¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io
Date

Re: Repairing plated through holes.

 

Mouser stocks several sizes of staking tools from Keystone that might come in handy. Not really cheap, but I have one of the sets (don't remember which one) that I use when I need to stake a terminal onto a PCB. Keystone part nos TL-20, TL-21 and TL-22. These are made primarily for staking turret terminals onto a PCB, but also work well with larfer eyelets and rivets.

Cheers,
Dave M

On Sat, Jun 16, 2018 at 04:48 am, lop pol wrote:


On Sat, Jun 16, 2018 at 03:27 am, David M wrote:


Search Ebay for Copper Via Rivets. They are available in several diameters
and lengths. You'll have to come up with your own method and tooling to
stake
them to the PCB, but shouldn't be difficult.
Small nails, small hole punch, etc. should to the trick. Yankee ingenuity
rules.

Cheers,
Dave M

On Fri, Jun 15, 2018 at 07:28 pm, lop pol wrote:


I would like to affordably put together the things needed to repair plated
through holes. Anyone have any thoughts? I'm asking because I dont want
to
buy this more than once. My pocket has not much room for mistakes right
now
and i seem to do a lot better asking here before buying. Thanks guys
Ok. That is what I will do. The rivets/eyelets are really cheap most under .25
cents a piece at Mouser. I have a bunch of different punches so I will figure
something out. Thanks for the input.


Re: Up to date capacitor list for Tek 2465A and 2465B scopes (2018)

 

Lots of parts leads to confusion, its all good. Once this is done it will be good for a year or two at least.

Here is another correction
C1113 180 ?F 40 ELCTLT 330 ?F 50
C1114 250 ?F 100 ELCTLT 330 ?F 50

Should be...
C1113 180 ?F 100 ELCTLT 330 ?F 160
C1114 250 ?F 100 ELCTLT 330 ?F 160

To the A5 listing please add:
C2010 33 ?f 10 ELCTLT 33 ?F 10

I have not looked at the A1 board yet, but I will let you know if I find anything there


Re: Tek 2337 (looks mint) on Goodwill.

 

Looks like it is very clean, If I still lived in Ft worth I'd run over and turn it on to check it out. has probes and all.

100MHz Analog Scope ( later 150MHz version I think) plus a built in DMM
More info over on Tekwiki

Godwill in the auction business ..who knew

-DC
manuals@...

On 6/16/2018 11:47 AM, David Berlind wrote:
I'm sure everyone here has seen one of these. But I haven't. Is it some sort of hack of a digital scope? It looks mint (and cool).



I have no affiliation with the seller.


--
Dave
Manuals@...
www.ArtekManuals.com


Re: Repairing plated through holes.

 

Pace made a "fused eyelet setter" .
This was an eyelet staking tool that ran high current, low voltage through the eyelet after the eyelet was staked and while the eyelet was still under compression.
This fused the flared part of the eyelet to the lands.
Motorola used eyeletted boards in some of their two way radios.
This was the major source of failure as the eyelet solder connection broke free of the lands.
The fix for this was to resolder all of the eyeletted connections on both the top and bottom of the board.

One source of PCB repair things is



Glenn

On 6/15/2018 10:28 PM, lop pol via Groups.Io wrote:
I would like to affordably put together the things needed to repair plated through holes. Anyone have any thoughts? I'm asking because I dont want to buy this more than onct setter"e. My pocket has not much room for mistakes right now and i seem to do a lot better asking here before buying. Thanks guys


--
-----------------------------------------------------------------------
Glenn Little ARRL Technical Specialist QCWA LM 28417
Amateur Callsign: WB4UIV wb4uiv@... AMSAT LM 2178
QTH: Goose Creek, SC USA (EM92xx) USSVI LM NRA LM SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license"


Tek 2337 (looks mint) on Goodwill.

 

I'm sure everyone here has seen one of these. But I haven't. Is it some sort of hack of a digital scope? It looks mint (and cool).



I have no affiliation with the seller.


Re: Repairing plated through holes.

stefan_trethan
 

You can still buy the pop rivet nuts, but much more $$ than that.
You said they use an adapter, I have never seen that. Around here you
need to buy a dedicated rivet nut tool for around $100 (which looks
much like a slightly modified $20 pop rivet tool).

I always speculated that they are set just like a pop rivet, but with
a threaded rod instead of the discardable metal shank in a regular
rivet.
My plan was, should I ever need to set a rivet nut, that I would take
a hard screw and put that in the rivet, and pull on the screw from the
outside with a washer and a nut.

Do you have any opinion if that plan might succeed? I have been
curious about it for years.

ST

On Sat, Jun 16, 2018 at 5:40 PM, Michael A. Terrell
<mike.terrell@...> wrote:
They also made the rivnut,which is an internally threaded version of a pop rivet that used an adapter to set with a pop rivet tool. I used them in some projects. I wished that I had boght more of them, surplus. They were 50 cents/100 in USM boxes.


Michael A. Terrell


Re: Repairing plated through holes.

 

They also made the rivnut,which is an internally threaded version of a pop rivet that used an adapter to set with a pop rivet tool. I used them in some projects. I wished that I had boght more of them, surplus. They were 50 cents/100 in USM boxes.


Michael A. Terrell

-----Original Message-----
From: "ebrucehunter via Groups.Io" <Brucekareen@...>
Sent: Jun 16, 2018 10:46 AM
To: [email protected]
Subject: Re: [TekScopes] Repairing plated through holes.


About 50 years ago I purchased an eyelet press from Kepco, a company that then sold printed circuit board materials. At the time they offered a variety of different diameter rivets, of different lengths, and in both copper and brass. The press came with a number of special tips needed for the different size rivets. I was later able to fabricate special tips to also set Vector, rivet-type terminals.


Unfortunately, as a previous contributor has remarked, these eyelets almost always eventually gave trouble if not carefully soldered to the foil on both sides. Even soldered eyelets sometimes gave trouble, probably because, as another has mentioned, you need a little space under the head for solder to properly flow.


In looking at the press, I saw something I had never noticed before. The press was made in Italy by United Shoe Machinery, the company that made eyeletting machinery for the shoe industry and the original developer of the pop-rivet.



Bruce, KG6OJI


Re: Repairing plated through holes.

 

The key is the eyelet technique only works with two layer boards. The art
techniques for rebuilding through holes and multi-layer boards but it's not
something you do yourself although I tried it a couple times by building my
own Electrolux replating set up.

On Fri, Jun 15, 2018, 10:38 PM Harvey White <madyn@...> wrote:

On Fri, 15 Jun 2018 20:33:52 -0700, you wrote:

On Fri, Jun 15, 2018 at 08:17 pm, Mark Goldberg wrote:


Do a search for "eyelet plated through hole" and you will find eyelets
designed for that purpose. I do not know what kind of tooling they may
or
may not require. I have also used just a piece of wire bent through the
hole and onto the pads on two sides, but that is not the "official"
method.

Regards,

Mark

I see the eyelets on mouser are cheap. What im really wondering about is
the staking tool and which one to get.
You could likely make one. You'd need something for an anvil, could
easily be made from a 4mm or so rod, put a handle on it. The "hammer"
part could be made from an automatic center punch with a different
punch part. They're made to be removed. You'd want a design that
curled the end of the rivet over.

Note that eyelets work best if they are either solid copper, or tinned
copper. I'm not sure that any other variety would even accept solder.
Also, you may have to drill out the hole a bit if you are trying to
get the same inside diameter. A third (and slightly fatal) problem is
that I can't see any way to make this work on multi-layer boards.
Double sided is the most. It could hold down lifting foil on a single
sided board, also.

A little lathe work (if you have one) and it might just come together.

Harvey





Re: Repairing plated through holes.

 

About 50 years ago I purchased an eyelet press from Kepco, a company that then sold printed circuit board materials. At the time they offered a variety of different diameter rivets, of different lengths, and in both copper and brass. The press came with a number of special tips needed for the different size rivets. I was later able to fabricate special tips to also set Vector, rivet-type terminals.


Unfortunately, as a previous contributor has remarked, these eyelets almost always eventually gave trouble if not carefully soldered to the foil on both sides. Even soldered eyelets sometimes gave trouble, probably because, as another has mentioned, you need a little space under the head for solder to properly flow.


In looking at the press, I saw something I had never noticed before. The press was made in Italy by United Shoe Machinery, the company that made eyeletting machinery for the shoe industry and the original developer of the pop-rivet.



Bruce, KG6OJI


Re: Flyke 3330b

 

No coffee yet this morning, I missed one of the 3s :-)

Paul

On Sat, Jun 16, 2018 at 10:27:42AM -0400, Artekmedia wrote:
Paul

The 3330B is a Voltage/Current Calibrator

-DC

manuals@...

On 6/16/2018 10:20 AM, Paul Amaranth wrote:
Are you sure about that model number? A fluke 330b comes up as a
clamp on hand held. Voltage standards were 332, 335, etc.

From the context I'm assuming you're talking about one of the
80 pound voltage standards.

There is a volt-nuts (metrology) subgroup on eevblog with discussions
on these along with some very knowledgeable people who can help. I
have a 332b, pretty amazing piece of equipment.

Get a manual and be extremetly careful when that thing is on if you
have the cover off. I think that inner cover can be at 1KV.

Paul

On Sat, Jun 16, 2018 at 09:57:22AM -0400, Artekmedia wrote:
Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to
Tektronix and HP groups) but worth a try. (Also spelling it FLUKE
instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, oliver johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a
fluke but i figured someone could point me in the correct direction for a
group that can help me , or if someone here knows anything about going
about fixing this voltage standard .



--
Dave
Manuals@...
www.ArtekManuals.com







--
Dave
Manuals@...
www.ArtekManuals.com






!DSPAM:5b25176616721668650994!
--
Paul Amaranth, GCIH | Rochester MI, USA
Aurora Group, Inc. | Security, Systems & Software
paul@... | Unix & Windows


Re: Flyke 3330b

 

Go over to where there is a TONNE of experience on a broader scale.


Re: Flyke 3330b

 

I am sure about the model 3330b is the correct info , i have a download for the manual but am having trouble with tracking down why it will not output , to me it seems as if the unit has no power at the mother board , there was a bad bux48a and a bad 2n2869 and a bad 250uf . I am stumped?
On Sat, Jun 16, 2018 at 10:20 AM, Paul Amaranth<paul@...> wrote: Are you sure about that model number?? A fluke 330b comes up as a
clamp on hand held.? Voltage standards were 332, 335, etc.

From the context I'm assuming you're talking about one of the
80 pound voltage standards.

There is a volt-nuts (metrology) subgroup on eevblog with discussions
on these along with some very knowledgeable people who can help.? I
have a 332b, pretty amazing piece of equipment.

Get a manual and be extremetly careful when that thing is on if you
have the cover off.? I think that inner cover can be at 1KV.

? Paul

On Sat, Jun 16, 2018 at 09:57:22AM -0400, Artekmedia wrote:
Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to
Tektronix and HP groups) but worth a try. (Also spelling it FLUKE
instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, oliver johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a
fluke but i figured someone could point me in the correct direction for a
group that can help me , or if someone? here knows anything about going
about fixing this voltage standard .



--
Dave
Manuals@...
www.ArtekManuals.com






!DSPAM:5b25107313919973180691!
--
Paul Amaranth, GCIH? | Rochester MI, USA? ? ? ? ? ? ?
Aurora Group, Inc.? |? Security, Systems & Software
paul@...? |? Unix & Windows


Re: Flyke 3330b

 

Paul

The 3330B is a Voltage/Current Calibrator

-DC

manuals@...

On 6/16/2018 10:20 AM, Paul Amaranth wrote:
Are you sure about that model number? A fluke 330b comes up as a
clamp on hand held. Voltage standards were 332, 335, etc.

From the context I'm assuming you're talking about one of the
80 pound voltage standards.

There is a volt-nuts (metrology) subgroup on eevblog with discussions
on these along with some very knowledgeable people who can help. I
have a 332b, pretty amazing piece of equipment.

Get a manual and be extremetly careful when that thing is on if you
have the cover off. I think that inner cover can be at 1KV.

Paul

On Sat, Jun 16, 2018 at 09:57:22AM -0400, Artekmedia wrote:
Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to
Tektronix and HP groups) but worth a try. (Also spelling it FLUKE
instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, oliver johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a
fluke but i figured someone could point me in the correct direction for a
group that can help me , or if someone here knows anything about going
about fixing this voltage standard .



--
Dave
Manuals@...
www.ArtekManuals.com






!DSPAM:5b25107313919973180691!
--
Dave
Manuals@...
www.ArtekManuals.com


Re: Fluke 3330b

 

Looks like it has moved to Groups.io

/g/Fluke-DMM


Michael A. Terrell

-----Original Message-----
From: Artekmedia <manuals@...>

Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to
Tektronix and HP groups) but worth a try. (Also spelling it FLUKE
instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, Oliver Johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a Fluke but i figured someone could point me in the correct direction for a group that can help me , or if someone here knows anything about going about fixing this voltage standard.


Re: Flyke 3330b

 

Are you sure about that model number? A fluke 330b comes up as a
clamp on hand held. Voltage standards were 332, 335, etc.

From the context I'm assuming you're talking about one of the
80 pound voltage standards.

There is a volt-nuts (metrology) subgroup on eevblog with discussions
on these along with some very knowledgeable people who can help. I
have a 332b, pretty amazing piece of equipment.

Get a manual and be extremetly careful when that thing is on if you
have the cover off. I think that inner cover can be at 1KV.

Paul

On Sat, Jun 16, 2018 at 09:57:22AM -0400, Artekmedia wrote:
Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to
Tektronix and HP groups) but worth a try. (Also spelling it FLUKE
instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, oliver johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a
fluke but i figured someone could point me in the correct direction for a
group that can help me , or if someone here knows anything about going
about fixing this voltage standard .



--
Dave
Manuals@...
www.ArtekManuals.com






!DSPAM:5b25107313919973180691!
--
Paul Amaranth, GCIH | Rochester MI, USA
Aurora Group, Inc. | Security, Systems & Software
paul@... | Unix & Windows


Re: Flyke 3330b

 

Thanos for the info , the spelling was i mistake?
On Sat, Jun 16, 2018 at 9:57 AM, Artekmedia<manuals@...> wrote: Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to
Tektronix and HP groups) but worth a try. (Also spelling it FLUKE
instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, oliver johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a fluke but i figured someone could point me in the correct direction for a group that can help me , or if someone? here knows anything about going about fixing this voltage standard .



--
Dave
Manuals@...
www.ArtekManuals.com


Re: Flyke 3330b

 

Oliver

there is a Yahoo group "Fluke_DMM" . Not real active (compared to Tektronix and HP groups) but worth a try. (Also spelling it FLUKE instead of FLYKE helps :-))

-DC
manuals@...

On 6/16/2018 9:31 AM, oliver johnson via Groups.Io wrote:
I have a fluke 3330b , have no output . I am not sure where to post for a fluke but i figured someone could point me in the correct direction for a group that can help me , or if someone here knows anything about going about fixing this voltage standard .


--
Dave
Manuals@...
www.ArtekManuals.com


Re: Help needed with no trace no beamfinder on 465 (not b)

 

Hello tek lovers ,had dame issue on a 465 ,but i have th¨¦ service manual ,this is ¨¤ should have Firts ,papier is betterave than on screen, issu was ¨¤ failed q 1472 pnp switching ,replaced with ¨¤ 2 n2904 and Works well !!

Envoy¨¦ depuis mon Redmi 4X

Le Albert Otten <aodiversen@...>, 16 juin 2018 1:42 PM a ¨¦crit :








Hi Keith,

Obviously my suggestions are too cryptically for you. I hesitate to
continue since in my opinion you should understand the purpose of test
measurements, to learn something and also to protect the equipment and
yourself from damage. But let me try to clarify my previous suggestions.
Q1418 is the heart of the oscillator. Oscillation occurs because of the
feedback from collector to base provided by the windings 6-7 and 8-9 of
the transformer. The voltage and current amplitude in these windings is
determined by average base current which has to come from the regulator
circuit, Q1416 emitter. The oscillator produces an up-transformed
alternating voltage in the HV winding between pins 23 and 5 which is
half-wave rectified by CR1421 to give a negative HV at C1421 to C1424 and
CRT cathode.
A fault could be that one of these HV caps (or even C1488) starts to leak
at say 70 V. That would put extra load on the oscillator in its start-up
phase and and prevent the amplitude to rise further.
Another fault could be that CR1421 starts to leak, with the same
consequence.
Both these fault would show up of you externally feed TP1423 with a high
enough negative voltage (scope disconnected from the power inlet of
course). I simply use my 576? curve tracer for this purpose but you have
to improvise something. A DC supply, preferably variable, would be needed,
with + to scope ground and ¨C to TP via a DMM. The DMM at say 200 V DC
range. The DMM is 10 M (usually), the load on TP is about 30 M. So the DMM
reading should be about 25% of the supply voltage and maintain that
percentage when the supply voltage is increased. When the percentage
starts to increase it indicates leakage somewhere, and you can also try to
estimate how heavy that leakage is.
Of course this would eliminate only a leakage fault in the branch from
CR1421.
A fault in the branche from pin 3 is not very likely, it¡¯s low voltage.
That leaves a fault in the transformer itself (any winding could be
involved) or in the HV multiplier.

Suppose you disconnect P1400. Then you can supply an AC voltage across the
collector winding between pin 1 of P1400 socket and the fuse holder
terminal. I use a sine wave function generator for this. The generator
output is w.r.t. generator ground, so I connect that side to the fuse and
the live side to mentioned pin 1. (Additionally the fuse side can be
shorted to scope ground to prevent any floating voltage levels inside the
scope.) Some people use an audio amplifier.
This way I can control the input frequency an voltage amplitude. I also
monitor the generator output current and the TP voltage. With all this
information I can check that the resonance frequency is about what it
should be (maybe 50 kHz for a 465? I didn¡¯t look it up) and also see if
stange things happen if I increase the amplitude.

L1419 and C1418/C1419 mainly serve to prevent oscillator signals to enter
15 V elsewhere. I think that even without these C¡¯s the oscillator should
still work.

Albert

On Fri, Jun 15, 2018 at 05:50 pm, Keith Ostertag wrote:


Thanks for your continued help Albert. I _might_ be able to cobble
together a
power supply in excess of 100V... but I am not understanding what that
will
tell me, possibly because I am unsure of your use for the term leakage
in this
context. Since TP1486 is in the secondary side of T1420 (correct?) how
can
anything I do with it eliminate T1420 as a fault? For instance if
T1420's
primary (inline with Q1418) is shorted or partially shorted? At least
that 's
the question that comes to my mind, since I don't understand the
circuit. What
if L1419 is shorted or partially shorted? And how does that eliminate
(or
indicate) the HV multiplier as the fault?

Let's see... you are suggesting that I might "drive the collector
winding from
an external source" by adding a negative voltage to TP1486 (which is
connected
to the T1420's secondary), is that correct? So if I had a +200VDC power
supply
I would connect the +200V positive terminal to ground and the negative
terminal to TP1486?

In the case I could obtain an power supply, where exactly would I
measure the
leakage current and what level would be the pass/fail amount? I assume
you
mean replace the fuse F1419 with an ammeter as I did before in order to
measure Q1418's collector current?

My only guess is... by adding voltage to the secondary, a working
primary must
likewise increase, thus increasing the collector current. So... if it
does not
increase Q1418's collector current then that suggests a problem with
L1419 or
T1420 (or C1419)?

And if that is close to being correct as a strategy, then wouldn't
slightly
increasing the positive voltage at the fuse test it similarly? I say
that
assuming T1420's secondary is unlikely to have been damaged by anything
I
might have shorted in front of it...

Oops... rereading your message I see:"Then you could see at which
frequency it
resonates and how it responds to increasing primary amplitude.", which I
somehow missed in the above...

Let me start over... Q1418's circuit drives the signal taken from Q1416
through T1420, multiplying it by whatever the turns ratio of T1420 is.
You are
suggesting I unplug Q1418 and replace it with a short? Then read the
current
through F1419? Not sure how we are to read a resonate frequency when we
are
not applying a sweep... Sorry, as you can see this is a bit over my
head...





Flyke 3330b

 

I have a fluke 3330b , have no output . I am not sure where to post for a fluke but i figured someone could point me in the correct direction for a group that can help me , or if someone here knows anything about going about fixing this voltage standard .


This may help find replacement transistors

 

This may help find replacement transistors. It also has a search function for surface mount ID codes to part numbers.