Re: 7623A vertical signal output


 

Mike,
On the similar 7603A, the VERT SIG OUT is specified as 0.5V per division of deflection into 1MEG, 25mV per division of deflection into 50 ohms.
--John Gord


On Sun, May 12, 2024 at 11:06 AM, mike zias / W3JRR wrote:
I am using the vertical signal output (back panel) on a 7623A. ?It seems to have a 2:1 attenuator. ?I am measuring the output with
a 1 meg input on another scope and find I get 1/2 the signal going into the 7623A vertical input. ?I haven't seen this specified in?
the online manual I use. ?This won't be a problem as long as it does it for all the inputs I select on the vertical plug-in. ?I will use Ch1, Ch1, and Add.
Is this attenuation normal and specified? ?

Join [email protected] to automatically receive all group messages.