Re: Can I install a 2465B motherboard in a 2445B chassis? What will I get?
Chuck Harris
The ROM's on the A5 board contain everything they need to
toggle quoted message
Show quoted text
know about being a 2467B, 2465B, 2445B, or 2455B. The A5 controller's are interchangeably the same. You can take one out of any member of the 24[4,5,6]5A or B family, and swap around. The only significant difference has to do with the separate Display board on A's, and early B's. You can swap a controller/display board from an A into a late B, and vice versa, if you want... [The "A" controllers have some ROM omissions for the extra diddly stuff that the "B" can do, and some changes in the calibration routines, battery voltage sensing...] The way the A5 controller knows what scope you have is by reading the P/J512, pin 11, and pin 32. Pin 11 goes with a jumper, W120, which if missing is a 2445B/2455B, if grounded, is a 2465B. P32 determines whether the scope is a 2465B, or a 2467B. The significant difference (board wise) between a 65, and a 67, is the HV power supply board, the 67 board has the control voltages necessary for the MCP. As a result, the 67 Board grounds P32, and passes the signal through the A1 motherboard to P/J512, P32, on the A5 Controller board. The only schematic where Tektronix forgot to erase this information is the analog control <2> schematic for B050000 and above in the 2465B/2467B manual (070-6863-01). Someone once said: "What a tangled web we weave, when first we practice to deceive." One of the girl poets, as I recall. Emily Dickinson maybe? I hope you won't use this information for evil. -Chuck Harris Raymond Domp Frank wrote: On Sat, Mar 3, 2018 at 02:59 pm, Sergey Kubushyn wrote:ISTR that enabling 500 ps/div horizontal speed to a 2445B just takes adding a jumper near the edge of the main board, at one of the connectors. Sorry, don't remember exactly where and I have no doc. where I am at the moment. I trust this'll ring a bell with someone on this forum who has the details. |
Re: Typo in 2901 SM or wrong value resistor?
I took a look and that's what I see. The R364 on the divider board is a 1k resistor and the manual calls out a 100 ohm resistor.
toggle quoted message
Show quoted text
DaveD On 3/3/2018 6:32 PM, Brian Bloom via Groups.Io wrote:
I'm replacing out-of-tolerance resistors in my 2901 time mark gen and ran across a 1k resistor where the SM says it should be 100R. The resistor in question is R364 on the divider board directly behind Q347. This isn't a model number dependent resistor, but my SN is B041433. |
Re: Can I install a 2465B motherboard in a 2445B chassis? What will I get?
On Sat, Mar 3, 2018 at 02:59 pm, Sergey Kubushyn wrote:
ISTR that enabling 500 ps/div horizontal speed to a 2445B just takes adding a jumper near the edge of the main board, at one of the connectors. Sorry, don't remember exactly where and I have no doc. where I am at the moment. I trust this'll ring a bell with someone on this forum who has the details. Raymond |
Re: Replacement of Tantalum cap removal of U450??
I would probably just clip the leads as you suggested and solder theWherever possible, I try and pull the leads by leaving a length while clipping, heating and pulling the remaining length, for fear of a remaining piece of wire dropping off inside while soldering the new part on top. Raymond |
Dating a 555
I¡¯m trying to approximate the year that my 555 was made. Maybe there¡¯s some way other than just looking at the date codes on components? What I know is the serial number is 601 and the center strip on the scope head is chrome not Tek blue. To be fair I haven¡¯t really poked around too much in the head as I¡¯m working my way the through power supply first. If you have any thoughts or suggestions I¡¯d love to hear them. Oh and the connections are all SO-239 which could be nothing more than an option but I don¡¯t know. Thanks in advance!
|
Re: App Notes
Gary
toggle quoted message
Show quoted text
Curious ..... what is the motivation in this day and age for Tektronix to keep a repository of 30+ year old applications notes? -DC manuals@... On 3/4/2018 12:36 AM, Gary Robert Bosworth wrote:
Thank you Ronald. I hope you can find them. Hard to believe they are not readily available through Tektronix. Why can't Tektronix maintain a repository? --
Dave Manuals@... www.ArtekManuals.com |
Re: App Notes
Thank you Ronald. I hope you can find them. Hard to believe they are not readily available through Tektronix. Why can't Tektronix maintain a repository?
toggle quoted message
Show quoted text
Gary On March 3, 2018, at 9:13 PM, Ronald Carlson <ronald.carlson.54@...> wrote:
Gary, About 6-7 years ago I found these application notes, or most of them, online. Some are definitive tutorials and give procedures directly applicable to the J16 and its many probes. Pg 7 of the 1981 revision of the instruction manual provides the titles of the application notes. As I recall, I found them in an obscure website and am hunting around for them. Will post links with what I find. Ron |
Re: App Notes
WOW did I miss the mark.
toggle quoted message
Show quoted text
First I thought I was replying to the HP group. Second I see where my website suggestion seems to have disappeared in a huge string of ¡°stuff¡¯. So, regardless if your interest is HP, TEK or actually numerous other subjects it is worth visiting archive dot org. Ken On 3Mar, 2018, at 10:30 PM, ken chalfant <kpchalfant@...<mailto:kpchalfant@...>> wrote:
Greetings, I am only catching part of this conversation but allow me to suggest: <><> Applying the search term "Hewlett Packard" brings up an enormous amount of material. BTW, applying the search term "Tektronix" also delivers amazing results. Regards, Ken On 3Mar, 2018, at 10:13 PM, Ronald Carlson <ronald.carlson.54@...<mailto:ronald.carlson.54@...><mailto:ronald.carlson.54@...>> wrote: Gary, About 6-7 years ago I found these application notes, or most of them, online. Some are definitive tutorials and give procedures directly applicable to the J16 and its many probes. Pg 7 of the 1981 revision of the instruction manual provides the titles of the application notes. As I recall, I found them in an obscure website and am hunting around for them. Will post links with what I find. Ron |
Re: App Notes
Greetings,
toggle quoted message
Show quoted text
I am only catching part of this conversation but allow me to suggest: www.archieve.org<> Applying the search term "Hewlett Packard" brings up an enormous amount of material. BTW, applying the search term "Tektronix" also delivers amazing results. Regards, Ken On 3Mar, 2018, at 10:13 PM, Ronald Carlson <ronald.carlson.54@...<mailto:ronald.carlson.54@...>> wrote:
Gary, About 6-7 years ago I found these application notes, or most of them, online. Some are definitive tutorials and give procedures directly applicable to the J16 and its many probes. Pg 7 of the 1981 revision of the instruction manual provides the titles of the application notes. As I recall, I found them in an obscure website and am hunting around for them. Will post links with what I find. Ron |
Re: App Notes
Gary,
About 6-7 years ago I found these application notes, or most of them, online. Some are definitive tutorials and give procedures directly applicable to the J16 and its many probes. Pg 7 of the 1981 revision of the instruction manual provides the titles of the application notes. As I recall, I found them in an obscure website and am hunting around for them. Will post links with what I find. Ron |
Re: What calibration items to buy.
From easy to difficult:
Almost all timebase calibration can be done with a 10 MHz function or signal generator and frequency counter. The fastest sweeps with magnification require a 100 MHz or faster source for linearity adjustment but this is seldom needed. Vertical calibration can be done with a function generator, AC voltmeter, and some attenuators but a calibration generator like a PG506 makes it much faster. Transient response calibration requires a fast reference level pulse generator. The PG506 fulfills this function up to at least 100 MHz but 350 MHz and higher require a faster generator like the tunnel diode pulser Tektronix recommends. Leo sells one which is suitable and more than fast enough: I have an SG503 for generating a leveled sine wave up to 250 MHz but I have not required it for an oscilloscope calibration yet. The SG504 does not cover low frequencies so both are needed for a full range. |
Re: Replacement of Tantalum cap removal of U450??
On Sat, 03 Mar 2018 16:46:32 -0800, you wrote:
Hi,The back of the board connects to the vertical CRT deflection pins. The little collet sockets pull straight off of the pins but you still have to be very careful. U450 includes a threaded base (as per photo in Service Manual) .... and I am unsure as to how to remove it ... or whether I should even try .... :) ..... please advise any ideas?????????I would probably just clip the leads as you suggested and solder the new capacitor on from the top of the board instead of messing with it and risking damage if you are not familiar with removal and assembly. I have not actually done this on my 7603 but I have on my 78xx mainframes. |
Re: 7623A + 7A26 + 7B53
On Fri, 2 Mar 2018 18:48:31 -0300, you wrote:
...Two timebases can indeed generate a raster in these mainframes. A raster is also useful for verifying the condition of the CRT phosphor. 3. The vertical plugin (7A26) seems to be working fairly... But anyA 7B53A can also be used as a calibrated vertical amplifier in XY mode. The clipping does sound like a problem with the 7A26 but try testing it in the right vertical slot also. 4. The "working" 7B53A is acting weird... It's sweeping backwards (fromI had a 7B92A which did this because the delayed timebase cam had separated from the ends. Aligning it and gluing it to the ends with plastic welding compound fixed it. Luckily you have two 7B53As so between the two you may have a complete set of replacement mechanical parts for any which are broken and the second one can serve as a model for reassembling the other. I can also see some upward bump of about 0.1div on the left end of theIt is usual to see some tiny ripples at the beginning of the sweep which are interference from the trigger but it sounds like you are seeing something else. Check the regulated low voltage power supply outputs for ripple. 6. The CRT... well that's where I have mixed feelings... For once, it'sKeep in mind that storage CRTs are not as bright as non-storage CRTs. I will post some pictures as soon as I can take.One of the slower single trace vertical amplifiers like the 7B15A or 7B16A would be convenient for diagnostics and in place of a standardizer. The single trace amplifiers are simpler inside providing easier access to the interface signals than the dual trace amplifiers. |
Re: Typo in 2901 SM or wrong value resistor?
Manual error to be sure ...in my manual the parts list say 100 and the schematic says 1K ( base of Q366) ..my money is on 1K being correct. At 100 ohms I am guessing would pull the base voltage to low and Q366 would be turned off all the time
toggle quoted message
Show quoted text
-DC manuals@... On 3/3/2018 8:32 PM, Brian Bloom via Groups.Io wrote:
I'm replacing out-of-tolerance resistors in my 2901 time mark gen and ran across a 1k resistor where the SM says it should be 100R. The resistor in question is R364 on the divider board directly behind Q347. This isn't a model number dependent resistor, but my SN is B041433. --
Dave Manuals@... www.ArtekManuals.com |
Re: Typo in 2901 SM or wrong value resistor?
It is a typo in the parts list. On the schematic it is 1k. The part number is for a 1k (315-0102-00 is a 1k).
toggle quoted message
Show quoted text
Regards. -----Original Message-----
From: Brian Bloom via Groups.Io <analogaddict013@...> To: TekScopes <[email protected]> Sent: Sat, Mar 3, 2018 8:33 pm Subject: [TekScopes] Typo in 2901 SM or wrong value resistor? I'm replacing out-of-tolerance resistors in my 2901 time mark gen and ran across a 1k resistor where the SM says it should be 100R. The resistor in question is R364 on the divider board directly behind Q347. This isn't a model number dependent resistor, but my SN is B041433. Does anyone have a 2901 they can pop the top on and have a look? Thanks, Brian |
Typo in 2901 SM or wrong value resistor?
I'm replacing out-of-tolerance resistors in my 2901 time mark gen and ran across a 1k resistor where the SM says it should be 100R. The resistor in question is R364 on the divider board directly behind Q347. This isn't a model number dependent resistor, but my SN is B041433.
Does anyone have a 2901 they can pop the top on and have a look? Thanks, Brian |
Re: Cheap differential probe ?
Hello,
May or may not be a fit for your application or budget, but I saw this on page 69 of the March 2018 RSGB magazine: Micsig DP10013 100 MHz high voltage differential probe, about GBP150 Also found on amazon.com for about $170 I do not have one, and have no affiliation with that company (I just subscribe to RSGB). Hope that it helps. Scott |