Keyboard Shortcuts
ctrl + shift + ? :
Show all keyboard shortcuts
ctrl + g :
Navigate to a group
ctrl + shift + f :
Find
ctrl + / :
Quick actions
esc to dismiss
Likes
Search
7623A vertical signal output
I am using the vertical signal output (back panel) on a 7623A. ?It seems to have a 2:1 attenuator. ?I am measuring the output with
a 1 meg input on another scope and find I get 1/2 the signal going into the 7623A vertical input. ?I haven't seen this specified in? the online manual I use. ?This won't be a problem as long as it does it for all the inputs I select on the vertical plug-in. ?I will use Ch1, Ch1, and Add. Is this attenuation normal and specified? ? |
Mike,
toggle quoted message
Show quoted text
On the similar 7603A, the VERT SIG OUT is specified as 0.5V per division of deflection into 1MEG, 25mV per division of deflection into 50 ohms. --John Gord On Sun, May 12, 2024 at 11:06 AM, mike zias / W3JRR wrote: I am using the vertical signal output (back panel) on a 7623A. ?It seems to have a 2:1 attenuator. ?I am measuring the output with |
Hello John, I don't understand that specification. ?What division is being referenced? ?However, somehow that spec. must translate to the division I am seeing. ?
On Sunday, May 12, 2024, 06:23:42 PM EDT, John Gord via groups.io <johngord@...> wrote:
Mike, On the similar 7603A, the VERT SIG OUT is specified as 0.5V per division of deflection into 1MEG, 25mV per division of deflection into 50 ohms. --John Gord On Sun, May 12, 2024 at 11:06 AM, mike zias / W3JRR wrote: I am using the vertical signal output (back panel) on a 7623A. ?It seems to have a 2:1 attenuator. ?I am measuring the output with |
Mike,
toggle quoted message
Show quoted text
The screen is divided into 8 vertical divisions.? A full scale (8 division) peak-to-peak waveform on the screen should generate a 4 volt peak-to-peak signal at the VERT SIG OUT jack. --John On Sun, May 12, 2024 at 04:48 PM, mike zias / W3JRR wrote:
|
Hello Jon,
? I don't need Full BW so I use the 1 meg input.?? Just to be sure I have this straight: ? ?Using another scope to view Vert.Sig.Out, full scale on my 7623A (8div) will map to full scale on the 2nd scope when that one is set for 500 m/div; 8 divisions maps to 8 divisions.? Using a signal generator into the 7623A, ch.1, and the vert.sig.out to my 2465A.? Set the generator for a 5 volt pk (sine wave). 7623A 5v/div? ? 2465A 0.5 v/div ? ? 1 div? ? ? ? ? ? ? ? ? 1 div? ? ? ? 5v:0.5v = 10:1 division ? ? ?7623A 2v/div? ?2565A 0.5 v/div ? ? 2.5 div? ? ? ? ? ? ? ?2.5 div? ? ? 5v:1.25v = 4:1 division ? ? That seems to be a rather screwy system.? I understand dividing down to send a signal to other test equipment but when you are trying to duplicate the operation of a simple 1A1 this does give some challenges.? I would like to find a schematic of that output. |
On Thu, May 16, 2024 at 08:05 AM, mike zias / W3JRR wrote:
Hello Jon,==== Isn't 1A1 operating the same way? The gain is specified to be ~10 only when Ch1 is set to 5mV/cm. If you set Ch1 to a different attenuation factor output will be attenuated same way as what you are seeing on 7623A. BTW, your 2465A also has Ch2 output at the back with 10mV/div into 50-ohm. Ozan Here is the actual text from 1A1: "Signal¡ªOutput Voltage and Gain¡ªWith Channel 1 set for a deflection factor of 0.005 v/cm, output signal voltage is about 50 mv/cm or a gain of approximately 10."? |
Mike: VERT OUT is NOT intended to duplicate 1A1 or other old PU.
Its main use is to cascade VERT OUT to same scope other PU Vert CH 1 IN. See OP and SERV manual for scope you have for exact spec as it IS not uniform for every scope with VERT OUT. Often the setup specified 50 mV or 10 mV on second stge to achieve a 10X increaded overall scale factor. Exactly what are you trying to acomplish? Cheers Amicalment Jon |
Thanks Jon,
? ?Well, I think I have the normalization factor figured out. ?It does limit the usefulness. ?I am going back through my old papers and videos? from Dr. Middlebrook (went to his seminar in 1984 and have his his 20 hr video). ?So I built the 709 amplifier circuit he used and I wanted? to duplicate the measurements. ?So, when vx or vy is relatively small he liked to use A1vx, A2vy, and A1vx-A2vy. ?The A factors being selected vertical v/cm controls of the 1A1. ?I am using an HP wave analyzer as Middlebrook did. ?Since A1 and A2 will likely be quite different the 7623A? vert. out is problematic. ? |
Dear Mike, bravo for the duplication of Dr Middlebrooks techniques.
We are of same vintage as I knew Middlebrook and his collaborator Slobodan CUK very well. We had visited at their Caltech Pasadena lab. Another fine approach was by Dean Venable, a pioneer in injection plot and use of computers in 1980s. Had Dean visit for a demo, very impressive! Nowadays we use a differential probe or amp. We like AM503 but also have the Yokogawa x100 x1000 differential probe. Exactly what is your application? Have an absolutely fantastic day! Jon |
Hi Jon,
? ?I have a copy of Dean Venables book: "Stability Analysis Made Simple".? I bought that for $30 back in the Mid 1980's.? Retired for several years now, this work is just as a hobbyist.? I do still teach on online class in first year physics for New York Empire State University.? If I win the lottery I will look at a Solatron 1250 or similar.? Otherwise, I am on an Ebay budget.? I see Venable equipment for sale but usually as "parts only".?? Having been burned before on Ebay I avoid that class of equipment.? My 7623A will work as long as I keep both channels on the same scale.? Using different scales is going to require a lot of thought (I was hoping to avoid that - lol).? Have a great day. |
to navigate to use esc to dismiss