Re: Help I got a Tracking Generator for my Tek 2710 but how do I install it?
I think my biggest hurdle is going to be getting the proper Semi Rigid cables that are needed I have no clue where to buy these........
=====================================
Item number 280970372957
____________________________________________
|
Hi David,
no square wave on the pins 10 and 7, both slopes.... it just measures a +4V dc, at both pins. Does it mean my U460 has died?
T.
toggle quoted message
Show quoted text
--- In TekScopes@..., David <davidwhess@...> wrote: That narrows the problem down to some point at or after transistor array U460. The good news is that the 2213 does not use a custom trigger IC so if the problem can be found, it should be relatively easy to fix.
With the trigger set to LINE and AUTO, adjust the LEVEL to midway and measure pin 10 of U480C and pin 7 of U480B. With both SLOPE settings, there should be at least a 300mV to 400mV square wave on each pin.
If there is, then check pin 2 of U480A. It should have a 300mV to 400mV square wave.
On Sun, 23 Sep 2012 10:51:11 -0000, "tjeenz" <tjeenz@...> wrote:
Hi David, still out there?
back from trip... I checked tp444: with trig on "line" i have a sinus with about 0.5V tt
if i connect a 4V top-top sinus at ch1 with trig on "int" i have 150mV sinus on tp444 (very small)
U460B, pin 2: norm: there is indeed a +0.65 - -0.65V which can be controlled by the trig-lev. pot. auto: sinus 50 mV (again with 4V tt at ch1). which seems very small...
in all situations the display does not show a "still" image.
tjeenz
--- In TekScopes@..., "tjeenz" <tjeenz@> wrote:
hi David, thx for the tip! i will check this as soon as possible (couple days out of town). i check tp444 before with ext. trigger and that was very small compared to the input signal. don't know the exact amplitude but much lower. probably back on saturday. tjeenz.
--- In TekScopes@..., David <davidwhess@> wrote:
I would start by looking at TP444 (middle of schematic 4) which should be a duplicate of the trigger input signal as selected by the source and coupling controls. At the very least, there should be a 60Hz sine wave at that point when the trigger source is set to line.
Then I would check pin 2 of U460B which should be a duplicate of the tri.gger level voltage as set by the trigger level control. In Auto mode this voltage will be between the peak levels of the trigger signal but in normal mode, it should be between +0.65V and -0.65V.
It would be nice to have a nicer scan of the service manual. I extracted and cleaned up schematic 4 from the one available online:
On Tue, 18 Sep 2012 11:00:41 -0000, "tjeenz" <tjeenz@> wrote:
Hi, Have a trigger problem with a 2213 tektronix.
In auto and line trigger i have a baseline. In norm not. Also ext. trigger fails (signal from function generator 4v pp and ch1 on 1v/div)
cannot get still image to see frequency, amplitude is ok. with signal on one of the channels, trig in auto i can almost get still image if freq is low enough with var hold-off. trig-level does nothing. Also the trigger led never blinks.
Checked all p.s. voltages. Are ok (no ripples), but according to service manual there should be some specific resistance on the voltages. These are way too low in my scope: 8.6v 114 ohm at tp500= 58 ohm +8.6v 95 ohm at W975 = 16 ohm +5v 330 ohm at W985 = 16 ohm +30v 905 ohm at W965 = 16 ohm +100v 12.5 kohm at W966 = 16 ohm
so i guess something is wrong somewhere. checked the big capacitors in the p.s, for 30V and +8.6V, both were in spec.
Are these resistance values in the manual a good reference for finding the faulty area?
regards, tjeenz
|
Re: Help I got a Tracking Generator for my Tek 2710 but how do I install it?
Robert - You will need to make these semi-rigid cables. I cannot imagine being able to find the correct cables unless you get lucky. Making the cables is not as hard as you may think. You will need the Type-N connector, SMA connectors and some compatible semi-rigid. The hard part is making the necessary bends. What I do is use a piece of coat hangar wire to make a 'template'. Some patience will be required to get this right. You can then use the template to build your cable. Then solder the connectors to the ends. Easier said than done, but not as hard as you may imagine. The finished cable may not 'look' the best but it will function just fine, I'm sure. I believe you will need to remove the front panel in order to mount the type-N connector.
Stay positive.. Jim
toggle quoted message
Show quoted text
--- In TekScopes@..., "dreadlk" <dreadlk@...> wrote: For starters I found this on ebay but it looks too short.
I am scouring the 2712 SVC Manual looking for the correct part number but so far no luck, it is diagramed but the part numbers are not listed for any of the cables.
--- In TekScopes@..., "dreadlk" <dreadlk@> wrote:
Thanks for the information, I agree with both of your assessments. It was easy to find the Cables in a flex format so I went out and bought some on day one. Now that I read your posts I know it won't work and my dilemma is that I have no idea where to find the Semi Rigid versions. I am going to do some searching but I have a feeling I am going to need a lot of luck on my side to make this happen. If the unit will error out it makes little sense for me to proceed further until I locate these cables so at this point I am going to mark out the plate for cutting and drilling based on Jims info and then when I know where to get the needed cables I will start the cutting and drilling.
If anybody has any of these cables and is willing to sell them please let me know or if you have any ideas on a source that might sell them the info would be greatly appreciated.
Robert KD8MJR
--- In TekScopes@..., "Rich Miller" <av8torrich@> wrote:
Hello Robert:
With regard to the rigid coax cable used to connect the TG to the front panel, I think you are going to want to use this. While at low frequency other types of cable may work fine, when you get into microwave you may run into a problem with the TG Gain Flatness Normalization. If the output of the TG is off by more than +/-3dBm (This is from memory so double check), it will show up as a fault when you first attempt to normalize.
Rich, AJ3G
From: TekScopes@... [mailto:TekScopes@...] On Behalf Of dreadlk Sent: Friday, September 21, 2012 2:33 PM To: TekScopes@... Subject: [TekScopes] Re: Help I got a Tracking Generator for my Tek 2710 but how do I install it?
Hi Jim
Thank you so much for the reply. I do have a VR BW board in that cavity and I received the TG with no cables at all, I will have to figure out how to obtain the needed parts and where to source them. I was wondering if flexible cables with SMA and Type N will work or does it have to be the rigid stuff? I obtained a copy of the 2712 SV Manual and am going to look through it tonight to try and get a better idea of what I am dealing with, I want to thank you for your post, it certainly clarified things for me, I hope I am up to the task of doing this Job as I must admit I am a bit worried about messing up my 2710 it has been absolutely a pleasure to use over the years and is in fantastic condition.
Let me ask you Jim is there any chance of having my proverbial cake and eating it too? I mean could I leave in the VR BW board and run cables to the Back of the unit and put the TG in a separate enclosure, kind of like a DIY 2707? I am just wondering if anybody has ever done that, I would imagine that it might not be practical but I just wanted to get your opinion.
Anyway Jim I am going to take out the VR BW board and see what I am dealing with. If you have any suggestions please post them, and thanks so much for your help.
Robert KD8MJR
--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com> , "Jim" <jflan@> wrote:
Hi Robert - The TG mounts in the cavity on the left side of the analyzer. This also is the same cavity which houses the Variable Resolution BW board. You can only use one or the other, not both. It wasn't clear whether you have the VR BW board or not.
There are two semi-rigid cables that will need to be made. One routes the RF to the front panel and has an SMA on one end and a Type N on the other. The other routes the 2nd LO (2 GHz) to the TG. Did you receive these cables? The ribbon cable will plug into the main power supply board to provide the interface. in fact, I just remembered that one end of the ribbon cable is crimped to a IDC connector that is *soldered* to the power supply board. The other end uses at typical IDC connector that will plug into the TG. The 2712 (not 2710) service manual has a block diagram that will aide in making the necessary connections. This manual is down loadable from the Tek.com site and will be a better reference for what you are wanting to do. You will have some work to do to get your TG up and running. Take your time and it will go smoothly.
Good luck. Jim WB5KYE
--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com> , "dreadlk" <dreadlk@> wrote:
I posted about a month ago about searching for an internal tracking
generator for my 2710 SA and I finally found one from an old Tektronix dealer.
Now I have this unit with Id No 119-3889-07 it's in a golden colored all
metal case with 3 female SMA connectors and a 10 pin 2 row IDC connector.
Luckily I found several old Mobo serial connectors that are the right
size but only use 9 of the 10 pins so I removed the wire and made a 10 wire straight pin to pin cable from the connector ends.
Now I open up my 2710 to locate grounds and power pins on this IDC
connector and have a WTF moment. Where the Heck does this thing go!! I remember a side section that was free but the unit won't fit and all the card slots are used up! Plus its just too big, so I have no clue how this fits into a 2710.
I can find nothing on the net about this Option 04 tracking generator,
it's not even listed in the 2710 service manual but I have seen several 2710 sold with Tracking Gens built in and the guy who sold it to me told me it works with the 2710 Arggggh....
Anybody have any idea what I should do next? I can't even locate the 10
pin connector that it supposed to plug into much less fit the unit into my 2710. Man I hope I did not waste all that money, I am having serious regrets now. Sorry if this comes of as a Crazy Rant but I should have done my homework before making this purchase, I am upset with myself.
Robert KD8MJR
|
Agreed, probably why they popped, undersized for the circuit. Sent on the TELUS Mobility network with BlackBerry
From: "Tom Jobe" <tomjobe@...>
Sender: TekScopes@...
Date: Sun, 23 Sep 2012 15:27:56 -0700 To: <TekScopes@...> ReplyTo: TekScopes@...
Subject: Re: [TekScopes] Re: WooHoo! 465B
?
You are way ahead of us on the rectifiers!
But... I think your biggest problem today?is going to be
an angry wife if you don't quit playing with that fine 465B
oscilloscope!
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 3:20
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
i did bend and shave to size for the holes on rectifiers for a 465M
fix. bigger is better when it comes to rectifiers.
--- In TekScopes@..., "Tom
Jobe" <tomjobe@...> wrote: > > The Tekscopes archive could
have saved you a lot of work Brad. > Your advice to go up on full wave
bridge voltage rating is excellent! > in my experience, the limiting
factor on a replacement bridge is the diameter of the leads coming out of the
bridge, and whether they will fit into the holes on the mainboard. You need to
lay the bridge over on an angle to get vertical clearance for a taller bridge,
as well as to gain access to solder the new bridge in place. > The
larger bridges have the leads on larger centers, but you can bend the leads
inward to fit the hole pattern in the mainboard. > tom jobe... >
> > > ----- Original Message ----- > From: Brad
King > To: TekScopes@...
> Sent: Sunday, September 23, 2012 3:01 PM > Subject: Re:
[TekScopes] Re: WooHoo! 465B > > > > I had to undo
the power transformer on mine and lift it up to access the other side of the
board to get at that 8v rectifier, replaced it with a garden variety nte 8
amp. > > > > Sent on the TELUS Mobility network
with BlackBerry > >
---------------------------------------------------------- > >
From: "Tom Jobe" <tomjobe@...> > Sender: TekScopes@...
> Date: Sun, 23 Sep 2012 14:59:34 -0700 > To: <TekScopes@...> >
ReplyTo: TekScopes@...
> Subject: Re: [TekScopes] Re: WooHoo! 465B > > >
> > I didn't notice if you said the low voltages were good or
not, but until every low voltage is good, you are wasting your time checking
anything else. > Another thing to look for is dead rectifiers. There are
a few of them right where the big PS caps are mounted and they are easy to
change without taking the scope apart. These are those full wave bridges all
in one big rectangular chip. > The Tekscopes Message archive has several
methods people have used successfully to change the rectifiers on the 4XX
scopes like the 465, 465B, 475, 474A, etc, if you don't have your own
method. > tom jobe... > > > > > -----
Original Message ----- > From: pdxareaid > To: TekScopes@...
> Sent: Sunday, September 23, 2012 2:04 PM > Subject: [TekScopes]
Re: WooHoo! 465B > > > > thanks for the tip. i did
not know tantalums smelled. > i'll probably hit big filter caps
anyway...i had a bad experience with them in a 465M among other PS components.
i have the service manual etc. need to read up and start checking things in
and around ps. afraid to leave it on and have failures cascade. quick voltage
measurements to start. > i understand failed tants are pretty easy to
spot. they fail shorted usually? is that right? > phil > >
--- In TekScopes@..., "Tom
Jobe" <tomjobe@> wrote: > > > > Hi Phil, > >
A free 465B is a wonderful thing! > > From my own experiences and
reading the Tekscopes group for years, maybe consider starting by checking the
resistance to ground at each power supply test point, then go looking for some
of those pesky tantalums in the 10uF to 47uF range on the shorted
supply(s). > > The tantalums make the bad smells when they die, more
so than the electrolytics. > > Maybe someone has a list of expected
resistances for the voltage test points on a 465B ? > > tom
jobe... > > > > > > > > ----- Original
Message ----- > > From: pdxareaid > > To: TekScopes@...
> > Sent: Sunday, September 23, 2012 1:42 PM > > Subject:
[TekScopes] WooHoo! 465B > > > > > > >
> I just picked up a 465B off Craigslist. Came with 10x probe. > >
> > the ad: > > "Unit was working, then stored for 3 or 4
years. When I turned it on recently, it died. There's a beam, but no sweep. It
worked well before that. I was hoping someone would want to fix it and put it
to good use. Free." > > > > Nice guy...engineer...just did
not want to deal with it. > > He mentioned electrical burning smell
when turning on. > > > > I got it home and spent 10 minutes
with it...opened it up and turned it on. Burning smell coming from PS section
I think. Whole screen glows off and on, no trace to speak of. > > It
is in excellent shape otherwise, calibration stickers all over it. >
> > > I am thinking PS filter caps to start, then tantalum and
electrolytic search after that. > > Phil >
> >
|
correct except not the wife...attorney. no fun. must not go out to scope, must not go out to scope...
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: You are way ahead of us on the rectifiers! But... I think your biggest problem today is going to be an angry wife if you don't quit playing with that fine 465B oscilloscope! tom jobe...
----- Original Message ----- From: pdxareaid To: TekScopes@... Sent: Sunday, September 23, 2012 3:20 PM Subject: [TekScopes] Re: WooHoo! 465B
i did bend and shave to size for the holes on rectifiers for a 465M fix. bigger is better when it comes to rectifiers.
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > The Tekscopes archive could have saved you a lot of work Brad. > Your advice to go up on full wave bridge voltage rating is excellent! > in my experience, the limiting factor on a replacement bridge is the diameter of the leads coming out of the bridge, and whether they will fit into the holes on the mainboard. You need to lay the bridge over on an angle to get vertical clearance for a taller bridge, as well as to gain access to solder the new bridge in place. > The larger bridges have the leads on larger centers, but you can bend the leads inward to fit the hole pattern in the mainboard. > tom jobe... > > > > ----- Original Message ----- > From: Brad King > To: TekScopes@... > Sent: Sunday, September 23, 2012 3:01 PM > Subject: Re: [TekScopes] Re: WooHoo! 465B > > > > I had to undo the power transformer on mine and lift it up to access the other side of the board to get at that 8v rectifier, replaced it with a garden variety nte 8 amp. > > > > Sent on the TELUS Mobility network with BlackBerry > > ---------------------------------------------------------- > > From: "Tom Jobe" <tomjobe@> > Sender: TekScopes@... > Date: Sun, 23 Sep 2012 14:59:34 -0700 > To: <TekScopes@...> > ReplyTo: TekScopes@... > Subject: Re: [TekScopes] Re: WooHoo! 465B > > > > > I didn't notice if you said the low voltages were good or not, but until every low voltage is good, you are wasting your time checking anything else. > Another thing to look for is dead rectifiers. There are a few of them right where the big PS caps are mounted and they are easy to change without taking the scope apart. These are those full wave bridges all in one big rectangular chip. > The Tekscopes Message archive has several methods people have used successfully to change the rectifiers on the 4XX scopes like the 465, 465B, 475, 474A, etc, if you don't have your own method. > tom jobe... > > > > > ----- Original Message ----- > From: pdxareaid > To: TekScopes@... > Sent: Sunday, September 23, 2012 2:04 PM > Subject: [TekScopes] Re: WooHoo! 465B > > > > thanks for the tip. i did not know tantalums smelled. > i'll probably hit big filter caps anyway...i had a bad experience with them in a 465M among other PS components. i have the service manual etc. need to read up and start checking things in and around ps. afraid to leave it on and have failures cascade. quick voltage measurements to start. > i understand failed tants are pretty easy to spot. they fail shorted usually? is that right? > phil > > --- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > > > Hi Phil, > > A free 465B is a wonderful thing! > > From my own experiences and reading the Tekscopes group for years, maybe consider starting by checking the resistance to ground at each power supply test point, then go looking for some of those pesky tantalums in the 10uF to 47uF range on the shorted supply(s). > > The tantalums make the bad smells when they die, more so than the electrolytics. > > Maybe someone has a list of expected resistances for the voltage test points on a 465B ? > > tom jobe... > > > > > > > > ----- Original Message ----- > > From: pdxareaid > > To: TekScopes@... > > Sent: Sunday, September 23, 2012 1:42 PM > > Subject: [TekScopes] WooHoo! 465B > > > > > > > > I just picked up a 465B off Craigslist. Came with 10x probe. > > > > the ad: > > "Unit was working, then stored for 3 or 4 years. When I turned it on recently, it died. There's a beam, but no sweep. It worked well before that. I was hoping someone would want to fix it and put it to good use. Free." > > > > Nice guy...engineer...just did not want to deal with it. > > He mentioned electrical burning smell when turning on. > > > > I got it home and spent 10 minutes with it...opened it up and turned it on. Burning smell coming from PS section I think. Whole screen glows off and on, no trace to speak of. > > It is in excellent shape otherwise, calibration stickers all over it. > > > > I am thinking PS filter caps to start, then tantalum and electrolytic search after that. > > Phil > > >
|
You are way ahead of us on the rectifiers!
But... I think your biggest problem today?is going to be
an angry wife if you don't quit playing with that fine 465B
oscilloscope!
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 3:20
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
i did bend and shave to size for the holes on rectifiers for a 465M
fix. bigger is better when it comes to rectifiers.
--- In TekScopes@..., "Tom
Jobe" wrote: > > The Tekscopes archive could
have saved you a lot of work Brad. > Your advice to go up on full wave
bridge voltage rating is excellent! > in my experience, the limiting
factor on a replacement bridge is the diameter of the leads coming out of the
bridge, and whether they will fit into the holes on the mainboard. You need to
lay the bridge over on an angle to get vertical clearance for a taller bridge,
as well as to gain access to solder the new bridge in place. > The
larger bridges have the leads on larger centers, but you can bend the leads
inward to fit the hole pattern in the mainboard. > tom jobe... >
> > > ----- Original Message ----- > From: Brad
King > To: TekScopes@...
> Sent: Sunday, September 23, 2012 3:01 PM > Subject: Re:
[TekScopes] Re: WooHoo! 465B > > > > I had to undo
the power transformer on mine and lift it up to access the other side of the
board to get at that 8v rectifier, replaced it with a garden variety nte 8
amp. > > > > Sent on the TELUS Mobility network
with BlackBerry > >
---------------------------------------------------------- > >
From: "Tom Jobe" > Sender: TekScopes@...
> Date: Sun, 23 Sep 2012 14:59:34 -0700 > To: <TekScopes@...> >
ReplyTo: TekScopes@...
> Subject: Re: [TekScopes] Re: WooHoo! 465B > > >
> > I didn't notice if you said the low voltages were good or
not, but until every low voltage is good, you are wasting your time checking
anything else. > Another thing to look for is dead rectifiers. There are
a few of them right where the big PS caps are mounted and they are easy to
change without taking the scope apart. These are those full wave bridges all
in one big rectangular chip. > The Tekscopes Message archive has several
methods people have used successfully to change the rectifiers on the 4XX
scopes like the 465, 465B, 475, 474A, etc, if you don't have your own
method. > tom jobe... > > > > > -----
Original Message ----- > From: pdxareaid > To: TekScopes@...
> Sent: Sunday, September 23, 2012 2:04 PM > Subject: [TekScopes]
Re: WooHoo! 465B > > > > thanks for the tip. i did
not know tantalums smelled. > i'll probably hit big filter caps
anyway...i had a bad experience with them in a 465M among other PS components.
i have the service manual etc. need to read up and start checking things in
and around ps. afraid to leave it on and have failures cascade. quick voltage
measurements to start. > i understand failed tants are pretty easy to
spot. they fail shorted usually? is that right? > phil > >
--- In TekScopes@..., "Tom
Jobe" wrote: > > > > Hi Phil, > >
A free 465B is a wonderful thing! > > From my own experiences and
reading the Tekscopes group for years, maybe consider starting by checking the
resistance to ground at each power supply test point, then go looking for some
of those pesky tantalums in the 10uF to 47uF range on the shorted
supply(s). > > The tantalums make the bad smells when they die, more
so than the electrolytics. > > Maybe someone has a list of expected
resistances for the voltage test points on a 465B ? > > tom
jobe... > > > > > > > > ----- Original
Message ----- > > From: pdxareaid > > To: TekScopes@...
> > Sent: Sunday, September 23, 2012 1:42 PM > > Subject:
[TekScopes] WooHoo! 465B > > > > > > >
> I just picked up a 465B off Craigslist. Came with 10x probe. > >
> > the ad: > > "Unit was working, then stored for 3 or 4
years. When I turned it on recently, it died. There's a beam, but no sweep. It
worked well before that. I was hoping someone would want to fix it and put it
to good use. Free." > > > > Nice guy...engineer...just did
not want to deal with it. > > He mentioned electrical burning smell
when turning on. > > > > I got it home and spent 10 minutes
with it...opened it up and turned it on. Burning smell coming from PS section
I think. Whole screen glows off and on, no trace to speak of. > > It
is in excellent shape otherwise, calibration stickers all over it. >
> > > I am thinking PS filter caps to start, then tantalum and
electrolytic search after that. > > Phil >
> >
|
That's what I had to do, bend the leads to fit. It was tight but it worked :) Sent on the TELUS Mobility network with BlackBerry
From: "Tom Jobe" <tomjobe@...>
Sender: TekScopes@...
Date: Sun, 23 Sep 2012 15:16:03 -0700 To: <TekScopes@...> ReplyTo: TekScopes@...
Subject: Re: [TekScopes] Re: WooHoo! 465B
?
The Tekscopes archive could have saved you a lot of work
Brad.
Your advice to go up on full wave bridge voltage rating is
excellent!
in my experience, the limiting factor on a replacement bridge
is the diameter of the leads coming out of the bridge, and whether they will fit
into the holes on the mainboard. You need to lay the bridge over on an angle to
get vertical clearance for a taller bridge, as well as to gain access to solder
the new bridge in place.
The larger bridges have the leads on larger centers, but you
can bend the leads inward to fit the hole pattern in the mainboard.
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 3:01
PM
Subject: Re: [TekScopes] Re: WooHoo!
465B
?
I had to undo the power transformer on mine and lift it up to access the
other side of the board to get at that 8v rectifier, replaced it with a garden
variety nte 8 amp.
Sent on the TELUS Mobility network with BlackBerry
Date: Sun, 23 Sep 2012 14:59:34 -0700
To: <TekScopes@...>
ReplyTo: TekScopes@...
Subject: Re: [TekScopes] Re: WooHoo! 465B
?
I didn't notice if you said the low voltages were good or
not, but until every low voltage is good, you are wasting your time checking
anything else.
Another thing to look for is dead rectifiers. There are a
few of them right where the big PS caps are mounted and they are easy to
change without taking the scope apart. These are those full wave bridges
all?in one big rectangular chip.
The Tekscopes Message archive?has several?methods
people have used successfully to change the rectifiers on the 4XX scopes like
the 465, 465B, 475, 474A, etc, if you don't have your own method.
tom jobe...
?
?
?
?
----- Original Message -----
Sent: Sunday, September 23, 2012 2:04
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
thanks for the tip. i did not know tantalums smelled. i'll probably
hit big filter caps anyway...i had a bad experience with them in a 465M
among other PS components. i have the service manual etc. need to read up
and start checking things in and around ps. afraid to leave it on and have
failures cascade. quick voltage measurements to start. i understand
failed tants are pretty easy to spot. they fail shorted usually? is that
right? phil
--- In TekScopes@...,
"Tom Jobe" <tomjobe@...> wrote: > > Hi Phil, > A
free 465B is a wonderful thing! > From my own experiences and reading
the Tekscopes group for years, maybe consider starting by checking the
resistance to ground at each power supply test point, then go looking for
some of those pesky tantalums in the 10uF to 47uF range on the shorted
supply(s). > The tantalums make the bad smells when they die, more so
than the electrolytics. > Maybe someone has a list of expected
resistances for the voltage test points on a 465B ? > tom
jobe... > > > > ----- Original Message -----
> From: pdxareaid > To: TekScopes@...
> Sent: Sunday, September 23, 2012 1:42 PM > Subject:
[TekScopes] WooHoo! 465B > > > > I just picked up
a 465B off Craigslist. Came with 10x probe. > > the ad: >
"Unit was working, then stored for 3 or 4 years. When I turned it on
recently, it died. There's a beam, but no sweep. It worked well before that.
I was hoping someone would want to fix it and put it to good use.
Free." > > Nice guy...engineer...just did not want to deal with
it. > He mentioned electrical burning smell when turning on. >
> I got it home and spent 10 minutes with it...opened it up and
turned it on. Burning smell coming from PS section I think. Whole screen
glows off and on, no trace to speak of. > It is in excellent shape
otherwise, calibration stickers all over it. > > I am thinking
PS filter caps to start, then tantalum and electrolytic search after
that. >
Phil >
|
i did bend and shave to size for the holes on rectifiers for a 465M fix. bigger is better when it comes to rectifiers.
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: The Tekscopes archive could have saved you a lot of work Brad. Your advice to go up on full wave bridge voltage rating is excellent! in my experience, the limiting factor on a replacement bridge is the diameter of the leads coming out of the bridge, and whether they will fit into the holes on the mainboard. You need to lay the bridge over on an angle to get vertical clearance for a taller bridge, as well as to gain access to solder the new bridge in place. The larger bridges have the leads on larger centers, but you can bend the leads inward to fit the hole pattern in the mainboard. tom jobe...
----- Original Message ----- From: Brad King To: TekScopes@... Sent: Sunday, September 23, 2012 3:01 PM Subject: Re: [TekScopes] Re: WooHoo! 465B
I had to undo the power transformer on mine and lift it up to access the other side of the board to get at that 8v rectifier, replaced it with a garden variety nte 8 amp.
Sent on the TELUS Mobility network with BlackBerry
------------------------------------------------------------------------------
From: "Tom Jobe" <tomjobe@...> Sender: TekScopes@... Date: Sun, 23 Sep 2012 14:59:34 -0700 To: <TekScopes@...> ReplyTo: TekScopes@... Subject: Re: [TekScopes] Re: WooHoo! 465B
I didn't notice if you said the low voltages were good or not, but until every low voltage is good, you are wasting your time checking anything else. Another thing to look for is dead rectifiers. There are a few of them right where the big PS caps are mounted and they are easy to change without taking the scope apart. These are those full wave bridges all in one big rectangular chip. The Tekscopes Message archive has several methods people have used successfully to change the rectifiers on the 4XX scopes like the 465, 465B, 475, 474A, etc, if you don't have your own method. tom jobe...
----- Original Message ----- From: pdxareaid To: TekScopes@... Sent: Sunday, September 23, 2012 2:04 PM Subject: [TekScopes] Re: WooHoo! 465B
thanks for the tip. i did not know tantalums smelled. i'll probably hit big filter caps anyway...i had a bad experience with them in a 465M among other PS components. i have the service manual etc. need to read up and start checking things in and around ps. afraid to leave it on and have failures cascade. quick voltage measurements to start. i understand failed tants are pretty easy to spot. they fail shorted usually? is that right? phil
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > Hi Phil, > A free 465B is a wonderful thing! > From my own experiences and reading the Tekscopes group for years, maybe consider starting by checking the resistance to ground at each power supply test point, then go looking for some of those pesky tantalums in the 10uF to 47uF range on the shorted supply(s). > The tantalums make the bad smells when they die, more so than the electrolytics. > Maybe someone has a list of expected resistances for the voltage test points on a 465B ? > tom jobe... > > > > ----- Original Message ----- > From: pdxareaid > To: TekScopes@... > Sent: Sunday, September 23, 2012 1:42 PM > Subject: [TekScopes] WooHoo! 465B > > > > I just picked up a 465B off Craigslist. Came with 10x probe. > > the ad: > "Unit was working, then stored for 3 or 4 years. When I turned it on recently, it died. There's a beam, but no sweep. It worked well before that. I was hoping someone would want to fix it and put it to good use. Free." > > Nice guy...engineer...just did not want to deal with it. > He mentioned electrical burning smell when turning on. > > I got it home and spent 10 minutes with it...opened it up and turned it on. Burning smell coming from PS section I think. Whole screen glows off and on, no trace to speak of. > It is in excellent shape otherwise, calibration stickers all over it. > > I am thinking PS filter caps to start, then tantalum and electrolytic search after that. > Phil >
|
The Tekscopes archive could have saved you a lot of work
Brad.
Your advice to go up on full wave bridge voltage rating is
excellent!
in my experience, the limiting factor on a replacement bridge
is the diameter of the leads coming out of the bridge, and whether they will fit
into the holes on the mainboard. You need to lay the bridge over on an angle to
get vertical clearance for a taller bridge, as well as to gain access to solder
the new bridge in place.
The larger bridges have the leads on larger centers, but you
can bend the leads inward to fit the hole pattern in the mainboard.
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 3:01
PM
Subject: Re: [TekScopes] Re: WooHoo!
465B
?
I had to undo the power transformer on mine and lift it up to access the
other side of the board to get at that 8v rectifier, replaced it with a garden
variety nte 8 amp.
Sent on the TELUS Mobility network with BlackBerry
Date: Sun, 23 Sep 2012 14:59:34 -0700
To:
ReplyTo: TekScopes@...
Subject: Re: [TekScopes] Re: WooHoo! 465B
?
I didn't notice if you said the low voltages were good or
not, but until every low voltage is good, you are wasting your time checking
anything else.
Another thing to look for is dead rectifiers. There are a
few of them right where the big PS caps are mounted and they are easy to
change without taking the scope apart. These are those full wave bridges
all?in one big rectangular chip.
The Tekscopes Message archive?has several?methods
people have used successfully to change the rectifiers on the 4XX scopes like
the 465, 465B, 475, 474A, etc, if you don't have your own method.
tom jobe...
?
?
?
?
----- Original Message -----
Sent: Sunday, September 23, 2012 2:04
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
thanks for the tip. i did not know tantalums smelled. i'll probably
hit big filter caps anyway...i had a bad experience with them in a 465M
among other PS components. i have the service manual etc. need to read up
and start checking things in and around ps. afraid to leave it on and have
failures cascade. quick voltage measurements to start. i understand
failed tants are pretty easy to spot. they fail shorted usually? is that
right? phil
--- In TekScopes@...,
"Tom Jobe" wrote: > > Hi Phil, > A
free 465B is a wonderful thing! > From my own experiences and reading
the Tekscopes group for years, maybe consider starting by checking the
resistance to ground at each power supply test point, then go looking for
some of those pesky tantalums in the 10uF to 47uF range on the shorted
supply(s). > The tantalums make the bad smells when they die, more so
than the electrolytics. > Maybe someone has a list of expected
resistances for the voltage test points on a 465B ? > tom
jobe... > > > > ----- Original Message -----
> From: pdxareaid > To: TekScopes@...
> Sent: Sunday, September 23, 2012 1:42 PM > Subject:
[TekScopes] WooHoo! 465B > > > > I just picked up
a 465B off Craigslist. Came with 10x probe. > > the ad: >
"Unit was working, then stored for 3 or 4 years. When I turned it on
recently, it died. There's a beam, but no sweep. It worked well before that.
I was hoping someone would want to fix it and put it to good use.
Free." > > Nice guy...engineer...just did not want to deal with
it. > He mentioned electrical burning smell when turning on. >
> I got it home and spent 10 minutes with it...opened it up and
turned it on. Burning smell coming from PS section I think. Whole screen
glows off and on, no trace to speak of. > It is in excellent shape
otherwise, calibration stickers all over it. > > I am thinking
PS filter caps to start, then tantalum and electrolytic search after
that. >
Phil >
|
i agree, all ps's are first check (except i just found a smoking tantalum). i have a few big rectifiers left over from a 465M fix. if they are a prob i'll be able to hit em pretty easily. i have not even had a volt meter on it yet. i do have a working scope to debug it. thanks.
toggle quoted message
Show quoted text
--- In TekScopes@..., "Tom Jobe" <tomjobe@...> wrote: I didn't notice if you said the low voltages were good or not, but until every low voltage is good, you are wasting your time checking anything else. Another thing to look for is dead rectifiers. There are a few of them right where the big PS caps are mounted and they are easy to change without taking the scope apart. These are those full wave bridges all in one big rectangular chip. The Tekscopes Message archive has several methods people have used successfully to change the rectifiers on the 4XX scopes like the 465, 465B, 475, 474A, etc, if you don't have your own method. tom jobe...
----- Original Message ----- From: pdxareaid To: TekScopes@... Sent: Sunday, September 23, 2012 2:04 PM Subject: [TekScopes] Re: WooHoo! 465B
thanks for the tip. i did not know tantalums smelled. i'll probably hit big filter caps anyway...i had a bad experience with them in a 465M among other PS components. i have the service manual etc. need to read up and start checking things in and around ps. afraid to leave it on and have failures cascade. quick voltage measurements to start. i understand failed tants are pretty easy to spot. they fail shorted usually? is that right? phil
--- In TekScopes@..., "Tom Jobe" <tomjobe@> wrote: > > Hi Phil, > A free 465B is a wonderful thing! > From my own experiences and reading the Tekscopes group for years, maybe consider starting by checking the resistance to ground at each power supply test point, then go looking for some of those pesky tantalums in the 10uF to 47uF range on the shorted supply(s). > The tantalums make the bad smells when they die, more so than the electrolytics. > Maybe someone has a list of expected resistances for the voltage test points on a 465B ? > tom jobe... > > > > ----- Original Message ----- > From: pdxareaid > To: TekScopes@... > Sent: Sunday, September 23, 2012 1:42 PM > Subject: [TekScopes] WooHoo! 465B > > > > I just picked up a 465B off Craigslist. Came with 10x probe. > > the ad: > "Unit was working, then stored for 3 or 4 years. When I turned it on recently, it died. There's a beam, but no sweep. It worked well before that. I was hoping someone would want to fix it and put it to good use. Free." > > Nice guy...engineer...just did not want to deal with it. > He mentioned electrical burning smell when turning on. > > I got it home and spent 10 minutes with it...opened it up and turned it on. Burning smell coming from PS section I think. Whole screen glows off and on, no trace to speak of. > It is in excellent shape otherwise, calibration stickers all over it. > > I am thinking PS filter caps to start, then tantalum and electrolytic search after that. > Phil >
|
trigger view did nothing. but i had it on long enough that i noticed a smoking tantalum. and it was even moaning! ikyn it was making noise when it started to smoke. it was in a cluster of 3 tants on the large rightmost board (viewed from the front) towards the rear...i don't even know what board that is yet but i will as soon as i can get to look at the service manual. phil
toggle quoted message
Show quoted text
--- In TekScopes@..., "pdxareaid" <public_email@...> wrote: arggg...i fogot to poke the trigger view. one last trip out there right now.
--- In TekScopes@..., "Brad King" <cpgixxer@> wrote:
Trigger view as far as I know detaches the vertical pre from the circuit. The service manuals have block diagrams to help segregate the circuit to narrow it down. Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:48:37 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
lol..i have not had time to do much of anything with it yet (only had it a couple of hours) but i can't resist dropping everything and running out to poke it with these tips rolling in...
when the screen is glowing, none of the controls seems to affect it and it generally lights up the whole screen.
using the beamfinder, i get a bright dot that jumps around and in and out of what looks like it might be trying to show me a reasonable beamfinder display, but severely shift off screen to the right. the position controls seem to affect it but not in a consistent way...but there is some effect. all of the beamfinder action appears to be on the right half of the screen. phil
ps. i'm going to be in trouble if i don't finish some other stuff today...i'll leave poking it again til tomorrow but will check in to soak up the tips for when i do get out there. thanks for all the help. keep em coming :-)
--- In TekScopes@..., "Brad King" <cpgixxer@> wrote:
The glow is because the trace isn't centered, could be a bad resistor or transistor in the vertical. Does it move h or v with the adjustment? Did you try pressing the trigger view button? Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:25:55 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
ah, the beamfinder...never was very fond of it. now maybe i see why it is there. thanks for the tip.
sweep rate does not seem to affect anything.
beamfinder: I got the usual bright connected 2 (i assume) point display but severely shifted to the right such that only 1/2 of the expected display is on the screen. curiously, when i tried that, and released the beamfinder, i got a bright point on the display rather than a full screen glow. then it went back to the glow. i tried the beamfinder a few more times and it seemed to start bringing something back from the dead. i even saw it flash what looked like it might have been a reasonable trace...but it disappeared very quickly. the burning smell was making me nervous so i shut it down.
are these useful clues?
--- In TekScopes@..., larrys@ wrote:
"pdxareaid" <public_email@> wrote:
Whole screen glows off and on, no trace to speak of. Does the rate of the on / off vary with the sweep speed setting? Sounds like you may have a trace, just offscreen. See if the beamfinder shows it too you.
-ls-
|
Just a note of caution ... Years ago we used many of these gold relays and another Aromat series in dark blue package (smaller than the DS). If you dropped them (dark blue) from 8" onto a bench surface (without mat), most were permanently damaged. The DS seems to be a very different construction but I would still handle with care. Den
toggle quoted message
Show quoted text
--- In TekScopes@..., "denyhstk" <denyhstk@...> wrote: I was also surprised by the garden variety relays. Tek must have characterized them and also, they must be fairly consistent from lot to lot.
Somewhere I read about poor return loss into the TDS input ports. Just for fun I examined mine and while I can't recall the numbers, they weren't that bad (for a scope).
Dan, Good point on the preheating of the substrates. I did this on one and then stopped. The substrates were still warm from removal of the relay leg "stubs" after cutting and from cleanup with solder-wick.
I'm not sure what metal shield (as return path) you'r referring to. I thought all connections to the hybrids were via the gold plated fuzz-buttons in the plastic carrier plates.
Den
--- In TekScopes@..., "mickeyaslam" <mickeyaslam@> wrote:
When I first saw the relays I was shocked that Tektronix is using cheap relays for supposedly high end test equipment. Same goes for electrolytic caps.
--- In TekScopes@..., "Dan Meeks" <dan-meeks@> wrote:
There was a known issue with some TDS scopes caused by poor solder on the metal shield on the ceramic attenuator boards. The legs on that shield provide the ground return for the relays. So a bad solder joint can cause all sorts of odd front-end issues.
Be aware that re-soldering is tricky because of the ceramic. If you can't pre-heat the board, be careful and don't apply too much heat from the soldering iron or it will crack. I have soldered a bunch of these, so it's not a super-high risky thing, but just be careful.
Dan
|
I had to undo the power transformer on mine and lift it up to access the other side of the board to get at that 8v rectifier, replaced it with a garden variety nte 8 amp. Sent on the TELUS Mobility network with BlackBerry
From: "Tom Jobe" <tomjobe@...>
Sender: TekScopes@...
Date: Sun, 23 Sep 2012 14:59:34 -0700 To: <TekScopes@...> ReplyTo: TekScopes@...
Subject: Re: [TekScopes] Re: WooHoo! 465B
?
I didn't notice if you said the low voltages were good or not,
but until every low voltage is good, you are wasting your time checking anything
else.
Another thing to look for is dead rectifiers. There are a few
of them right where the big PS caps are mounted and they are easy to change
without taking the scope apart. These are those full wave bridges all?in
one big rectangular chip.
The Tekscopes Message archive?has several?methods
people have used successfully to change the rectifiers on the 4XX scopes like
the 465, 465B, 475, 474A, etc, if you don't have your own method.
tom jobe...
?
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 2:04
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
thanks for the tip. i did not know tantalums smelled. i'll probably hit
big filter caps anyway...i had a bad experience with them in a 465M among
other PS components. i have the service manual etc. need to read up and start
checking things in and around ps. afraid to leave it on and have failures
cascade. quick voltage measurements to start. i understand failed tants are
pretty easy to spot. they fail shorted usually? is that
right? phil
--- In TekScopes@..., "Tom
Jobe" <tomjobe@...> wrote: > > Hi Phil, > A free 465B
is a wonderful thing! > From my own experiences and reading the
Tekscopes group for years, maybe consider starting by checking the resistance
to ground at each power supply test point, then go looking for some of those
pesky tantalums in the 10uF to 47uF range on the shorted supply(s). >
The tantalums make the bad smells when they die, more so than the
electrolytics. > Maybe someone has a list of expected resistances for
the voltage test points on a 465B ? > tom jobe... > >
> > ----- Original Message ----- > From: pdxareaid
> To: TekScopes@...
> Sent: Sunday, September 23, 2012 1:42 PM > Subject: [TekScopes]
WooHoo! 465B > > > > I just picked up a 465B off
Craigslist. Came with 10x probe. > > the ad: > "Unit was
working, then stored for 3 or 4 years. When I turned it on recently, it died.
There's a beam, but no sweep. It worked well before that. I was hoping someone
would want to fix it and put it to good use. Free." > > Nice
guy...engineer...just did not want to deal with it. > He mentioned
electrical burning smell when turning on. > > I got it home and
spent 10 minutes with it...opened it up and turned it on. Burning smell coming
from PS section I think. Whole screen glows off and on, no trace to speak
of. > It is in excellent shape otherwise, calibration stickers all over
it. > > I am thinking PS filter caps to start, then tantalum and
electrolytic search after that. > Phil >
|
I didn't notice if you said the low voltages were good or not,
but until every low voltage is good, you are wasting your time checking anything
else.
Another thing to look for is dead rectifiers. There are a few
of them right where the big PS caps are mounted and they are easy to change
without taking the scope apart. These are those full wave bridges all?in
one big rectangular chip.
The Tekscopes Message archive?has several?methods
people have used successfully to change the rectifiers on the 4XX scopes like
the 465, 465B, 475, 474A, etc, if you don't have your own method.
tom jobe...
?
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 2:04
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
thanks for the tip. i did not know tantalums smelled. i'll probably hit
big filter caps anyway...i had a bad experience with them in a 465M among
other PS components. i have the service manual etc. need to read up and start
checking things in and around ps. afraid to leave it on and have failures
cascade. quick voltage measurements to start. i understand failed tants are
pretty easy to spot. they fail shorted usually? is that
right? phil
--- In TekScopes@..., "Tom
Jobe" wrote: > > Hi Phil, > A free 465B
is a wonderful thing! > From my own experiences and reading the
Tekscopes group for years, maybe consider starting by checking the resistance
to ground at each power supply test point, then go looking for some of those
pesky tantalums in the 10uF to 47uF range on the shorted supply(s). >
The tantalums make the bad smells when they die, more so than the
electrolytics. > Maybe someone has a list of expected resistances for
the voltage test points on a 465B ? > tom jobe... > >
> > ----- Original Message ----- > From: pdxareaid
> To: TekScopes@...
> Sent: Sunday, September 23, 2012 1:42 PM > Subject: [TekScopes]
WooHoo! 465B > > > > I just picked up a 465B off
Craigslist. Came with 10x probe. > > the ad: > "Unit was
working, then stored for 3 or 4 years. When I turned it on recently, it died.
There's a beam, but no sweep. It worked well before that. I was hoping someone
would want to fix it and put it to good use. Free." > > Nice
guy...engineer...just did not want to deal with it. > He mentioned
electrical burning smell when turning on. > > I got it home and
spent 10 minutes with it...opened it up and turned it on. Burning smell coming
from PS section I think. Whole screen glows off and on, no trace to speak
of. > It is in excellent shape otherwise, calibration stickers all over
it. > > I am thinking PS filter caps to start, then tantalum and
electrolytic search after that. > Phil >
|
I was also surprised by the garden variety relays. Tek must have characterized them and also, they must be fairly consistent from lot to lot.
Somewhere I read about poor return loss into the TDS input ports. Just for fun I examined mine and while I can't recall the numbers, they weren't that bad (for a scope).
Dan, Good point on the preheating of the substrates. I did this on one and then stopped. The substrates were still warm from removal of the relay leg "stubs" after cutting and from cleanup with solder-wick.
I'm not sure what metal shield (as return path) you'r referring to. I thought all connections to the hybrids were via the gold plated fuzz-buttons in the plastic carrier plates.
Den
toggle quoted message
Show quoted text
--- In TekScopes@..., "mickeyaslam" <mickeyaslam@...> wrote: When I first saw the relays I was shocked that Tektronix is using cheap relays for supposedly high end test equipment. Same goes for electrolytic caps.
--- In TekScopes@..., "Dan Meeks" <dan-meeks@> wrote:
There was a known issue with some TDS scopes caused by poor solder on the metal shield on the ceramic attenuator boards. The legs on that shield provide the ground return for the relays. So a bad solder joint can cause all sorts of odd front-end issues.
Be aware that re-soldering is tricky because of the ceramic. If you can't pre-heat the board, be careful and don't apply too much heat from the soldering iron or it will crack. I have soldered a bunch of these, so it's not a super-high risky thing, but just be careful.
Dan
|
arggg...i fogot to poke the trigger view. one last trip out there right now.
toggle quoted message
Show quoted text
--- In TekScopes@..., "Brad King" <cpgixxer@...> wrote: Trigger view as far as I know detaches the vertical pre from the circuit. The service manuals have block diagrams to help segregate the circuit to narrow it down. Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@...> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:48:37 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
lol..i have not had time to do much of anything with it yet (only had it a couple of hours) but i can't resist dropping everything and running out to poke it with these tips rolling in...
when the screen is glowing, none of the controls seems to affect it and it generally lights up the whole screen.
using the beamfinder, i get a bright dot that jumps around and in and out of what looks like it might be trying to show me a reasonable beamfinder display, but severely shift off screen to the right. the position controls seem to affect it but not in a consistent way...but there is some effect. all of the beamfinder action appears to be on the right half of the screen. phil
ps. i'm going to be in trouble if i don't finish some other stuff today...i'll leave poking it again til tomorrow but will check in to soak up the tips for when i do get out there. thanks for all the help. keep em coming :-)
--- In TekScopes@..., "Brad King" <cpgixxer@> wrote:
The glow is because the trace isn't centered, could be a bad resistor or transistor in the vertical. Does it move h or v with the adjustment? Did you try pressing the trigger view button? Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:25:55 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
ah, the beamfinder...never was very fond of it. now maybe i see why it is there. thanks for the tip.
sweep rate does not seem to affect anything.
beamfinder: I got the usual bright connected 2 (i assume) point display but severely shifted to the right such that only 1/2 of the expected display is on the screen. curiously, when i tried that, and released the beamfinder, i got a bright point on the display rather than a full screen glow. then it went back to the glow. i tried the beamfinder a few more times and it seemed to start bringing something back from the dead. i even saw it flash what looked like it might have been a reasonable trace...but it disappeared very quickly. the burning smell was making me nervous so i shut it down.
are these useful clues?
--- In TekScopes@..., larrys@ wrote:
"pdxareaid" <public_email@> wrote:
Whole screen glows off and on, no trace to speak of. Does the rate of the on / off vary with the sweep speed setting? Sounds like you may have a trace, just offscreen. See if the beamfinder shows it too you.
-ls-
|
Trigger view as far as I know detaches the vertical pre from the circuit. The service manuals have block diagrams to help segregate the circuit to narrow it down. Sent on the TELUS Mobility network with BlackBerry
From: "pdxareaid" <public_email@...>
Sender: TekScopes@...
Date: Sun, 23 Sep 2012 21:48:37 -0000 To: <TekScopes@...> ReplyTo: TekScopes@...
Subject: [TekScopes] Re: WooHoo! 465B
?
lol..i have not had time to do much of anything with it yet (only had it a couple of hours) but i can't resist dropping everything and running out to poke it with these tips rolling in...
when the screen is glowing, none of the controls seems to affect it and it generally lights up the whole screen.
using the beamfinder, i get a bright dot that jumps around and in and out of what looks like it might be trying to show me a reasonable beamfinder display, but severely shift off screen to the right.
the position controls seem to affect it but not in a consistent way...but there is some effect.
all of the beamfinder action appears to be on the right half of the screen.
phil
ps. i'm going to be in trouble if i don't finish some other stuff today...i'll leave poking it again til tomorrow but will check in to soak up the tips for when i do get out there.
thanks for all the help. keep em coming :-)
--- In TekScopes@..., "Brad King" <cpgixxer@...> wrote:
>
> The glow is because the trace isn't centered, could be a bad resistor or transistor in the vertical. Does it move h or v with the adjustment? Did you try pressing the trigger view button?
> Sent on the TELUS Mobility network with BlackBerry
>
toggle quoted message
Show quoted text
> -----Original Message-----
> From: "pdxareaid" <public_email@...>
> Sender: TekScopes@...
> Date: Sun, 23 Sep 2012 21:25:55
> To: < TekScopes@...>
> Reply-To: TekScopes@...
> Subject: [TekScopes] Re: WooHoo! 465B
>
>
> ah, the beamfinder...never was very fond of it. now maybe i see why it is there. thanks for the tip.
>
> sweep rate does not seem to affect anything.
>
> beamfinder:
> I got the usual bright connected 2 (i assume) point display but severely shifted to the right such that only 1/2 of the expected display is on the screen.
> curiously, when i tried that, and released the beamfinder, i got a bright point on the display rather than a full screen glow. then it went back to the glow.
> i tried the beamfinder a few more times and it seemed to start bringing something back from the dead. i even saw it flash what looked like it might have been a reasonable trace...but it disappeared very quickly.
> the burning smell was making me nervous so i shut it down.
>
> are these useful clues?
>
>
> --- In TekScopes@..., larrys@ wrote:
> >
> > "pdxareaid" <public_email@> wrote:
> > > Whole screen glows off and on, no trace to speak of.
> >
> > Does the rate of the on / off vary with the sweep speed setting?
> > Sounds like you may have a trace, just offscreen. See if the
> > beamfinder shows it too you.
> >
> > -ls-
> >
>
|
lol..i have not had time to do much of anything with it yet (only had it a couple of hours) but i can't resist dropping everything and running out to poke it with these tips rolling in...
when the screen is glowing, none of the controls seems to affect it and it generally lights up the whole screen.
using the beamfinder, i get a bright dot that jumps around and in and out of what looks like it might be trying to show me a reasonable beamfinder display, but severely shift off screen to the right. the position controls seem to affect it but not in a consistent way...but there is some effect. all of the beamfinder action appears to be on the right half of the screen. phil
ps. i'm going to be in trouble if i don't finish some other stuff today...i'll leave poking it again til tomorrow but will check in to soak up the tips for when i do get out there. thanks for all the help. keep em coming :-)
toggle quoted message
Show quoted text
--- In TekScopes@..., "Brad King" <cpgixxer@...> wrote: The glow is because the trace isn't centered, could be a bad resistor or transistor in the vertical. Does it move h or v with the adjustment? Did you try pressing the trigger view button? Sent on the TELUS Mobility network with BlackBerry
-----Original Message----- From: "pdxareaid" <public_email@...> Sender: TekScopes@... Date: Sun, 23 Sep 2012 21:25:55 To: <TekScopes@...> Reply-To: TekScopes@... Subject: [TekScopes] Re: WooHoo! 465B
ah, the beamfinder...never was very fond of it. now maybe i see why it is there. thanks for the tip.
sweep rate does not seem to affect anything.
beamfinder: I got the usual bright connected 2 (i assume) point display but severely shifted to the right such that only 1/2 of the expected display is on the screen. curiously, when i tried that, and released the beamfinder, i got a bright point on the display rather than a full screen glow. then it went back to the glow. i tried the beamfinder a few more times and it seemed to start bringing something back from the dead. i even saw it flash what looked like it might have been a reasonable trace...but it disappeared very quickly. the burning smell was making me nervous so i shut it down.
are these useful clues?
--- In TekScopes@..., larrys@ wrote:
"pdxareaid" <public_email@> wrote:
Whole screen glows off and on, no trace to speak of. Does the rate of the on / off vary with the sweep speed setting? Sounds like you may have a trace, just offscreen. See if the beamfinder shows it too you.
-ls-
|
I do believe that the tantalums often fail shorted, at least
when they don't blow up.
Maybe one of the more knowledgeable?members will weigh in
on this question.
?
Flip the scope over on its back and look for the few dipped
tantalums that have multiple colors on them. One of the problem
tantalums?is along the back edge of the mainboard near the high voltage
shield. I think there is another one or two under that HV shield as well as one
out in the middle of the mainboard. The tantalums that are all one color seem to
be more reliable. Some of the problem tantalums are rated for 20 volts on a 15
volt supply, so go up on the voltage rating?if you replace any of
them.
tom jobe...
?
?
?
toggle quoted message
Show quoted text
----- Original Message -----
Sent: Sunday, September 23, 2012 2:04
PM
Subject: [TekScopes] Re: WooHoo!
465B
?
thanks for the tip. i did not know tantalums smelled. i'll probably hit
big filter caps anyway...i had a bad experience with them in a 465M among
other PS components. i have the service manual etc. need to read up and start
checking things in and around ps. afraid to leave it on and have failures
cascade. quick voltage measurements to start. i understand failed tants are
pretty easy to spot. they fail shorted usually? is that
right? phil
--- In TekScopes@..., "Tom
Jobe" > > Hi Phil, > A free 465B
is a wonderful thing! > From my own experiences and reading the
Tekscopes group for years, maybe consider starting by checking the resistance
to ground at each power supply test point, then go looking for some of those
pesky tantalums in the 10uF to 47uF range on the shorted supply(s). >
The tantalums make the bad smells when they die, more so than the
electrolytics. > Maybe someone has a list of expected resistances for
the voltage test points on a 465B ? > tom jobe... > >
> > ----- Original Message ----- > From: pdxareaid
> To: TekScopes@...
> Sent: Sunday, September 23, 2012 1:42 PM > Subject: [TekScopes]
WooHoo! 465B > > > > I just picked up a 465B off
Craigslist. Came with 10x probe. > > the ad: > "Unit was
working, then stored for 3 or 4 years. When I turned it on recently, it died.
There's a beam, but no sweep. It worked well before that. I was hoping someone
would want to fix it and put it to good use. Free." > > Nice
guy...engineer...just did not want to deal with it. > He mentioned
electrical burning smell when turning on. > > I got it home and
spent 10 minutes with it...opened it up and turned it on. Burning smell coming
from PS section I think. Whole screen glows off and on, no trace to speak
of. > It is in excellent shape otherwise, calibration stickers all over
it. > > I am thinking PS filter caps to start, then tantalum and
electrolytic search after that. > Phil >
|
The glow is because the trace isn't centered, could be a bad resistor or transistor in the vertical. Does it move h or v with the adjustment? Did you try pressing the trigger view button? Sent on the TELUS Mobility network with BlackBerry
From: "pdxareaid" <public_email@...>
Sender: TekScopes@...
Date: Sun, 23 Sep 2012 21:25:55 -0000 To: <TekScopes@...> ReplyTo: TekScopes@...
Subject: [TekScopes] Re: WooHoo! 465B
?
ah, the beamfinder...never was very fond of it. now maybe i see why it is there. thanks for the tip.
sweep rate does not seem to affect anything.
beamfinder:
I got the usual bright connected 2 (i assume) point display but severely shifted to the right such that only 1/2 of the expected display is on the screen.
curiously, when i tried that, and released the beamfinder, i got a bright point on the display rather than a full screen glow. then it went back to the glow.
i tried the beamfinder a few more times and it seemed to start bringing something back from the dead. i even saw it flash what looked like it might have been a reasonable trace...but it disappeared very quickly.
the burning smell was making me nervous so i shut it down.
are these useful clues?
--- In TekScopes@..., larrys@... wrote:
>
> "pdxareaid" <public_email@...> wrote:
> > Whole screen glows off and on, no trace to speak of.
>
> Does the rate of the on / off vary with the sweep speed setting?
> Sounds like you may have a trace, just offscreen. See if the
> beamfinder shows it too you.
>
> -ls-
>
|