Re: Help with R7903 needed, some success...
Hi Keith,
Yes, that CRT heater current should have a trapezoid or square shaped waveform. When looking at the schematics, the 6.3V CRT heater supply is generated through transformer TR1310B (schematics #11) This secondary side gets driven by the inverter circuitry in the power supply, see schematics # 9. As there is current flowing, at least the CRT heaters seem to be intact and can be ticked off your check list. Good Luck, Magnus |
Re: 7000 series on screen graphics question 7704A versus 7633
--- In TekScopes@..., "Brian J. Henry" <bhenry@...> wrote:
Brian, thanks for the reply, would you know if the later board is retro fittable to my earlier scope? Thanks. |
Re: 7000 series on screen graphics question 7704A versus 7633
13/07/2012 11:27
Is the astigmatism adjustment internal on these David? If I look closely it almost appears to be trying to show two numerals / letters at the same time, possibly the figure 1 (one)? As it's otherwise 100% and new old stock I'd quite like to get this facet of its operation fixed. Thanks. -- Best Regards, Chris Wilson. mailto: chris@... |
Re: Help with R7903 needed, some success...
Albert
Hi Keith,
toggle quoted message
Show quoted text
And there is also no or hardly any response to the Intensity pot?? Note that the sweep must be running for Intensity effect, otherwise TP1876 will remain at the low 9V level. That Vgk = -142 V would suppress the beam for sure. Can you check for proper continuity in P1675? CR1676 to slider R1674, and CR1680 to TP1876 (via 10k). Maybe you can measure the DC voltages from P1675 pins 2 and 3 to ground during 'scope operation? (This is "safe", not at the -3kV level). Pin 2 should reflect Grid Bias setting, and pin 3 the Intensity setting at TP1876. That 142 V seems to be more than the theoretical extreme (130V - 9V). That's why I think of a fault in P1675. I suppose the neons DS1687 and DS1688 are invisible? These might glow and limit Vgk to -142 V. Albert Hi Albert- With my DMM neg probe on the -3kV TP and the pos probe on the bottom of R1719 (goes to pin 3 of CRT) I get about -142 volts. This is with the Readout off. |
Re: I have a type 585 oscope how much is it worth
sipespresso
There are many reasons a person might value a scope. When a person cares about the beauty of an instrument and what it represents historically, its utility as an oscilloscope for everyday use becomes a secondary factor, if a factor at all.
toggle quoted message
Show quoted text
Also, a lot of interesting concepts are embodied in old scopes. For example, distributed amplifiers, equivalent-time analog sampling, tunnel diode triggering, and storage CRTs. These things tickle my mind. I own some scopes just so I can marvel at them. Basically, they are Faberge eggs for an electronics enthusiast. That said, I do like to bring my old scopes into proper calibration. Sometimes a decision has to be made, whether to functionally restore or just clean up a scope. Historic integrity is likely to be lost in a functional restoration of an early Tek scope. It is a tough decision. I have no rigid position on the issue. -Kurt For example, I recently got a 514. I'm pretty sure that I'd need to do extensive replacement of internal parts to get it working properly. It is never clear what's the right thing to do in this situation and every person has to make their own decision. For the 514, I'm going to leave it alone. Other times, I couldn't resist the urge to do functional restoration. My 567 was just too interesting not to repair. So I replaced the electrolytics and got it fully operational. I wouldn't blame somebody if they chose, in the same situation, to leave the scope unmolested. --- In TekScopes@..., "Steve" <ditter2@...> wrote:
|
Re: 7000 series on screen graphics question 7704A versus 7633
The readout font on my 7623B and 7934 oscilloscopes looks nearly identical to that of the 7704A. The 7623B and the 7934 use a newer design of the readout board that generates the characters using a single custom ROM IC instead of the several different custom ICs used on the earlier board. With the exception of my 7854, all of my other 7000 series oscilloscopes have a readout font that looks similiar, but much sharper to that of the 7633. I believe that all of the very late production scopes had the newer readout board installed, but I can't confirm that.
Brian Henry |
Re: TDS 684a keeps turning ON and OFF
Hi Jay,
toggle quoted message
Show quoted text
Thanks for your suggestion but it didn't work. I am pretty sure power supply is the problem. Maybe something has to do with start up circuit, since power supply tries to start and then shuts off. One thing I noticed is that the voltage on the input bulk cap (470uF, 450V) is not stable and jumps ups and down(should have been stable DC). Maybe something has to do with power factor correction circuit. not sure need to do further investigation. Maybe something not working in the OV and OC protection circuit. mickey --- In TekScopes@..., "jayw_comark" <jayw_comark@...> wrote:
|
Re: 7603 Help for jumping readouts
In the mainframe cal procedure, specifically for the vertical output
toggle quoted message
Show quoted text
amplifier, look for an adjustment called thermal compensation. It may help. It would be a good idea to check the HF response if this is tweaked.. -ls- "photo692002" <photo692002@...> wrote: My R7603 has developed a problem where the readouts jump around a lot |
Re: 7603 Help for jumping readouts
Rob
I have had similar on a few of my mainframes. More pronounced in some with
toggle quoted message
Show quoted text
7A12's and the traces DC offset/positioned high or low on the CRT , etc.. (I have not 100% decided it's the same issue). In any event, I got marked improvement in the most gross case by fixing the loose card slot side. Not sure what they're called but the little plastic pieces that go on both sides of the card slot. In the other case I had not tightened the readout card screws well so a grounding issue I can but assume. That said they still all still do it with the 7A12's (I actually have 4 that do the same thing). I therefore assume that there is some feedback pathway that exists in general between verticals and the Mainframe but TEK fixed it in other/later plug-ins. I can but assume that there may be a fix on the mainframe side as well. As a suggestion, perhaps look at your serial number and later ones with specific emphasis on any cap size changes on the readout card, etc. Decoupling caps in particular? In any event, I will be highly interested in what the peoples with the true expertise have to say about this one. As I truly am only sharing experience vs. any real insight I know. I hope this finds you well Rob -----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf Of photo692002 Sent: Thursday, July 12, 2012 7:07 PM To: TekScopes@... Subject: [TekScopes] 7603 Help for jumping readouts My R7603 has developed a problem where the readouts jump around a lot based on the input signal frequency and amplitude, signal trace is fine. With no input, readouts are steady. This readout problem occurs with several different vertical and horizontal plugins, problem appears to be in the 7603 frame. Power supplies measure well and look good on scope, dont think they are involved. I have been unable to determine if the readout board or the vertical or horizontal amp board is involved as I do not have spares. Has anyone seen this problem and resolved it? Larry ------------------------------------ Yahoo! Groups Links |
Re: 7000 series on screen graphics question 7704A versus 7633
Looks like the astigmatism needs adjusting on the 7633
toggle quoted message
Show quoted text
Dave -----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf Of chrismwilsonuk Sent: 13 July 2012 00:17 To: TekScopes@... Subject: [TekScopes] Re: 7000 series on screen graphics question 7704A versus 7633
|
Re: Tek 468 problem
Hi Jimmy, and thank you for that information. Checked U571 - no problem, checked U558 op-amp faulty waveform present. In between is U352, a LF351 op-amp, this proved to be faulty and the source of the oscillation. New LF351 and scope now working fine again. Just needs a good service and thorough clean. Im going to keep this one, its a lovely bit of kit and is so useable for the things Im interested in. Thanks to everyone who took time to read this thread.
toggle quoted message
Show quoted text
Bill --- In TekScopes@..., "jtjewell83" <jtjewell83@...> wrote:
|
Re: Help with R7903 needed, some success...
This is almost exactly what I saw when I measured the heater current
on my 7904 with a P6021 set at 10mA/mV and 5mV/div. Peak to peak was 23mV on mine so 230mA peak to peak. On Thu, 12 Jul 2012 18:02:12 -0000, "keithostertag" <keitho@...> wrote: Still trying to confirm whether the CRT is good or not, or why there's no activity from it. |
Re: I need CRT for 305DMM
Bob Koller
Martin, I think I have one. I will look within the next few days. From: Martin Mehlhose To: tekscopes@... Sent: Thursday, July 12, 2012 3:13 PM Subject: [TekScopes] I need CRT for 305DMM
?
I need a CRT for my Tek 305DMM greetings from germany Martin |
7603 Help for jumping readouts
My R7603 has developed a problem where the readouts jump around a lot based on the input signal frequency and amplitude, signal trace is fine. With no input, readouts are steady. This readout problem occurs with several different vertical and horizontal plugins, problem appears to be in the 7603 frame.
Power supplies measure well and look good on scope, dont think they are involved. I have been unable to determine if the readout board or the vertical or horizontal amp board is involved as I do not have spares. Has anyone seen this problem and resolved it? Larry |
Re: 7000 series on screen graphics question 7704A versus 7633
Rob
The 7633 shown does look like a storage mode is on. The one I have is a bit
toggle quoted message
Show quoted text
sharper, etc. in non-storage mode about the same in storage mode/s. That said, the storage capable scopes in general (7633 and 7844) are not as sharp on the readout functions as the non-storage counter parts. (they are however for sure better than yours...except again in storage mode/s..... not sure what to make of that) Hopefully somewhat helpful. Rob -----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf Of larrys@... Sent: Thursday, July 12, 2012 6:33 PM To: TekScopes@... Subject: Re: [TekScopes] 7000 series on screen graphics question 7704A versus 7633 The 7633 is normal 7K R/O, looks like the astig may be a bit off. The 7704A doesn't look like the one I have, nor the one I had as a bench scope in my Tek days. More like a 7854 than 7K R/O. -ls- David DiGiacomo <daviddigiacomo@...> wrote: You messed up the second link, but anyway, it's because the 7633 has awrote:
------------------------------------ Yahoo! Groups Links |
Re: Help with R7903 needed, some success...
keithostertag
Hi Albert- With my DMM neg probe on the -3kV TP and the pos probe on the bottom of R1719 (goes to pin 3 of CRT) I get about -142 volts. This is with the Readout off.
toggle quoted message
Show quoted text
If I am reading the manual and the troubleshooting guide correctly, the control grid is normally 40-50 volts more negative than the cathode supply, and may cutoff if 65 volts more negative. Correct? So maybe the bias is so far off that it is blanking the CRT? Adjusting R1674 moves it only about +- 2 volts. Hmm... If this is true, does that point to the HV board A18? That's a very difficult board for access to components. Just randomly testing a few components on the HV board I haven't found anything obviously bad, yet. Your thoughts? Thanks, Keith Ostertag --- In TekScopes@..., "Albert" <aodiversen@...> wrote:
|
Re: 7000 series on screen graphics question 7704A versus 7633
I've seen what appear to be two "styles" of 7704A RO.? Earlier ones (actually, most) seem to have larger, "fatter" characters, while the newest ones tend to be much smaller, tighter, and do look like the second example.? It seems to be more than just adjustments.? I wonder if the change in "font" could be traced to different revisions of the RO board. ? -Dave
From: larrys@...
To: TekScopes@... Sent: Thursday, July 12, 2012 4:32:35 PM Subject: Re: [TekScopes] 7000 series on screen graphics question 7704A versus 7633 ? The 7633 is normal 7K R/O, looks like the astig may be a bit off. |
Re: 7000 series on screen graphics question 7704A versus 7633
The 7633 is normal 7K R/O, looks like the astig may be a bit off.
toggle quoted message
Show quoted text
The 7704A doesn't look like the one I have, nor the one I had as a bench scope in my Tek days. More like a 7854 than 7K R/O. -ls- David DiGiacomo <daviddigiacomo@...> wrote: You messed up the second link, but anyway, it's because the 7633 has a |
Re: 7000 series on screen graphics question 7704A versus 7633
--- In TekScopes@..., Chris Wilson <chris@...> wrote:
and Sorry for broken link, and thanks for the reply |
Re: 7000 series on screen graphics question 7704A versus 7633
You messed up the second link, but anyway, it's because the 7633 has a
toggle quoted message
Show quoted text
storage tube. On Thu, Jul 12, 2012 at 4:28 PM, Chris Wilson <chris@...> wrote:
|