Re: Reverse part lockups
Please find my comments embedded. My text refers to the text immediately above it:
On Mon, Aug 31, 2020 at 07:39 PM, Dallas Smith wrote: I fully agree. I think making them full OCR is different than making them searchable.Absolutely: - "Full OCR" implies converting a printed (text-)document to a character-by-character computer-redable and -modifiable document. - "Searchable" means that text can be searched on a computer, even if the text is (or seems) part of an image on the computer screen. I think what most of us are looking for is just text searchable, not full-OCR, meaning the document will not be editable or otherwise modified as text. Bookmarks may be added to an unlocked document by a user with the right tool. I sent Raymond an unlocked version on the 155 RPR for him to checkout to see ifThe 155-document that I received is searchable (as I got it from Dallas) but hardly compressable further. It's only 11 MB, that's small by comparison with some of the others. Are there more RPR files for other parts available somewhere, such as the 156 parts?Yes and no. Several are available on Tekwiki, here: Unfortunately, 156 isn't one of them (yet?). Several other interesting ones are available though, like 151 (Transistors). That's a big one (234 MB). I thought that might be caused by the images (each page is an image) being uncompressed or only slightly compressed in the PDF. With the rather poor quality of the images, the right way to "OCR" them is to use the method of adding an invisible text layer put "behind" the images and searching that layer. A hit will cause the (approximate) spot on the image to be highlighted. Most OCR programs can prepare the text layer and PDF-readers see it. I decided to try 151 and have been able to make it text-searchable and compress it a bit (to 196 MB), with very little loss of quality. The two operations aren't linked, so keeping best available quality, not compressed further, would be an option. If the files in their text-searchable form are put back on Tekwiki locked, a user wouldn't be able to compress. And I would be willing to process them. Raymond is looking for ways to make large PDF's more usable to us. IfThe example of 151 makes it likely that I would be successful at the others as well. To do that, I would need to know if the intended operation is allowed, IOW is not prohibited by someone's copyright (claim) or (Tek's) release permission. Dave Brown suggested such prohibition. Perhaps after processing, I could submit the files to Dave (Tek Museum) for them to lock and upload. And Dave Brown? Raymond |
TM500 series plastic face plate repair; solvents?
In the world of modeling and slot cars (and any other place styrene is used) there is a repair technique whereby one uses liquid styrene modeling solvent to "melt" scrap plastic into a goo that can be applied as a filler to holes, missing sections, or even build up new features that didn't previously exist.? Once the solvent outgasses, the new plastic can be worked the same as the original.? Slot car folks do this all the time, because kids were encouraged to "modify" their cars, leaving endless cut wheel wells in their wake.? A really good repair is almost invisible, but usually given away by dirt/slight discoloring in the repair.
Last night I was wondering if the same technique could be applied to all those broken corners on the TM500 plugins.? The plastic isn't styrene, and while the styrene solvent did dissolve the plastic, it became chalky and unusable, even the next morning.?? Has anyone experimented with this before?? Know what type of plastic was used?? It wouldn't be good for large missing sections (too much time involved), but smaller missing sections could easily be repaired if the right solvent was found. -Dave |
Re: Strange Tek2440 issue
Acquisition stopped, doesn't happen. I can go all the way to 2ns/div.
toggle quoted message
Show quoted text
Normal trigger but trigger set to +or- 90v (not triggering), it happens. Single, doesn't happen, can go all the way to 2ns/div, can even trigger it (in single) and it doesn't happen. The probe was hooked to the calibrator for the triggers. Normally doesn't matter if there's a signal there or not. Thanks! Vince. On 08/31/2020 01:43 PM, Siggi wrote:
Hey Vince, --
K8ZW |
Re: 1S1 sampling unit; (also 7B53AN & 7A18 plugins)
Hi,
toggle quoted message
Show quoted text
I have a copy of the IO-14 scope schematic as 4 gif files. I could email it to you if you are interested, but I don't think it would be easy (or useful) to try to connect it to a 1S1. You really need a 500-series Tek scope to use it. Others may have different ideas. Colin. -----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of dc888@... Sent: 31 August 2020 06:04 To: [email protected] Subject: [TekScopes] 1S1 sampling unit; (also 7B53AN & 7A18 plugins) Hi all, just acquired the above (but missed out on the actual TEK scope), all at a local recycling centre. Would like to buy a TEK scope, working or not, in which I can use the 1S1. Is anyone prepared to dispose of a suitable TEK scope? Plus carriage to Oz at cost. Alternatively there is a Heathkit IO-14 (8MHz, valves/tubes, triggers to 2.5 MHz) on its way to me. Maybe I could use the 1S1 in this, with suitable connections? I have the IO-14 schematic, but Heathkit Manuals have now been commercialised. Any advice appreciated. Thanks & Regards David C. |
Re: Strange Tek2440 issue
Hey Vince,
toggle quoted message
Show quoted text
I have no idea what's going on there, though I'd suggest looking at the power supplies as a first step. You can also - perhaps - diagnose this a little ways by playing with trigger and acquisition settings. My guess would be that a power supply is collapsing when the sample clock exceeds some particular rate, though that's of course just a guess. Does this still occur when the acquisition is stopped? If not, you can maybe narrow this down by e.g. going to NORM sweep/trigger mode, and setting up a trigger that never hits. Does it still happen when acquisition is started but no trigger fires? How about under a single-shot acquisition in AUTO mode, etc. Siggi On Sun, Aug 30, 2020 at 5:37 PM Vince Vielhaber <vev@...> wrote:
|
Re: Reverse part lockups
Dallas Smith
Sorry all for any confusion, was just looking for someway to reverse part lookup. I have a large number of Tek PC Boards to use for parts but take a lot of time to find an available part from. Was not aware of the TekWiki RPR's file supplied by the Tek Museum, couldn't find them from the home page index. In my opinion the RPR files are somewhat useless without the ability to search them. I think making them full OCR is different than making them searchable. I send Raymond a unlocked version on the 155 RPR for him to checkout to see if it could be compressed more. Are there more RPR files for other parts available somewhere, such as the 156 parts?
toggle quoted message
Show quoted text
With permission I would be willing to unlock all the PRP files for Raymond to process. Raymond is looking for ways to make large PDF's more usable to us. If Raymond can not make them searchable I would be willing to spend the four or more hours to do so if legal, but will still be unlocked. Any thoughts Dennis? On Thu, Aug 27, 2020 at 09:21 PM, Dallas Smith wrote:
|
1S1 sampling unit; (also 7B53AN & 7A18 plugins)
Hi all,
just acquired the above (but missed out on the actual TEK scope), all at a local recycling centre. Would like to buy a TEK scope, working or not, in which I can use the 1S1. Is anyone prepared to dispose of a suitable TEK scope? Plus carriage to Oz at cost. Alternatively there is a Heathkit IO-14 (8MHz, valves/tubes, triggers to 2.5 MHz) on its way to me. Maybe I could use the 1S1 in this, with suitable connections? I have the IO-14 schematic, but Heathkit Manuals have now been commercialised. Any advice appreciated. Thanks & Regards David C. |
Re: 4-pin Lemo connector used on end of cable for 4041 PD keyboard
On Mon, Aug 31, 2020 at 06:04 PM, Gary Robert Bosworth wrote:
Hi Gary, I checked. It is indeed the connector that I thought. It is the same Tek used for the power supply connection of probes like the P6201 and P6202A. You'll find the same sockets on 'scopes like the 7104, 7854, 7904(A). It is Tek part no. 131-0778-00. The 1101 PSU has four sockets for them. I checked and double-checked the Lemo part no: FFA.0S.304.CLAC42. Digikey (USA) currently has 5 in stock @ a lot of money, something like USD 30. I have only a few left (as sets with sockets, actually, have a few more sockets and counter-plugs) and sending from the Netherlands may be costly, maybe about (half) as much as the Digikey price. That would leave little left for me if I'd want to "compete". Have a look and if you still want me to send one (Digikey total cost, order cost), let me know your ZIP code, I'll check postal costs and we can go from there. Raymond |
Re: Strange Tek2440 issue
Repet on, Repet off, no difference.
toggle quoted message
Show quoted text
Vince. On 08/31/2020 01:42 AM, Roger Evans via groups.io wrote:
100nsec per div and faster sounds like the switch from real time sampling to equivalent time interpolation (300Msample/sec real time). You can turn off the equivalent time mode by going to the Acquire menu and setting 'Repet' to Off. If that enables you to get to faster sweep speeds then the problem is clearly in the circuitry that does the equivalent time interpolation but I would need a thorough read of the manual to go into any detail! --
K8ZW |
Re: TDS3054 Replacement of DALLAS DS1742W (U640), data content in NV-RAM
some of the chinese fakes do actually work and identical to the original but since their date code is also old they ma not last that long
I also got several fakes that didnt work until I got this one that worked as yours does. The fakes that didnt work are visibly larger in size than the original one I also got a fake one that worked perfectly fine as DS1742 but not as a DS1742W ! but I am surprised that why your clock is not working. The chip that I got worked. But as I said the date code is already 12 years old so it may not last assuming that it is a real date code! as far as I know the hours are stored at those two places. when I changed both to a new number the scope showed the correct # of hours |
Re: TDS3054 Replacement of DALLAS DS1742W (U640), data content in NV-RAM
Hello all,
thank you very much for your replies. Here is my list of actions (so far): 1. The 'chinese part' worked ok in a programmer (readable, writeable, reasonable content in the registers 0x7F8...0x7FF), also see my question A. below 2. I desoldered the old DS1742W from the TDS3054 mainboard and replaced it with socket strips 3. I programmed the chinese DS1742W with all 0xFF in the address range 0x000...0x7F7 (not touching the registers). 4. I put the chinese DS1742W into the socket and powered the scope up for a couple of minutes and to have a look into the system menus. - date/time appeared as '6 Apr 22 2:40:80' (not counting up) - Cal menu text: 'The last factory calibration occurred 0 (operating) hours ago on 19-Dec-2001 - Cal menu text: 'The last SPC occured 0 (operating) hours ago on 8-Nov-2006' - System Errors: 15, Display Errors: 15 (so no errors on other subunits) - Total operating time: 0 hours - Number of Power Ups: 206 Also see my question B. below 5. After shutting down the scope I read out the chinese DS1742W on the programmer to see which entries had been changed from 0xFF to something else. Following is the RAM content by address: 0x000 ... 0x297, mixture of random-looking values 0x298 ... 0x61F, all bytes 0x00 0x620 ... 0x63F, all bytes 0xFF 0x640 ... 0x6DB, mostly 0x00 but also non-zero values in between 0x6DC ... 0x7DF, all bytes 0xFF 0x7E0...0x7E3, 4 bytes: 0x00 0x00 0x00 0x04 0x7E4 ... 0x7E7, 4 bytes: 0x00 0x00 0x00 0x04 0x7E8...0x7EB, 4 bytes: 0x00 0x00 0x00 0x04 0x7EC ... 0x7EF, 4 bytes: 0x00 0x00 0x00 0x00 0x7F0 ... 0x7F7, all bytes 0xFF 0x7F8...0x7FF, register map as described in datasheet. Also see my question C. below Questions and remarks: A. What 'special characteristics' can be found with a chinese fake chip? Is it an empty housing containing nothing but only rows of pins sticking out? My chinese version had the exact physical appearance of the housing and the same look of the potting from the bottom side. Weight: Chip from TDS3054: 7,21g Chinese DS1742W: 7,38g B. The clock cannot be running because the /OSC-bit register 0x7F9 bit7 is set to '1'. The scope obviously does not reprogram this bit. The number of power ups seems to be stored somewhere else but not in the DS1742W. The value of '206' is ok because it was at '203' when the scope operated correctly the last time before the DS1742W failed. C. The scope seems to write nearly into the whole memory, the content at 0x000 ... 0x297 seems to be 'mystery' (or perhaps random garbage). The value (hex) 00 00 00 04 appears 3 times. Could it be that the total operating time is stored 3 times (and not twice as written by AMIRB)? The value 4 is reasonable because the scope was switched on for about 4 minutes. Any remarks on the meaning of the other values? Thanks again! Best regards -Roland |
Re: Reverse part lockups
On Sun, Aug 30, 2020 at 04:48 PM, Raymond Domp Frank wrote:
I'm giving this another try in this thread. This message is directed at Dave Brown (Tek Museum): 1. Do you consider these files free to unlock for the purpose of OCR'ing, possibly reducing in size, and re-uploading to TekWiki or not, either because of a copyright claim or a necessary/missing release by Tektronix not allowing this use? 2. If you consider them not free for the above purpose, will you propose a way forward to make them available? Raymond |
Re: 4-pin Lemo connector used on end of cable for 4041 PD keyboard
Raymond: Thank you for all the trouble you are going thru. If the
connector is the correct one (it is only about 5/16" diameter with a "D"-shaped 4-pin insulator inside), name your price and I will pay through any means you desire. I will contact you in a day or so to remind you of my needs. Bye for now. Gary On Sat, Aug 29, 2020 at 3:54 PM Raymond Domp Frank <hewpatek@...> wrote: On Sun, Aug 30, 2020 at 12:38 AM, Gary Robert Bosworth wrote:--4-pin Gary Robert Bosworth grbosworth@... Tel: 310-317-2247 |
Re: 5441 rack / bench conversion document
From Tim P (UK)
toggle quoted message
Show quoted text
Many thanks for that - as I suspected, it is necessary to unsolder the delay line and re-attach to the interface board. :-( Otherwise a topological conundrum involving a fourth (spacial) dimension !!! Tim On Mon, 31 Aug 2020 at 16:04, zenith5106 <hahi@...> wrote:
On Sun, Aug 30, 2020 at 08:29 PM, Tim Phillips wrote:between rack / bench formats ? |
Re: Tektronix TDS3PRT Printer
Since a previous message to this group I have been seeking an explanation for the scarcity of of available information on tthe TDS3PRT printer and the reason for its discontinuance. I have been hoping to find the manual or instructions furnished with this printer and any software that might have been furnished with it. What I have learned so far is that according to a Wikipedia article:
"Reports began surfacing in the 2000s finding the oestrogen-related chemical bispheonol A ("BPA") [is] mixed in with thermal (and some other) papers. While the health concerns are very uncertain, various health and science oriented political pressure organizations such as the Environmental Working Group have pressed fro these versions to be pulled from market." This may be the explanation. The TDS3PRT printers bear Sony/Tektronix labels with J prefixes. A source confirmed they were manufactured by Sony. Sony has manufactured several high-quality thermal printer models for medical applications such as ultrasound scans. Another source indicated the TDS3PRT used 110 mm paper rolls. Sony also employed this width paper rolls in their printers and offered at least three grades from normal to highly glossy, UPP-110S, UPP-110HD, and UPP-110HG. These remain available from eBay and Amazon sources and may be compatible with the Tektronix printer. I have not found any source for the Tektronix 016-1907-00 paper (5 rolls). Sony, being the manufacturer, may have nixed public release of the manuals. One former seller mentioned the manual was available on apexwaves.com, but the file has apparently been pulled down. Are any of these printers to be found in the hands of members of this group? Can anyone provide further information, particularly a source for the manual? Bruce, KG6OJI |
Re: 5441 rack / bench conversion document
On Sun, Aug 30, 2020 at 08:29 PM, Tim Phillips wrote:
They are here now: They are scanned from microfiche using a photo scanner so unfortunately the quality doesn't get any better than that, sorry. /H?kan |
Re: TDS3054 Replacement of DALLAS DS1742W (U640), data content in NV-RAM
A DS1742W from China? Chances are really high it's a counterfeit. Here's an eevblog thread that has some pictures of a China fake vs real:
The dimensions are a giveaway, at least on that one. Further in that thread, there's a link to someone who made an adapter board for a DS1744WP, which is still available from reputable sources (Mouser, Digikey, etc). Functionally, it's the same as the DS1742W, but with more SRAM that would never be accessed. The adapter board fits TDS3000 scopes. I ended up doing surgery on my old DS1742W to connect an external battery. I'd recommend not throwing away your old DS1742W, and install a DIP socket. -mark |
Re: Tek 7854 intermittent issues
On Mon, Aug 31, 2020 at 01:12 AM, Nick Corvid wrote:
And in response to Albert, you are correct that it's missingHi Nick, the dot pattern is not what I expected. It seems that still linear interpolation is done to fill missing horizontal positions. I expected to see no dots at all during some wide horizontal gaps. Anyway, in my opinion it's clear that data are missing. In your new photos there is also a negative outlier and, very strange, a negative trace near the end of the sweep which looks like a inverted portion of the correct waveform. In the correct waveform the rising edges end more or less rounded while the falling edges are sharp. Wrong sign bit? I thought I would see horizontal gaps when I do a too fast single sweep acquisition. But my 7854 simply refuses AQS with a nasty beep. Albert |
Re: Strange Tek2440 issue
100nsec per div and faster sounds like the switch from real time sampling to equivalent time interpolation (300Msample/sec real time). You can turn off the equivalent time mode by going to the Acquire menu and setting 'Repet' to Off. If that enables you to get to faster sweep speeds then the problem is clearly in the circuitry that does the equivalent time interpolation but I would need a thorough read of the manual to go into any detail!
Regards, Roger |