Re: 7904 pre-260000 PSU crackling
On Jan 5, 2025, at 06:34 , James via groups.io <james@...> wrote:Are you sure it's in the PSU? Does it change if you bump the HV lead? I've had a tic from my 7704A that sounds like a static crackle if I move the HV lead with a nonconductive tool. Have not been able to make that go away. I have a 7904 that's a bit older than yours, and IIRC it had a crackle that went away after disconnecting the CRT anode connector, shorting it to the frame, cleaning it, and applying some silicone grease. It still has a PSU tick when starting, but I understand this is common. Adam |
Re: 7904 pre-260000 PSU crackling
Thanks Zen, much appreciated.
I checked the main caps fairly cursorily for capacitance, which was textbook value, but didn't check them at proper operating voltage for issues. I will certainly do so when I pull the PSU apart. I forgot to mention that the crackling continues (decaying) for a minute or two after switch off, maybe that helps narrow the culprit down? Thanks for the info about the resistances, understood! Re the neon, my poor description, it flickers dimly (like most neons do) when the psu is switched on and very quickly stops after it is switched off. It doesn't blink though if that matters. I note that the reservoir caps don't hold charge for any time after switch off. The switch, like all Tek switches, looks built to last forever. It functions fine thankfully. My understanding is that the PSU in the 7904A is rather different but have no experience to back that up. |
Re: 7904 pre-260000 PSU crackling
I have a 7904A with similar issues on the bench also with a VERY early serial number 507. In my case I have confirmed that there is issues with the main bulk filters they are arcing internally. 200 Vdc cap is being fed by a 100ma 200 Vdc current source is building to about 70 ish Vdc then arcing down to about 50vdc. It Is causing a static like tic when it arcs over. This one is blowing the mains fuse.
toggle quoted message
Show quoted text
For the resistance testing these measurements are not an exact spec but more like a gross find the short. This is due to the meter used for the measurements. All Ohms ranges have a voltage and current applied to the circuit and this can throw off the readings. Some meters top out at 3 Vdc some like my Keithly go up to 10. There is no way of knowing what active components are turning on at the higher voltage. Exact is not needed there but in the ball park is fine. 25Khz is the switching frequency of the supply. 4 is the power switch good? The neon should stop eventually it can take a bit given the health of the bulk caps if the power switch is leaking current, it will keep the caps charged due to line -> switch -> rectifier -> bulk caps. This would however be a REALLY odd failure mode. Zen -----Original Message-----
From: [email protected] <[email protected]> On Behalf Of James via groups.io Sent: Sunday, January 5, 2025 9:34 AM To: [email protected] Subject: [TekScopes] 7904 pre-260000 PSU crackling Happy New Year everyone! I have a 7904 s/n 215841. I got it at an auction in the UK where I live, where there don't seem to be many of this series available, and it was FILTHY when I picked it up. Knowing the tantalum issue I immediately replaced as many of the PSU tants as I could get at without dismantling the unit completely, and removed as much of the grime as I could which was like oily black soot and covered all the circuit boards - got a surprise when I discovered that the HV line was bright pink! Now it works, the PSU comes on and the voltages are within spec, screen is bright (albeit not a patch on my old 545 - is this normal?). The issue is that almost immediately there is a slight ticking noise inside the PSU. It starts as a fairly regular tick (every second or two) but after ten or twenty minutes has become more of a quiet crackling noise. This is far lower in volume than eg. a start up tick. There was no visible light from arcing when looking in a darkened room but I didn't want to stick my face too close, also there is no discernable burning smell. After the crackling develops the voltages remain steady, the screen doesn't change. I dug into the PSU, lord what a job, to look for anything I could find but apart from plenty of grime didn't find any tell-tale carbon arcs or blackened components. I had a look inside the cap multiplier but, again, no obvious arcs. The neon that is on the inverter board had stopped working so I replaced that, changed out the 470 Ohm sense resistors R1429 and 1439 which were burnt but still near-ish to spec, reassembled, and tried again, more in hope than expectation. No change. My questions are: 1. Can anyone point me at the likely source of the crackling? I am fearing it might be the transformer futzing out, but have no idea, after pulling the PSU apart, how I could test it reliably (and safely). If you think it is that, can it be rewound? I can't do this sadly, but I do know a couple of professional trafo winders. If the symptoms scream "cap multiplier" at you I do have a used spare that is labelled good, bought from Sphere's old stocks but, before replacing it haphazardly, I wanted to ask those with more experience than me how to proceed. That's all of you I suspect... 2. The service manual suggests testing impedances on the LV regulator board test points with the power wiring looms connected and disconnected , however their advice is to investigate if the measurements are "significantly lower". But these are for s/n 260000 and up. My PSU is not much different to the circuits in the Tabalabs .pdf. However, my measurements differ from the specs given in table 4-3 as follows: +50V TP with wiring loom connected - spec 3.4k Ohm, measured 2.95 kOhm. Is this significantly lower? Disconnected, the regulator resistance rises to 23.6 kOhm. -50V TP with wiring loom connected - spec 270 Ohm, measured 3.74kOhm. Disconnected, the resistance is 4.34 kOhm. This appears significantly higher but is it an issue. Thoughts? 3. The manual suggests testing the Pre-reg power supply across R1287. The waveform is more or less like that illustrated in 4-4 (A) showing correct operation. Looking at the burst voltages at the test points given in table 4-4 there are stable voltages with little evidence of bursts, and they come at 25 kHz rather than 100 Hz intervals, I assume this is normal? 4. The new neon bulb on the inverter board is supposed to blink to indicate charge on the reservoir caps. Mine flickers dimly and continuously. Is this normal or does it indicate something wrong? Any and all help, advice, and pointers gratefully received. |
Re: Adapter PCBs for 148-003x-00 relay replacement
After seeing some interest, I decided to populate at least 1 panel with all SMT components. This might also be the way selling them to you. This enables me to paste and reflow solder a whole panel which makes soldering of the SOT363 rectifier component a lot easier than soldering a single tiny PCB by hand.
As said in my initial post, I'm still waiting for the relays to arrive. Then I'd also like to do some measurements to see how good the solution performs in comparison to the original relay. Therefore, I think it takes at least another 2-3 weeks until I can give you some feedback. At work I have an 8GHz VNA (which is hopefully not blocked by someone else), this might be helpful when exploring HF-properties of the whole contraption. Regarding the 148-0063-00 relay. I had to google the pinout of this thing, but I think this should also be doable with a little solder bridge between adjacent pins. Best Regards, Wolfgang |
7904 pre-260000 PSU crackling
Happy New Year everyone!
I have a 7904 s/n 215841. I got it at an auction in the UK where I live, where there don't seem to be many of this series available, and it was FILTHY when I picked it up. Knowing the tantalum issue I immediately replaced as many of the PSU tants as I could get at without dismantling the unit completely, and removed as much of the grime as I could which was like oily black soot and covered all the circuit boards - got a surprise when I discovered that the HV line was bright pink! Now it works, the PSU comes on and the voltages are within spec, screen is bright (albeit not a patch on my old 545 - is this normal?). The issue is that almost immediately there is a slight ticking noise inside the PSU. It starts as a fairly regular tick (every second or two) but after ten or twenty minutes has become more of a quiet crackling noise. This is far lower in volume than eg. a start up tick. There was no visible light from arcing when looking in a darkened room but I didn't want to stick my face too close, also there is no discernable burning smell. After the crackling develops the voltages remain steady, the screen doesn't change. I dug into the PSU, lord what a job, to look for anything I could find but apart from plenty of grime didn't find any tell-tale carbon arcs or blackened components. I had a look inside the cap multiplier but, again, no obvious arcs. The neon that is on the inverter board had stopped working so I replaced that, changed out the 470 Ohm sense resistors R1429 and 1439 which were burnt but still near-ish to spec, reassembled, and tried again, more in hope than expectation. No change. My questions are: 1. Can anyone point me at the likely source of the crackling? I am fearing it might be the transformer futzing out, but have no idea, after pulling the PSU apart, how I could test it reliably (and safely). If you think it is that, can it be rewound? I can't do this sadly, but I do know a couple of professional trafo winders. If the symptoms scream "cap multiplier" at you I do have a used spare that is labelled good, bought from Sphere's old stocks but, before replacing it haphazardly, I wanted to ask those with more experience than me how to proceed. That's all of you I suspect... 2. The service manual suggests testing impedances on the LV regulator board test points with the power wiring looms connected and disconnected , however their advice is to investigate if the measurements are "significantly lower". But these are for s/n 260000 and up. My PSU is not much different to the circuits in the Tabalabs .pdf. However, my measurements differ from the specs given in table 4-3 as follows: +50V TP with wiring loom connected - spec 3.4k Ohm, measured 2.95 kOhm. Is this significantly lower? Disconnected, the regulator resistance rises to 23.6 kOhm. -50V TP with wiring loom connected - spec 270 Ohm, measured 3.74kOhm. Disconnected, the resistance is 4.34 kOhm. This appears significantly higher but is it an issue. Thoughts? 3. The manual suggests testing the Pre-reg power supply across R1287. The waveform is more or less like that illustrated in 4-4 (A) showing correct operation. Looking at the burst voltages at the test points given in table 4-4 there are stable voltages with little evidence of bursts, and they come at 25 kHz rather than 100 Hz intervals, I assume this is normal? 4. The new neon bulb on the inverter board is supposed to blink to indicate charge on the reservoir caps. Mine flickers dimly and continuously. Is this normal or does it indicate something wrong? Any and all help, advice, and pointers gratefully received. |
Burned diode on CRT driver board CSA803A
Dear all,
The screen of my CSA803A developed a problem. It showes flyback lines, low intensity and I could not get it calibrated properly. The size of the picture, focus and convergence are o.k.. After removing the CRT driver board A8 (671-1027-01), it turned out that a (zener?) diode is completely burned and also a 100uF/50V elco did not look healthy. This diode is immediately below the metal box shielding the flyback transformer and alongside what seems to be a very big tantalum elco. Any help in finding the type of this diode is very much appreciated! Kind regards, Gerard |
Re: Looking for 067-0521-01
Ditto. I found a TU-7 on ebay for a reasonable price on ebay awhile back
after searching for several years. DaveD KC0WJN On Sun, Jan 5, 2025 at 07:47 Eric Boyle via groups.io <k0elb= [email protected]> wrote: Just keep an eye out on eBay for them, I bought a 1M1 for $40 plus |
Re: Tek T-130 LC - add mains/EMC filter or not?
The130 uses two low frequency oscillators, 100..150 kHz. No modern line filter works below the FCC, or other minimum frequency.
The instrument is not affected by incoming EMI nor creates conducted EMI. I would forget the filter and just install a plain IEC recepticle Jon |
Tek T-130 LC - add mains/EMC filter or not?
As my Tek 130 has lost its NEMA 5-15 inlet, and been butchered to use an IEC inlet.
I'm tidying up the rear panel and adding a small anodised plate to cover theunsightly square-peg-in-round-hole issue, but/and I have the option to either: (a) Use an integrated IEC filter (e.g. a 1A Schaffner), or, (b) Add some basic XY filtering across a non-filtered IEC, or, (c) Add a chassis mount power filter. (a) is neatest but needs further minor chassis butchering, (b) very easy, and (c) doable but less original. Before I consume any cycles investigating, do people here think I need to consider adding filtering to this particular instrument? |
Re: Looking for 067-0521-01
On 4 Jan 2025 9:25 pm, John Williams wrote:
I have not seen the history of these fixtures, the TU-7, 1M1, and 067-0521-01. I would like to believe the products developed out of in-house test fixtures used to initially calibrate new units, or in repair to test customers oscilloscopes before shipping them back.I would think that very possible. HP developed a three-piece VTVM calibration system for use in the factory and by field techs. The calibration generator was originally assigned a five digit specification number (a system used in-house) but later released to the public in the catalog with a conventional three-digit plus one letter model number. -- Dale H. Cook, GR/HP/Tek Collector, Roanoke/Lynchburg, VA |
Re: Looking for 067-0521-01
I have not seen the history of these fixtures, the TU-7, 1M1, and 067-0521-01. I would like to believe the products developed out of in-house test fixtures used to initially calibrate new units, or in repair to test customers oscilloscopes before shipping them back. If anyone has some information on this, I would appreciate hearing about it.
|
Re: 7904 strange problems
I started looking at the vertical interface board and how the 7A18s could not work here while 7A26s are fine, then took some wild goose chases that were dead ends. I decided to make sure it was the 7904 and not the 78A18s again. I went to try them in another 7904 - one that I had fixed previously. I fired it up and it immediately went crazy due to a HV problem inside the PS unit. Not due to the 7A18s, but plain old HV breakdown I think because it's been sitting in the garage by the front doors for a long time, in the damp and cold. I never had one bit of HV trouble on this one or the others through the summer during all the repairs. So, I'm down one unit again. This is why I need so many mainframes, just to be sure of some being usable.
Anyway, I tried again with another unit that fired up OK, and the 7A18s worked, so the problem was this particular 7904, but still a mystery. After much more fooling around with it, I finally found there was no +50V supply ONLY on the plug-in connectors. The +50 was all over the place, and in fact is distributed from the main interface to everything else, so this was very unusual. There are no filter chokes or resistors or anything else in series to the connectors. Long story short, the connection failed at the lead and plated through hole at the plus end of the big Ta bypass capacitor for that supply. Either I or someone else long ago had pulled that cap to check or replace it, and the bond to the inner layer run that apparently feeds only the plug-in connectors broke down on re-soldering, unbeknownst to me. I believe this is a four-layer board, which always has potential grief on soldering difficult parts or locations. This looked very bad, the kind of thing that can make it a junker due to having to rip everything apart so far down just to figure it out and repair the problem. This is the worst part of the whole scope, accessibility-wise. After some difficult measurements I found that the connectors were all still tied to each other at least, so I just had to go around the broken hole to re-hook to +50V and the cap lead, without having to take everything apart. I managed to tack onto the connector that comes right from the supply (a sure thing), then to the cap and then the pin 19 on the right vertical, and voila, back in business. I proofed it with the 7A18s, all good apparently. So now it all seems to work with all slots and plugs so far. That crashing when the 5V light is unloaded is still there - a strange one but maybe something also related to the main interface. I checked to make sure it wasn't shorted to the regular +5V - it's OK there. Next is to figure out the return path(s). I first thought I need to check all the mods that are supposed to be done in the PS, but remembered that I had worked on it extensively and almost certainly did all those already, before shelving it and later finally finding the bad HV multiplier last week. It is getting close, and I think I'll have it figured out soon. Once I get this one done, I'll have to fix the other one that just crapped out. It seems like I always have at least one that's busted - it's hard to keep up. Ed |
Re: Adapter PCBs for 148-003x-00 relay replacement
I'd like to have some of those boards. I have two 7A13s and would like to replace all the relays. I also have a 7A12 which uses 8 of them.
toggle quoted message
Show quoted text
In addition I need two of the "polarized" relays (148-0063-00). I cobbled up something out of two small relays that worked but getting two small relays in the small compartment is challenging. Please let me know. Thanks, Barry - N4BUQ ----- Original Message -----
From: "w.m.d.koch via groups.io" <w.m.d.koch@...> Hello all, |
Re: odd connectors on an 11801B
They operate a test equipment calibration service and sales from a
brick and mortar so it's a step up from an ebay seller but yes that price is still absurd for what it is. What drives me nuts is that they are experts so they should know this isn't the conventional version of 1M and they think it's part of 1T which it obviously is not. The original listing I shared was $397.00 with about $120 for shipping. Which is closer to sanity. I am unclear on how they ended up making two listings for the same item. On Fri, Jan 3, 2025 at 5:10?PM John Williams via groups.io <books4you4@...> wrote:
|