¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io
Date

Re: 7854: Excessive stored trace brightness.

 

Do you have a dual time base, a 7B87, or a 7B80/7B85 (or equivalent)
which you could use to check if the A intensified by B display is
correct?

I wonder if your readout troubles are related. The stored trace
intensify command could be controlled through the readout or from the
same place the readout is controlled from. Verifying it should be as
simple as disabling that function by pulling a transistor but I did a
cursory search of the service manual and did not find out it.

You might try fiddling with the contrast adjustment on the front of
the oscilloscope to see if that affects anything. I know mine has
trouble with the intensified display when using some dual time bases.

On Mon, 9 Jul 2012 12:36:13 -0400, David Holland
<david.w.holland@...> wrote:

Yup. Blanking looks good to me...

I'll double check the pot, and make certain its smooth. It looks
like I have some reading to do in the manual....(this is certainly
going to be umm.. interesting)

David


On Mon, Jul 9, 2012 at 11:32 AM, <larrys@...> wrote:
Are the parts that are supposed to be blanked getting blanked?
-ls-

David Holland <david.w.holland@...> wrote:
Hello all,

I'm kind of getting around to poking at my 7854 scope.

The stored trace brightness is excessive, and the adjustment won't go
down any further. I'm reasonably certain the pot is good, since
it'll smoothy vary over what range it does have. (From "too bright",
to "cr*p, I'd better turn that down before the CRT gets burnt").
Normal display, and readout's are of a more reasonable intensity, and
have goodly range. (Readout isn't working correctly, but that's
not the problem I want to concentrate first on.)

Before I crack the covers, any suggestions of which parts of the
service manual would be most relevant to read...

thanks in advance...

David


------------------------------------

Yahoo! Groups Links



------------------------------------

Yahoo! Groups Links



Re: Video triggering of 'scopes in general

 

Denis:

I'll email some photos: the pages are larger than my scanner bed.

John

--- In TekScopes@..., Denis <xyzzx_adv@...> wrote:

John,
?? I would like to request a copy of the 453 Mod 127C Schematic.?? I still have some NTSC gear around and want to see how Tek implemented sync separation, for possible inclusion into a 453A.
??Thanks,
Denis


________________________________


Re: 7854: Excessive stored trace brightness.

 

Yup. Blanking looks good to me...

I'll double check the pot, and make certain its smooth. It looks
like I have some reading to do in the manual....(this is certainly
going to be umm.. interesting)

David

On Mon, Jul 9, 2012 at 11:32 AM, <larrys@...> wrote:
Are the parts that are supposed to be blanked getting blanked?
-ls-

David Holland <david.w.holland@...> wrote:
Hello all,

I'm kind of getting around to poking at my 7854 scope.

The stored trace brightness is excessive, and the adjustment won't go
down any further. I'm reasonably certain the pot is good, since
it'll smoothy vary over what range it does have. (From "too bright",
to "cr*p, I'd better turn that down before the CRT gets burnt").
Normal display, and readout's are of a more reasonable intensity, and
have goodly range. (Readout isn't working correctly, but that's
not the problem I want to concentrate first on.)

Before I crack the covers, any suggestions of which parts of the
service manual would be most relevant to read...

thanks in advance...

David


------------------------------------

Yahoo! Groups Links



------------------------------------

Yahoo! Groups Links



Re: Help with R7903 needed, some success...

keithostertag
 

Thanks. I did pull out the HV probe and measured the HV TP on the z-axis board- -2985V. This is with two plugins loaded. I will try your other suggestions later, gotta run right now for other responsibilities...

But if anyone would like to recommend tracking the crt problem, please do so and I will get back to it later this afternoon or tonight.

Keith Ostertag

--- In TekScopes@..., David <davidwhess@...> wrote:

The regulated voltages at P1896 all look fine. Note that the pin 2
voltage at 129.5 is right where it should be after you adjusted R1293
for what I assume is +55 at TP1625.

The unregulated voltages look fine to me also. The fact that they are
a little high may be do to low plug-in load. The unregulated voltages
are marked with nominal values.

Others here probably have better suggestions for working with the CRT
but I would do a sanity check of TP1635 (next to U1635) which should
be within millivolts of zero if the -3kV supply regulation is working
correctly. Then I would check TP1876 (at the bottom of schematic 11)
which is the output of the z-axis amplifier and marked 7.6 volts.

At this point if everything seems normal, I would break out the high
voltage probe and check the -3kV supply. There is a test point shown
off of pin 2 of the CRT (again at the bottom of schematic 11) but it
is not numbered.

On Mon, 09 Jul 2012 16:06:02 -0000, "keithostertag"
<keitho@...> wrote:

Thanks David. Here are my current readings after adjusting R1293/TP1625 and R1513/-50V:

At P1896:
Pin 2 (+130V): 129.5
Pin 3 (+50V): 50.0
Pin 4 (+15V): 15.07
Pin 5 (+5V): 5.02
Pin 6 (-15V): -14.98
Pin 7 (-50V): -50.0

At P1452 (red):
Pins 1&2 (-17V): -18.78
Pin 3 (+54V): 58
Pins 4&5 (+17V): 18.78
Pin 6 (-54V): -58

+5L at pin3 of P1417 (blue) = 5.06

So I still need to check some caps to resolve those high unregulated lines at P1452.

But, I've got no crt activity at all, and the above are close enough to at least have some crt, right? So I guess I need to start looking around for why I have no crt. No BF, no readout, nada. Just grat illumination.

BTW- my "Control Illumination" button on the front panel won't stay pushed in.

Thanks,
Keith Ostertag


Re: Help with R7903 needed, some success...

 

The regulated voltages at P1896 all look fine. Note that the pin 2
voltage at 129.5 is right where it should be after you adjusted R1293
for what I assume is +55 at TP1625.

The unregulated voltages look fine to me also. The fact that they are
a little high may be do to low plug-in load. The unregulated voltages
are marked with nominal values.

Others here probably have better suggestions for working with the CRT
but I would do a sanity check of TP1635 (next to U1635) which should
be within millivolts of zero if the -3kV supply regulation is working
correctly. Then I would check TP1876 (at the bottom of schematic 11)
which is the output of the z-axis amplifier and marked 7.6 volts.

At this point if everything seems normal, I would break out the high
voltage probe and check the -3kV supply. There is a test point shown
off of pin 2 of the CRT (again at the bottom of schematic 11) but it
is not numbered.

On Mon, 09 Jul 2012 16:06:02 -0000, "keithostertag"
<keitho@...> wrote:

Thanks David. Here are my current readings after adjusting R1293/TP1625 and R1513/-50V:

At P1896:
Pin 2 (+130V): 129.5
Pin 3 (+50V): 50.0
Pin 4 (+15V): 15.07
Pin 5 (+5V): 5.02
Pin 6 (-15V): -14.98
Pin 7 (-50V): -50.0

At P1452 (red):
Pins 1&2 (-17V): -18.78
Pin 3 (+54V): 58
Pins 4&5 (+17V): 18.78
Pin 6 (-54V): -58

+5L at pin3 of P1417 (blue) = 5.06

So I still need to check some caps to resolve those high unregulated lines at P1452.

But, I've got no crt activity at all, and the above are close enough to at least have some crt, right? So I guess I need to start looking around for why I have no crt. No BF, no readout, nada. Just grat illumination.

BTW- my "Control Illumination" button on the front panel won't stay pushed in.

Thanks,
Keith Ostertag


Re: Help with R7903 needed, some success...

keithostertag
 

Thanks David. Here are my current readings after adjusting R1293/TP1625 and R1513/-50V:

At P1896:
Pin 2 (+130V): 129.5
Pin 3 (+50V): 50.0
Pin 4 (+15V): 15.07
Pin 5 (+5V): 5.02
Pin 6 (-15V): -14.98
Pin 7 (-50V): -50.0

At P1452 (red):
Pins 1&2 (-17V): -18.78
Pin 3 (+54V): 58
Pins 4&5 (+17V): 18.78
Pin 6 (-54V): -58

+5L at pin3 of P1417 (blue) = 5.06

So I still need to check some caps to resolve those high unregulated lines at P1452.

But, I've got no crt activity at all, and the above are close enough to at least have some crt, right? So I guess I need to start looking around for why I have no crt. No BF, no readout, nada. Just grat illumination.

BTW- my "Control Illumination" button on the front panel won't stay pushed in.

Thanks,
Keith Ostertag


Re: 7854: Excessive stored trace brightness.

 

Are the parts that are supposed to be blanked getting blanked?
-ls-

David Holland <david.w.holland@...> wrote:

Hello all,

I'm kind of getting around to poking at my 7854 scope.

The stored trace brightness is excessive, and the adjustment won't go
down any further. I'm reasonably certain the pot is good, since
it'll smoothy vary over what range it does have. (From "too bright",
to "cr*p, I'd better turn that down before the CRT gets burnt").
Normal display, and readout's are of a more reasonable intensity, and
have goodly range. (Readout isn't working correctly, but that's
not the problem I want to concentrate first on.)

Before I crack the covers, any suggestions of which parts of the
service manual would be most relevant to read...

thanks in advance...

David


------------------------------------

Yahoo! Groups Links



Re: 7854: Excessive stored trace brightness.

 

I would not count on the potentiometer being good. With age the wiper
in that type tends to become open and intermittent which can be
repaired by cleaning it. I need to clean those potentiometers on my
7854.

I would start by doing the power supply and z-axis calibration
outlined in the service manual.

The readout and storage intensity controls are pretty simple
(schematic 17, 13, 34, to 13) but I wonder if the problem is caused by
the stored intensity intensify function being stuck on. I am not sure
how that is handled.

On Mon, 9 Jul 2012 10:33:40 -0400, David Holland
<david.w.holland@...> wrote:

Hello all,

I'm kind of getting around to poking at my 7854 scope.

The stored trace brightness is excessive, and the adjustment won't go
down any further. I'm reasonably certain the pot is good, since
it'll smoothy vary over what range it does have. (From "too bright",
to "cr*p, I'd better turn that down before the CRT gets burnt").
Normal display, and readout's are of a more reasonable intensity, and
have goodly range. (Readout isn't working correctly, but that's
not the problem I want to concentrate first on.)

Before I crack the covers, any suggestions of which parts of the
service manual would be most relevant to read...

thanks in advance...

David


Re: Help with R7903 needed, some success...

 

On Mon, 09 Jul 2012 13:53:45 -0000, "keithostertag"
<keitho@...> wrote:

The service manual says to adjust TP1625 for +55 volts plus-or-minus 25 volts. What exactly does that mean? I've never seen such a wide range given for a single adjustment. Does it mean that any voltage between 20-80 volts is OK? Is it implying that once you adjust it to 55 volts that you can expect it to vary between 20-80 depending on temperature/load? Or maybe that if all other voltages are correct then you can leave it at any voltage between 20-80? Or?
The voltage at TP1625 is controlled by U1635 to regulate the -3kV
supply and indirectly the CRT acceleration voltage. It will vary with
the total load on the power supply. As long as it is within range,
the -3kV supply will be directly held in regulation. Tektronix was
conservative here since TP1625 can actually swing from about 128 volts
to -48 volts when controlling the -3kV supply.

The -3kV supply and CRT acceleration voltage have to be tightly
controlled because they directly affect the horizontal and vertical
deflection. The 7903/7904 is a little different from my other 7000
series oscilloscopes since these voltages are fixed while in my others
they have a separate adjustment built around U1635.

R1292 controls the output of the 130 volt supply and indirectly all of
the other unregulated outputs including the high voltage supply. It
might be better to adjust it until the 130 volt supply is dead on and
then check TP1625 to see if it is within range. If not, then adjust
R1292 until TP1625 is within range. It might be informative to know
the voltage of the 130 volt supply when TP1625 is adjusted which is
actually mentioned in the manual. It should be 130 volts +/- 5.2
volts.

I'm finding (so far) that adusting it to 55 when cold, it then proceeds to slowly decrease (as low as 30 volts) as it warms up. After warming up for an hour and re-adjusting it to 55 it still fluctuates slightly. Though -50V TP is steady. Since I am (evidently) not finished replacing caps I expect to revisit this adjustment several times...
Since the -50 volt supply is separately regulated, it is not going to
reveal much unless the -54 volt supply is low.


7854: Excessive stored trace brightness.

 

Hello all,

I'm kind of getting around to poking at my 7854 scope.

The stored trace brightness is excessive, and the adjustment won't go
down any further. I'm reasonably certain the pot is good, since
it'll smoothy vary over what range it does have. (From "too bright",
to "cr*p, I'd better turn that down before the CRT gets burnt").
Normal display, and readout's are of a more reasonable intensity, and
have goodly range. (Readout isn't working correctly, but that's
not the problem I want to concentrate first on.)

Before I crack the covers, any suggestions of which parts of the
service manual would be most relevant to read...

thanks in advance...

David


Re: 545B Repair

 

¿ªÔÆÌåÓý

On TV sets, this is a sign of a bad CRT.

On 2012-07-08 2:12 PM, yurighisio wrote:

Hello,

I have repaired the low voltage power supply, the problem was in a 6080 tube V737A. I have replaced this tube and voltage are ok. For future repair and check purpose, have anyone the current for each voltage ?
I have performed calibration, it seem to work well.
A strange problem appears on crt when the trace intensity increase, past the control middle point, trace thickens and glow down.
I have checked -1700V and +8300V they don't change with knob moving, control grid voltage (pin 3) change accordingly with intensity control from -1900 to -1700.

Juri



Re: Help with R7903 needed, some success...

keithostertag
 

Still working on this R7903 repair, still at this point concentrated on the PS. The PS is no longer in tic mode- yea!

The service manual says to adjust TP1625 for +55 volts plus-or-minus 25 volts. What exactly does that mean? I've never seen such a wide range given for a single adjustment. Does it mean that any voltage between 20-80 volts is OK? Is it implying that once you adjust it to 55 volts that you can expect it to vary between 20-80 depending on temperature/load? Or maybe that if all other voltages are correct then you can leave it at any voltage between 20-80? Or?

I'm finding (so far) that adusting it to 55 when cold, it then proceeds to slowly decrease (as low as 30 volts) as it warms up. After warming up for an hour and re-adjusting it to 55 it still fluctuates slightly. Though -50V TP is steady. Since I am (evidently) not finished replacing caps I expect to revisit this adjustment several times...

BTW- thank you David (once again) for your explanation last post. I am still replacing caps! and still no crt (yet).

Except for last week during which I was waiting for parts I have been working on this project several hours a day 5/days/wk for about 3 months! Thanks to everyone for their help!

Keith Ostertag


Re: Video triggering of 'scopes in general

Tim Phillips
 

¿ªÔÆÌåÓý

from Tim P (UK)
Interesting ! I have an'add-on' Sync Seperator box (015-0016-00) which
outputs selected line or field sync to the delayed sweep trigger.
I had an old analog camera with the lens covered, just to play with it.
?
There was also the Type 'V' plug-in for the 500 series that never got to market.
Tim
?


Re: Just got a tek 555 for free, and in need of some advice!

 

PL259 or UHF.? Seen a lot in HAM radio. ? Here is an adapter to BNC: http://www.amazon.com/BNC-Female-PL259-Male-Adapter/dp/B000VGUCTM


--- On Mon, 7/9/12, dutchbosstard <childwhereugo@...> wrote:

From: dutchbosstard
Subject: [TekScopes] Re: Just got a tek 555 for free, and in need of some advice!
To: TekScopes@...
Date: Monday, July 9, 2012, 3:44 AM

?



I do not know the name of the style of screw-in connector, not a BNC. they look like these:




can someone give me the name for those things? i guess i should be happy that the 2 probes that came with the scope have the same strange looking connector. the timebases have BNC connectors.



Re: Bench test the 7904 power supply.

 

Hi Joe,

No worries, I think the following message from G?ran shall get you started: See???

Cheers,

Magnus

P.S. This might be an old hat to you, but I learned that the 7904 and the 7904A-Version use a different power supply design.
See http://w140.com/tekwiki/wiki/7904? This provides a service note on the power supply and the 7904 Service Manual, with schematics.
References: Section 4-8 and following in the 7904 Service Manual.



TDS 540 scope needs help

 

Hi all:

i would be very very appreciated if you can tell me how to manage

these 2 ics.(acq board u501&u502,sn:tek 234 0718 40)

i got no 250Mhz signs from these 2 ics.

they are dead?

i approved the pic:



regards


Re: Just got a tek 555 for free, and in need of some advice!

 

ill look into 3-in-one oil, i think ive seen some before around the machine shop at my university.

I see that you painted the carts of your 555 scopes? this might come off as another dumb question, but are the two provisions for plug-ins on the 555 cart just there as a holding rack for 2 extra plug-ins, or is it actually possible to somehow connect those plugins on the cart to the actual scope?

I appreciate your offer, and as it turns out, i might actually need a part. the day i turned it on, at my friends house, I dropped one of the white little things that the outside panel screws get screwed into, i just tried to find it in the parts list, but I dont know what their exact name is.

-Robby

--- In TekScopes@..., Keith Knox <hptekman@...> wrote:

Hi Robby,
??
At the present time I am working on 4 Tek 555 scopes. I have 3 of them working, 555 #4 is not working yet but will be soon.
??
As for the fans I use a drop of 3-in-ONE oil and find that it works very well. Fans must spin freely or they will not run or they will run slowly.
??
Test points for LV power supplys are on the bottom of power supply unit.
??
Because the Tek 555 is a dual beam scope, there are two -1350 HV adjustments. One for each cathode.?? The HV Anode lead that plugs into the CRT will provide about +8650 volts.
This totals out to 10KV accelerating voltage.
??
When you get your Tek 555 working,?? remember that the upper beam will only write to the upper 4 cm of the crt and the lower beam will only write to the lower 4 cm of the crt.
??
I have some spare parts for 555 scopes.?? Let me know if there is anything you need.
??
Have attached??photos of my three working 555 scopes.
??
Keith


--- On Fri, 7/6/12, dutchbosstard <childwhereugo@...> wrote:


From: dutchbosstard <childwhereugo@...>
Subject: [TekScopes] Re: Just got a tek 555 for free, and in need of some advice!
To: TekScopes@...
Date: Friday, July 6, 2012, 10:01 PM



??



awesome. so when i get it back to my house ill carefully hit it with 40 psi. im guessing that shouldnt damage any tubes as long as im careful. My air compressor is sanborn reciprocating compressor - ive never noticed any oil in the lines, only a small bit of moisture from when i drain the tank all the way. it also has a air filter on the input.

so i guess ill just utilize a dry cloth until i get the scope back to my house.

a few questions about the test points - when the time comes that i power up the scope, where are the test points for the low voltages from the power supply? are they in the PSU box? or are they somewhere inside the scope? also, when i opened up the case yesterday i found the -1350V test point for the CRT, but I was under the impression that the total accelerating voltage for the crt is 10kV. where is that test point located, and should i worry about it?

also what fuses should i see for the main fuse, and the filament fuse in the PSU box?

thanks,

Robby

--- In TekScopes@..., tubesnthings@ wrote:


Valid concerns! Set air pressure no higher than necessary, say below 40
psi.
Use caution and common sense - 40 year old dust doesn't belong in your
lungs - do it outside and wear a mask.



In a message dated 7/6/2012 1:43:15 P.M. Pacific Daylight Time,
childwhereugo@ writes:

since i have an air compressor at home, should i try blowing out all the
dust? Ive heard that you can do this to electronic equipment, but i am kinda
terrified to do this since i dont want to have anything become
unconnected, or cause anything to short out, or worse yet - damage a tube.

are these valid concerns, or am i overacting? my air compressor is usually
set to 100 PSI.


Re: Just got a tek 555 for free, and in need of some advice!

 

first I would like to say thank you for the very clear advice.

I have been slowly going through the manual since the day i got the scope, but i definitely need to read it thoroughly instead of just kind of skimming it. Yes, i did read the part in the manual about running it with all plug-ins installed. If anyone would like to know, I believe my scope may be a 1967 model, since there is a little information diagram on the bottom of the power supply that has written that it was inspected in '67. It also has the newer 21A and 22A time bases, and the vertical plugins that came with it are the type 53/54K and the type M 4 trace plugin. I do not know the name of the style of screw-in connector, not a BNC. they look like these:




can someone give me the name for those things? i guess i should be happy that the 2 probes that came with the scope have the same strange looking connector. the timebases have BNC connectors.

as far as the fans go, would it really be necessary to soak them in diesel fuel, if i happen to get em turning after disassembly? I know that electric motors and generators often use oil-lite bearings, i've never heard of the bearings losing their oil, but then again I dont know much about oil-lite other than it feels strange, looks strange, and has an interesting smell to it. However, if i cannot get the 555 fan to get turn nicely after taking it apart, i will surely follow your advice word for word.

as far as solder goes, I did indeed read where it says in the power supply to use the silver bearing solder. what I am confused on is: is this solder flux cored, or do i need to solder it with external flux (ive only ever soldered with 60/40 flux cored solder) or does this specific solder need flux? my last solder question is, should I buy this tektronix solder off of ebay?



as far as V799, the thermonic diode, I just did a little reading about it online, could i look up the datasheet to find the pins, remove it and measure the resistance of the filament to make sure its not blown that way?

thanks,

Robby

--- In TekScopes@..., "Morris" <vilgotch@...> wrote:

Hi,

Let me add some comments from an experienced Tek scope (including a 555 scoposaurus) restorer to the excellent advice you have had so far:

I gather you are not experienced with vacuum tube equipment although you have worked with high voltage before. I assume you know how to be careful around gutsy power supplies, there's more than enough juice in the 555 to kill you.

At this stage you have had the scope running after a fashion so you can probably assume that the CRT is OK, you know the time delay relay works, no electrolytics have blown up, no smoke got released and the power supplies are at least producing enough to a give you a green flash. That's all very encouraging.

What I would do now is to stop and read the manual thoroughly so you become familiar with what's there, where the test points are and how the thing works. Then I would give it as thorough a clean as possible as has been discussed. The next thing to do before anything else is to get the fan working. The best way to do that is to remove it altogether which involves unsoldering the leads from the tagstrip, taking careful note of where to reconnect them. If there is no little roll of silver solder with the scope you need to go and get some - the commercially available silver bearing solder stuff is not exactly the same as the Tek solder but is preferable to ordinary 60/40 which should not be used.

Taking the fan motor apart is pretty straightforward, just make sure you record where everything goes so you can reassemble it. The bearings can be soaked in thin oil (such as 3-in-1), diesel fuel or kerosene to clean all the gunk off and then lubed with something thicker. Car engine oil is probably OK. Once it spins freely, you can put it back. Eventually you might do the same for the fan in the power supply but for now just put some oil into the oil holes on the bearings.

Once both fans are running you can run the scope again and check the LV power supply voltages at the test points. Start with the -150 which is the master reference for everything else. You don't need to measure the HV just yet but make sure the heater voltages are right. The heater supply is regulated using V799 in the power supply, a special thermal sensing diode type 2AS15A. If it's failed with an open filament then the heater voltage will be low to pretty well all the tubes in the scope. If V799's not lighting up you will need to suspend all operations until you can find a replacement (plenty available on eBay).

Incidentally I missed whether you mentioned which timebase & vertical plugins are installed. Please let us know. Don't try to run the scope without plugins.

Once you've done all that, get back to the brains trust here and I'm sure you will get excellent advice on how to proceed further.

Good Luck!!

Morris

--- In TekScopes@..., "dutchbosstard" <childwhereugo@> wrote:

so yesterday i got real brave and decided to power the scope up. the power supply had the proper 5 amp fuse, but the heater supply had an 8 amp fuse instead of the 7 amp one. i checked the fan in the power supply and i tried turning it by hand and it seemed to spin pretty well, but the larger fan in the 555 was harder to turn, and i also had to remove its dust screen since the dust screen was pressed in, preventing the fan's movement. i powered it on with all the covers off and the PSU fan came on right away, but i noticed that the 555's fan wasnt powering on. all the tubes started to glow and i heard a relay click and then i started fiddling trying to find a beam. i rotated both of the horizontal display knobs and i couldnt get a trace, but every time i turned the know, a trace would appear for a slit second. both beams did this. after the scope had teen turned on for a whole minute, i decided to shut the thing off since the fan in the 555 didnt come off. the fan in the PSU box also had originally started rotating rather slow, but kept gaining speed until after about 30 seconds it started rotating extremely fast.

now this might be a dumb question, but could I have damaged anything by powering it on with a 8amp heater supply fuse instead of the listed 7amp? and also, is there any difference between a "slo-blow" fuse and a "time-delay" fuse?

also, what should i do about the fan in the 555? should i disconnect it and try to power it up with 115v on its own?


Re: Video triggering of 'scopes in general

 

Look at the LM1881 video sync IC for more ideas on good ole NTSC stuff. This is one of the most elaborate, yet simple to use ICs for decoding and triggering of video signals, and National's datasheet gives a very nice technical explanation of the whole thing.

A while back, I had planned to modify a 7K timebase plug-in with one of these for high performance video triggering, but NTSC went obsolete before I got around to it. This also was right after I finally acquired a really nice genlock box and color bar/pattern generator.

My advice on this is to not get too carried away with NTSC, other than for fun as a learning experience or work of art.

Ed

--- In TekScopes@..., Denis <xyzzx_adv@...> wrote:

John,
?? I would like to request a copy of the 453 Mod 127C Schematic.?? I still have some NTSC gear around and want to see how Tek implemented sync separation, for possible inclusion into a 453A.
??Thanks,
Denis


________________________________
From: John S <John@...>
To: TekScopes@...
Sent: Friday, March 4, 2011 9:21 AM
Subject: [TekScopes] Video triggering of 'scopes in general


??
- All the old TV standards adopted around the world have an odd numbers of lines in a "complete" picture, which is defined as a TV Frame. (25Hz Europe, c.30Hz USA)
- To conserve RF bandwidth, the signal is transmitted in sequential Fields, each of which contains half the lines (odd lines Field 1, even lines Filed 2 etc), as required to assemble the full TV Frame.
- This is known as interlace scanning, and gives rise to display artefacts such as line-twitter on strong horizontal black-white transitions.
- As the odd number of lines is clearly not divisible by 2, odd fields terminate (theoretically) half way through a line, and even fields commence similarly.
- The problem confronted by TV engineers is to get the 'scope to trigger stably (ideally) at the FRAME rate, which is half the field rate. This allows a 'scope with a good dual-time-base delayed sweep to select a particular line. (The line can be a bit dim, as the 'scope is only writing for c. 60uS in every c. 40mS).
- I have a 453 with TV triggering ( MOD127C ): Tek used to supply special graticules (which I also have) based around standardised TV test & measurement waveforms.
- broadcasters use a few of the "blank" lines in the field blanking area* to transmit the test waveforms so an entire transmission chain can be quality monitored

I can supply a copy of the MOD127C 453 handbook description to anyone who's interested.

John

* the "field blanking" lines contain synchronising pulses which are "broader" than on normal video lines, and the shift in DC level is used by the sync separators in the TV to lock the frame oscillator to field rate. They were kept free of video in order to allow the CRT scan time to return to the top of the screen without losing active picture content.

--- In mailto:TekScopes%40yahoogroups.com, Gala Dragos <gala_dragos@> wrote:

Hi!

I've worked with a TD2002B recently. It had a trigger mode named "Video". It was used to display PAL/SECAM/NTSC video signals and lines. How do I do that with Tektronix 475 or 465 ?

Thanks.


Bench test the 7904 power supply.

 

I know this has been asked before.. and the answer is always something like "We cover this a number of times" But frankly I can't find any reference to it. What I am trying to do is test a couple of 7904 power supply's that are not connect to the mainframe. Normally I would be able to just fire up the power supply and measure the voltages and go on with life. But these switchers must need some sort of load or special cable or?? What does one have to do to be able to bring up a 7904 power supply if it's completely remote from the frame?

Thank you for your responses and I apologize, but please point me to the right post of file if it's been covered before.

Joe AG4QC