Re: 7633 fan possible to repair?
John Griessen
On 03/16/2018 09:01 AM, Fabio Trevisan wrote:
Well, not only it has a fan, but it was, like your very sticky.This matches the way old Husqvarna sewing machines get. Cleaning with solvent such as kerosene, then applying Triflow lube is best for a good repair that lasts. Just adding Triflow can be enough in some cases. |
Re: 7633 fan possible to repair?
Hello Lop,
Apparently we will walk hand in hand with our two recently acquired 76xx scopes (mine is a 7623A). When I opened up mine, the very slow fan was the first thing I realized. Before I opened it, I was just so happy that I thought the 7623A didn't even have a fan (oh man, that 464 fan sound is annoying!). Well, not only it has a fan, but it was, like your very sticky. Upon initial application of WD40, it didn't come loose immediately, as it's common when we use WD40, so powerful WD40 is at loosening these things. But it seems that the oil / grease residue of this particular fan gets really, really sticky... almost like tar. I`m sorry that I didn't take pictures of the process, but here's what I did: 1. Firstly, remove the fan from the motor's shaft. It comes off easily, but don't risk pulling it out by the blades. Insert a flat screwdriver between the motor frame and the fan and use it as lever, so to push the fan out. 2. On the two sides of the motor, there are two small aluminum baskets that are help in place by two bent tabs. they hold inside them two generous tabs of felt, that are there to store the lubricant that keeps the sintered (porous) brass wet. Unbend the tabs so to make the aluminum baskets loose. They wont come out because they're on the inner side, between the fan's frame and the rotor, but making them loose will allow you to apply WD40 generously inside them. This will soften the old oil, as well as wet the lint tabs. 3. Then, apply WD40 the closest as possible to the joint between the shaft and the bearing, so to create some pressure for the WD40 to enter the gap. I used my fingers to hold around the WD40 needle, the shaft, and the bearing plate, so to allow only space for the WD40 to go in but little space for it to come out, in an attempt to increase the pressure into the gap. I don't know if it really helps, but it gives you an idea of how desperate I was, because even WD40 was having troubles to loose this thing. 4. At last, to help accelerate the process, I attached the shaft to my drill machine, and applied some continuous rotation while I kept pouring in WD40. After a couple of minutes, the shaft was turning completely loose. As a last measure, knowing that WD40 oil is quite volatile and don't last too long (to remain as a more permanent lubricant), I applied fine machine oil (sewing machine oil) to the felt tabs, sparingly, so to soak them completely. I let the excess drip on paper towel, then I put the aluminum baskets back in place, holding them firmly from the back and re-bending the tabs. It's important to hold them tight in place WHILE you are bending the tabs, so that they hold firmly into place, otherwise they will rattle and make noise. Wish you luck. Rgrds, Fabio |
Re: TEK 212
Hi Lee,
toggle quoted message
Show quoted text
You are in luck. There is no reason to waste your time making a battery pack. This company makes a perfect duplicate of the original battery pack and they do not charge much for it. Anyone with one of these scopes should write their contact information down. I have used them for years. They are excellent. You wouldn't know their replacement from the original except their replacement batteries hold a greater charge than the original Tek batteries. This info is in the archives somewhere. But here is the information on the battery replacement company again: Dennis Tillman W7PF -----Original Message-----
From: Dennis Tillman [mailto:dennis@...] Sent: Sunday, March 15, 2015 12:20 PM To: 'radiogeo@...' <radiogeo@...>; 'goodappl@...' <goodappl@...> Subject: Tek 212 and 214 Battery Pack This is the link to the company that makes up perfect battery pack replacements for the tek 212 and 214 scopes. They are so inexpensive and so perfect it isn't worth it to try to make your own. The only unusual thing about them is you have to ask for a quote and wait for a reply since they only make these when they get an order for them. They don't stock them. Use the "contact us" link at the bottom of the page to send them an email request. Be sure to specify that you want product #5149. From: KeepIt SimpleStupid [mailto:keepitsimplestupid@...] Sent: Tuesday, March 05, 2013 1:36 PM To: dennis@... Subject: Re: Tek 214 battery Pack Here is a link: It's kinda tough getting there. It's under Medical and Scientific batteries Scott -----Original Message----- From: [email protected] [mailto:[email protected]] On Behalf Of Leslie Austin Sent: Wednesday, March 14, 2018 8:03 AM To: [email protected] Subject: [TekScopes] TEK 212 Hi guys. I am new to this forum, and am starting off with a wanted ad, I hope nobody minds. I have a 212 'scope which I bought nearly 20 years ago. The batteries were quite dead, so I removed them, and the charging circuit, and wired in a socket for a 12v external battery. However, I have now obtained enough cells of suitable type for new battery packs and want to reinstate them as before. When I removed the original packs, I turned two insulating posts as dummies to maintain the scope's physical integrity. Now I will NOT have thrown away the empty packs and the short cable interconnects, but repeated searches have failed to locate them. Does anybody have an old 212 battery pack, with or without contents and connecting leads I can get my hands on? Les. -- Dennis Tillman W7PF TekScopes Moderator |
7633 fan possible to repair?
The 7633 I just got running had a stiff fan that needed poking to get started. It was replaced with a fan from my other now parts scope. The the scope and fan are so clean it almost looks like the unit was hardly used. My theory is the fan just stiffened up from years of sitting. Is there anyway to repair this thing? I would like to have a backup. I wanted to wait and ask before I started tearing into it. /g/TekScopes/photo/37195/3?p=Name,,,20,1,0,0
|
Re: Tek 317 odd trace behavior after warm-up
No specific experience with a 317, but did you replace electrolytics as
well? Its old enough to give it consideration. Id also check power supplies (voltage and ripple) while its ¡°bad¡± Bad caps would be my first thought for something of that age But beyond that I cant make much more in the way of useful suggestions David On Thu, Mar 15, 2018 at 12:02 PM CZ4A via Groups.Io <artofruin1285= [email protected]> wrote: Hello everyone,Sent from Mobile Annoyance thingy, please pardon any typos |
Re: Another New Guy
Hi Benjamin, So you want to relearn electronics but don't want to fix the 475 to learn ? That is confusing.
1) If you go through the 475 S/M from page-1 to last page , you will find the problem and fix the unit. Since this old scope is new to you, this will give you a record of readings you take and get you familiar with it and teach you too. A DMM will find MOST problems in a scope. If you have one scope you will eventually need a second to fix the other. 2) Get a copy of the 1989 Tek manual; TROUBLESHOOTING YOUR OSCILLOSCOPE ; GETTING DOWN TO BASICS , # 068-0315-00. Still ;the best out there. I got mine from K04bb.com , though a few websites have it. It says do NOT remove the scope cover until you have isolated the fault to a section. Also, the front panel is the best tool you have for troubleshooting... The 88 page manual is a STUDY-manual, not something to flip-through. When you use it correctly , it will find the problem in your scope ! But , this haphazard , go here , go there , try this , try that , BS is the Wrong Method. 3) Spend an hour or so , trying every knob / switch / dial and documenting that it is working......and working correctly. Make a list or photocopy the User Guide for your notes. Let's say, ......95% of problems a scope has will be demonstrated by the front panel settings and you can open the schematic and isolate to the part or ckt , and you have not even taken the cover off yet ! ! ! Drives me crazy to read , ' I have no Hor Trace" only to find out later ' oh yeh, the trigger level was off and I never bothered to select Auto-Mode " ! |
Re: 485 main vertical gain
Hey Lop Pol,
I don't have the manual of the 485 (besides the page you posted), but it seems you guessed it, apparently it's calling for measuring the voltage between the 2 points... This is what's called a differential voltage measurement (as opposed to absolute measurement, which is taken in relation to ground... or in other words, while you hold the ground lead at the ground level = 0V). Rgrds, Fabio P.S. In time, careful about taking differential voltage measurements if you're using a BENCH voltmeter (a voltmeter that is powered by mains, and not by battery). A bench voltmeter usually have its ground lead internally connected to the ground (all the way through its power cord to the building's ground, or earth). Those voltmeters are not suitable to perform differential voltage measurements in other equipment that are - just as well - grounded. You would end up shorting to ground the point of the circuit where you connect the ground (black) lead. |
Lot of 75 Tek transistors for sale in Brazil - No afiliation
Hello Group,
I was browsing the local auction site from Brazil (not evil bay... but just as evil) and just saw this item: There are 75 transistors in the lot, all listed in the description. It's selling for BRL 50,00 which is roughly USD 15,00 Maybe someone is looking for some hard to find transistor? In case you have interest, I can liaise if you need. Rgrds, Fabio |
Re: Another New Guy
This is the spot to get some help.
toggle quoted message
Show quoted text
After reading threads for a few years, my thoughts on how to help people help you: Get the service manual(Check, good start) Get a good picture of the scope, there are a "lot"!!! of knob position combinations that have to be considered. For example, It does not sound like it, but a check to be sure your not in horizontal external would be good. If there are not signs of mechanical damage, cycling the switches on a scope that has been sitting can help, flipping it in and out of horizontal external a few times to see if you have some switch dirt that will loosen up. A dirty Beam find button has a history of fooling people. Another thing is to give us an idea of what other tools you have to work with. Any scope history? Serial number? Some idea of your general location. Then it's to the power supply, do some reading first From an educational perspective "getting a crappy digital toy" depending on what it is, to look into the scope might have some merit. Take some time and don't be in a hurry John -----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Benjamin A Sent: Monday, March 12, 2018 3:58 PM To: [email protected] Subject: [TekScopes] Another New Guy I'm very new at this, so please be gentle and use small words! I haven't done real electronics since college, but recently I've been bitten by the bug and wanted to relearn all the things I've forgotten both for myself, and for work (I teach programming and coach robotics!). Since I read that a good, old analog scope was preferable to a super cheap digital, I found a 475 on Craigslist. Bringing it home, I started learning how it worked and was starting to get really into it... When the horizontal sweep died. I was fiddling with the delay settings (super cool!) and the trace disappeared, instead lighting up something like the whole lower half erratically. I powered it off, and let it sit for a minute before powering it on, and it worked briefly before the sweep quit entirely leaving me with just a dot in the middle of the screen. I still get vertical movement when I connect the probe to the calibrator, and the dot can be moved with both the vertical and horizontal adjustment. There's also an issue with channel one-- the voltage display seems way too low. So I've read some of the service manual and some of the threads on here and people seem to imply that the low voltage power supply is the first thing to check, followed by the horizontal amp, but... My real question is: should I even be trying this right now? Should I stash this scope away, get a crappy digital toy to learn from and tackle this in the future? That would be a bummer, but I don't want to either make things worse or hurt myself (although I know enough to stay away from the mains/high voltage sections). Is there a way to find a local wiz to help learn how to troubleshoot it? I'd love to learn the ropes, but I don't want to get in over my head without some kind of lifeline! |
Re: femtosecond fusion
Craig Sawyers
60fs is a short pulse for sure! Shortest pulse thing I designed (about 15 years ago) was frequency quadrupling 250fs pulses with 30% efficiency. It was not straightforward. 250ps is a pulse of light only 75um long. 60fs is 15um long - far thinner than a human hair! The trickiest thing is measuring and verifying the pulse length, because there is no detector/electronics that works remotely that fast. Craig |
Re: Tektronix fail
If you are a member of the group, you should see a photos button on the left side when you access the group with the web interface. You can create an album and put your pictures there. Then copy the URL of the picture and send it in a message so we can see it. Please note that we need the name you give the picture so it can be found.
toggle quoted message
Show quoted text
Do you have a copy of the service manual for your scope? The 2465B manual should have the same power supply as your 2445B. If not, look around on the tekwiki site. -----Original Message-----
From: Qualit¨¦ SAV <iphonequalite@...> To: TekScopes <[email protected]> Sent: Thu, Mar 15, 2018 12:02 pm Subject: [TekScopes] Tektronix fail Hello Thank you for welcoming me to your home. I am French and I have a breakdown on my tektronix 2445B. A capacitor of the smoke supply as well as a resistor. For the capacitor no problem but for the resistance m, impossible to read the value !! Do you have an idea ???? I am new on this group and I can not put a picture to show you the capacitor and resistance incriminated !!! The capacitor at the input of the power supply is 0.068microFarad X2 and 250v but the resistance ???? |
Re: Back feet for 335 scope
If you have an original foot, I can make a mold and pour plastic feet. I've already done it for some older Yaesu and Kenwood rigs.
toggle quoted message
Show quoted text
Vince. On 03/15/2018 12:39 PM, Bert Haskins wrote:
--
Michigan VHF Corp. |
Re: Back feet for 335 scope
Jeff Davis
Hi Bert,
That would be great. Even though the foot has the same part number on the 314, it would be good to confirm correct fit on that unit. Let's wait until I get results from Lee - I actually sent him several variants to try to get the best fit. Once that's determined, I can send you a sample of that size. PM me your address, please. Jeff ________________________________ From: [email protected] <[email protected]> on behalf of Bert Haskins <bhaskins@...> Sent: Thursday, March 15, 2018 9:39 AM To: [email protected] Subject: Re: [TekScopes] Back feet for 335 scope On 3/15/2018 11:07 AM, Jeff Davis wrote: Hello all,If you would like another tester, I would be happy to volunteer. I have a 314 and a 335. Thanks, Bert |
Re: Back feet for 335 scope
On 3/15/2018 11:07 AM, Jeff Davis wrote:
Hello all,If you would like another tester, I would be happy to volunteer. I have a 314 and a 335. Thanks, Bert |