¿ªÔÆÌåÓý

Re: HP 8566B ROMs available


John Miles
 

¿ªÔÆÌåÓý

Interesting point.? I?just checked my service manuals, and it appears that there is indeed an?attenuator-correction ROM in the 8566B.? The HP?guys didn't cut any corners on this piece of gear!??It?is a separate part installed on A12, the front-panel keyboard assembly.? (A12U8, to be specific.)?
?
The zipfile I posted contains only the main operating firmware, from the 68000 CPU board.? I'm pretty sure that the images aren't serialized or otherwise tied to the specific instrument they came from.
?
-- john KE5FX

-----Original Message-----
From: elie shvarzman [mailto:eliesh@...]
Sent: Saturday, September 18, 2004 5:32 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP 8566B ROMs available

do'es its conteining the step attenuator correction table?
this corrections are depand on the preticuler step attenuator in the instrument so its different from instrument to instrument.