote: Reply to sender . Reply to group . Reply via Web Post . All Messages (6) . Top ^ 3f Re: DC504 counter display Mon Feb?18,?2013 11:17?pm (PST) . Posted by: "Glenn Little WB4UIV" glittle_29445 If it displays a "8" as a "0" and a "0" as a "U" it is almost certainly a driver issue. If it were a bad display segment the "0" would display as a "0" if the "8" displayed as a "0" . Do all positions display the same incorrect digits? 73 Glenn WB4UIV At 11:52 PM 2/18/2013, you wrote: Anyone familiar with the TM500 counter, model DC504 ? I just picked one up & I noticed a couple of the digits are missing the middle horizontal segment. So, it displays an "8" as a "0" , or a "0" as a "U" . In your experience, is this a display problem or one of the chips driving the display ? HankC Reply to sender . Reply to group . Reply via Web Post . All Messages (6) . Top ^ 4a Manual / schematic for Leader 524S counter? Mon Feb?18,?2013 9:46?pm (PST) . Posted by: "Dave C" davec2468 Anybody have the user guide for this counter? (Leader user guides always include the schematic.) Thanks, Dave Reply to sender . Reply to group . Reply via Web Post . All Messages (2) . Top ^ 4b Manual / schematic for Leader 524S counter? Mon Feb?18,?2013 9:48?pm (PST) . Posted by: "Dave C" davec2468 Um... that's a model 824S counter. Dave -=-=-=- Anybody have the user guide for this counter? (Leader user guides always include the schematic.) Thanks, Dave Reply to sender . Reply to group . Reply via Web Post . All Messages (2) . Top ^ 5a Tek 485 Mon Feb?18,?2013 11:12?pm (PST) . Posted by: "John Polakowski" johnpolakow I have two Tektronix scopes: a 485 and a 2465. Both of them need a little TLC, but I'm going to focus on the 485 at the moment as it is non-functional. I'm not an experienced electronics tech, but I'm not a complete novice either. I really like the blue phosphor of the 485. For the moment the traces will show up, but the traces don't respond to input on either channel. I think the problem lies with the attenuator cards, but I'm not 100% sure thats the only problem. This I know for sure: if I take my signal gen and hook up its output to the place on the vertical amplifier where the attenuator plugs in, the signal is displayed on the scope. Another strange symptom is that on Channel 1, if the impedance is on the 50 Ohm setting, the trace is there, but on the 1 MegaOhm setting it disappears. Any thoughts guys? Have any experience with this sort of thing? I've taken out the channel 1 attenuator and have it in my hands right now. Where should I start troubleshooting? Thanks! John Reply to sender . Reply to group . Reply via Web Post . All Messages (10) . Top ^ Visit Your Group View All Topics Create New Topic 18 New Members 8 New Photos We are making changes based on your feedback, Thank you ! Submit Feedback The Yahoo! Groups Product Blog Check it out! CHANGE SETTINGS TERMS OF USE UNSUBSCRIBE |
|
Re: IC programmers slightly OT
Hi Mark,
toggle quoted message
Show quoted text
I have not seen any Xeltec programmer over many years. However, if I am allowed to make a recommendation: I can honestly recommend the products from batronix.com. I have their BX40, and it works like a charm. Oh, I can see that they are a Distributor of Xeltec - interestingly enough... Cheers, Magnus --- In TekScopes@..., "Mark Wendt (Contractor)" <mark.wendt@...> wrote:
|
|
Re: 2246 Mod A - no trace
Hi Nick,
toggle quoted message
Show quoted text
There is currently no trace on any of the channels after installing the IC socket at U174. However, the beam finder works as do all the menus and also the vertical time base cursors and readouts. I suspect something with the motherboard now at or near U174. Using your plan, I will apply an identical calibration signal to Ch2 and Ch3 and compare what happens to the signal using my R7704 as a test scope. More to come on this ... Yes, I will be careful as you say to not blow up something else along the way! Priya. --- In TekScopes@..., "nielsentelecom@..." <nielsentelecom@...> wrote:
|
|
Re: Tektronix 2465 capacitors etc.
¿ªÔÆÌåÓýChange Q1222. It should be full off and the collector should be up around
90 volts or so. However, Q1223 could also be pulling that line down so change it
also. You could lift the collector of Q1223 and see if the supply comes up just
to rule that out. Also (2), is R1212 ok? If it were open, the same issue would
be there.
?
Tom
?
?
?
?
|
|
Re: DC504 counter display
Mark Wendt (Contractor)
¿ªÔÆÌåÓýOn 2/18/2013 11:52 PM, HankC wrote:
What happens when the digit is a "5" or a "6"?? How about when it is a "9" or a "7"?? Is it only the top and middle horizontal segments? Mark |
|
Re: What use for a 640 Ohm 1x Probe?
Mark Wendt (Contractor)
So spake Mr Alex Gates... ;-)
toggle quoted message
Show quoted text
Mark On 2/18/2013 5:50 PM, Alex wrote:
640 ohms ought to be enough for anyone. |
|