Hello,
After a long period of non use of my MMS 70k I decided to use it again to measure some NF with the 70875A DLP software. A new calibration of the spectrum was done OK. I loaded the 70875A DLP software and after some manipulation I realized I could not use anymore the software. Don't know what I did but impossible to return to the spectrum functionalities. When I try to preset the instrument (green button) or the Menu function I got USERLOCK ON and impossible anymore to use the instrument. How to go to the USERLOCK OFF back again and/or to reset the instrument? I'm really stucked on this situation with the impossibility to use it again. Thanks for your help.
-- Pierre-Fran?ois, F5BQP JN18AN
|
USERLOCK OFF par HPIB. On Monday, April 20, 2020, 08:55:34 PM GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...> wrote:
Hello,
After a long period of non use of my MMS 70k I decided to use it again to measure some NF with the 70875A DLP software. A new calibration of the spectrum was done OK. I loaded the 70875A DLP software and after some manipulation I realized I could not use anymore the software. Don't know what I did but impossible to return to the spectrum functionalities. When I try to preset the instrument (green button) or the Menu function I got USERLOCK ON and impossible anymore to use the instrument. How to go to the USERLOCK OFF back again and/or to reset the instrument? I'm really stucked on this situation with the impossibility to use it again. Thanks for your help.
-- Pierre-Fran?ois, F5BQP JN18AN
|
Ok fine JF, but how to proceed when you have
connected a computer to the HPIB ??
How to send the command ?
73's
pf
Le 20/04/2020 à 23:19, jfphp via
groups.io a écrit?:
toggle quoted message
Show quoted text
USERLOCK OFF par HPIB.
On Monday, April 20, 2020, 08:55:34 PM GMT+2,
Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hello,
After a long period of non use of my MMS 70k I decided to
use it again to measure some NF with the 70875A DLP
software.
A new calibration of the spectrum was done OK.
I loaded the 70875A DLP software and after some
manipulation I realized I could not use anymore the
software.
Don't know what I did but impossible to return to the
spectrum functionalities.
When I try to preset the instrument (green button) or the
Menu function I got USERLOCK ON and impossible anymore to
use the instrument.
How to go to the USERLOCK OFF back again and/or to reset
the instrument?
I'm really stucked on this situation with the
impossibility to use it again.
Thanks for your help.
--
Pierre-Fran?ois, F5BQP
JN18AN
-- Pierre-Fran?ois, F5BQP JN18AN
|
Hi Pierre-Fran?ois
?
Are you actually going to send commands to
the? analyser if so why
?
I can see the need to do screen dumps to a
computer? also? dater streams of the displayed information
?
Paul
?
Ok fine JF, but how to proceed when
you have connected a computer to the HPIB ??
How to send the command ?
73's
pf
toggle quoted message
Show quoted text
Le 20/04/2020 à 23:19, jfphp via groups.io a écrit?:
On Monday, April 20, 2020, 08:55:34 PM
GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hello,
After a long period of non use of my MMS 70k I decided to use it again to measure
some NF with the 70875A DLP software.
A new calibration of the spectrum was done OK.
I loaded the 70875A DLP software and after some manipulation I realized I could
not use anymore the software.
Don't know what I did but impossible to return to the spectrum functionalities.
When I try to preset the instrument (green button) or the Menu function I got USERLOCK ON
and impossible anymore to use the instrument.
How to go to the USERLOCK OFF back again and/or to reset the instrument?
I'm really stucked on this situation with the impossibility to use it again.
Thanks for your help.
--
Pierre-Fran?ois, F5BQP
JN18AN
?
No virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886 - Release Date: 08/14/18
Internal Virus Database is out of date.
-- Pierre-Fran?ois, F5BQP JN18AN
|
Hi Paul,
I would like to reset my analyser as I'm into a deadlock within
the DLP sequence and I cannot do anything to exit.
Even the Preset Instrument button (green button) shows me the
message USERLOCK ON when I press it.
pf
Le 21/04/2020 à 10:56, Paul Bicknell a
écrit?:
Hi Pierre-Fran?ois
?
Are you
actually going to send commands to
the? analyser if so why
?
I can see the
need to do screen dumps to a
computer? also? dater streams of the displayed
information
?
Paul
?
Ok fine
JF, but how to proceed when
you have connected a computer to the HPIB ??
How to send the command ?
73's
pf
Le 20/04/2020 à 23:19, jfphp via
groups.io a écrit?:
On
Monday, April 20, 2020, 08:55:34 PM
GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hello,
After a long period of non use of my MMS
70k I decided to use it again to measure
some NF with the 70875A DLP software.
A new calibration of the spectrum was
done OK.
I loaded the 70875A DLP software and
after some manipulation I realized I
could
not use anymore the software.
Don't know what I did but impossible to
return to the spectrum functionalities.
When I try to preset the instrument
(green button) or the Menu function I
got USERLOCK ON
and impossible anymore to use the
instrument.
How to go to the USERLOCK OFF back again
and/or to reset the instrument?
I'm really stucked on this situation
with the impossibility to use it again.
Thanks for your help.
--
Pierre-Fran?ois, F5BQP
JN18AN
?
No
virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886 -
Release Date: 08/14/18
Internal Virus Database is out of date.
--
Pierre-Fran?ois, F5BQP
JN18AN
-- Pierre-Fran?ois, F5BQP JN18AN
|
Programing manual 70900-90284 p. 2.550 Check HPIB address of the 70900A/B !
On Tuesday, April 21, 2020, 11:11:02 AM GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...> wrote:
Hi Paul, I would like to reset my analyser as I'm into a deadlock within the DLP sequence and I cannot do anything to exit. Even the Preset Instrument button (green button) shows me the message USERLOCK ON when I press it. pf Le 21/04/2020 à 10:56, Paul Bicknell a écrit?: Hi Pierre-Fran?ois ? Are you actually going to send commands to the? analyser if so why ? I can see the need to do screen dumps to a computer? also? dater streams of the displayed information ? Paul
toggle quoted message
Show quoted text
From: [email protected] [mailto:[email protected] ] On Behalf Of Pierre-Fran?ois (f5bqp_pfm) Sent: 21 April 2020 06:17 To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Problem with HP MMS 70000, USERLOCK ON ? Ok fine JF, but how to proceed when you have connected a computer to the HPIB ?? How to send the command ? 73's pf Le 20/04/2020 à 23:19, jfphp via groups.io a écrit?: On Monday, April 20, 2020, 08:55:34 PM GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...> wrote: Hello, After a long period of non use of my MMS 70k I decided to use it again to measure some NF with the 70875A DLP software. A new calibration of the spectrum was done OK. I loaded the 70875A DLP software and after some manipulation I realized I could not use anymore the software. Don't know what I did but impossible to return to the spectrum functionalities. When I try to preset the instrument (green button) or the Menu function I got USERLOCK ON and impossible anymore to use the instrument. How to go to the USERLOCK OFF back again and/or to reset the instrument? I'm really stucked on this situation with the impossibility to use it again. Thanks for your help. -- Pierre-Fran?ois, F5BQP JN18AN ? No virus found in this message. Checked by AVG - Version: 2016.0.8048 / Virus Database: 4793/15886 - Release Date: 08/14/18 Internal Virus Database is out of date. -- Pierre-Fran?ois, F5BQP JN18AN
-- Pierre-Fran?ois, F5BQP JN18AN
|
Thank you JF, will keep you informed .
73's
pf
Le 21/04/2020 à 13:37, jfphp via
groups.io a écrit?:
toggle quoted message
Show quoted text
Programing manual 70900-90284 p. 2.550
Check HPIB address of the 70900A/B !
On Tuesday, April 21, 2020, 11:11:02 AM GMT+2,
Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hi Paul,
I would like to reset my analyser as I'm into a
deadlock within the DLP sequence and I cannot do
anything to exit.
Even the Preset Instrument button (green button) shows
me the message USERLOCK ON when I press it.
pf
Le 21/04/2020
à 10:56, Paul Bicknell a écrit?:
Hi Pierre-Fran?ois
?
Are you
actually going to send commands to the? analyser
if so why
?
I can see
the need to do screen dumps to a computer? also?
dater streams of the displayed information
?
Paul
?
Ok
fine JF, but how to proceed when you have
connected a computer to the HPIB ??
How to send the command ?
73's
pf
Le 20/04/2020 à 23:19, jfphp via
groups.io a écrit?:
On Monday, April 20, 2020, 08:55:34
PM GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hello,
After a long period of non use of my MMS
70k I decided to use it again to measure
some NF with the 70875A DLP software.
A new calibration of the spectrum was
done OK.
I loaded the 70875A DLP software and
after some manipulation I realized I
could not use anymore the software.
Don't know what I did but impossible to
return to the spectrum functionalities.
When I try to preset the instrument
(green button) or the Menu function I
got USERLOCK ON and impossible anymore
to use the instrument.
How to go to the USERLOCK OFF back again
and/or to reset the instrument?
I'm really stucked on this situation
with the impossibility to use it again.
Thanks for your help.
--
Pierre-Fran?ois, F5BQP
JN18AN
?
No virus found in
this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886 -
Release Date: 08/14/18
Internal Virus Database is out of date.
--
Pierre-Fran?ois, F5BQP
JN18AN
--
Pierre-Fran?ois, F5BQP
JN18AN
-- Pierre-Fran?ois, F5BQP JN18AN
|
Hi JF,
I could USERLOCK OFF with my HP computer using your method.
I've tried to load again the 70875A DLP loaded into my 70900B and
faced out to the same USERLOCK ON !
I guess the program is corrupted (ERR2001), tomorrow I'll try to
reload it with my original personality module.
May be due to the 70900B battery which could be out of gas .
Will see.
Thanks again
Pierre-Fran?ois, F5BQP
?
Le 21/04/2020 à 13:37, jfphp via
groups.io a écrit?:
toggle quoted message
Show quoted text
Programing manual 70900-90284 p. 2.550
Check HPIB address of the 70900A/B !
On Tuesday, April 21, 2020, 11:11:02 AM GMT+2,
Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hi Paul,
I would like to reset my analyser as I'm into a
deadlock within the DLP sequence and I cannot do
anything to exit.
Even the Preset Instrument button (green button) shows
me the message USERLOCK ON when I press it.
pf
Le 21/04/2020
à 10:56, Paul Bicknell a écrit?:
Hi Pierre-Fran?ois
?
Are you
actually going to send commands to the? analyser
if so why
?
I can see
the need to do screen dumps to a computer? also?
dater streams of the displayed information
?
Paul
?
Ok
fine JF, but how to proceed when you have
connected a computer to the HPIB ??
How to send the command ?
73's
pf
Le 20/04/2020 à 23:19, jfphp via
groups.io a écrit?:
On Monday, April 20, 2020, 08:55:34
PM GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hello,
After a long period of non use of my MMS
70k I decided to use it again to measure
some NF with the 70875A DLP software.
A new calibration of the spectrum was
done OK.
I loaded the 70875A DLP software and
after some manipulation I realized I
could not use anymore the software.
Don't know what I did but impossible to
return to the spectrum functionalities.
When I try to preset the instrument
(green button) or the Menu function I
got USERLOCK ON and impossible anymore
to use the instrument.
How to go to the USERLOCK OFF back again
and/or to reset the instrument?
I'm really stucked on this situation
with the impossibility to use it again.
Thanks for your help.
--
Pierre-Fran?ois, F5BQP
JN18AN
?
No virus found in
this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886 -
Release Date: 08/14/18
Internal Virus Database is out of date.
--
Pierre-Fran?ois, F5BQP
JN18AN
--
Pierre-Fran?ois, F5BQP
JN18AN
-- Pierre-Fran?ois, F5BQP JN18AN
|
Hi JF & the Gang,
After reloading a fresh DLP NF 70875A from my original ROM
personality module I got the same error ERR 2001 and entered once
again into the USERLOCK ON dead lock.
I was again obliged to send the USERLOCK OFF over the HPIB bus and
the analyzer function started to work fine back after a CAL ALL .
So I suspect a battery out of gas within the 70900B (but not sure
...), could someone let me know the type of battery I will have to
purchase ?
Thanks
73's
Pierre-Fran?ois, F5BQP
Le 21/04/2020 à 13:53, Pierre-Fran?ois
(f5bqp_pfm) a écrit?:
toggle quoted message
Show quoted text
Thank you JF, will keep you informed .
73's
pf
Le 21/04/2020 à 13:37, jfphp via
groups.io a écrit?:
Programing manual 70900-90284 p. 2.550
Check HPIB address of the 70900A/B !
On Tuesday, April 21, 2020, 11:11:02 AM GMT+2,
Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hi Paul,
I would like to reset my analyser as I'm into a
deadlock within the DLP sequence and I cannot do
anything to exit.
Even the Preset Instrument button (green button)
shows me the message USERLOCK ON when I press it.
pf
Le
21/04/2020 à 10:56, Paul Bicknell a écrit?:
Hi Pierre-Fran?ois
?
Are
you actually going to send commands to the?
analyser if so why
?
I
can see the need to do screen dumps to a
computer? also? dater streams of the displayed
information
?
Paul
?
Ok fine JF, but how to proceed
when you have connected a computer to the HPIB
??
How to send the command ?
73's
pf
Le 20/04/2020 à 23:19, jfphp via
groups.io a écrit?:
On
Monday, April 20, 2020, 08:55:34 PM
GMT+2, Pierre-Fran?ois (f5bqp_pfm) <pfmonet@...>
wrote:
Hello,
After a long period of non use of my
MMS 70k I decided to use it again to
measure some NF with the 70875A DLP
software.
A new calibration of the spectrum was
done OK.
I loaded the 70875A DLP software and
after some manipulation I realized I
could not use anymore the software.
Don't know what I did but impossible
to return to the spectrum
functionalities.
When I try to preset the instrument
(green button) or the Menu function I
got USERLOCK ON and impossible anymore
to use the instrument.
How to go to the USERLOCK OFF back
again and/or to reset the instrument?
I'm really stucked on this situation
with the impossibility to use it
again.
Thanks for your help.
--
Pierre-Fran?ois, F5BQP
JN18AN
?
No virus found in
this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886 -
Release Date: 08/14/18
Internal Virus Database is out of date.
--
Pierre-Fran?ois, F5BQP
JN18AN
--
Pierre-Fran?ois, F5BQP
JN18AN
--
Pierre-Fran?ois, F5BQP
JN18AN
-- Pierre-Fran?ois, F5BQP JN18AN
|
I then decided to open my 70900B, have not done this since many years when I upgraded the firmware with the latest from my previous 70952B. At that time I uploaded this latest firmware on KO4BB web site. So I tested the battery and found 3.074v for a 3v battery, so my problem isn't this. The battery model not replaced by me is a Panasonic BR-2/3A . So the problem seems to be more complex than expected, may be a faulty memory ?? Question: Is there somewhere a memory test that could be run to test the memories of the 70900B memory CPU card? Any suggestion to the person having experience with the 70900B logic ?
The problem is just when I load an HP DLP (70875A) within the analyzer I got this problem, otherwise in spectrum analyzer mode the MMS is working fine.
pf
-- Pierre-Fran?ois, F5BQP JN18AN
|
And I measured the effective retention voltage on the vdd of the TC551001CF-70L memories, it's at 2.864v , the datasheet says it should be at 2v minimum to keep up the data, so definitively it's not a battery problem, probably a faulty memory. Just an assumption. A memory test would be a good solution to identify if it's such a problem or something else.
-- Pierre-Fran?ois, F5BQP JN18AN
|
Hi have you looked at the service manual
?
?
I then decided to open my 70900B, have not done this since many years
when I upgraded the firmware with the latest from my previous 70952B.
At that time I uploaded this latest firmware on KO4BB web site.
So I tested the battery and found 3.074v for a 3v battery, so my problem isn't
this.
The battery model not replaced by me is a Panasonic BR-2/3A .
So the problem seems to be more complex than expected, may be a faulty memory
??
Question: Is there somewhere a memory test that could be run to test the
memories of the 70900B memory CPU card?
Any suggestion to the person having experience with the 70900B logic ?
The problem is just when I load an HP DLP (70875A) within the analyzer I got
this problem, otherwise in spectrum analyzer mode the MMS is working fine.
pf
--
Pierre-Fran?ois, F5BQP
JN18AN
No virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886 - Release Date: 08/14/18
Internal Virus Database is out of date.
|
1. Are you sure the program did run once or is it the first time you try to run this DLP into this 70900B CPU (coming from an OSA) ? 2. There is no way to test the internal memory 3. Try to erase all the progs in memory : do not use "erase" but "purge" with every name in the DLP catalog (read text under PERASE, PURGE and ERASE commands in the prog manual) 4. Compatibility of this DLP with the OSA 70900 board ??? Try another "normal" board 5. If not running with a SA board, DLP faulty (seen with a receiver ROM personality) On Wednesday, April 22, 2020, 12:09:24 PM GMT+2, Paul Bicknell <paul@...> wrote:
Hi have you looked at the service manual ?
toggle quoted message
Show quoted text
From: [email protected] [mailto: [email protected] ]On Behalf Of Pierre-Fran?ois(f5bqp_pfm) Sent: 22 April 2020 09:40 To: [email protected] Subject: Re:[HP-Agilent-Keysight-equipment] Problem with HP MMS 70000, USERLOCK ON ? I then decided to open my 70900B, have not done this since many yearswhen I upgraded the firmware with the latest from my previous 70952B. At that time I uploaded this latest firmware on KO4BB web site. So I tested the battery and found 3.074v for a 3v battery, so my problem isn'tthis. The battery model not replaced by me is a Panasonic BR-2/3A . So the problem seems to be more complex than expected, may be a faulty memory?? Question: Is there somewhere a memory test that could be run to test thememories of the 70900B memory CPU card? Any suggestion to the person having experience with the 70900B logic ?
The problem is just when I load an HP DLP (70875A) within the analyzer I gotthis problem, otherwise in spectrum analyzer mode the MMS is working fine.
pf
-- Pierre-Fran?ois, F5BQP JN18AN No virus found in this message. Checked by AVG - Version: 2016.0.8048 / Virus Database: 4793/15886 - Release Date: 08/14/18 Internal Virus Database is out of date.
|
Hi,
Yes for sure !
73's
pf
Le 22/04/2020 à 12:13, Paul Bicknell a
écrit?:
Hi have you
looked at the service manual
?
?
I then decided to open my 70900B, have not
done this since many years
when I upgraded the firmware with the latest from
my previous 70952B.
At that time I uploaded this latest firmware on
KO4BB web site.
So I tested the battery and found 3.074v for a 3v
battery, so my problem isn't
this.
The battery model not replaced by me is a
Panasonic BR-2/3A .
So the problem seems to be more complex than
expected, may be a faulty memory
??
Question: Is there somewhere a memory test that
could be run to test the
memories of the 70900B memory CPU card?
Any suggestion to the person having experience
with the 70900B logic ?
The problem is just when I load an HP DLP (70875A)
within the analyzer I got
this problem, otherwise in spectrum analyzer mode
the MMS is working fine.
pf
--
Pierre-Fran?ois, F5BQP
JN18AN
No virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database: 4793/15886
- Release Date: 08/14/18
Internal Virus Database is out of date.
-- Pierre-Fran?ois, F5BQP JN18AN
|
1) Yes, it was running during many years !
2) Too bad ...
3) I'll re-install the CPU+Mem? boards and as I unplugged the
Memory board from the CPU where the battery is located I'm sure
the old programs are gone.
4) Not necessary it worked for years, this isn't a compatibility
of the DLP with the last OSA firmware.
If really needed I could reinstall the old PROMS as I installed
PLCC sockets several years ago with the new firmware.
I would like to avoid to do this because time consuming.
5) As I discovered this problem with the already loaded DLP since
years and then with a fresh load (this morning) from the original
personality I'm pretty sure than the DLP isn't faulty.
For the moment I would suspect a faulty RAM, will continue to
investigate.
73's
pf
Le 22/04/2020 à 12:43, jfphp via
groups.io a écrit?:
toggle quoted message
Show quoted text
1. Are you sure the program did run once or is it the first
time you try to run this DLP into this 70900B CPU (coming from
an OSA) ?
2. There is no way to test the internal memory
3. Try to erase all the progs in memory : do not use "erase" but
"purge" with every name in the DLP catalog (read text under
PERASE, PURGE and ERASE commands in the prog manual)
4. Compatibility of this DLP with the OSA 70900 board ??? Try
another "normal" board
5. If not running with a SA board, DLP faulty (seen with a
receiver ROM personality)
On Wednesday, April 22, 2020, 12:09:24 PM GMT+2, Paul
Bicknell <paul@...> wrote:
Hi
have you looked at the service manual
?
?
I then decided to open my 70900B, have
not done this since many yearswhen I upgraded
the firmware with the latest from my previous
70952B.
At that time I uploaded this latest firmware on
KO4BB web site.
So I tested the battery and found 3.074v for a
3v battery, so my problem isn'tthis.
The battery model not replaced by me is a
Panasonic BR-2/3A .
So the problem seems to be more complex than
expected, may be a faulty memory??
Question: Is there somewhere a memory test that
could be run to test thememories of the 70900B
memory CPU card?
Any suggestion to the person having experience
with the 70900B logic ?
The problem is just when I load an HP DLP
(70875A) within the analyzer I gotthis problem,
otherwise in spectrum analyzer mode the MMS is
working fine.
pf
--
Pierre-Fran?ois, F5BQP
JN18AN
No virus found in this message.
Checked by AVG -
Version: 2016.0.8048 / Virus Database:
4793/15886 - Release Date: 08/14/18
Internal Virus Database is out of date.
-- Pierre-Fran?ois, F5BQP JN18AN
|
PROBLEM SOLVED !
After the memory card was unplugged from the CPU card which has retention the battery and then plugged again the problem disappeared. Conclusion the ERASE ALL of the analyzer doesn't clear properly the internal memory, the only way to do a real clear is to disassemble the memory card from the CPU board of the 70900B. I presume also for the A version.
Thanks to all of you who helped me. 73's pf -- Pierre-Fran?ois, F5BQP JN18AN
|
Hi Pierre
?
40 manuals a r now on the HP 70000 Group site? along with a folder
called ?
Notes for? MMS And in there is an
XL sheet for? 2 of the key boards I did last year and a few more bits
of handy information
?
The folder 70000 MMS had a very
comprehensive catalogue that is worth printed out and binding ??unfortunately it
is missing 2 pages but the catalogue is a mine of information
?
Regards Paul
?
PROBLEM SOLVED !
After the memory card was unplugged from the CPU card which has retention the
battery and then plugged again the problem disappeared.
Conclusion the ERASE ALL of the analyzer doesn't clear properly the internal
memory, the only way to do a real clear is to disassemble the memory card from
the CPU board of the 70900B.
I presume also for the A version.
Thanks to all of you who helped me.
73's
pf
--
Pierre-Fran?ois, F5BQP
JN18AN
|
Hi,
Has anyone managed to get a copy of the HP70875A on a card?
Hope to hear, thanks,
Bas.
|