¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io
Date
Re: App Notes
Gary, About 6-7 years ago I found these application notes, or most of them, online. Some are definitive tutorials and give procedures directly applicable to the J16 and its many probes. Pg 7 of the
By Ron · #145230 ·
Re: What calibration items to buy.
From easy to difficult: Almost all timebase calibration can be done with a 10 MHz function or signal generator and frequency counter. The fastest sweeps with magnification require a 100 MHz or faster
By David Hess · #145229 ·
Re: Replacement of Tantalum cap removal of U450??
The back of the board connects to the vertical CRT deflection pins. The little collet sockets pull straight off of the pins but you still have to be very carefu
By David Hess · #145228 ·
Re: 7623A + 7A26 + 7B53
Two timebases can indeed generate a raster in these mainframes. A raster is also useful for verifying the condition of the CRT phosphor. A 7B53A can also be used as a calibrated vertical amplifier in
By David Hess · #145227 ·
Re: Typo in 2901 SM or wrong value resistor?
Great, thanks!! Now after replacing 80% of the resistors, I can finally put this thing back together!! -Brian
By Brian Bloom · #145226 ·
Re: Typo in 2901 SM or wrong value resistor?
Manual error to be sure ...in my manual the parts list say 100 and the schematic says 1K ( base of Q366) ..my money is on 1K being correct. At 100 ohms I am guessing would pull the base voltage to low
By Dave / NR1DX · #145225 ·
Re: Typo in 2901 SM or wrong value resistor?
It is a typo in the parts list. On the schematic it is 1k. The part number is for a 1k (315-0102-00 is a 1k). Regards.
By Tom Miller · #145224 ·
Typo in 2901 SM or wrong value resistor?
I'm replacing out-of-tolerance resistors in my 2901 time mark gen and ran across a 1k resistor where the SM says it should be 100R. The resistor in question is R364 on the divider board directly
By Brian Bloom · #145223 ·
Re: Cheap differential probe ?
Hello, May or may not be a fit for your application or budget, but I saw this on page 69 of the March 2018 RSGB magazine: Micsig DP10013 100 MHz high voltage differential probe, about GBP150 Also
By Scott Scheirman · #145222 ·
Replacement of Tantalum cap removal of U450??
Hi, Those of you who were following my "7603 working ... then blank" Thread will know that the problem was a shorted C492 Tanatalum cap on the Vertical Amp board. To replace this cap, I could simply
By tinkera123 · #145221 ·
Re: What calibration items to buy.
If the SG is leveled, then it can check frequency response. There are several 7000 series plugins (not 7xxx) that can do this, with differing frequency responses. That ought to allow you to use an
By Harvey White · #145220 ·
Re: Tek manuals saved from the dumpster - free for postage
Hi guys, It would be better if you expressed interest via private email rather than on the list. I subscribe in digest form so direct emails make it easier for me to keep track of who expresses
By Steve Goldstein · #145219 ·
Re: Can I install a 2465B motherboard in a 2445B chassis? What will I get?
A5 ROM has nothing to do with calibration -- that is stored in NVRAM -- but it adds those higher sweep rates that 2445B doesn't have i.e. it changes the scope identity. It is no biggie -- you simply
By Sergey Kubushyn · #145218 ·
New file uploaded to [email protected]
Hello, This email message is a notification to let you know that a file has been uploaded to the Files area of the [email protected] group. *File:* TG501 schematics.pdf *Uploaded By:* Robert Hay
By [email protected] Notification <TekScopes+notification@...> · #145217 ·
Re: Tek manuals saved from the dumpster - free for postage
On Sat, 3 Mar 2018, Steve Goldstein wrote: I would take one SD-24 set and SD-14 Service Reference if they are still available. > Here are some manuals I saved from the heap over the years. I don't
By Sergey Kubushyn · #145216 ·
Re: Can I install a 2465B motherboard in a 2445B chassis? What will I get?
Thanks for the answer.? That is what I expected but needed and expert opinion.? I don't mind the problem with the A5 ROM because it will probably need replacement anyway for battery problems.?
By machineguy59 · #145215 ·
Tek manuals saved from the dumpster - free for postage
Here are some manuals I saved from the heap over the years. I don't need them, do you? Strongly prefer local pickup (I'm a bit north of Boston, MA) but will ship by Media Mail within the US.
By Steve Goldstein · #145214 ·
Re: What calibration items to buy.
OT: Although not all that one might want for calibrating scopes, a good start would be a reasonably good Rf signal generator, not necessarily PG or SG modules ... Some sellers on eBay appear to
By garp66 · #145213 ·
Re: Can I install a 2465B motherboard in a 2445B chassis? What will I get?
It will give you a full featured 2465B, not a poor man's one. All parts other than A1 board and front panel label are exactly the same for 2445B and 2465B. Ah, the A5 ROM is different too. --- *
By Sergey Kubushyn · #145212 ·
Re: What calibration items to buy.
That (and price) is why I guess I don't have either. Oh well. Do have the PG506 and TG601, as well as an SG502 and 503. The 504 has eluded me. Harvey
By Harvey White · #145211 ·