low amplitude readings below 10MHz on a 8591E


 

Hi All,
I came into the possession of a 8591E and after calibration,
I found out that when I connect my marconi 2019A to the SP and set it to 1Mhz and 0dBm
the peak is hardly to see amongst the noise, while around 20Mhz it about 24dBm below the reference level and
50Mhz it is as it should be.?
Does anybody know what is happening here and what to do about it?

Gr. Kees


 

Hi

This has all the signs of a bad solder joint or connector,

G Edmonds


 

¿ªÔÆÌåÓý

Hi there,

The key to this is that the problem is below 10mhz.

The first calibration point is at 12mhz.

From what I remember the FM coil in the YTO? does the below 10mhz steps.

The main coil in the YTO must be ok as the CAL ?passes.

Go into the cal menu and find SERVICE ?CAL and do the tests as shown the user manual

Let us know what you find

?

Regards

?

Mike g4wam

?

?


 

Just a thought - started with my 8591E losing low end cal signal response with remaining good readings on the high end.? It was the result of a dying YIG.? Started out initially slowly but gradually signals started losing amplitude up the frequency range.? Eventually the cal signals went quiet with just noise.? Was a real head banger until the YIG died.
--
WA4WHL


 

Hi, for laughs and giggles?I just did a new cal and now I got a GAIN FAIL op the AMPTD.

Gr Kees


Op za 18 mrt 2023 om 11:11 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi there,

The key to this is that the problem is below 10mhz.

The first calibration point is at 12mhz.

From what I remember the FM coil in the YTO? does the below 10mhz steps.

The main coil in the YTO must be ok as the CAL ?passes.

Go into the cal menu and find SERVICE ?CAL and do the tests as shown the user manual

Let us know what you find

?

Regards

?

Mike g4wam

?

?


 

¿ªÔÆÌåÓý

This is now leaning towards a YTO failure¡ªnot good.

Definitely do the CAL tests as per the user guide though

?

Mike

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 12:49
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi, for laughs and giggles?I just did a new cal and now I got a GAIN FAIL op the AMPTD.

?

Gr Kees

?

?

Op za 18 mrt 2023 om 11:11 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi there,

The key to this is that the problem is below 10mhz.

The first calibration point is at 12mhz.

From what I remember the FM coil in the YTO? does the below 10mhz steps.

The main coil in the YTO must be ok as the CAL ?passes.

Go into the cal menu and find SERVICE ?CAL and do the tests as shown the user manual

Let us know what you find

?

Regards

?

Mike g4wam

?

?


 

Hi,
Are we talking about the user manual 08590-90301, because there is not much cal data in it
other than the normal stuff. I also have the installation, verification & Operation manual.
Or are there other manual I need?
?
Gr. Kees


 

¿ªÔÆÌåÓý

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 


got it and will start reading.?
Any recommendations about what paragraph?

Gr Kees


Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

¿ªÔÆÌåÓý

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

Hi Mike,
I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.
After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

Still the lower end is terrible ( low amplitude ). I've done the service diag and?
1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?
The rest has the same picture as shown in the?8590-90316 manual.
So according to the manual it's in the A7 board. does somebody have a spare will to sell
or is too premature.

Gr Kees



Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

¿ªÔÆÌåÓý

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

Hi Mike its?
10dBm above the ref lvl. with 10 db ATTEN

Gr. Kees

Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

¿ªÔÆÌåÓý

That suggests a very strong LO? signal ¨Ca good thing so probably not a YTO fault,

Just to clarify..

?

CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST? ?

?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:50
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike its?

10dBm above the ref lvl. with 10 db ATTEN

?

Gr. Kees

?

Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

hi Mike,

CAL FREQUENCY passes yes

CAL AMPTD ¨Cpasses with CAL link in place?

After switching it on tonight? ( sched was about 13 C) it failed, but with the marconi 300Mhz at -20dBm it did,

after that with? the CAL link it passed?the amptd?CAL

?

CONFIDENCE TEST? ? STEP GAIN ATTEN FAIL -6000? SQR 110


Will do it again after a 1 hr warmup


Gr. Kees



Op ma 20 mrt 2023 om 21:46 schreef Mike Lockley via <mikelockley=[email protected]>:

That suggests a very strong LO? signal ¨Ca good thing so probably not a YTO fault,

Just to clarify..

?

CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST? ?

?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:50
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike its?

10dBm above the ref lvl. with 10 db ATTEN

?

Gr. Kees

?

Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

¿ªÔÆÌåÓý

Gr Kees,

?

My ?observation from what you are saying is that the unit is getting too cold ¨Cthese units should be at room temp all the time.

The YTO LO does not like getting towards condensation levels ¨Cthey have no heater built in to the YIG assy.

Also, what is the state of the internal 3.7v battery?

There should be a date on the back of the SA with the replacement date .

If the battery is failing or failed? you will get CAL fail problems

There is a lot of info on YouTube about battery replacement.

?

Regards

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 21 March 2023 18:24
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

hi Mike,

CAL FREQUENCY passes yes

CAL AMPTD ¨Cpasses with CAL link in place?

After switching it on tonight? ( sched was about 13 C) it failed, but with the marconi 300Mhz at -20dBm it did,

after that with? the CAL link it passed?the amptd?CAL

?

CONFIDENCE TEST? ? STEP GAIN ATTEN FAIL -6000? SQR 110

?

Will do it again after a 1 hr warmup

?

Gr. Kees

?

?

Op ma 20 mrt 2023 om 21:46 schreef Mike Lockley via <mikelockley=[email protected]>:

That suggests a very strong LO? signal ¨Ca good thing so probably not a YTO fault,

Just to clarify..

?

CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST? ?

?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:50
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike its?

10dBm above the ref lvl. with 10 db ATTEN

?

Gr. Kees

?

Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

HI Mike,
battery date almost not readable but it looks like the last part of the year is 32 so should?be 83, 93
so I'll replace it after viewing some video's.
After replacement I'll see what the state is like.
The moisture in the shed is never above 60% so no moisture issues.


Tx for the help?
Gr. Kees

Op di 21 mrt 2023 om 20:00 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

?

My ?observation from what you are saying is that the unit is getting too cold ¨Cthese units should be at room temp all the time.

The YTO LO does not like getting towards condensation levels ¨Cthey have no heater built in to the YIG assy.

Also, what is the state of the internal 3.7v battery?

There should be a date on the back of the SA with the replacement date .

If the battery is failing or failed? you will get CAL fail problems

There is a lot of info on YouTube about battery replacement.

?

Regards

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 21 March 2023 18:24
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

hi Mike,

CAL FREQUENCY passes yes

CAL AMPTD ¨Cpasses with CAL link in place?

After switching it on tonight? ( sched was about 13 C) it failed, but with the marconi 300Mhz at -20dBm it did,

after that with? the CAL link it passed?the amptd?CAL

?

CONFIDENCE TEST? ? STEP GAIN ATTEN FAIL -6000? SQR 110

?

Will do it again after a 1 hr warmup

?

Gr. Kees

?

?

Op ma 20 mrt 2023 om 21:46 schreef Mike Lockley via <mikelockley=[email protected]>:

That suggests a very strong LO? signal ¨Ca good thing so probably not a YTO fault,

Just to clarify..

?

CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST? ?

?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:50
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike its?

10dBm above the ref lvl. with 10 db ATTEN

?

Gr. Kees

?

Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

¿ªÔÆÌåÓý

Gr Kees,

There 2 battery types ¨Ccylindrical and disc shaped.

The older is cylindrical ¨Cit will probably be on its last gasps so check the voltage before you remove it.

It is not backed up by a charger circuit., therefore when you change it you will lose all the factory set up data so make sure you read through the restore pages procedure and have all the data to hand,

The options if you have any will recover themselves? except for the gate option ¨Cbut I have the restore procedure for that.

?

?

Regards

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 21 March 2023 20:09
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

HI Mike,

battery date almost not readable but it looks like the last part of the year is 32 so should?be 83, 93

so I'll replace it after viewing some video's.

After replacement I'll see what the state is like.

The moisture in the shed is never above 60% so no moisture issues.

?

?

Tx for the help?

Gr. Kees

?

Op di 21 mrt 2023 om 20:00 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

?

My ?observation from what you are saying is that the unit is getting too cold ¨Cthese units should be at room temp all the time.

The YTO LO does not like getting towards condensation levels ¨Cthey have no heater built in to the YIG assy.

Also, what is the state of the internal 3.7v battery?

There should be a date on the back of the SA with the replacement date .

If the battery is failing or failed? you will get CAL fail problems

There is a lot of info on YouTube about battery replacement.

?

Regards

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 21 March 2023 18:24
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

hi Mike,

CAL FREQUENCY passes yes

CAL AMPTD ¨Cpasses with CAL link in place?

After switching it on tonight? ( sched was about 13 C) it failed, but with the marconi 300Mhz at -20dBm it did,

after that with? the CAL link it passed?the amptd?CAL

?

CONFIDENCE TEST? ? STEP GAIN ATTEN FAIL -6000? SQR 110

?

Will do it again after a 1 hr warmup

?

Gr. Kees

?

?

Op ma 20 mrt 2023 om 21:46 schreef Mike Lockley via <mikelockley=[email protected]>:

That suggests a very strong LO? signal ¨Ca good thing so probably not a YTO fault,

Just to clarify..

?

CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST? ?

?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:50
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike its?

10dBm above the ref lvl. with 10 db ATTEN

?

Gr. Kees

?

Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via <mikelockley=[email protected]>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of display)

What value ?is it?

Mike

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 20 March 2023 19:28
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi Mike,

I've backup the constants ( paper) and used my marconi frequency?generator?to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again and this time it succeeded.

?

Still the lower end is terrible ( low amplitude ). I've done the service diag and?

1. the MAIN Coil driver?is much steeper and starts higher at the -20dbm horizontal line?

The rest has the same picture as shown in the?8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a spare will to sell

or is too premature.

?

Gr Kees

?

?

?

Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via <mikelockley=[email protected]>:

Hi

?in your case ?from 238-247

if you haven¡¯t done it already? ..201 onwards? ?backing up correction constants

?

One thing to check is the level of your LO feedthrough amplitude ?..the signal at 0hz on the left hand edge of the display

Do frequency 0hz? span 1mhz¡­ and see what the level of the LO is? --good ones are around -10db or better.

?

By the way you¡¯re 8591? is a bit different on the front end than the 8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube

?

Regards

Mike g4wam

?

?

?

?

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 17:47
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?


got it and will start reading.?

Any recommendations about what paragraph?

?

Gr Kees

?

?

Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via <mikelockley=[email protected]>:

You need 08590-90316? from? 2001

8590 series Service Guide

?

Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam

?

?

From: [email protected] [mailto:[email protected]] On Behalf Of KeesHofwegen
Sent: 18 March 2023 14:41
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

?

Hi,

Are we talking about the user manual 08590-90301, because there is not much cal data in it

other than the normal stuff. I also have the installation, verification & Operation manual.

Or are there other manual I need?

?

Gr. Kees


 

Given that it is so trivial to connect a backup supply in
parallel before you remove the old, but still holding data
cell, why wouldn't you?

Usually, non chargeable backup batteries are connected to
the memory's logic supply through a diode, and the 5V (or whatever)
logic supply is also connected to the memory's logic supply
through a diode, just connect the backup battery to the junction
of those two diode's cathodes, before you remove the original and
solder in the new.

Be warned that most soldering irons have tips that are grounded
to the power line bonding circuit.

-Chuck Harris


On Wed, 22 Mar 2023 14:13:27 -0000 "Mike Lockley via groups.io"
<mikelockley@...> wrote:
Gr Kees,

There 2 battery types ¨Ccylindrical and disc shaped.

The older is cylindrical ¨Cit will probably be on its last gasps so
check the voltage before you remove it.

It is not backed up by a charger circuit., therefore when you change
it you will lose all the factory set up data so make sure you read
through the restore pages procedure and have all the data to hand,

The options if you have any will recover themselves except for the
gate option ¨Cbut I have the restore procedure for that.





Regards

Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 21 March 2023 20:09 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



HI Mike,

battery date almost not readable but it looks like the last part of
the year is 32 so should be 83, 93

so I'll replace it after viewing some video's.

After replacement I'll see what the state is like.

The moisture in the shed is never above 60% so no moisture issues.





Tx for the help

Gr. Kees



Op di 21 mrt 2023 om 20:00 schreef Mike Lockley via groups.io
<mikelockley@...>:

Gr Kees,



My observation from what you are saying is that the unit is getting
too cold ¨Cthese units should be at room temp all the time.

The YTO LO does not like getting towards condensation levels ¨Cthey
have no heater built in to the YIG assy.

Also, what is the state of the internal 3.7v battery?

There should be a date on the back of the SA with the replacement
date .

If the battery is failing or failed you will get CAL fail problems

There is a lot of info on YouTube about battery replacement.



Regards

Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 21 March 2023 18:24 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



hi Mike,

CAL FREQUENCY passes yes

CAL AMPTD ¨Cpasses with CAL link in place?

After switching it on tonight ( sched was about 13 C) it failed, but
with the marconi 300Mhz at -20dBm it did,

after that with the CAL link it passed the amptd CAL



CONFIDENCE TEST ? STEP GAIN ATTEN FAIL -6000 SQR 110



Will do it again after a 1 hr warmup



Gr. Kees





Op ma 20 mrt 2023 om 21:46 schreef Mike Lockley via groups.io
<mikelockley@...>:

That suggests a very strong LO signal ¨Ca good thing so probably not
a YTO fault,

Just to clarify..



CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST ?



Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 20 March 2023 19:50 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



Hi Mike its

10dBm above the ref lvl. with 10 db ATTEN



Gr. Kees



Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via groups.io
<mikelockley@...>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of
display)

What value is it?

Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 20 March 2023 19:28 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



Hi Mike,

I've backup the constants ( paper) and used my marconi frequency
generator to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again
and this time it succeeded.



Still the lower end is terrible ( low amplitude ). I've done the
service diag and

1. the MAIN Coil driver is much steeper and starts higher at the
-20dbm horizontal line

The rest has the same picture as shown in the 8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a
spare will to sell

or is too premature.



Gr Kees







Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via groups.io
<mikelockley@...>:

Hi

in your case from 238-247

if you haven¡¯t done it already ..201 onwards backing up correction
constants



One thing to check is the level of your LO feedthrough amplitude
..the signal at 0hz on the left hand edge of the display

Do frequency 0hz span 1mhz¡­ and see what the level of the LO is
--good ones are around -10db or better.



By the way you¡¯re 8591 is a bit different on the front end than the
8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube



Regards

Mike g4wam













From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 18 March 2023 17:47 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E




got it and will start reading.

Any recommendations about what paragraph?



Gr Kees





Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via groups.io
<mikelockley@...>:

You need 08590-90316 from 2001

8590 series Service Guide



Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam





From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 18 March 2023 14:41 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



Hi,

Are we talking about the user manual 08590-90301, because there is
not much cal data in it

other than the normal stuff. I also have the installation,
verification & Operation manual.

Or are there other manual I need?



Gr. Kees








 

As it is so simple to restore the data after the battery is removed ,I personally would not like to poke a soldering iron around in the A16 pcb.
That¡¯s my opinion anyway
Regards

Mike g4wam

-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Chuck Harris
Sent: 22 March 2023 17:31
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on a 8591E

Given that it is so trivial to connect a backup supply in
parallel before you remove the old, but still holding data
cell, why wouldn't you?

Usually, non chargeable backup batteries are connected to
the memory's logic supply through a diode, and the 5V (or whatever)
logic supply is also connected to the memory's logic supply
through a diode, just connect the backup battery to the junction
of those two diode's cathodes, before you remove the original and
solder in the new.

Be warned that most soldering irons have tips that are grounded
to the power line bonding circuit.

-Chuck Harris


On Wed, 22 Mar 2023 14:13:27 -0000 "Mike Lockley via groups.io"
<mikelockley@...> wrote:
Gr Kees,

There 2 battery types ¨Ccylindrical and disc shaped.

The older is cylindrical ¨Cit will probably be on its last gasps so
check the voltage before you remove it.

It is not backed up by a charger circuit., therefore when you change
it you will lose all the factory set up data so make sure you read
through the restore pages procedure and have all the data to hand,

The options if you have any will recover themselves except for the
gate option ¨Cbut I have the restore procedure for that.





Regards

Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 21 March 2023 20:09 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



HI Mike,

battery date almost not readable but it looks like the last part of
the year is 32 so should be 83, 93

so I'll replace it after viewing some video's.

After replacement I'll see what the state is like.

The moisture in the shed is never above 60% so no moisture issues.





Tx for the help

Gr. Kees



Op di 21 mrt 2023 om 20:00 schreef Mike Lockley via groups.io
<mikelockley@...>:

Gr Kees,



My observation from what you are saying is that the unit is getting
too cold ¨Cthese units should be at room temp all the time.

The YTO LO does not like getting towards condensation levels ¨Cthey
have no heater built in to the YIG assy.

Also, what is the state of the internal 3.7v battery?

There should be a date on the back of the SA with the replacement
date .

If the battery is failing or failed you will get CAL fail problems

There is a lot of info on YouTube about battery replacement.



Regards

Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 21 March 2023 18:24 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



hi Mike,

CAL FREQUENCY passes yes

CAL AMPTD ¨Cpasses with CAL link in place?

After switching it on tonight ( sched was about 13 C) it failed, but
with the marconi 300Mhz at -20dBm it did,

after that with the CAL link it passed the amptd CAL



CONFIDENCE TEST ? STEP GAIN ATTEN FAIL -6000 SQR 110



Will do it again after a 1 hr warmup



Gr. Kees





Op ma 20 mrt 2023 om 21:46 schreef Mike Lockley via groups.io
<mikelockley@...>:

That suggests a very strong LO signal ¨Ca good thing so probably not
a YTO fault,

Just to clarify..



CAL FREQUENCY passes

CAL AMPTD ¨Cpasses with CAL link in place?

CONFIDENCE TEST ?



Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 20 March 2023 19:50 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



Hi Mike its

10dBm above the ref lvl. with 10 db ATTEN



Gr. Kees



Op ma 20 mrt 2023 om 20:43 schreef Mike Lockley via groups.io
<mikelockley@...>:

Gr Kees,

Did you take a look at the 0hz feedthrough amplitude ¨C(left side of
display)

What value is it?

Mike







From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 20 March 2023 19:28 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



Hi Mike,

I've backup the constants ( paper) and used my marconi frequency
generator to have the amptd NOT fail.

After that the cal signal of the 8591E was used to do an amptd again
and this time it succeeded.



Still the lower end is terrible ( low amplitude ). I've done the
service diag and

1. the MAIN Coil driver is much steeper and starts higher at the
-20dbm horizontal line

The rest has the same picture as shown in the 8590-90316 manual.

So according to the manual it's in the A7 board. does somebody have a
spare will to sell

or is too premature.



Gr Kees







Op za 18 mrt 2023 om 20:20 schreef Mike Lockley via groups.io
<mikelockley@...>:

Hi

in your case from 238-247

if you haven¡¯t done it already ..201 onwards backing up correction
constants



One thing to check is the level of your LO feedthrough amplitude
..the signal at 0hz on the left hand edge of the display

Do frequency 0hz span 1mhz¡­ and see what the level of the LO is
--good ones are around -10db or better.



By the way you¡¯re 8591 is a bit different on the front end than the
8590e series , yours is more prone to mixer failure but can be fixed .

Lots of info on you tube



Regards

Mike g4wam













From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 18 March 2023 17:47 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E




got it and will start reading.

Any recommendations about what paragraph?



Gr Kees





Op za 18 mrt 2023 om 18:19 schreef Mike Lockley via groups.io
<mikelockley@...>:

You need 08590-90316 from 2001

8590 series Service Guide



Are large tome but worth downloading or buying

¸é±ð²µ²¹°ù»å¡¯²õ

Mike g4wam





From: [email protected]
[mailto:[email protected]] On Behalf Of
KeesHofwegen Sent: 18 March 2023 14:41 To:
[email protected] Subject: Re:
[HP-Agilent-Keysight-equipment] low amplitude readings below 10MHz on
a 8591E



Hi,

Are we talking about the user manual 08590-90301, because there is
not much cal data in it

other than the normal stuff. I also have the installation,
verification & Operation manual.

Or are there other manual I need?



Gr. Kees