Re: 2445A calibration
Thank you, Chuck,
I haven't touched the capacitors yet. I've tried all 3 AWGs and got the same result. CAL01 had to run again this time with 7834a, since the steps with the extenal DSO scope were not aligned correctly. Now it's fine, but still LIMIT error. I'll try using the high resolution HDO6034 scope at work to check the AWG output before connecting to 2445a in CAL02. But I can assume that it will be fine since all the equipement at work is fresh out the calibration lab. |
Re: 2445A calibration
To be more concrete I've thrown to IDA the 2465a ROM files. 160-3302-09 & 160-3303-09 are the part number downloaded from here:
The code structure is quite interesting. The bank switching routines are in the first 2K RAM working area. I've quickly found the voltage reference setting routine: /g/TekScopes/photo/74577/22?p=Name,,,20,1,20,0 See the comments in the code. The 0xAC=172d code thrown in the DAC sets the reference voltage of 1.36V: 1ma*(4095-172)*4/4096=3831uA. Solving for x=TP2421 voltage: x*14.2KOhm/13.0KOhm + 681Ohm*(4095-172)*0.004A/4096 = -x x=(4095-172)*0.004A/4096*681Ohm/(1Ohm+14200Ohm/13000Ohm) = -1.247V The above assumes that the DAC is calibrated for the 1mA reference current. Here are some other examples of the DAC and the counter access: /g/TekScopes/photo/74577/21?p=Name,,,20,1,20,0 /g/TekScopes/photo/74577/20?p=Name,,,20,1,20,0 The code is complex in structure due to the ugly banking implementation. There are similar sections in all four 32K ROM pages. DAC is accessed all over the place in byte fashion and also by direct 16 bit X or even S register store. The DAC work area in the first 2K of RAM is also used heavily and many locations are duplicated. If completely stuck, then I'll try tracing the LIMIT error, but looking just at the ADC routine complexity (uses almost 200 bytes of various runtime data to do the conditional branching) likely it will consume all the time for several months... |
Re: 2445A calibration
Chuck Harris
Hi Maxim,
toggle quoted message
Show quoted text
The normalizer is just a resistor and a trimmer capacitor in parallel, but in series with the center conductors of a male and a female BNC connector. It is built in a little Pomona box. You can do the same thing with a scope probe, using the probe's trimmer capacitor to square up the waveform on one channel, and the C205 to square it up on the other. I doubt it needs adjustment, assuming that you haven't moved it already. That CAL03 throws a LIMIT on CH3 and 4 makes me even more certain that your problem is your generator. -Chuck Harris maxim.vlasov@... wrote: Chuck, |
Re: 7000-series power-hungry plug-ins
Chuck Harris
If you have ever used a 7805, or LM309, LM319, or LM340, ... you
toggle quoted message
Show quoted text
have already used current foldback. -Chuck Harris tinkera123 wrote: Okay, understand that now. I have played with/built many linear supplies, but none with much protection circuitry. |
Re: 2445A calibration
Chuck Harris
PHK is an open source legend. His disassembler is available
for free on github. PHK wrote the NTP routines for openbsd. -Chuck Harris Harvey White wrote: On Sat, 13 Oct 2018 16:05:14 -0400, you wrote:... PHK, Poul-Henning Kamp wrote a nice one that actually analyzes the codeHmmm, some people want money for those things, not that I begrudge him |
Re: 2445A calibration
Chuck Harris
There are many disassemblers that actually emulate the target
toggle quoted message
Show quoted text
code to determine the exact way it operates... all in software. CAL01 doesn't have much to do with CAL02. Which attenuator settings give you the RANGE error when running CAL02? -Chuck Harris maxim.vlasov@... wrote: Chuck, |
Re: 7000-series power-hungry plug-ins
Okay, understand that now. I have played with/built many linear supplies, but none with much protection circuitry.
toggle quoted message
Show quoted text
Thank you, Chuck. Cheers, Ian In regulated supplies, there are often over current protection |
Re: 2445A calibration
On Sat, 13 Oct 2018 16:05:14 -0400, you wrote:
Harvey White wrote:<snip> I probably would have liked the ones for the 8085 and the 6502, justPossibly, most of the older processor's pods have been long since scrapped,The largest problem is that the disassembler probes for given microprocessorsNow for the 6800 and such, you could find the most up-to-date probe, because. Every once and while, I need to look at some of those old ones. Right now it's the 6802 that I need to look at. I found a few, they want DOS.A couple, actually. Sourceforge, and github are littered with them,Hmmm, got one for the 6800 series? I have core images... I have one that I wrote that runs under windows and is gui driven. Hmmm, some people want money for those things, not that I begrudge him the income, but I tend to go to open source. That's what mine does, although I get tired of writing tools to make tools.... ...The one in the DM5010 isn't, I think, and actually has internal RAM.IIRC, the 6802 did more integration of the RAM into the processor. ITek tended to use the 68B02, which is a 6802 without the internal RAM. For analysis cases, all it does (of course) is to change the location of some of the RAM. Harvey
|
Re: 2445A calibration
Chuck Harris
Harvey White wrote:
Indeed, but I find that the LA disassemblers are problematic. It is difficultThe state of LA disassembler probes never reached a point where usefulThere are some situations where the LA can be of use, but mostly for to tell with a piplined processor whether the memory reads you are processing are reading data, or reading instructions. Without a probe specific to the processor that can make those determinations, you are just looking at data translated to assembler mnemonics... not useful. The 8086 was the first commonly used microprocessor to suffer this problem... and solution. Intel included extra pads that brought out pipeline tracking information for emulators and logic analyzers. Tracking specific events under specific conditions is where a LA shines. Possibly, most of the older processor's pods have been long since scrapped,The largest problem is that the disassembler probes for given microprocessorsNow for the 6800 and such, you could find the most up-to-date probe, but some are certain to still exist. I just ditched some 6502, Z80, and 8085 emulators... They probably are still available from the recycler if anyone is interested. A couple, actually. Sourceforge, and github are littered with them,Hmmm, got one for the 6800 series? I have core images... particularly the 6809, which is an upfeatured 6802 used in the automotive industry. PHK, Poul-Henning Kamp wrote a nice one that actually analyzes the code by executing it. He used it to disassemble the code in some HP counters. ... IIRC, the 6802 did more integration of the RAM into the processor. ITek tended to use the 68B02, which is a 6802 without the internal RAM. -Chuck Harris |
PG 506 IS IT NECESSARY FOR CALIBRATING?
toggle quoted message
Show quoted text
---------- Original Message ---------- |
Re: 2445A calibration
On Sat, Oct 13, 2018 at 08:40 AM, <maxim.vlasov@...> wrote:
... Actually, the 167xx doesn't always need a special adapter. There's a good thread on HP/Agilent disassemblers here: In post #3 in that thread there's an example of using a 16702B logic analyzer to probe a 6802 processor using flying lead probes directly to the processor pins (no adapter). The thing being probed was a Tek 2465. On the 2445A/2465A and 2465B/2465B, Tek uses bank-switched memory, making it even more fun to untangle the code. The 2465 does not have bank switching. (Maybe IDA handles bank switching?) If someone wants to head down the disassembly path, here's some info on how the banks are handled, specifically for a 2465A, but I think the B series is the same: I've done some disassembly of the code along with real-time tracing to figure out how some things work. It's time consuming and some of the code seems a bit hack-y as it was ported from one platform to the next. -mark |
Re: 468 horizontal jitter
On Sat, Oct 13, 2018 at 09:51 AM, Jeff Urban wrote:
Thanks Jeff. There is jitter on the A gate out with no input just a free running trace. I had to put my 465 in 10x to see it but it does match the jitter on the 468. So... This looks like its going to be a fun one to figure out. |
Re: vintageTEK museum releases Replaceable Parts Registry (RPR)
On Sat, Oct 13, 2018 at 09:05 AM, walter shawlee wrote:
I notice it doesn't seem to tie to the general main page at all.Yes I was wondering about thaat too ! Such a nice resource... that's not even visible/accessible, kinda defeats the purpose of making it public ^^ I guess the link to that RPR page might fit at the bottom of the main page, under the section labelled " Manuals, Catalogs and Other Publications " ?! Thanks for all the scanned sections so far, and for all those to come :-) Vincent Trouilliez |
Re: 468 horizontal jitter
In case you haven't, it is important to check if it is a centering problem or a triggering problem. Even in auto or free run it is still triggered, so if it is the problem could be all the way back at the ramp generator.
If the trace itself is moving and not the sweep (I hope you got that...) then I would more look for hash on a PS line somewhere. Of course that applies to the ramp generator as well but different PS legs of course. |
Re: 2445A calibration
On Sat, 13 Oct 2018 01:47:20 -0400, you wrote:
Or, "Square Wave".To be a bit pedantic, to me, square wave doesn't talk about the levels, so where's 0 and where's Vmax...? bet there's lots of opinions on that one. Harvey
|
Re: 2445A calibration
On Sat, 13 Oct 2018 09:55:50 -0400, you wrote:
I have been designing and debuging embedded systems for more thanThere are some situations where the LA can be of use, but mostly for tracking specific events under specific conditions, IMHO. Now for the 6800 and such, you could find the most up-to-date probe, so that becomes less of a factor for that generation of processors, I'd think. Hmmm, got one for the 6800 series? I have core images... IIRC, the 6802 did more integration of the RAM into the processor. I think that the DM5010 uses one as well. Harvey
|
Re: vintageTEK museum releases Replaceable Parts Registry (RPR)
walter shawlee
Thanks so much for another great RPR scan, the 311 set is very handy to have!
I can hardly wait for that 366 scan, it will answer a lot of questions. the page link is here: I notice it doesn't seem to tie to the general main page at all. thank you for all the hard work! -walter -- Walter Shawlee 2 Sphere Research Corp. 3394 Sunnyside Rd. West Kelowna, BC, V1Z 2V4 CANADA Phone: +1 (250-769-1834 -:- We're all in one boat, no matter how it looks to you. (WS2) All you need is love. (John Lennon) But, that doesn't mean other things don't come in handy. (WS2) |