Chuck Harris
 

As I recall, there was a variation of the 7D20 that was called
the 7D20T, that fit in its own chassis, and was used in conjunction
with computers to gather data.

A picture is shown on page 338 of the 1984 TEK Products catalog.

-Chuck Harris



Richard Steedman wrote:

I have a 7D20 with a couple of modifications or options added. Can anyone provide some information about these?

The first is a board sitting on top of board A11 above the vacant IC sockets U1230, U1330 and U1430. It has a connector P200, connected to J100 on the CCD board A5. It has three ICs - a dual 4-input NOR gate, a quad analog switch/mux and a dual op-amp. One of the 7D20 photos on Barry Johnson's website looks like it has the same mod [()]().

The second is a small board sitting behind the front panel in-between boards A6 and A7. It's connector has flying leads soldered directly to pins of the CCD chips on the underside of board A5.

I have posted photos [here](/g/TekScopes/album?id=253175).

The 7D20 has a serial number in the B085000's and is running firmware F1.03. Any insights gratefully appreciated!




Re: WTD - P6201 Probe tips (Tek P/N 206-0200-00)

 

On 08/09/2020 11:57, Raymond Domp Frank wrote:
Hi Phil,
I'm pretty certain I have a few, NOS. How "many" do you need?
Shipment from the Netherlands to the UK.
Let me know if not found locally.

Raymond
At least one, ideally two. NL->UK shipping is fine, I can't see that
being too expensive!

How much are you asking for the probe tips?

I've found a few ebay listings in the US but either the price or the
shipping tends to be the killer (or both).

Thanks,
--
Phil.
philpem@...


Help Troubleshooting 466, No +15VDC

 

This 466 worked great for years, then suddenly I smelled something burning, and the fan stopped. All low voltages good on interface board, *except* for +15VDC. Reads about 4VDC, (and about 4 ohms to ground with power off....). If I leave power on past about 15 sec, I can smell something burning again, so I only power up long enough the check the voltages.
I'm thinking tantalum cap, but wondering a good way to isolate the offender? Assume the the +15VDC goes all over the scope, to other boards, so would be a bear to lift all tant caps one at a time. Thinking of breaking the trace somewhere near the PS source. Any other tips or likely suspects?

Thanks in advance!


Re: Tek TDS694C advice pls

fauffing
 

Have you used the P6158 probes with this scope, and if you have, what are the drawbacks of the P6158 compared to the P6249?

Thanks,
- Frank


466 Interface Bd 15V-65V short

 

Troubleshooting an issue with my 466, where both traces would get blurry (unfocus) intermittently. The problem appeared to be mechanical in nature, as it responded to taps on the case.
Opened it up and found a chalky, powdery bridge between C1737 (65V) and the 15V test point on the interface board. Cleaned it up, and appears to have cleared the problem. Has anyone seen this before?


7D20 with unknown options or mods

 

I have a 7D20 with a couple of modifications or options added. Can anyone provide some information about these?

The first is a board sitting on top of board A11 above the vacant IC sockets U1230, U1330 and U1430. It has a connector P200, connected to J100 on the CCD board A5. It has three ICs - a dual 4-input NOR gate, a quad analog switch/mux and a dual op-amp. One of the 7D20 photos on Barry Johnson's website looks like it has the same mod .

The second is a small board sitting behind the front panel in-between boards A6 and A7. It's connector has flying leads soldered directly to pins of the CCD chips on the underside of board A5.

I have posted photos here.

The 7D20 has a serial number in the B085000's and is running firmware F1.03. Any insights gratefully appreciated!


Re: P6249 Oscillation

fauffing
 

Does the probe calibrate? I am new to the 694C and have two probes that calibrate and one which does not. All have oscillation similar to what you've described, with the more pronounced oscillation on the probe which will not calibrate (and I cannot effectively set offset according to documentation on this probe either). Does anyone have a screenshot of a 694C on the calibration pins at various scales they'd be willing to share, so I can use for a comparison? I really don't know what to reasonable expect from a working vs. a non-working P6249 probes.


Re: WTD - P6201 Probe tips (Tek P/N 206-0200-00)

 

Hi Phil,
I'm pretty certain I have a few, NOS. How "many" do you need?
Shipment from the Netherlands to the UK.
Let me know if not found locally.

Raymond


WTD - P6201 Probe tips (Tek P/N 206-0200-00)

 

Hi folks,

I've just picked up a P6201 probe kit which is almost complete -- except for the probe tips. Sadly the one on the probe is a bit worn.

Does anyone have any spare P6201 probe tips (Tek P/N 206-0200-00) for sale?

Cheers
Phil.


Re: 10 EASY WAYS TO TROUBLESHOOT WITH AN OSCILLOSCOPE

 

Good day,?
while the author certainly had good intentions, the recommendation to probe live utility power per sections 2 and 3 without proper safety??precautions or links to the known safety standards & best practices?is unbelievably?risky and unnecessary.?Why pose such a risk for beginners? Why the author fails to?recommend some?low voltage probing as a start?is beyond comprehension.
Cheers,
Magnus

Pete,

Some (many young?) people think "scopemeters" are the Real McCoy, having
never used a traditional oscilloscope and may think no one else alive
has either.

OTOH, The digital heads only think in 1s and 0s and somehow conceive
that oscilloscopes are no longer needed because analog is dead and every
signal running around inside a digital device is a clean high or low
state (1 or 0). (I do agree with them that antenna theory is at least
partly witchcraft)

Many years ago (1980s?)I introduced the head of the electronics
department at the best of the bunch area community college to a
semiconductor curve tracer. He didn't know they existed. To his credit I
could see a bright light flashing right through his eyes when I
explained what he was looking at, and he was a relatively recent EE!
Quite some time later, he informed me that he had requisitioned one for
his department, got it in their lab, played with it for a few days, and
procured several more. Curve tracing various types of semiconductors,
and documenting the meaning of what is displayed became a required part
of the degree path for his students. I suspect his graduate placement
rate went up considerably. Sadly, he is only a few years my junior and
may be on his way out, perhaps already retired. I can only hope he had a
hand in choosing his replacement.

Anyway, my point is that, previously, this same CC had been graduating
students with an associate degree in electronics with the barest
consideration of discrete component behavior beyond resistors and
capacitors. Linear was nearly all opamps and digital was all logic
probes on engineered student experiment platforms that had been
carefully designed to produce precise, repeatable results. No Surprises!
With that sort of mentality driving many of our schools, how can we
expect the general public, even the smart ones, to know they don't know?

It took me many years of study, and more than a few years learning, to
realize how little I know. I don't even poopoo my friend who is in
search of perpetual motion and/or free energy. I just tell him to build
it and show me it working and let me play with it in my lab. If it still
works after an inquisition, we'll become rich and famous unless, of
course, were killed first to suppress it.

Thomas Garson
Aural Technology, Ashland, OR
By my calculation, the? dynamic range of the universe is roughly 679dB,
which is approximately 225 bits, collected at a rate 1.714287514x10^23 sps.

On 9/7/20 6:24 PM, saipan59 (Pete) wrote:
Looking at the author's site, it *appears* that he has good intentions, and enjoys learning about a wide variety of things, but lacks actual expertise/experience (at least in electronics).
He may not realize that a scope is NOT essentially an auto-ranging DMM, and that advising newbies to make AC line measurements is a Bad Idea.

Pete


Re: Tek 4654M; No Trace, No +5V, No +95V; Help

 

Quick up-date. Disconnected the multiplier feed/connection to T550 pin 10. With this connection broken fuse F558 still blew upon powering up the scope. Am thinking that U550 has issues with more than just the multiplier. Was thinking of going back over some of caps again to see if one of the caps that goes to ground was bad and I missed it the first time around.


Re: 1S1 sampling unit; (also 7B53AN & 7A18 plugins)

David Collier
 

A bit off -topic: today picked up Fairchild 765H Portascope with 74-13A timebase and 76-02 25MHz dual trace plugins, with probe, all for A$10!
Transistors with quite a few nuvistor tubes; EHT power via EL86 and 12AU7 tubes. Standard of construction not far behind TEK, if at all, with five fuses.
Cleaning it up before gradually powering up in due course. Won't take 1S1 TEK plug in though.


Re: 10 EASY WAYS TO TROUBLESHOOT WITH AN OSCILLOSCOPE

 

Pete,

Some (many young?) people think "scopemeters" are the Real McCoy, having never used a traditional oscilloscope and may think no one else alive has either.

OTOH, The digital heads only think in 1s and 0s and somehow conceive that oscilloscopes are no longer needed because analog is dead and every signal running around inside a digital device is a clean high or low state (1 or 0). (I do agree with them that antenna theory is at least partly witchcraft)

Many years ago (1980s?)I introduced the head of the electronics department at the best of the bunch area community college to a semiconductor curve tracer. He didn't know they existed. To his credit I could see a bright light flashing right through his eyes when I explained what he was looking at, and he was a relatively recent EE! Quite some time later, he informed me that he had requisitioned one for his department, got it in their lab, played with it for a few days, and procured several more. Curve tracing various types of semiconductors, and documenting the meaning of what is displayed became a required part of the degree path for his students. I suspect his graduate placement rate went up considerably. Sadly, he is only a few years my junior and may be on his way out, perhaps already retired. I can only hope he had a hand in choosing his replacement.

Anyway, my point is that, previously, this same CC had been graduating students with an associate degree in electronics with the barest consideration of discrete component behavior beyond resistors and capacitors. Linear was nearly all opamps and digital was all logic probes on engineered student experiment platforms that had been carefully designed to produce precise, repeatable results. No Surprises! With that sort of mentality driving many of our schools, how can we expect the general public, even the smart ones, to know they don't know?

It took me many years of study, and more than a few years learning, to realize how little I know. I don't even poopoo my friend who is in search of perpetual motion and/or free energy. I just tell him to build it and show me it working and let me play with it in my lab. If it still works after an inquisition, we'll become rich and famous unless, of course, were killed first to suppress it.

Thomas Garson
Aural Technology, Ashland, OR
By my calculation, the dynamic range of the universe is roughly 679dB,
which is approximately 225 bits, collected at a rate 1.714287514x10^23 sps.

On 9/7/20 6:24 PM, saipan59 (Pete) wrote:
Looking at the author's site, it *appears* that he has good intentions, and enjoys learning about a wide variety of things, but lacks actual expertise/experience (at least in electronics).
He may not realize that a scope is NOT essentially an auto-ranging DMM, and that advising newbies to make AC line measurements is a Bad Idea.
Pete


Re: 10 EASY WAYS TO TROUBLESHOOT WITH AN OSCILLOSCOPE

 

please research the author first before debating the article !
Why? This is unscientific. Debatable advice is debatable no matter who utters it.

EJP


Re: Lot of 10 scopes (400s & 2400s) with 5 scopemobiles

Roy Morgan
 

Tek folks,

The seller as acknowledged my offer and I will be making arrangements to get the lot.

Most likely I will store them near Greenfield till they can be sorted and moved on to folks who want them.

I'll keep the list posted.


Roy Morgan
K1LKY Western Mass
607-765-8867


-----Original Message-----
From: [email protected] <[email protected]> On Behalf Of Roy Morgan
Sent: Monday, September 07, 2020 8:20 PM
To: [email protected]
Subject: Re: [TekScopes] Lot of 10 scopes (400s & 2400s) with 5 scopemobiles

Tek folks,

I am working to put together a deal for these scopes (I am on western mass in striking distance of the NH seller)

Goal is to get the lot and pass along scopes to interested list members, covering just costs. )

I will make offer to buy the lot and keep the list posted.

Roy

Roy Morgan
K1LKY Western Mass

On Sep 7, 2020, at 10:36 AM, David Berlind <david@...> wrote:

?For sale in NH for $450. I am not affiliated with the seller.


Re: 10 EASY WAYS TO TROUBLESHOOT WITH AN OSCILLOSCOPE

 

Looking at the author's site, it *appears* that he has good intentions, and enjoys learning about a wide variety of things, but lacks actual expertise/experience (at least in electronics).
He may not realize that a scope is NOT essentially an auto-ranging DMM, and that advising newbies to make AC line measurements is a Bad Idea.

Pete


Re: Lot of 10 scopes (400s & 2400s) with 5 scopemobiles

Roy Morgan
 

Tek folks,

I am working to put together a deal for these scopes (I am on western mass in striking distance of the NH seller)

Goal is to get the lot and pass along scopes to interested list members, covering just costs. )

I will make offer to buy the lot and keep the list posted.

Roy

Roy Morgan
K1LKY Western Mass

On Sep 7, 2020, at 10:36 AM, David Berlind <david@...> wrote:

?For sale in NH for $450. I am not affiliated with the seller.


Re: Tek 4654M; No Trace, No +5V, No +95V; Help

 

Hi Barry
Thanks for the prompt response. I'll let you know how I make out.
Harrison N1FAM


Re: Tek 4654M; No Trace, No +5V, No +95V; Help

 

Michael. Thank you for your pictures and explanation. Your repair looks very professional and obviously was effective. Unfortunately I don't have access to a 3D printer for the enclosure. I am thinking maybe I can use a craft store plastic container or a small pill bottle. I am not familiar with the Tek X Ref catalog but will search for it on line. Building a replacement circuit will be new for me. I haven't done that in the past. I am good at parts replacement. So circuit building will be new for me. From your pictures it appears your new circuit is spliced into the red line going to the CRT. Thanks again for the information. I was thinking of shelving the 465M because I didn't have a replacement U550 but you motivated me to perhaps carry on. My functional scope is a Tek TDS460A. The 465M is strictly a bench and learning exercise/project. The scope was given to me as excess and not working. I do have a "parts" 455 but I was trying to save it so that if I got educated and repaired the 465M I might be able to bring the 455 back just for the exercise. The 455 U550 doesn't seem any better so I haven't considered swapping it out. It did look good at one point but no longer. It appears to be blowing the F736 fuse. Don't know if that is the result of my probing or if something just gave up once it was powered up for a bit. Before it was blowing the F736 fuse It didn't have a trace but did have a beam finder. The 455 also spent some time under water years ago so that certainly didn't help it. Enough on the 455. Thanks again for the information. Harrison N1FAM


Re: Tek 4654M; No Trace, No +5V, No +95V; Help

 

Hi Harrison,

Yes, that all sounds correct.

From the symptoms your scope is exhibiting as compared to mine and others, the multiplier is likely arcing inside when it gets up to working voltage. As far as I know, this is fairly common failure in the 465M's multiplier.

Good luck with it.

Thanks,
Barry - N4BUQ

----- Original Message -----
From: "Harrison" <buma7@...>
To: [email protected]
Sent: Monday, September 7, 2020 6:30:40 PM
Subject: Re: [TekScopes] Tek 4654M; No Trace, No +5V, No +95V; Help

Barry, Thank you for the picture and explanation. I believe it is pins 8, 9,
and 10 of T550 that relate to U550. It looks like pin 8 goes to the PCB.
Pins 9 and 10 go to U550. I wanted to confirm that to disconnect the
multiplier from T550 it was just the end connection (the one closest to the
board edge) that you disconnected, That would be the U550 connection that
mates with T550 Pin 10. Is that correct? Once I unsolder that connection I
would replace the F558 fuse and power up the unit to see if the fuse doesn't
blow and the 95V bus is back. Harrison N1FAM