Tektronix 465B horizontal display issue, and a fix
I'll try to keep this brief, but this is an end to end diagnosis and repair description. Maybe this will help someone in the future who has a 465B or similar scope with a similar problem? Experienced
By
Keith
·
#208600
·
|
Re: Tektronix 2213A CRT question
The info on TekWiki is obviously wrong.
154-0838-00 was used only 2213 and 2215.
154-0861-00 was used in 2213A, 2215A, 2235 and 2236.
Check out this old post on the same subject:
By
zenith5106
·
#208599
·
|
Re: Tektronix 2213A CRT question
Hi Dave,
Can't answer your question on the substitution,
but I believe that I have a good used CRT for the scope that you
have. Let me know off the list if you would like it and we'll make a
deal.
By
Kim Herron
·
#208598
·
|
Re: No Power, 2246 Oscilloscope
It is a well known issue in the 224x series scopes power supply that any original rectifier diodes marked ZM or ZS tend to get leaky or fail over time. There is a total of 17 diodes that should be
By
Alex
·
#208597
·
|
Re: 2784
Hi all,
and thank you Mark and Gerald for the files and the hints. I found out, that the EMI Filter is blown. But I think it's not so easy to repair, because it's filled with epoxy.
Perhaps I will
By
Jürgen W
·
#208596
·
|
Re: SC503 Interesting Wave Form
The input is a sine wave. The wave form cause is a single component.
By
Barry Breaux
·
#208595
·
|
Re: No Power, 2246 Oscilloscope
That didn’t take you very long. You get an A+.
By
WA2SCL_Jeff
·
#208594
·
|
Re: Is there an alternative to the 151-0703-00 NPN's from the 492 power supply board?
Hi chaps.
I suppose this is the final update, as found a shorted 47?F 20v cap (C6033) on the A62 board, and now the 492 appears to be working again (although I don't yet know how to use it, to
By
James55
·
#208593
·
|
Re: Replacement Feet for Tektronix 465, 465B, 466, 468, 475 and 475A Oscilloscopes
Hi Benny,
I'm interested in replacing the rear feet on my 'ol Tektronics 475. Do you still print them? Price on 4?
Thanks
Stan W8SRD
By
sdegroff
·
#208592
·
|
Tektronix 2213A CRT question
Hello All,
Back Story:
Bought a cheap 2213A off Marketplace with fuzzy display issue,
Fixed issue by replacing focus resistors,
discovered screen damage in that there was trace distortion in lower
By
TugboatDave
·
#208591
·
|
Re: I need help fixing a Tek 3B3 time base
#chat-notice
Are the voltage’s adjusted with the VR’s on the plug-in Or is it an adjustment inside the scope itself?
By
henrybanneker@...
·
#208590
·
|
Re: I need help fixing a Tek 3B3 time base
#chat-notice
What are???
Nothing quoted for context.
Glenn
--
-----------------------------------------------------------------------
Glenn Little ARRL Technical Specialist QCWA LM
By
Glenn Little
·
#208589
·
|
SC503 Interesting Wave Form
Hello,
I posted a picture in photo of a waveform being drawn by an DC502. I asked if anyone wanted to guess the cause. I know the answer, but take a guess if you want. I will post the component
By
Barry Breaux
·
#208588
·
|
Re: Need Parts - Tex 468 Scope
Chris,
I could use the case if it’s available.
I’m in Wichita Ks 67212
Craig
By
Craig Cramb
·
#208587
·
|
Re: Need Parts - Tex 468 Scope
Where are you located? I think I have a complete (maybe working?) 468 in the pile of extras.
By
Chris Wilkson
·
#208586
·
|
Re: I need help fixing a Tek 3B3 time base
#chat-notice
Is it adjusted using some of the VR‘s on the plug-in
By
henrybanneker@...
·
#208585
·
|
Re: I need help fixing a Tek 3B3 time base
#chat-notice
Yes they are
By
henrybanneker@...
·
#208584
·
|
Re: Tek TDS684A hacks and screen swap advice please?
Great reply from Jared, covers pretty much everything. I would add, unlocking 1GHz means you probably need to re-do parts of the calibration. Also I question the actual usefulness of that extra
By
fenugrec
·
#208583
·
|
Re: Tek TDS684A hacks and screen swap advice please?
According to both the manual and catalog from that time the TDS684A
doesn't have any options that can be unlocked. Also the full 1GHz bandwidth
applies only to attenuator settings 10 mV/div and
By
zenith5106
·
#208582
·
|
Re: Tek TDS684A hacks and screen swap advice please?
The TDS684A is definitely a 1GHz scope, running at 5G Samples/sec.
Here's info on replacing the CRT with an
By
Jared Cabot
·
#208581
·
|