¿ªÔÆÌåÓý

Date

Re: HP-IB to PC Printer for 54501 Scope?

John Miles
 

Try the 7470.exe application at .
If it works, I'd appreciate a heads-up so I can add the 54501A to the list
of supported instruments.

-- john KE5FX

-----Original Message-----
From: Brooke Clarke [mailto:brooke@...]
Sent: Wednesday, September 22, 2004 1:02 PM
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] HP-IB to PC Printer for 54501 Scope?


Hi:

I have an HP 54501A scope and would like to get a copy of the
screen to a printer and/or a file on my PC. I have an NI GP-IB
card and HP-IB cables. Is there a free program for doing this?

Thanks,

Brooke Clarke, N6GCE
--
w/Java
w/o Java




HP-IB to PC Printer for 54501 Scope?

 

Hi:

I have an HP 54501A scope and would like to get a copy of the screen to a printer and/or a file on my PC. I have an NI GP-IB card and HP-IB cables. Is there a free program for doing this?

Thanks,

Brooke Clarke, N6GCE
--
w/Java
w/o Java


Re: XFMR part

 

Dave,

Thanks for looking that info up. This transformer is probably from
something else.

Some extra info that might help identify it is that the terminals come
out of the transformer case in five round "bulkheads" lettered "A" to
"E". The "A" group has four terminals which I think are a pair of
input windings (for 120 or 240, I guess). The next three groups are
all five terminals each and the last group is just two terminals. Each
grouping is numbered sequentially from 1. I've looked at several
different HP scope manuals on the BAMA web site and cannot find this
number but nor can I find any schematics that use a letter-number
terminal identification scheme.

Brad

--- In hp_agilent_equipment@..., "Dave Brown"
<tractorb@i...> wrote:
Brad
The std 180 mainfrme power xfmr (according to my handbook, for s/ns
prefix 975- and above) is 9100-1129 (I think-may not be reading this
correctly). This is a dual voltage 115/230 primary. The alternative,
for slightly lower mains etc(e.g. 100 volt) is 9100-1109. No mention
of 9100-0154, but it may have been in earlier models.
From the cct the xfmr secondary winding details as follows-
10vdc/50mA, 122.5vdc/300mA, 25vdc/1.3A, 19.5vdc/1.1A, 120vdc/170mA
and 6.3vac/500mA, 4.5vac/120mA.
Note the spec in terms of dc out for most windings.
73
Dave, ZL3FJ


----- Original Message -----
From: "Brad Hernlem" <alibhernlem@y...>
To: <hp_agilent_equipment@...>
Sent: Tuesday, September 21, 2004 1:24 AM
Subject: [hp_agilent_equipment] XFMR part #9100-0154



I have a large transformer with
part number 9100-0154 for which I seek pinout and rating data. I
think that it MAY have come from a 180 series oscilloscope.

Can anyone here help me out?

TNX

Brad




------------------------ Yahoo! Groups


Yahoo! Groups Links








logsa

Ahlbrecht, Andreas
 

¿ªÔÆÌåÓý

Hi Group,

is anyone aware of problems with the logsa site?
I'm getting a 'page cannot be displayed' on

and just a blank page at

Anyone else having trouble accessing these sites?
Is it just me?

Thanks
Andi


Re: XFMR part

Dave Brown
 

Brad
The std 180 mainfrme power xfmr (according to my handbook, for s/ns prefix 975- and above) is 9100-1129 (I think-may not be reading this correctly). This is a dual voltage 115/230 primary. The alternative, for slightly lower mains etc(e.g. 100 volt) is 9100-1109. No mention of 9100-0154, but it may have been in earlier models.
From the cct the xfmr secondary winding details as follows-
10vdc/50mA, 122.5vdc/300mA, 25vdc/1.3A, 19.5vdc/1.1A, 120vdc/170mA and 6.3vac/500mA, 4.5vac/120mA.
Note the spec in terms of dc out for most windings.
73
Dave, ZL3FJ

----- Original Message -----
From: "Brad Hernlem" <alibhernlem@...>
To: <hp_agilent_equipment@...>
Sent: Tuesday, September 21, 2004 1:24 AM
Subject: [hp_agilent_equipment] XFMR part #9100-0154



I have a large transformer with
part number 9100-0154 for which I seek pinout and rating data. I
think that it MAY have come from a 180 series oscilloscope.

Can anyone here help me out?

TNX

Brad




------------------------ Yahoo! Groups

Yahoo! Groups Links








FW: Fraud

John Miles
 

¿ªÔÆÌåÓý

Forwarded from a Hong Kong acquaintance who just got burned big-time by a?lumber salesman?in Quebec.? I don't have any more details about the alleged fraud, or how?the payment was made, but it's?a big enough deal to justify posting a heads-up.
?
-- john KE5FX
?

-----Original Message-----
From:??(name redacted)?
Sent: Monday, September 20, 2004 2:54 AM
To: 'John Miles'
Subject: Fraud

Hi John

?

Just thought you might get a laugh from hearing that I recently paid USD 6200 to a guy in Quebec Canada for 3 planks of wood which were supposed to be an Agilent E4432B Signal generator. It was eBay ¡°bait¡± I should have walked away from.

?

He actually shipped the goods before payment, thinking that would make me feel more comfortable. The shipping time would allow him to collect his money and disappear. Well it worked, he got the money and I got three expensive planks of wood!

?

Pass this around to your friends. If anyone comes across a guy named Konstantinos Tabiris, watch out. His email for what its worth is hns0909@....

?

Now I have to find a more ¡°legitimate¡± seller and in the mean time, hope to recover my money somehow¡­..

?

Cheers

(name redacted)

?


Re: HP 10511A Spectrum Generator

 

--- In hp_agilent_equipment@..., "Dave Brown"
<tractorb@i...> wrote:
From the 1968 catalog-
Hello Dave,

thank you very much for those informations about the 10511A

Jorgen
dj0ud


Re: New file uploaded to hp_agilent_equipment

phila_renewal
 

To avi49il:

Thanks very much for the file. I haven't had a chance to unpack it
yet, but with an 8559A that seems to be acting "squirrely," on at
least one of the freq. ranges, I know it will be a huge help.

Thanks again,
-Keith


--- In hp_agilent_equipment@...,
hp_agilent_equipment@... wrote:

Hello,

This email message is a notification to let you know that
a file has been uploaded to the Files area of the
hp_agilent_equipment
group.

File : /HP-8559A-OM-SM.part2.rar
Uploaded by : avi49il <zipi-u@z...>
Description : hp-8559-manual/part2

You can access this file at the URL:
OM-SM.part2.rar

To learn more about file sharing for your group, please visit:


Regards,

avi49il <zipi-u@z...>


XFMR part

 

I have a large transformer with
part number 9100-0154 for which I seek pinout and rating data. I
think that it MAY have come from a 180 series oscilloscope.

Can anyone here help me out?

TNX

Brad


Re: HP 10511A Spectrum Generator

Dave Brown
 

From the 1968 catalog-
Input 25-50 MHz(10-75 useful) with 1-3 volts ex 50 ohm source
Output 50 ohms nominal and -19 dBm minimum for any harmonic number between 1 and 10
Which doesn't quite stack up with the quoted output frequency range as 50-500 Mhz but I guess it's near enough!
Intended as an accessory for the 5100A.
Cheers
Dave Brown, ZL3FJ
Christchurch, NZ

----- Original Message -----
From: "Jorgen Lund-Nielsen" <dj0ud@...>
To: <hp_agilent_equipment@...>
Sent: Monday, September 20, 2004 10:15 AM
Subject: [hp_agilent_equipment] HP 10511A Spectrum Generator


Helo,

is there any information about this, i think, a comb generator,
available in this group?
I expect it's a SRD-Based device, but i didn't knew any working
conditions for it, i.e. (max) input power, input frequency range,
output frequency range ( ... rise time of the combs) and so on.
Could you please have a look in your Catalogues (i expext the end of the
sixties) for some data for this unit?

Best regards,

Jorgen
dj0ud



------------------------ Yahoo! Groups

Yahoo! Groups Links






HP 10511A Spectrum Generator

 

Helo,

is there any information about this, i think, a comb generator,
available in this group?
I expect it's a SRD-Based device, but i didn't knew any working
conditions for it, i.e. (max) input power, input frequency range,
output frequency range ( ... rise time of the combs) and so on.
Could you please have a look in your Catalogues (i expext the end of the
sixties) for some data for this unit?

Best regards,

Jorgen
dj0ud


Re: HP 8566B ROMs available

 

¿ªÔÆÌåÓý

do'es its conteining the step attenuator correction table?
this corrections are depand on the preticuler step attenuator in the instrument so its different from instrument to instrument.
?


Re: HP 8566B ROMs available

John Miles
 

¿ªÔÆÌåÓý

Interesting point.? I?just checked my service manuals, and it appears that there is indeed an?attenuator-correction ROM in the 8566B.? The HP?guys didn't cut any corners on this piece of gear!??It?is a separate part installed on A12, the front-panel keyboard assembly.? (A12U8, to be specific.)?
?
The zipfile I posted contains only the main operating firmware, from the 68000 CPU board.? I'm pretty sure that the images aren't serialized or otherwise tied to the specific instrument they came from.
?
-- john KE5FX

-----Original Message-----
From: elie shvarzman [mailto:eliesh@...]
Sent: Saturday, September 18, 2004 5:32 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP 8566B ROMs available

do'es its conteining the step attenuator correction table?
this corrections are depand on the preticuler step attenuator in the instrument so its different from instrument to instrument.


Re: HP8601A Generator/Sweeper

Jose V. Gavila
 

Hello!

Talking about HP8601As.... I have a couple of them (50 Ohm versions; be sure the one you find is not a 75 Ohm version!) but, unfortunately, one has a bad output power amplifier. I wonder if somebody has a parts unit which could act as a donnor of that part.

Thanks!

JOSE
----------------------------------------------------------------------
73 EB5AGV / EC5AAU - JOSE V. GAVILA
La Canyada - Valencia (SPAIN) - Loc: IM99SM

Vintage Radio:
Vintage Test Equipment:
European Boatanchors List:


HP 8566B ROMs available

John Miles
 

I've submitted the EPROM firmware from my HP8566B to Bruce Lane's archive:

ftp://ftp.bluefeathertech.com/pub/electronics/testgear/HP/firmware/8566B/

Filename is hp8566b_9_29_86.zip . These image files are from my own 8566B
with serial prefix 27xx; I don't know if they will work in earlier or later
instruments.

See guidelines for access to Bruce's archive below.

-- john KE5FX

-----Original Message-----
From: Bruce Lane [mailto:kyrrin@...]
Sent: Saturday, September 18, 2004 12:56 PM
To: jmiles@...

-=-=-=-=-=-=-=-=-=-=-

I run a good-sized FTP archive that currently contains, among other things,
EPROM (firmware) images for Tektronix test gear that they no longer make or
support. This is done to provide a repository of image files that one can
use to make a new EPROM if one should fail (yes, they do fail on occasion).

By sending me an archive with 8566B EPROM images in it, John Miles has made
it clear that doing the same for HP stuff would probably be a good idea. So
be it! I welcome contributions for the archive subject to the following
conditions.

(1) Any firmware or software sent should be for equipment that HP/Agilent
no longer manufactures or supports.

(2) If what you're sending is related to current equipment, it must be
publicly available for download from HP/Agilent as well. In other words,
HP/Agilent needs to have given permission, in some way, for its release.

(3) Submissions should be compressed with ZIP or a similar scheme. Files
with .exe extensions will be rejected by my firewall, so choose carefully.
;-) Send submissions to kyrrin (at) bluefeathertech dot com.

(4) My spam filters have been known to be overzealous. If you try to send
me something, and it bounces, try sending me a note (WITHOUT the file
attachment) to kc7gr at qsl dot net (my backup address). Include the bounce
notice, and I will see what I can do about punching a hole in the filters so
you'll be able to get through.

For downloading, the host is ftp.bluefeathertech.com. Log on as 'anonymous'
(minus the quotes), and send your E-mail addy as your password. The path is
/pub/electronics/testgear, and you will see subdirectories below that.

Please keep in mind the following limitations.

(1) Download speed for anonymous logins is capped at 128K, and there's a
limit of five simultaneous logins permitted. This is done to make the most
of a limited bandwidth pipe (my DSL line). If you're not able to get in at
any one time, please try again later.

(2) Certain ISPs and IP ranges may be blocked from accessing the archive at
all, due to ongoing spam and virus transmissions. If you find you cannot
access the archive, drop me a note with details of what you're looking for.
If I have it, I'll send it along.

Thanks much.


-=-=-=-=-=-=-=-=-=-=-=-
Bruce Lane, Owner & Head Hardware Heavy,
Blue Feather Technologies --
kyrrin (at) bluefeathertech do/t c=o=m
"If Salvador Dali had owned a computer, would it have been equipped with
surreal ports?"


Re: HP8601A Generator/Sweeper

Robert Fincher
 

Hello again Andi,
Good choice. Though if that's all you intend to do,
then a marker gen might be superfluous. It's of more
use when you need to set up a pass-band across a
fairly wide frequency range, such as aligning a TV IF
strip (who does that these days) or in certain
high-end radio receivers where a broader intermediate
stage passband is required. I have a Racal RA17
receiver with such a stage, and the marker gen made
quite a difference. Basically, the marker inserts
"blips" on the screen trace at frequencies you select.
The 8600 has 5 markers, all of which can be set using
the digital frequency read-out. so you can assess very
quickly how the passband's shape is coming up. So, I'd
recommend the 8601 for a start - see how you like it,
and then look our for a marker gen if yiu think you
need it. In Australia, good 8601's are still
commanding high prices - typically in the range AUD200
to 500 (USD150 to 350). Good luck.
Rob
--- "Ahlbrecht, Andreas" <ahlbra@...>
wrote:
Hello Rob,

thanks for the quick answer!
I want to use the 8601 for alignment of broadcast
type radios.
However, I must admit that I, in theory, know how to
to just
that but am still not relly clear about what a
marker generator
exaclty is and how it works. I know that it can be
set up, for
example, to see exactly where the center frequency
is located
in the filter response curve on the 'scope, but
there it ends.

Can you or anyone else advise or point me to some
resource on
the net where this is explained? (Did a quick google
search
but wasn't successful.)

(You see, I'm an eager-to-learn hobbyist - not a
professional.)

Many thanks!
Andi

--- "Robert Fincher" wrote:
Hi there,
I've had an HP8601A for about 5 years. It worked
perfectly when I got it, and it still does. It's a
very nice instrument, and very useful for all manner
of things, particularly filter alignment. It's also
nice to set it up to measure the frequency response
of
your 'scope (up to 110 MHz). to get best use of the
8601, you should try and get its matching marker
generator - the HP8600. Makes testing a pleasure.
Cheers
Rob Fincher VK3BRF

--- "Ahlbrecht, Andreas" <ahlbra@...>
wrote:
Hi Group,

I'm about to get a HP8601A RF Generator / Sweeper
which I _think_ is basically working.
However, I would like to know if anyone has good
or
bad experience with this device.
(Is it prone to failure, does it perform well,
does
it need frequent re-calibration, ...)

Also I would like to ask if anyone has operator
and/or
service manual for this unit.

A quick response, if possible, would be very much
appreciated!
Find local movie times and trailers on Yahoo! Movies.


436A heads (8481A etc) and repair

Mark KAHRS
 

Just out of curiosity, has anyone attempted to do this at home?
Or is this another "Don't try this at home kids" kinda thing.


Re: HP 3505 Microwave Switch

 

Hi Lennart:

My guess is that it's a PIN diode switch. Get out your DMM and set it to the "diode" function and try all possible paths (both directions) between the case, input center pin and output center pin to see where the diode(s) are located. The most probable configuration is a couple of PIN diodes with their cathodes grounded and the anodes connected to both center pins. Applying a positive current (not a voltage) to the center pin on either side using a bias-T will turn on the diodes and give you some isolation y reflecting the power.

Have Fun,

Brooke Clarke, N6GCE

--
w/Java
w/o Java


Date: Fri, 17 Sep 2004 11:41:03 -0000
From: "odenman9" <odenman9@...>
Subject: Re: HP 3505 Microwave Switch

I have found out that the frequency range is 0.5 to 12 GHZ but I still
dont know how to drive it. I have uploaded a picture to the Photos section in this group
Regards
Lennart


New file uploaded to hp_agilent_equipment

 

Hello,

This email message is a notification to let you know that
a file has been uploaded to the Files area of the hp_agilent_equipment
group.

File : /HP-8559A-OM-SM.part2.rar
Uploaded by : avi49il <zipi-u@...>
Description : hp-8559-manual/part2

You can access this file at the URL:


To learn more about file sharing for your group, please visit:


Regards,

avi49il <zipi-u@...>


Re: HP 3505 Microwave Switch

 

I have found out that the frequency range is 0.5 to 12 GHZ but I still
dont know how to drive it. I have uploaded a picture to the Photos
section in this group
Regards
Lennart

--- In hp_agilent_equipment@..., "Roger Westenskow"
<hored@e...> wrote:
Have you received any replies on this switch?
I have quite a few HP units, and components.
If I had a bit of info on the switch you need I
may be of some help.
Do you have a picture, or page from a PDF file?
I will help if I can
Thanks
Roger / N7NGY
----- Original Message -----
From: odenman9
To: hp_agilent_equipment@...
Sent: Saturday, August 14, 2004 4:39 AM
Subject: [hp_agilent_equipment] HP 3505 Microwave Switch


Hi,
Is there anyone who has information about: HP 3505 Microwave
Switch?
I have not found it in any of my HP catalogues. It has an
uncommon
part number without an A or B.
Regards
Lennart



Yahoo! Groups Sponsor
ADVERTISEMENT





--------------------------------------------------------------------
----------
Yahoo! Groups Links

a.. To visit your group on the web, go to:


b.. To unsubscribe from this group, send an email to:
hp_agilent_equipment-unsubscribe@...

c.. Your use of Yahoo! Groups is subject to the Yahoo! Terms of
Service.