Keyboard Shortcuts
ctrl + shift + ? :
Show all keyboard shortcuts
ctrl + g :
Navigate to a group
ctrl + shift + f :
Find
ctrl + / :
Quick actions
esc to dismiss
Likes
Search
Help needed with no trace no beamfinder on 465 (not b)
I had been working on this 465 (SN B175976) for a few days. I had replaced a burned out input resistor on one channel and tightened a shaft coupling. After that it was more or less working, but I was checking something else (can't recall) when I noticed that suddenly I had no trace, and pressing the beamfinder shows nothing. So I suspect that I accidentally shorted something. A visual inspection shows nothing obviously burned or open.
My trouble shooting and electronic skills are pretty minimal, so I could really use some help fixing this. I do have another scope, a recently acquired 2213A. I'm using a 465 manual that is close but not exactly for my scope- the manual's Tek number is tek-070-1330-00 (some other manuals seem to cover a later model- B200000 and up). I can't find a trouble shooting chart in this manual that I read other people referring to. I have been reading through a past thread that seems related: /g/TekScopes/topic/7656979#126934, and using that as a starting point I have checked the following: All low voltage supplies are within tolerance, and ripple on each seems close to within tolerance. Fuse F1419 is good. Tantalum C1419 seemed fine, but I went ahead and replaced it with a 100uF 50v electrolytic. I do have a high voltage probe, so I checked TP1423, it measures only -82V (should be -2450V) Q1418 seems good. Following a comment by David Hess in the previous thread, I looked at the collector with my 2213A- it shows a slightly distorted wave approx 1.5 volts p-p (Horiz on my 2213A set to 10uS). It's base voltage measures about .47vdc using a VOM. David's comment was: "If you have a secondary working oscilloscope, the collector of Q1418 on schematic 10 will show if the high voltage inverter is operating." Albert suggested to check C1416- mine measures .48vdc, way off from what it should be. Alberts past comment was: "The best indication for output of the oscillator circuit is the voltage at C1416, shown as -4.5 V in the schematic. Roughly speaking the AC peak-peak voltage across the base winding in T1420 is twice this voltage. Since the average voltage across that winding is zero (neglecting the small resistive loss) you should also measure about -4.5 V at the base of Q1418. But a measurement load at the base might influence the oscillation." The collector of Q1404 measures about 3.3v, should be around 4.4v? I don't know what the above indicates, or what to do next. tmillerdems had said: "First thing I would do is lift the input lead from the HV multiplier that connects to T1420 transformer. That will isolate the HV multiplier that is known to cause HV problems. The regulation comes from sampling the -2450 volt cathode supply. You should still get some display if you get the cathode supply going sans PDA supply. " I don't see where/how to locate the "input lead from the HV multiplier that connects to T1420 transformer" - can I access that from the main board? Both T1420 and the HV multiplier are located beneath (and hidden by) the main board, correct? I also cannot locate CR1421 (using the image on page 160 of the manual). If either the T1420 or the HV multiplier are toast then that probably makes the scope unrepairable (at least by me, I'm guessing). Oddly enough, I also have 464 that I recently acquired with the same problem (no beam)- at least I didn't cause that one! Thanks if you can help, Keith Ostertag |
Hello Keith,
toggle quoted message
Show quoted text
you can also disconnect the ground strap of the HV multiplier, which pokes out from between the main board and the vertical amplifier board underneath the HV cover on the main board. Tt is soldered to the main board ground plane. Desolder it, and put it out of the way so it won't spark or anything.... wrap heat shrink on it or so. If the HV multiplier is toast, you should see a trace with it disconnected. I have replaced the HV multiplier on my 465. Removal of the vertical amplifier board is required to do this. The procedure is in the service manual and is fairly straightforward. I can't offer much help beyond this, because this is the only repair I know on Tek scopes ;-) My repair thread is here: /g/TekScopes/topic/7654225#107633 Hope this helps, Ian. On Wed, May 23, 2018 at 12:45 AM, Keith Ostertag <stuff@...> wrote:
I had been working on this 465 (SN B175976) for a few days. I had replaced --
One man's "magic" is another man's engineering. "Supernatural" is a null word. -- Excerpt from the notebooks of Lazarus Long, from Robert Heinlein's "Time Enough for Love" |
So... if I remove that ground strap, and I still don't see anything on the screen (not even using the beamfinder)... that suggests the HV multiplier is NOT the fault that's causing my problem?
Is this the ground strap you mean? /g/TekScopes/photo/49286/0?p=Name,,,20,1,0,0 Doesn't seem likely to me, as it isn't under the HV cover, but near the front end... but it is the only ground strap visible on that side of the main board on my scope. Here is a photo of the area under the HV cover, and I don't seen anything resembling a ground strap coming up between the "main board and the vertical amplifier board" as you describe: /g/TekScopes/photo/49286/1?p=Name,,,20,1,0,0 |
The ground strap you show is not the one. The 0 ohm resistor next to P1445 and the nylon nut is the ground for the HV multiplier.
toggle quoted message
Show quoted text
----- Original Message -----
From: "Keith Ostertag" <stuff@...> To: <[email protected]> Sent: Sunday, May 27, 2018 6:44 PM Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b) So... if I remove that ground strap, and I still don't see anything on the screen (not even using the beamfinder)... that suggests the HV multiplier is NOT the fault that's causing my problem? Is this the ground strap you mean? /g/TekScopes/photo/49286/0?p=Name,,,20,1,0,0 Doesn't seem likely to me, as it isn't under the HV cover, but near the front end... but it is the only ground strap visible on that side of the main board on my scope. Here is a photo of the area under the HV cover, and I don't seen anything resembling a ground strap coming up between the "main board and the vertical amplifier board" as you describe: /g/TekScopes/photo/49286/1?p=Name,,,20,1,0,0 |
You need to match the manual to your scope for starters.
toggle quoted message
Show quoted text
There are two different manuals ....;. what is the serial number on your scope? -DC manuals@... On 5/27/2018 7:30 PM, Keith Ostertag wrote:
Also, one manual says in order to isolate the HV multiplier one needs to unsolder "dummy resistor"... huh? illustration points to near R1449,R1447, R1448... but maybe that's also not relevant to my scope? --
Dave Manuals@... www.ArtekManuals.com |
My 465 SN is B175976.
I am looking at a manual marked "tek-070-1330-00" (236 pages, with yellow addendum pages added in the back), as well as a manual for later version 465's (070-1861-00, B250000 & up, says Rev B, June 1975, 291 pages). I think the first is the correct one, but it is more limiting (for instance, no trouble shooting flowchart). |
toggle quoted message
Show quoted text
----- Original Message -----
From: "Keith Ostertag" <stuff@...> To: <[email protected]> Sent: Sunday, May 27, 2018 9:24 PM Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b) My 465 SN is B175976. I am looking at a manual marked "tek-070-1330-00" (236 pages, with yellow addendum pages added in the back), as well as a manual for later version 465's (070-1861-00, B250000 & up, says Rev B, June 1975, 291 pages). I think the first is the correct one, but it is more limiting (for instance, no trouble shooting flowchart). |
PDA = Post Deflection Acceleration.
toggle quoted message
Show quoted text
Bruce On 28 May 2018 at 16:26 Keith Ostertag <stuff@...> wrote: |
P1445 is the 4 pin connector right next to the jumper that disconnects the HV multiplier ground. The PDA cable is the high voltage that goes to the front of the CRT tube. You just need to pull it our about an inch to unload the multiplier output.
toggle quoted message
Show quoted text
Once you do this and the -2450 VDC cathode comes back, you have found the bad multiplier. Regards, Tom ----- Original Message -----
From: "Keith Ostertag" <stuff@...> To: <[email protected]> Sent: Monday, May 28, 2018 12:26 AM Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b) Thanks! That's jumper P1445, I believe. What's the "PDA lead "? I don't recognize what PDA stands for... |
Hi Keith,
toggle quoted message
Show quoted text
as others have already shown, that's not the one. Apparently I was mistaken. The ground lead doesn't come out from between the two boards. Here's a picture of what it looks like when the vertical board is removed: The ground lead pokes through the board. My mistake, apologies. Others have already shown what to remove to isolate the HV multiplier. All of them have way more experience than I have. Good luck, Ian. On Mon, May 28, 2018 at 12:44 AM, Keith Ostertag <stuff@...> wrote:
So... if I remove that ground strap, and I still don't see anything on the --
One man's "magic" is another man's engineering. "Supernatural" is a null word. -- Excerpt from the notebooks of Lazarus Long, from Robert Heinlein's "Time Enough for Love" |
Yes the 1330 manual is the correct manual for your unit. while the 1861 manual can be useful at times, as it pertains to your unit? you need to be very careful . Some circuits in the later manual are very different and some are essentially unchanged from the earlier version.
toggle quoted message
Show quoted text
good luck -dc manuals@... On 5/27/2018 9:24 PM, Keith Ostertag wrote:
My 465 SN is B175976. --
Dave Manuals@... www.ArtekManuals.com |
Chuck Harris
Post Deflection Accelerator. It is the Anode cap at
toggle quoted message
Show quoted text
the edge of the CRT's funnel usually to the left of the screen. -Chuck Harris Keith Ostertag wrote: Thanks! That's jumper P1445, I believe. |
Is this the PDA? /g/TekScopes/photo/49286/2?p=Name,,,20,1,0,0
I was thinking it was the end of the HV mulitplier... |
Yes, that is the connector for the HV lead that goes to the CRT Post Deflection Accelerator (PDA) electrode. The PDA is a conducting coating on the inside of the tube funnel that starts after the horizontal deflection plates.. You should be aware that the outer coating of the tube forms a capacitor that holds a high voltage charge for a long time. If you disconnect the HV connector, be careful that the lead does not come in contact with any circuitry as it will damage it.
toggle quoted message
Show quoted text
----- Original Message -----
From: "Keith Ostertag" <stuff@...> To: <[email protected]> Sent: Monday, May 28, 2018 9:09 AM Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b) Is this the PDA? /g/TekScopes/photo/49286/2?p=Name,,,20,1,0,0 I was thinking it was the end of the HV mulitplier... |
Yes, that is the connector for the HV lead that goes to the CRT Post Deflection Accelerator (PDA) electrode. The PDA is a conducting coating on the inside of the tube funnel that starts after the horizontal deflection plates.. You should be aware that the outer coating of the tube forms a capacitor that holds a high voltage charge for a long time. If you disconnect the HV connector, be careful that the lead does not come in contact with any circuitry as it will damage it.
toggle quoted message
Show quoted text
----- Original Message -----
From: "Keith Ostertag" <stuff@...> To: <[email protected]> Sent: Monday, May 28, 2018 9:09 AM Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b) Is this the PDA? /g/TekScopes/photo/49286/2?p=Name,,,20,1,0,0 I was thinking it was the end of the HV mulitplier... |
Thanks you for that clarification!
OK, so I unplugged P1445 and the PDA lead, then powered up and gave it some time to warm up. Nothing on the screen even when beamfinder pressed, and TP1423 HV test point still reads about -87 volts. So... that means the HV mulitplier is not dead? Or at least it means something else is causing the problem? |
No, no, no. You do not unplug P1445. That stays put.
toggle quoted message
Show quoted text
Did you look at the picture I posted yesterday? I circled the part you need to UNSOLDER to open the ground to the HV Multi. ----- Original Message -----
From: "Keith Ostertag" <stuff@...> To: <[email protected]> Sent: Monday, May 28, 2018 11:25 AM Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b) Thanks you for that clarification! OK, so I unplugged P1445 and the PDA lead, then powered up and gave it some time to warm up. Nothing on the screen even when beamfinder pressed, and TP1423 HV test point still reads about -87 volts. So... that means the HV mulitplier is not dead? Or at least it means something else is causing the problem? |
to navigate to use esc to dismiss