¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io
Date

Re: Cleaning dirty pots on 5xx series plug-ins

 

--- In TekScopes@..., tubesnthings@... wrote:

Dan;
There's a bunch of material on this subject in the forum's archives.
More was added, recently. Check it out.
Bernd
OK. Sounds interesting. I see links to "Files", "Photos", "Database", and others but no "archives." Would someone please point me to the forum archives.
Thanks,
Tom


Re: WTB: Readout IC (155-0021-01), or entire readout board

Rob
 

¿ªÔÆÌåÓý

Yep Good Luck. It wasn¡¯t too long ago that a list of vendors was delineated and discussed here. Wish I could remember. You may want to run a limited search for say the last 9-12 months. Anyway, Talon is worth remembering as well as a Greek company and of course classic Tek page¡­. and ¡­..and¡­

?

Well you get the point. I hope this finds you well.

Rob

?

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of chrismh_somenumber
Sent: Thursday, February 28, 2013 10:55 PM
To: TekScopes@...
Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout board

?

?

Thanks, I didn't know about that site. I checked Sphere and Tucker.

I ordered one from SMC. Hopefully he has it.

Chris

--- In TekScopes@..., "Rob" wrote:
>
> Looks like the chip can be obtained here:
>
>
>
>
>
>
>
> I did not search further at the other "normal parts vendors" (Sphere, Talon,
> etc.)... No particular association with SMC in a monetary way. I am however
> a satisfied customer and I do chat with the owner from time to time.
>
>
>
> Anyway, not 100% sure he has 155-0021-01 left nor how his price compares to
> others out there. Personally I have done enough business with SMC that if he
> has it I stop my search and purchase just based on my time being valuable
> and results of enough price shopping I am satisfied.
>
>
>
> Anyway, hopefully helpful. Sorry for rambling on so.
>
> Rob
>
>
>
> From: TekScopes@... [mailto:TekScopes@...] On Behalf
> Of chrismh_somenumber
> Sent: Thursday, February 28, 2013 8:35 PM
> To: TekScopes@...
> Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout
> board
>
>
>
>
>
> The readout board part # is 670-1900-XX. Looks like it was used in lots of
> scopes, not just those I listed.
>
> Thanks
>
> --- In TekScopes@... ,
> "chrismh_somenumber" wrote:
> >
> > By swapping in the above IC from another scope, I've determined that the
> custom IC on my 7844 has died.
> >
> > Would anyone happen to have one they would part with, or have an entire
> working readout board from a 7844/7904A?
> >
> > Thanks,
> > Chris
> >
>


Re: WTB: Readout IC (155-0021-01), or entire readout board

chrismh_somenumber
 

Thanks, I didn't know about that site. I checked Sphere and Tucker.

I ordered one from SMC. Hopefully he has it.

Chris

--- In TekScopes@..., "Rob" <rgwood@...> wrote:

Looks like the chip can be obtained here:







I did not search further at the other "normal parts vendors" (Sphere, Talon,
etc.)... No particular association with SMC in a monetary way. I am however
a satisfied customer and I do chat with the owner from time to time.



Anyway, not 100% sure he has 155-0021-01 left nor how his price compares to
others out there. Personally I have done enough business with SMC that if he
has it I stop my search and purchase just based on my time being valuable
and results of enough price shopping I am satisfied.



Anyway, hopefully helpful. Sorry for rambling on so.

Rob



From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of chrismh_somenumber
Sent: Thursday, February 28, 2013 8:35 PM
To: TekScopes@...
Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout
board





The readout board part # is 670-1900-XX. Looks like it was used in lots of
scopes, not just those I listed.

Thanks

--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com> ,
"chrismh_somenumber" wrote:

By swapping in the above IC from another scope, I've determined that the
custom IC on my 7844 has died.

Would anyone happen to have one they would part with, or have an entire
working readout board from a 7844/7904A?

Thanks,
Chris


Re: WTB: Readout IC (155-0021-01), or entire readout board

Rob
 

¿ªÔÆÌåÓý

Looks like the chip can be obtained here:

?

?

I did not search further at the other ¡°normal parts vendors¡± (Sphere, Talon, etc.)... ?No particular association with SMC in a monetary way. I am however a satisfied customer and I do chat with the owner from time to time. ?

?

Anyway, not 100% sure he has 155-0021-01 left nor how his price compares to others out there. Personally I have done enough business with SMC that if he has it I stop my search and purchase just based on my time being valuable and results of enough price shopping I am satisfied.

?

Anyway, hopefully helpful. Sorry for rambling on so.

Rob ?

?

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of chrismh_somenumber
Sent: Thursday, February 28, 2013 8:35 PM
To: TekScopes@...
Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout board

?

?

The readout board part # is 670-1900-XX. Looks like it was used in lots of scopes, not just those I listed.

Thanks

--- In TekScopes@..., "chrismh_somenumber" wrote:
>
> By swapping in the above IC from another scope, I've determined that the custom IC on my 7844 has died.
>
> Would anyone happen to have one they would part with, or have an entire working readout board from a 7844/7904A?
>
> Thanks,
> Chris
>


Re: WTB: Readout IC (155-0021-01), or entire readout board

chrismh_somenumber
 

The readout board part # is 670-1900-XX. Looks like it was used in lots of scopes, not just those I listed.

Thanks

--- In TekScopes@..., "chrismh_somenumber" <hogancm@...> wrote:

By swapping in the above IC from another scope, I've determined that the custom IC on my 7844 has died.

Would anyone happen to have one they would part with, or have an entire working readout board from a 7844/7904A?

Thanks,
Chris


WTB: Readout IC (155-0021-01), or entire readout board

chrismh_somenumber
 

By swapping in the above IC from another scope, I've determined that the custom IC on my 7844 has died.

Would anyone happen to have one they would part with, or have an entire working readout board from a 7844/7904A?

Thanks,
Chris


Re: Could please any one give a clue about this nice pulsed line generator??

 

BNC have poor performance above 1.8 to 2GHz so they may slow the risetime of
the pulse. SMA is a much better choice, and this is what Tek used in the
later versions of the HAPG. But at the time it was originally built SMA was
probably not available. GR874 is capable of >8GHz. Soldering the RG-147
directly to the PC board is OK, but long lengths of coax make poor delay
lines (charge lines). They suffer from "dribble up" effects. The HAPGs use
special copper hard line for their internal delay lines. If the RG-147 is
short (a few feet) it "dribble up" won't matter.

Read the Theory of Operation in the manual to learn how a length of charge
line works. The pulse width is going to be twice as long as the length of
the line since the pulse has to travel out to the end and return to the
beginning before the avalanche, which is using the charge stored in the
line, can turn off. If you know the velocity of propagation of the
electrical signal in the particular 50 Ohm coax you are using (it must be 50
ohm) and you multiply the length times the velocity of propagation and then
double the result that should give you an idea of how wide the pulse will
be.

Dennis

-----Original Message-----
From: iglesia_cristiana_arpas_eternas, Sent: Thursday, February 28, 2013
3:22 PM

Hi Albert..probably I put some meters of RG147 as fixed(solder) internal
line in order to obtain a 100nS pulse width.
What is the secret about this GR connectors ?, I have the temptation to
replace by a BNC, any wrong with this change?.
Gabriel.
--- In TekScopes@..., "Albert" <aodiversen@...> wrote:

Correction: in my early XF unit the charge line is soldered to the board.
Probably a peltola jack has been removed there. In my later XH unit the
board and the line have SMA connectors.
Albert

--- The cable has a peltola connector at one end. I have another unit in
which the cable is soldered to the board.
---
Albert


Re: Could please any one give a clue about this nice pulsed line generator??

 

Hi Albert..probably I put some meters of RG147 as fixed(solder) internal line in order to obtain a 100nS pulse width.
What is the secret about this GR connectors ?, I have the temptation to replace by a BNC, any wrong with this change?.
Gabriel.

--- In TekScopes@..., "Albert" <aodiversen@...> wrote:

Correction: in my early XF unit the charge line is soldered to the board. Probably a peltola jack has been removed there. In my later XH unit the board and the line have SMA connectors.
Albert


--- The cable has a peltola connector at one end. I have another unit in which the cable is soldered to the board.
---
Albert


Re: Could please any one give a clue about this nice pulsed line generator??

 

I will be surprised if there is a manual since it does not even have a part
number.
Sometimes you have to settle for the next best thing.
Dennis

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of iglesia_cristiana_arpas_eternas
Sent: Thursday, February 28, 2013 3:18 PM
To: TekScopes@...
Subject: [TekScopes] Re: Could please any one give a clue about this nice
pulsed line generator??


Very useful info indeed..thanks a lot Dennis.
I'm not feel "safe men" until obtain the right info about..I just like to
have "all" the service manual of my gear.
I was send a mail to Artekmanual , but not response yet.
Regards Gabriel.
--- In TekScopes@..., "Dennis Tillman" <dennis@...> wrote:

I think you are being overly concerned about your purchase not working.
These are rugged instruments and not likely to fail. The PV509 manual
is close enough to tell you everything you will need to know about
your purchase. But here is some additional information on the HAPG
that I kept from years ago.



From: "Craig Sawyers" <c.sawyers@>

Date: Tue May 4, 2004 2:25 pm

Subject: Wierd Pulse generator craig_sawyers



Hi Group



Well I took delivery of the strange "High Amplitude Pulse Generator".
This

is a TM500 style plug-in, and I bought it kind of out of curiosity,
since I

had seen nothing like it. It also has no identifying type number at
all -

so it was either a development unit, or an internal use only test
fixture -

hopefully one of the ex-Tek guys on the list will recognize it - I've
put

some pics on the yahoo gallery.



This is the only thing I've seen that uses the strange attenuator that
the

7A29 uses. This has a huge bandwidth - way above 1GHZ, and probably
*way*

above so that it doesn't cramp the 7A29's style. I suspect that the
7A29

has a 1.5GHz bandwidth, so the attenuator is possible approaching 5GHz.



It also has a rigid charge line inside, estimated at around 3m (10') long.

This line isn't secured at all, and is prevented from shorting to the
back

of the PCB with a layer of self-adhesive plastic film - very
prototype-ish.

Date on the PCB is copyright 1978, and component date codes indicate
late

1982 build.



The unit is basically a pre-trigger pulse generator that uses an
avalanche

transistor to give a fast rising pulse. The pre-trigger can be
adjusted with a front

panel control from 50-125ns early, so that a sampling time base gets
time to

crank itself into action.



Using a 7104/7A29 gives a measured 410ps rise, implying a rise time of
the

pulse gen itself is 240ps (because the 7104/7A29 has a measured rise
time of

330ps).



Preshoot is 2%, and overshoot is un-measurably small, the very
slightest

ripple on the pulse top. One of the cleanest rises I have ever seen.
Front

panel says rise time <=300ps. Pulse length as expected (from the
internal

charge line) is about 30ns and the rep rate is 110kHz. Amplitude can
be

adjusted from 250mV to 25V in switched steps, with a variable control too.



Pretrigger gives a 2.5V pulse 250ns long with a 1.4ns rise.



I'll try it out on a sampler (S4) later and get a better grip on the
pulse

rise and what the pre-trigger does on a 7T11. I always find triggering
a

7T11 from external sources a bit fiddly, not surprising when you read
the

manual in detail and see that each triggering mode has very different

frequency and amplitude characteristics.



This is a comparison between the "Unknown Unit" and a Type 111
pre-trigger

pulse generator:



111

===

Fixed output level of at least 10V, not precise

<500ps rise

Switchable polarity

2ns pulse length without external charge line

Rep rate variable from 10Hz to 100kHz

Pre-trigger 30-250ns early, 300ns duration, <5ns rise, +10V



The unknown one

===============

Switched and variable output level from 250mV to 25V

<300ps rise

Fixed polarity

30ns pulse length without external charge line

Fixed rep rate of 110kHz (probably 100kHz nominal)

Pre-trigger 50-125ns early (measured 30-130ns), 250ns duration, 1.4ns
rise,

+2.5V

From: TekScopes@... [mailto:TekScopes@...] On
Behalf Of iglesia_cristiana_arpas_eternas
Sent: Tuesday, February 26, 2013 3:26 PM



Here the link to this almost new pulsed generator, at first glance is
a transmission line pulser with 300ps ..but no model or number at all.
I just brought it with the hope work right...but a good service manual
will be very appreciated and gratefully.
Link to Pulse Gen.
<
232112
187?autorefresh=true&pt=LH_DefaultDomain_0&hash=item3a7e99ce3b&nma=tru
e&si=w
4%252BkkAkTYIq9sTkPYLC8H98HEpU%253D&orig_cvip=true&rt=nc&_trksid=p2047
675.l2
557>

Regards Gabriel.

_._,___



------------------------------------

Yahoo! Groups Links


Re: Could please any one give a clue about this nice pulsed line generator??

 

Very useful info indeed..thanks a lot Dennis.
I'm not feel "safe men" until obtain the right info about..I just like to have "all" the service manual of my gear.
I was send a mail to Artekmanual , but not response yet.
Regards Gabriel.

--- In TekScopes@..., "Dennis Tillman" <dennis@...> wrote:

I think you are being overly concerned about your purchase not working.
These are rugged instruments and not likely to fail. The PV509 manual is
close enough to tell you everything you will need to know about your
purchase. But here is some additional information on the HAPG that I kept
from years ago.



From: "Craig Sawyers" <c.sawyers@>

Date: Tue May 4, 2004 2:25 pm

Subject: Wierd Pulse generator craig_sawyers



Hi Group



Well I took delivery of the strange "High Amplitude Pulse Generator". This

is a TM500 style plug-in, and I bought it kind of out of curiosity, since I

had seen nothing like it. It also has no identifying type number at all -

so it was either a development unit, or an internal use only test fixture -

hopefully one of the ex-Tek guys on the list will recognize it - I've put

some pics on the yahoo gallery.



This is the only thing I've seen that uses the strange attenuator that the

7A29 uses. This has a huge bandwidth - way above 1GHZ, and probably *way*

above so that it doesn't cramp the 7A29's style. I suspect that the 7A29

has a 1.5GHz bandwidth, so the attenuator is possible approaching 5GHz.



It also has a rigid charge line inside, estimated at around 3m (10') long.

This line isn't secured at all, and is prevented from shorting to the back

of the PCB with a layer of self-adhesive plastic film - very prototype-ish.

Date on the PCB is copyright 1978, and component date codes indicate late

1982 build.



The unit is basically a pre-trigger pulse generator that uses an avalanche

transistor to give a fast rising pulse. The pre-trigger can be adjusted with
a front

panel control from 50-125ns early, so that a sampling time base gets time to

crank itself into action.



Using a 7104/7A29 gives a measured 410ps rise, implying a rise time of the

pulse gen itself is 240ps (because the 7104/7A29 has a measured rise time of

330ps).



Preshoot is 2%, and overshoot is un-measurably small, the very slightest

ripple on the pulse top. One of the cleanest rises I have ever seen. Front

panel says rise time <=300ps. Pulse length as expected (from the internal

charge line) is about 30ns and the rep rate is 110kHz. Amplitude can be

adjusted from 250mV to 25V in switched steps, with a variable control too.



Pretrigger gives a 2.5V pulse 250ns long with a 1.4ns rise.



I'll try it out on a sampler (S4) later and get a better grip on the pulse

rise and what the pre-trigger does on a 7T11. I always find triggering a

7T11 from external sources a bit fiddly, not surprising when you read the

manual in detail and see that each triggering mode has very different

frequency and amplitude characteristics.



This is a comparison between the "Unknown Unit" and a Type 111 pre-trigger

pulse generator:



111

===

Fixed output level of at least 10V, not precise

<500ps rise

Switchable polarity

2ns pulse length without external charge line

Rep rate variable from 10Hz to 100kHz

Pre-trigger 30-250ns early, 300ns duration, <5ns rise, +10V



The unknown one

===============

Switched and variable output level from 250mV to 25V

<300ps rise

Fixed polarity

30ns pulse length without external charge line

Fixed rep rate of 110kHz (probably 100kHz nominal)

Pre-trigger 50-125ns early (measured 30-130ns), 250ns duration, 1.4ns rise,

+2.5V

From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of iglesia_cristiana_arpas_eternas
Sent: Tuesday, February 26, 2013 3:26 PM



Here the link to this almost new pulsed generator, at first glance is a
transmission line pulser with 300ps ..but no model or number at all.
I just brought it with the hope work right...but a good service manual will
be very appreciated and gratefully.
Link to Pulse Gen.
<
187?autorefresh=true&pt=LH_DefaultDomain_0&hash=item3a7e99ce3b&nma=true&si=w
4%252BkkAkTYIq9sTkPYLC8H98HEpU%253D&orig_cvip=true&rt=nc&_trksid=p2047675.l2
557>

Regards Gabriel.

_._,___


Re: Cleaning dirty pots on 5xx series plug-ins

 

¿ªÔÆÌåÓý

Dan;
There's a bunch of material on this subject in the forum's archives.
More was added, recently. Check it out.
Bernd
?
In a message dated 2/27/2013 8:25:32 A.M. Pacific Standard Time, kaboomdk@... writes:

?

Howdy folks,

? I have another cleaning question. ?I've got a bunch of type L and type CA (and other) 5-series plug ins and the variable attenuator pots have become "dirty". ?They are intermittent over much of their range, and usually not useful in the "cal" position.

? ?Is there any Tek-net wisdom particular to these pots I should follow in cleaning them? ?What is their failure mechanism?

? I have cleaned these with tuner cleaner or other freon based sprays in the past, and the effect is definitely not permanent as I find I have to clean them aga in (albeit a few years later). ?Is there any benefit to taking them out of the plug-ins, sonicating them in a cleaning bath and re-installing them? ?Anything I should avoid?

? Thanks again for the help.

? Dan


Re: Could please any one give a clue about this nice pulsed line generator??

 

¿ªÔÆÌåÓý

I think you are being overly concerned about your purchase not working. These are rugged instruments and not likely to fail. The PV509 manual is close enough to tell you everything you will need to know about your purchase. But here is some additional information on the HAPG that I kept from years ago.

?

From: "Craig Sawyers"

Date: Tue May 4, 2004? 2:25 pm

Subject: Wierd Pulse generator? craig_sawyers

?

Hi Group

?

Well I took delivery of the strange "High Amplitude Pulse Generator". This

is a TM500 style plug-in, and I bought it kind of out of curiosity, since I

had seen nothing like it. It also has no identifying type number at all -

so it was either a development unit, or an internal use only test fixture -

hopefully one of the ex-Tek guys on the list will recognize it - I've put

some pics on the yahoo gallery.

?

This is the only thing I've seen that uses the strange attenuator that the

7A29 uses. This has a huge bandwidth - way above 1GHZ, and probably *way*

above so that it doesn't cramp the 7A29's style. I suspect that the 7A29

has a 1.5GHz bandwidth, so the attenuator is possible approaching 5GHz.

?

It also has a rigid charge line inside, estimated at around 3m (10') long.

This line isn't secured at all, and is prevented from shorting to the back

of the PCB with a layer of self-adhesive plastic film - very prototype-ish.

Date on the PCB is copyright 1978, and component date codes indicate late

1982 build.

?

The unit is basically a pre-trigger pulse generator that uses an avalanche

transistor to give a fast rising pulse. The pre-trigger can be adjusted with a front

panel control from 50-125ns early, so that a sampling time base gets time to

crank itself into action.

?

Using a 7104/7A29 gives a measured 410ps rise, implying a rise time of the

pulse gen itself is 240ps (because the 7104/7A29 has a measured rise time of

330ps).

?

Preshoot is 2%, and overshoot is un-measurably small, the very slightest

ripple on the pulse top. One of the cleanest rises I have ever seen. Front

panel says rise time <=300ps. Pulse length as expected (from the internal

charge line) is about 30ns and the rep rate is 110kHz. Amplitude can be

adjusted from 250mV to 25V in switched steps, with a variable control too.

?

Pretrigger gives a 2.5V pulse 250ns long with a 1.4ns rise.

?

I'll try it out on a sampler (S4) later and get a better grip on the pulse

rise and what the pre-trigger does on a 7T11. I always find triggering a

7T11 from external sources a bit fiddly, not surprising when you read the

manual in detail and see that each triggering mode has very different

frequency and amplitude characteristics.

?

This is a comparison between the "Unknown Unit" and a Type 111 pre-trigger

pulse generator:

?

111

===

Fixed output level of at least 10V, not precise

<500ps rise

Switchable polarity

2ns pulse length without external charge line

Rep rate variable from 10Hz to 100kHz

Pre-trigger 30-250ns early, 300ns duration, <5ns rise, +10V

?

The unknown one

===============

Switched and variable output level from 250mV to 25V

<300ps rise

Fixed polarity

30ns pulse length without external charge line

Fixed rep rate of 110kHz (probably 100kHz nominal)

Pre-trigger 50-125ns early (measured 30-130ns), 250ns duration, 1.4ns rise,

+2.5V

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of iglesia_cristiana_arpas_eternas
Sent: Tuesday, February 26, 2013 3:26 PM

Here the link to this almost? new pulsed generator, at first glance is a transmission line pulser with 300ps ..but no model or number at all.
I just brought it with the hope work right...but a good service manual will be very appreciated and gratefully.
?

Regards Gabriel.

_._,___


Re: 7k Readout Board Swap

chrismh_somenumber
 

Thanks for the info. I'm going to give it a shot this evening.

Chris

--- In TekScopes@..., David <davidwhess@...> wrote:

The list I have says that the 7844 and 7104 use the same readout board
which is Tektronix part number 670-1900-XX so as far as I know, you
can just swap the boards.

There is a PDF called Tektronix7xxx_Assemblies.pdf in the files
section which lists which assemblies are used in which oscilloscopes.

On Thu, 28 Feb 2013 03:45:16 -0000, "chrismh_somenumber"
<hogancm@...> wrote:

I have a 7844 that works fine except for the complete lack of readout.

In looking at voltages and signals around the U2126 timer chip, all is OK except for pins 7 and 10. I have a junker 7104 that uses the same IC (and its readout works). I was going to swap the 7844's IC into the 7104, but the 7104's is soldered in.

I'm wondering if I can just swap the entire 7844 board into the 7104.

I have looked at the circuit diagrams for both and they are largely the same. The signals in and out are the same, they even use the same jumper designations.

Has anyone tried this before?

Thanks,
Chris


Re: Could please any one give a clue about this nice pulsed line generator??

Albert
 

Correction: in my early XF unit the charge line is soldered to the board. Probably a peltola jack has been removed there. In my later XH unit the board and the line have SMA connectors.
Albert

--- The cable has a peltola connector at one end. I have another unit in which the cable is soldered to the board.
---
Albert


Re: Could please any one give a clue about this nice pulsed line generator??

Albert
 

Hi Gabriel,

Your HAPG version is a rather early one. I have one with a sticker (not the original label) saying SN160. The boards are exactly the same version XF, with the same hand written signature GUB. The only difference in board components seems to be that yours has a variable R75. Further mine has an internal open ended charge line. You might make the same modification, with the advantage that you can use the GR connector to export the pretrigger signal to the front panel. Simply reconnect the GR connector to the empty socket (about 2 cm away from Q60, to the rear side). Pretrigger lead time can be adjusted with R75. Best avalanching can be set by R100. R10 sets -18V IIRC. C102 influences rise time and/or aberrations.
The internal charge line is a rigid 20 ns cable, probably very high quality. It might be difficult to find something like that. The cable has a peltola connector at one end. I have another unit in which the cable is soldered to the board. I suppose the internal charge line was a later modification, copied from the PG509.
The most vulnerable component in your unit might be the attenuator. Same problems as in the 7A29.

Albert


Re: 7000 Series Interface Socket Cover Changing

Tom Jobe
 

?
Thanks George,
This morning I was thinking that I should add more comments to those pictures while it was all fresh in my mind, so I?edited them to explain things a bit better.
Turns out that the character limit for the photo's name?is fairly short but the photo's description can be a little longer.
tom jobe...
?
?

----- Original Message -----
From: Glydeck
Sent: Thursday, February 28, 2013 7:10 AM
Subject: Re: [TekScopes] Re: 7000 Series Interface Socket Cover Changing

?

Tom,

Great photos and an excellent addition to the documentation of the 7K scope.

Thanks!

George

On Feb 27, 2013, at 7:30 PM, "Tony Jacobson" <tonyman@...> wrote:

?

Indeed!? Very timely for me, I think.

I just bought a used 7k rigid extender () that appears to be missing these plastic side covers.? Can anyone verify if the connector on the 067-0589-00 rigid extender is the same as t hose used on the 7k mainframe backplane?

If so, does anyone have a spare pair of these covers they might part with for a reasonable price?

Thanks,

Tony

(By the way, I have been lurking on here for a couple of years now.? I think this is my first serious post.? Thanks to all of y ou for sharing your invaluable knowledge.? I am learning a lot!)

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of keithostertag
Sent: Wednesday, February 27, 2013 7:11 PM
To: TekScopes@...
Subject: [TekScopes] Re: 7000 Series Interface Socket Cover Changing

?

Good photos Tom! Very helpful, useful addition to the archives!

Keith Ostertag


Re: 7000 Series Interface Socket Cover Changing

 

Tom,

Great photos and an excellent addition to the documentation of the 7K scope.

Thanks!

George

On Feb 27, 2013, at 7:30 PM, "Tony Jacobson" <tonyman@...> wrote:

?

Indeed!? Very timely for me, I think.

?

I just bought a used 7k rigid extender () that appears to be missing these plastic side covers.? Can anyone verify if the connector on the 067-0589-00 rigid extender is the same as t hose used on the 7k mainframe backplane?

?

If so, does anyone have a spare pair of these covers they might part with for a reasonable price?

?

Thanks,

?

Tony

?

(By the way, I have been lurking on here for a couple of years now.? I think this is my first serious post.? Thanks to all of y ou for sharing your invaluable knowledge.? I am learning a lot!)

?

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of keithostertag
Sent: Wednesday, February 27, 2013 7:11 PM
To: TekScopes@...
Subject: [TekScopes] Re: 7000 Series Interface Socket Cover Changing

?

?

Good photos Tom! Very helpful, useful addition to the archives!

Keith Ostertag


Re: Quality of FG 502 Waveforms

 

Try different sweep speeds to see what the envelope looks like, then try line triggering to see if it will lock to the envelope. If it does, then supply ripple or ground loops are the likely problem. If it does not, then look for oscillations in the power supplies.

I would suspect the "20V" supply you mentioned, but first check the manual to see if it's supposed to be regulated or not. Sometimes "raw" supplies are labeled with the nominal voltage under certain conditions, may have lots of ripple, and are subject to drastic change - all normal. If it's a regulated supply, it should stay very close to nominal under a wide range of conditions, have very little line frequency ripple, and should certainly be adjustable to the right setpoint value if there's a pot for that purpose.

Bad electrolytic capacitors are the prime suspects whenever supply ripple problems arise. In this case, I would blame a power supply fault until you get that 20V right - especially since you've already found a failed part in that section. On TM500s, also try different plug-in compartments to make sure it's not a mainframe pass transistor or connector problem.

Ed

--- In TekScopes@..., "Philip" <ndpmcintosh@...> wrote:

I have been studying and working with an FG 502 function generator. I recently went through the adjustment procedure and was able to do everything that did not require a distortion analyzer or spectrum analyzer. Actually I did those adjustments too but eyeballed the sine shape and used my best judgement as what gave the best appearance.

I am not convinced it is operating properly though. I uploaded three picture and perhaps someone who has experience with this unit can tell me if this is as good as it gets or if there is an indication of a fault or two somewhere.



The closeup of the sine shows (barely) that it appears to be composed of a number of signals closely bunched together.

The low frequency square shows that there are a number of tops stacked upon each other. This does not appear at high frequency.

Finally, the closeup of the high frequency square shows what it looks like at high frequency.

I also am unable to adjust the 20V power supply below 21.9 V. I replaced one cracked resistor on the front end of the power supply and am continuing to troubleshoot that problem.

Thanks for any insights.

Phil...


Re: Quality of FG 502 Waveforms

 

I was going to suggest checking both power supply outputs for ripple
with the AC range of a multimeter or an oscilloscope set to AC
coupling but I think you found the problem. That fuzzy output is
exactly what I would expect with AC ripple on the power supply.

I assume you mean one of the diodes in bridge rectifier CR600 is bad?
There is nothing special about the diodes. I would definitely double
check C620 and C680 out of circuit with something more than an ESR
meter.

I would fix the ripple before worrying about not being able to adjust
the voltage via R625 but this is one of those cases though where I
might use a variac to bring the input voltage up gradually while
watching the +20V and -20V regulated outputs to make sure they go into
regulation. This is not one of Tektronix's better power supply
designs. U615 or maybe U664 could have been damaged.

On Thu, 28 Feb 2013 02:45:22 -0000, "Philip" <ndpmcintosh@...>
wrote:

I don't have a differential comparator to check the ripple in the way the manual specifies, but I did throw a 1x probe on it and saw a huge amount of what looks to be sawtooth ripple. That is why I have been focusing on the power supply. The filter capacitors check good and I suspect perhaps a problem with the +20V supply rectifier.

The scope and TM500 are indeed plugged into the same outlet via extension cord and outlet strip.

Phil...


Check the power supply rails for ripple. A fat horizontal trace like those images show is typical of power supply ripple. Loose grounds and shield connections also can cause it, but if your shields and scope are all properly connected, ripple is the most likely problem.
Are your DM500 unit and your scope plugged into the same outlet? If not, you might be experiencing a ground loop. Plug both the TM500 and the scope into the same AC receptacke and see if the trace clears up.

Cheers,
Dave M


Re: Quality of FG 502 Waveforms

 

--- In TekScopes@..., "Philip" <ndpmcintosh@...> wrote:

I have been studying and working with an FG 502 function generator. I recently went through the adjustment procedure and was able to do everything that did not require a distortion analyzer or spectrum analyzer. Actually I did those adjustments too but eyeballed the sine shape and used my best judgement as what gave the best appearance.

I am not convinced it is operating properly though. I uploaded three picture and perhaps someone who has experience with this unit can tell me if this is as good as it gets or if there is an indication of a fault or two somewhere.



The closeup of the sine shows (barely) that it appears to be composed of a number of signals closely bunched together.

The low frequency square shows that there are a number of tops stacked upon each other. This does not appear at high frequency.

Finally, the closeup of the high frequency square shows what it looks like at high frequency.

I also am unable to adjust the 20V power supply below 21.9 V. I replaced one cracked resistor on the front end of the power supply and am continuing to troubleshoot that problem.

Thanks for any insights.

Phil...
I don't know the magnitudes of the waveform amplitudes (noise and signal pk-pk), but it looks like a classic ground loop problem to me: TM500 mainframe grounded through ground pin in power cable, as the scope is too. Ground through the BNC cable makes one huge! ground loop. Any dynamic magnetic field passing through the loop induces a current into what is essentially a shorted transformer secondary turn. Current is only limited by impedance of the conductors, and could reach a few dozen amps!. This induces a voltage across the impedances, with the largest possibly being the BNC cable ground. Ground at the front panel of the FG502 does not equal ground at the vertical input connector of the scope. Resulting potential difference is superimposed on the displayed waveform.

If you have a differential amplifier, try using it, without plugging the BNC cable into to it. Connect the + input to the BNC cable connector center pin, and the ¨C input to the BNC ground shell. If you don't have a diff amp, try quasi differential using two scope channels, added with one inverted. Make sure both as set to the same V/div.

- Steve