Re: Tektronix 2246 - Text Readout Broken
In order to replace the resistor, I think I'll need to remove everything so I can access the back:
On instruments with serial numbers below B040909, examine the circuitry near A10U802 (the Horizontal
By
Aneesh Raghavan
·
#208557
·
|
Re: Is there an alternative to the 151-0703-00 NPN's from the 492 power supply board?
Hi James,
The PSU schematic is f.i. at pdf page 225 492/492P Spectrum Analyzer Volume 2 Service manual ( 070-3784-01Tekwiki).
Albert
By
Albert Otten
·
#208556
·
|
Re: TDS210
??? On my TDS220 I changed the backlight lamp to a LED strip.
??? Mark Jordan, PY3SS
By
Mark Jordan
·
#208555
·
|
TDS210
Hi All
Anyone know what the correct voltage is at the input of L6 on the Power Supply. Its the supply circuit for the display CCFL. I'm getting -4.6v.
My display is a dull shade of orange with no
By
Bob Kendrick
·
#208554
·
|
Photo IMG_2525.jpeg uploaded
#photo-notice
The following photos have been uploaded to the SC503 - interesting waveform! ( /g/TekScopes/album?id=300884 ) photo album of the [email protected] group.
* IMG_2525.jpeg (
By
Group Notification <noreply@...>
·
#208553
·
|
Re: 278x calibration.
Hi Gerald,
after several years, it seems I found the right help for my 2784 Tektronix Spectrum Analyzer.
I would be very interested in your Processor Extender Card with serial Port.
Please let me
By
Jürgen W
·
#208552
·
|
Re: 278x Spectrum Analyser
Hi Gerald
I am repairing a Tektronix 2784. Therefore i'm searching for the Processor ExtenderBoard with the SerialPort or a Circuit Diagram for this.
I know it's a 85C30 Dual Serial Interface from
By
Jürgen W
·
#208551
·
|
2784
Hi all,
I am repairing a Tektronix 2784. Therefore i'm searching for the Processor ExtenderBoard with the SerialPort or a Circuit Diagram for this.
I know it's a 85C30 Dual Serial Interface from Zilog
By
Jürgen W
·
#208550
·
|
Re: Is there an alternative to the 151-0703-00 NPN's from the 492 power supply board?
Thank you Mark.
I can actually source these locally.
Looking at the board, the pin layout isn't as obvious as I thought and I haven't been able to find any kind of schematic for the 492 PSU so
By
James55
·
#208549
·
|
Re: A bewitched voltage divider?
I tried simply using two VOMs. The presence of the HV probe does not change the read value at U1540 pin 4 even one millivolt.
So there are two possibilities: the HV probe does not say the truth or the
By
Gianni Becattini
·
#208548
·
|
Re: Is there an alternative to the 151-0703-00 NPN's from the 492 power supply board?
James,
A BUV48A will work and is heavier capacity. Mouser has these in stock for $6,76.
Mark
By
Mark Vincent
·
#208547
·
|
Re: Is there an alternative to the 151-0703-00 NPN's from the 492 power supply board?
An NPN is an NPN is an NPN, no?
Therefore, it must be possible to use a more common package of transistor and bodge a solution to this unobtainium nonsense?
Ok, we don't know the Gain, but we do
By
James55
·
#208546
·
|
Is there an alternative to the 151-0703-00 NPN's from the 492 power supply board?
Hi all.
Can't seem to find these anywhere.
According to Sphere, these are,
'HARD to find 49xP Series Inverter Transistors! - 2N6587, TO-61, NPN, 10A/450V' - (SOLD OUT)
Is these a
By
James55
·
#208545
·
|
Re: A bewitched voltage divider?
Greetings, all.
It occurs to me that maybe this is a simpe loading problem.
The HV probe is likely to have a resistnace of the order of 100M. With 1600V, the current is 16uA — not
By
David Knight
·
#208544
·
|
Re: TekCATS
THANKS!
By
Gianni Becattini
·
#208543
·
|
Re: TekCATS
It is still important that an 11300 non-A with F/W 3.1 is treated as an A-scope in CATS.
http://www.hakanh.com/dl/docs/hardtofind/11k_CATS_2.19.pdf
/H?kan
By
zenith5106
·
#208542
·
|
Re: TekCATS
Sorry, my fault. It works. I was running in the wrong directory.
By
Gianni Becattini
·
#208541
·
|
Re: TekCATS
Thank you Hakan,
I tried but its still says “LOG ERROR-> CAN’T FIND EIS RCS VERSION NUMBER”
Actually this is not set and the EIS RCS control variable says NOT SET.
Is there a way to set it?
By
Gianni Becattini
·
#208540
·
|
Re: TekCATS
Select 11302A and ignore the error message and it will work.
/H?kan
By
zenith5106
·
#208539
·
|
TekCATS
Hi, it's me again... :)
I was going to go on with the 11302 checking and tried to connect to TekCATS 2.19, but in the meantime I had updated the firmware to the 3.1 version and TekCATS refuses to
By
Gianni Becattini
·
#208538
·
|