Re: Tektronix TDS8000 - Backup calibration parameters and other important stuff
Hi i am Charlyd on the evvblog i thought i jump in here i guess i missed this perticular part, because i am a member in groups.io here..and very interrested in exporting the eeprom content on the 7000/7000B first of all keep up the good work in exploring your equiptment.
@ozan you say and that is correct as far as i know. the cal. const. are in the eeprom. but the script extract it from the nvram how is that.... (sorry i send my reply in parts , because this is the third time i had to start retyping all here) |
Re: Is anyone out there still repairing refurbishing restoring 500 series oscilloscopes?
Hi John,
Thanks for offering to share. Do you have many 7000-series knobs and potentiometers? I'm also looking for a correct graticule, light filter and CRT bezel for my model 511 and model 422, which aren't exactly in the 500-series. If nothing else, I might have to make my own or use parts from similar models. I think I own or have seen just about all of the Tektronix analog scopes in a 60-mile radius, lol. I sure have made a lot of eBay purchases for how recent I got into the restoration hobby. I'd hate to know how much I've paid in just shipping for parts and tubes. |
Re: Fixing a vintage Tek 465M scope: Bad HV multiplier
On Sun, Jan 12, 2025 at 04:18 PM, richard ramacieri wrote:
Schematic shows 52V typical with 116V supply voltage, 56V is not excessive especially if your mains voltage is a bit above 120V. Ozan |
Re: Fixing a vintage Tek 465M scope: Bad HV multiplier
Unregulated voltage is higher than what us really "says". For example, the +5V unreg on my scope is around +11V, which then gets regulated to +5V. According to my service manual this is normal
For your scope, the service manual doesn't give a voltage for that, so I would recommend you check the line input settings. If those are good the 56V is probably normal. I would also recommend checking the regulation circuit for the +32V regulated as a short there can pop the fuse. Speaking of shorts, a shirt can by anywhere on the +32V rail, so you'll have to do some power tracing. Another question, is it blowing F558 or F736? If it's blowing F558, it's likely a problem with the HV transformer or the transistors/diodes that drive it. Benjamin |
Re: Fixing a vintage Tek 465M scope: Bad HV multiplier
Hi.. I have a 465m with HV section blowing the fuse... Found that the 32v unreg source voltage to be 56 volts !! ..A look at the diagram show the transfo, then a bridge and a 1000 mfd cap rated at 75 vdc...Is it normal to have 56 vdc instead of 32v unreg...I find this very unusual !! Please let me know..
|
Tek 465m..High voltage fuse blow (1/4 A).
Hi... My name is Richard, I recently acquired a 465m scope in mint condition..The problem is that the fuse in the high voltage PS burn instantly when the power is applied..Check all diodes, transistors, capacitor to be ok.. Didn't check the U555 so far... Voltage are ok, 5v and -5v and 32 volts, but as I went to check the 32 volts unreg use to supply the HV, I was surprise to see a reading of 56 volts.. As I look at the disgramm, this voltage come directly out of a bridge, filtered by a 1000mfd/75volts (wich voltage value is odd compare to 56 volts)..My question, is 56 volts normal at thispoint ?
|
Re: 067-1261-00
toggle quoted message
Show quoted text
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of W7HPW Radio via groups.io Sent: Friday, 10 January 2025 10:41 am To: [email protected] Subject: [TekScopes] 067-1261-00 Calibration Fixture Flexible Extender for 11000 series scopes, excess to my needs, like new I don't think it has ever been used, anybody want it? Mark |
Re: 7904A looking for good home
Hi all:
To add a bit to the surplus equipment discussion. My employer (large aerospace company) had a surplus store for decades, which had most everything from soup to nuts (almost literally). I was fortunate enough to acquire much of my Tek, HP, Fluke, and other test equipment from there for relatively reasonable, and sometimes fantastic prices. Like my 7834 got it for $65, and it works(!). Most of the test gear was either surplus to company needs, or had been replaced by something newer. Back around 2000, as I was apprenticing to be an Analog ASIC designer, the company purged a large number of electronics labs (“moving up the proverbial “value chain” just another bean counter cost cutting move), and huge amounts of test equipment would arrive for fairly good prices. One of my visits found me looking at an HP 213 pulse generator (the tube one with HV output), which an older couple had 2 of in their cart. Sadly, I was not able to convince them to let go of one of them. To me, it seemed they really did not know what they had, but oh well. Sadly, that surplus store went away not too long after, likely on someone’s hit list for cost reductions. It is now an online entity, but not suited to mere mortals as myself, as you need to buy in quantity ( a tub skid of wire cutters, let’s say), as opposed to singly as in the old physical store. Regards to everyone saving useful gear from wasteful landfills!! Randy Tek 555, 7834, 7633, 7623, HP, Fluke, Systron Donner, Analogic (Data Precision 6200 Analyzer), ESI, Power Designs, Dynatech Nevada, and miscellaneous. On Mon, Nov 25, 2024 at 10:57?AM Clark Foley via groups.io <clarkfoley= [email protected]> wrote: It’s always interesting to see where the thread of a topic goes. This one |
Re: probes for vacuum tubes scopes
Question to Jean Paul re: x-y distortion measurements method you mentioned.
I have an HP334 Distortion Analyzer, and in the HP manual, it tantalizingly mentions a way to view distortion products. BUT there are no connection details shown. I suspect it was left out in error. I’d really appreciate any info anyone has related to the x-y measurement technique with the HP 334 (or similar). This has baffled me for years. Thanks in advance, Randy Tek 555, 7834, 7633, 7623, 7D20, and a bunch of HP, Fluke, and other instrumentation. On Wed, Jan 1, 2025 at 6:28?PM Jean-Paul via groups.io <jonpaul= [email protected]> wrote: Probing the AE-35 unit in 2001 A space odyssey |
Re: Tds520 have a "striga" trouble.Need help!
Hi,
After long hours of measurements, cleaning etc. I had a final look at the acquisition board again and identified a suspiciuos via close to one of the infamous SMD caps ( see pic ). I traced it and it turned out it was eaten up by the electrolyte. After replacing the track with a wire the scope booted w/o errors and passed SPC. What do we learn ? Check all vias close to caps and pray they are not hidden underneath a part. The broken connection carries the TRIGCC"B" signal between U1551 pin 26 and U624 pin 14. One possible hint : SPC showed the following errors : 0x51efee4 (V main): TRIG"beta" trigGain[ CH1 ] 0.0 set to coldStart value 0x51efee4 (V main): TRIG"beta" dcOffset[ CH1 ] -1.6991698 set to coldStart value Best regards Hans |
Re: TDS544a logicTrigConfDiag STRIGA
Hi Simon,
Thanks for the response. After long hours of measurements, cleaning etc. I had a final look at the acquisition board again and identified a suspiciuos via close to one of the infamous SMD caps ( see pic ). I traced it and it turned out it was eaten up by the electrolyte. After replacing the track with a wire the scope booted w/o errors and passed SPC. What do we learn ? Check all vias close to caps and pray they are not hidden underneath a part. The broken connection carries the TRIGCC"B" signal between U1551 pin 26 and U624 pin 14. One possible hint : SPC showed the following errors : 0x51efee4 (V main): TRIG"beta" trigGain[ CH1 ] 0.0 set to coldStart value 0x51efee4 (V main): TRIG"beta" dcOffset[ CH1 ] -1.6991698 set to coldStart value Maybe you should give it another try ? Best regards Hans |
Tektronix TDS744A power supply repair - part help needed
Hello,
some time ago I have acquired TDS744A in unknown condition. Whet it arrived it didn't turn on. I have checked main fuse and it was blown. I have dismantled the scope and found that power supply had catastrophic failure, specifically in section of standby power supply. Both NTCs exploded and created complete mess. I was checking what else was bad, BU508A power transistor had B-C short, CR10, CR11 diodes and VR1 zener shorted. I have cleaned all dirt and swapped blown components, including VR4 and VR5 TVS diodes and famous C17 capacitor. However power supply was still not working. Then I have been checking T3 transformer and found that it is unfortunately broken. It has inter-winding short on primary winding (using my Bob Parker ring tester). I borrowed similar scope from my friend who was kind to allow me to temporarily desolder T3 transformer from his scope and use it in mine. After the swap my power supply started to work without any problem. This confirmed that really T3 is a culprit. I have also measured inductance, good transformer had 18mH on primary winding, bad one had just 1mH. I would like to ask your help if anyone has power supply for parts and would be willing to desolder and send T3 transformer (I will cover all costs). It has label 22900937 REV M? (it is small 3x3x3cm cube). I have been checking pictures on internet and the same transformer was used in several different Tektronix TDS scope power supply models, not just in TDS744A. Thank you very much. Vladimir |
Re: Tek 7704 scope cover
Will your 7704A cover fit on 79xx 'scopes? If so, I'd be interested in
having several (4 or 5) of them. DaveD KC0WJN On Sun, Jan 12, 2025 at 00:40 evan foss via groups.io <evanfoss= [email protected]> wrote: Hi Folks, |
Re: DPO/DSA 70000 PSU Repair
Hello Guy, i have one, with floating malfunction sometimes it starts, sometimes it goes to fail (red light) state. With/without load no matters.
It's starts 100% when i reconnect the hot line, till green led blinks. I'm trying to add capacitance, on hot and on cold sides, on controller side, on FET driver side, looks like that all electrolytic caps are OK. Also I tested them for ESR. It's quite complicated, i have no such experience on diagnosing that class of PSU. Can you help me with diagrams /oscillograms ? I have also a topic Tektronix DPO\DSA 70000 series Artesyn AFE 1200 Power Supply repair. On eevblog.com Best, Maksim |
Tek 7704 scope cover
Hi Folks,
I made a 3D printed cover for the front of my Tek 7704A. I was worried about accidentally harming the switches while I move it off/on the shelf. I could clean it up and share it but that would mean a large reprint to test it and I wanted to gauge interest before wasting that much plastic and electricity. PS: I am planning to do one for the 11801 as well. Though I will probably have to line that with copper tape or something. Evan |
Re: Tektronix SCD5000 Schematics
Alex,
There might be a few bread crumbs to follow to get you somewhere. It has been 35 years since I have seen an SCD5000 but I believe that the power supply assembly (the box) is similar or the same as used in the 1130x(A) scopes. The SCD5000 has additional and unique converter boards to develop other voltages but the assembly with the fan housing was initially the same as the 1130x(A). Consult the early 1130x schematics at the Tekwiki page. You might be able to determine if it is a match or a close cousin. Clark |
Re: Is anyone out there still repairing refurbishing restoring 500 series oscilloscopes?
Hi all. Been some time since I have replied to anything here.
Just a quick note to say that I have a nearly pristine 555 purchased years back for a ridiculously low price (local pickup for me). It has the purple phosphor (high speed, for photos?). Beautiful clean trace, but then, only 33MHz. I’m not able at the moment to do any electronics (medical issues) but hopefully that will change in the next 9 months or so. I would be interested in parts such as the panel mounted ground jacks. I also have a few plugins: CA, D, W, 1A7, 1A5. Love the 500 scopes- true works of art. Btw: I also have 7623, 7633, and 7834 scopes - love them too! On Thu, Dec 19, 2024 at 1:59?PM Jack Reynolds via groups.io <jackandladyreynolds@...> wrote: I thought it meant people eating tasty Animals1 |