Re: Help needed with no trace no beamfinder on 465 (not b)


 

Yes, that is the connector for the HV lead that goes to the CRT Post Deflection Accelerator (PDA) electrode. The PDA is a conducting coating on the inside of the tube funnel that starts after the horizontal deflection plates.. You should be aware that the outer coating of the tube forms a capacitor that holds a high voltage charge for a long time. If you disconnect the HV connector, be careful that the lead does not come in contact with any circuitry as it will damage it.

----- Original Message -----
From: "Keith Ostertag" <stuff@...>
To: <[email protected]>
Sent: Monday, May 28, 2018 9:09 AM
Subject: Re: [TekScopes] Help needed with no trace no beamfinder on 465 (not b)


Is this the PDA? /g/TekScopes/photo/49286/2?p=Name,,,20,1,0,0

I was thinking it was the end of the HV mulitplier...

Join [email protected] to automatically receive all group messages.