Geo,

These ¡°peaks¡± look to be counting noise. The difference between Kb1 and Kb2 is just 0.013 while the peak difference is 0.08 keV

Dud

From: [email protected] [mailto:[email protected]] On Behalf Of GEOelectronics@...
Sent: Saturday, October 24, 2020 6:15 AM
To: [email protected]
Subject: Re: [XRF] Si pin xrf ..1st go

?

Charles,

In the chart below, ignore the intensity%- it has no meaning in this context, but the X-Ray transition energies do.

.

?

Check out the different K lines on the chart, what you are seeing as 2 peaks in your scan probably is 2 peaks. Ka1 is the one to the right (higher energy side)

?

When you have a tall photopeak like in your scan, also look 1.74 keV below to see the silicon "Si Escapae Peak", Normally it is very small.

?

Mn-57

?

X-Ray Emissions:

?

? ? Energy? ???? ? ?Intensity ???? Assignment?

? ? ?(keV)? ??????? ? ? ? (%)? ? ??????? ? ? ? ? ? ??

?

? ? ?7.112???????? ? ? 0.0000? ? ??? ? ?Fe Kb4

? ? ?7.108???????? ? ? 0.0010? ? ??? ? ?Fe Kb5

? ? ?7.058???????? ? ? 1.8300? ? ??? ? ?Fe Kb1

? ? ?7.058???????? ? ? 0.9400? ? ??? ? ?Fe Kb3

? ? ?6.404???????? ? ?15.4000? ? ?? ? ?Fe Ka1

? ? ?6.391???????? ? ? 7.8000? ? ??? ? ?Fe Ka2

? ? ?6.266???????? ? ? 0.0000? ? ??? ? ?Fe Ka3

? ? ?0.792???????? ? ? 0.0160? ? ??? ? ?Fe Lb3

? ? ?0.792???????? ? ? 0.0100? ? ??? ? ?Fe Lb4

? ? ?0.717???????? ? ? 0.2000? ? ??? ? ?Fe Lb1

? ? ?0.708???????? ? ? 0.0018? ? ??? ? ?Fe Lb6

? ? ?0.705???????? ? ? 0.2600? ? ??? ? ?Fe La1

? ? ?0.705???????? ? ? 0.0290? ? ??? ? ?Fe La2

? ? ?0.628???????? ? ? 0.0220? ? ??? ? ?Fe Lh

? ? ?0.615???????? ? ? 0.0320? ? ??? ? ?Fe Ll

?


From: "Charles David Young" <charlesdavidyoung@...>
To: "XRF" <[email protected]>
Sent: Saturday, October 24, 2020 7:36:34 AM
Subject: Re: [XRF] Si pin xrf ..1st go

?

Using a convex mount for your buttons is very interesting.? I may have to try that.

?

Is your input rate always so high?? Mine is typically in the 10 to 20 range.? Have you done the Tune Slow/Fast Thesholds?

?

The attached plot shows a bifurcation of the FeKa1 peak that should go away with higher counts.

?

Charles

?

On Sat, Oct 24, 2020 at 5:07 AM Mike L. <loughlin3@...> wrote:

Taray,

Thanks for the post.? I am in the process of coming up with a similar Am241 device.? It is great that you found that Nickel line with your meteorite, that is useful information also.

Mike L.

?

Join [email protected] to automatically receive all group messages.