¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io
Date
Locked Re: Dennis Tillman
I can't wait until Dennis is back so Chuck Harris can be reinstated. larry
By Lawrance A. Schneider · #190925 ·
Re: 7904 problems
Hi, yes I agree. I had a similar 7904 problem years ago. In tic mode there was also a hissing sound. The problem was the HV multiplier. G?ran
By G?ran Krusell · #190924 ·
Re: 7904 problems
A minimum load is indeed required, in the order of 50 W. The cause of tick mode might also be in the HT wiring. In an old message #64991 I wrote the following: "My 7904 stayed in tick mode because of
By Albert Otten · #190923 ·
Re: 7904 problems
Hi all, I'm battling away with the two 7904s. The one with the faulty CRT had a functional power supply. The other one was completely dead despite having intact fuses. I transplanted the working power
By Morris Odell · #190922 ·
Re: 7A19 Fuse (later model, on attenuator - thin film?)
Hi,? I looked in 7a19 and 7a24 manuals and fuse listed for f100 (7a24) and f10 (7a19) are the same with part # 159-0118-00, which doesn't agree with your number.? Both plug-ins have 2 spare.? I
By Carl Hallberg · #190921 ·
Re: 7A19 Fuse (later model, on attenuator - thin film?)
The Mouser surface mount fuse idea has merit, Harvey. I would solder it on the lower ceramic substrate, bridging the blown thin film. It would still need to meet the 0.3 ohms (or less) spec. to keep
By Cliff Carrie · #190920 ·
Re: 7A19 Fuse (later model, on attenuator - thin film?)
I think I may have answered my own question. A glass fuse rated at 0.2A and 1 1/4" long has a resistance of about 6.75 ohms. To get down to 0.3 ohms using the 0.2A rated wire from such a fuse would
By Cliff Carrie · #190919 ·
Re: 7A19 Fuse (later model, on attenuator - thin film?)
Hmm, I haven't tried at all. you might be able to do the following: 1) design a PC board with the fuse shape.? Be sure you panelize it when you send it off, you'll get lots for very little. The fuse
By Harvey White · #190918 ·
Re: 7A19 Fuse (later model, on attenuator - thin film?)
Has anyone tried to open and repair one of these ceramic-based stripline fuses?? This thread has been dormant for some time. A 0.3 ohm fuse resistance makes sense in a 50 ohm terminated input. It
By Cliff Carrie · #190917 ·
Re: Troubleshooting 465Bs With Dead Channel 2
Tom, See if CR1796 is bad (shorted). Clean the pins and sockets for the transistors and IC. The pots may need cleaning as well. It is possible a transistor is bad. See about the voltages at points
By Mark Vincent · #190916 ·
Re: Probe tips P6133 and P6137
Hi all, thanks for the information provided. I decided to modify the P6133 probes into 1x probes: I opened both the broken probe tips, and replaced the burned hybrid with a zero Ohm replacement. I
By satbeginner · #190915 ·
Re: 2712 Spectrum Analyser, manual VR RBW filter level DAC setting on A13.
Hi Jaap, Did you manage to get information on converting a 2715 to 2712? I have a 2715 and tried to change the resistors on the motherboard and now it says "Misconfigured Spectrum Analyser" on the
By arainhoneves · #190914 ·
Re: Troubleshooting 465Bs With Dead Channel 2
The channel 2 amp seems to be saturating full + signal. Now why? Since everything works fine with channel 1, then that part of the switch and the vertical amp are ok. So either the preamp is
By Harvey White · #190913 ·
Troubleshooting 465Bs With Dead Channel 2
I am troubleshooting a 465Bs which suddenly lost the channel 2 display trace. I have the appropriate Tek service manual and test equipment including another working 465B. My current troubleshooting
By Tom Phillips · #190911 ·
Re: 577 curve tracer and low-current looping
Thanks Andy. If true, that would explain the 7CT1N's cleaner trace at the sweep begin and end points too. Yeah, flying leads on sensitive measurements is a no-go for several reasons.
By Precaud · #190910 ·
Re: 212, 213, was Re: [TekScopes] New kid on the block...
Looks good. And I happen to have some of those cells and some holders to use. Gonna have to look and see if I have any boost converters in my box of modules... -- Member of the toughest, meanest,
By Roy J. Tellason, Sr. · #190909 ·
Re: A sick 465B
Mark- thank you...my "B" is low definitely will have to verify. Ren¨¦e
By Ren¨¦e · #190908 ·
Re: 577 curve tracer and low-current looping
I can confirm the same behaviour on my 577 also. As I recall, the 7CT1N has a triangle-wave generator for the collector voltage, at a higher frequency than the line frequency, sine wave (or near sine
By Andy Warner · #190907 ·
Re: A sick 465B
Renee, My 465 is <250,000 and my B is high. The parts for yours may be different than what I posted. As you know, look at the S/N to make sure yours matches the schematic. I know some have commented
By Mark Vincent · #190906 ·
Re: A sick 465B
Mark- thanks for the list, what is your serial#? hi or lo? ( I too am re doing a 465B) Ren¨¦e
By Ren¨¦e · #190905 ·