Trivia- can you identify these probes/probe accessories circa 1960 (or before)?
keithostertag
This past weekend I bought a 310A in a leather carrycase with several probes and probe accessories. Nice! The 310A will be a future project- haven't even tried to power it up yet. Included several types of probe tips I may never find a use for.
In the case were a few things I can't identify. If you can, please also tell if there is a proper way (or improper way) to use them, and maybe why they aren't (evidently) used/manufactured anymore. Photo A: some type of scope input connector (not sure if you'd call it a probe): Photo B: some kind of probe accessory/tip (the black screw-on connectors were packaged for/by IBM if that's any clue): Photo C: evidently a Tek probe, but no markings other than "600V Max, 10 Meg, 14uuF, Atten 10X") around the base end. Cable is a Amphenol No. 621-685: What's the teflon screw intended for? The access for compensation is on the other side. Cheers, Keith Ostertag |
Re: 2205 channel 1 problem
Here's what I got for voltages. Values in brackets are what is printed on the schematic.
Q102e: +4.84 [-4.7] Q102b: +4.12 [not printed, but assumed about +4.3] Q102c: -3.45 [-4.2] Q103e: +4.53 [-4.7] Q103b: common to Q102b +4.12 Q103c: -4.44 [-4.2] Q104e: -3.85 [-4.9] Q105e: -5.26 [-4.9] U30pin2: +0.58 [0v] Clearly something is messed up here. |
Re: Tek 602 Display Unit
Dave Brown
Thanks Curly- doubt I'll ever need to put one in there but interesting anyway!
toggle quoted message
Show quoted text
Cheers DaveB, NZ ----- Original Message -----
From: "curlytronics" <curlytronics@...> To: <TekScopes@...> Sent: Monday, July 02, 2012 12:34 PM Subject: [TekScopes] Re: Tek 602 Display Unit My 602 fan is a square boxer, roughly 3.5in x 3.5in x 1.5in. Curly --- In TekScopes@..., "Dave Brown" <tractorb@...> wrote:
------------------------------------ Yahoo! Groups Links |
Re: 7B92A with tunnel diodes vs. version without
On Sun, 1 Jul 2012 20:59:15 -0600, David DiGiacomo
<daviddigiacomo@...> wrote: On Sun, Jul 1, 2012 at 5:29 PM, raymonddompfrank <r.domp.frank@...> wrote:I only have the later 7B92 plug-ins or I would do some tests. BothRecently, I acquired a 7B92A that in contrast with the unit that I had, turns out to have no (discrete?) tunnel diodes in the trigger circuits but instead, uses hybrids packaged like the ones in e.g. a 7904A's vertical amp. I have seen references to this newer version of the 7B92A but haven't been able to find any description of the reasons why and any possible consequences. I would guess (and could appreciate) lower cost, higher reliability?No, there are no tunnel diodes in the late version. I thought the versions include the HF Sync mode. I like them because they trigger more cleanly than the 7B53A which is to be expected. I thought they used monolithic ICs instead of hybrids also but that could still be the case. Tektronix may have used that package for an IC simply because it was the highest frequency IC package available to them. It is too bad there are no details on it. |
Re: Tek 547
Hi all,
toggle quoted message
Show quoted text
I agree, the 547 is an absolute classic and I still use one of mine as the preferred bench scope even though I have plenty of higher performance scopes if I need them. I hope someone local agrees to take yours and use it as it was meant to be used. They will outlast me and I only hope I can find someone to cherish my two 547s one day. Morris --- In TekScopes@..., "fred" <fredschneider@...> wrote:
|
Re: 7B92A with tunnel diodes vs. version without
On Sun, Jul 1, 2012 at 5:29 PM, raymonddompfrank <r.domp.frank@...> wrote:
Recently, I acquired a 7B92A that in contrast with the unit that I had, turns out to have no (discrete?) tunnel diodes in the trigger circuits but instead, uses hybrids packaged like the ones in e.g. a 7904A's vertical amp. I have seen references to this newer version of the 7B92A but haven't been able to find any description of the reasons why and any possible consequences. I would guess (and could appreciate) lower cost, higher reliability?No, there are no tunnel diodes in the late version. I thought the trigger circuits were monolithic ICs, but maybe not. I have found the early version to be faster, but others disagree. |
Re: Help with R7903 needed, some success...
On Mon, 02 Jul 2012 01:09:13 -0000, "keithostertag"
<keitho@...> wrote: More photos of traces, if you could look over and help explain I would appreciate it.What is the voltage at the anode of VR1297 and what is the -17 volt supply feeding it? If that reference is off then the +130 volt supply will be off and TP1625 will be off indicating that the -3KVolt supply may be out of regulation. If the +130 volt supply is in regulation, then I would start looking for a problem on the high voltage side or more bad electrolytics. After looking around for quite sometime I found that I am only getting a little more than 10VAC p-p off one side of T1235. Shouldn't I be getting more than twice that? Or am I misinterpreting that?10VACpp seem about right. The 20 volt zener diode VR1279 is just there to protect the IC. The output of T1235 is a current so the peak to peak voltage is made up of the 7.5 volt shunt regulator at pin 6 of U1275, 1.8 volts from CR1288 to CR1290, the low voltage drop across current sense resistor R1287, and 1.2 volts from bridge rectifier CR1280 to CR1283. That all adds up to about 10.5 volts peak to peak. I am getting about 300V p-p at TP1238, TP1231 and TP1234. is that too low?T1235 is used as a current transformer. The circuit making up the shunt regulator on the 55 turn output looks like a short to the 1 turn primary. The secondary is limited to about 10 Vpp so the primary will be 10/55 or about 180 mVpp. BTW- using my DMM, U1275 pin 6 is about 7.06V (should be 7.5) and pin 7 is -1.92 (should be -2). Voltage across the 20V zener VR1279 is only about 9.75V.The zener is just there to protect the IC. |
Re: 7B92A with tunnel diodes vs. version without
sipespresso
The early 7B92A had amplifiers in the signal path before the tunnel diodes. Amplifiers help keep the tunnel diodes safe, but they tend to reduce trigger bandwidth. Contrast this with the 5T1A of a decade earlier, where the trigger signal is passively coupled to the "trigger recognition" tunnel diodes. The 5T1A is electrically fragile because of this design.
toggle quoted message
Show quoted text
Also, the fastest triggering is really "HF sync" mode, not true bistable-multivibrator mode. HF sync mode exists in several Tek products using passive coupling of the trigger input to the tunnel diode (e.g., 7T11, 5T3, 519). -Kurt --- In TekScopes@..., "raymonddompfrank" <r.domp.frank@...> wrote:
|
Re: Help with R7903 needed, some success...
keithostertag
More photos of traces, if you could look over and help explain I would appreciate it.
I finally got the PS out of tic mode, though I have to "warm it up" first by powering it on with a heavy load on +5L for a few minutes, turn it off, then reconnect P1417 immediately and power back up. So now I can test the PS in somewhat normal mode (no CRT yet.) As recommended, I began to look at the power supply calibration procedure. But I got stopped at the first step- TP1625 to ground measures less than 11 volts. According to the SM it should be 55V plus-or-minus 25V (should that be .25V?). Adjusting R1231 all the way CW only gives about 11.3V. After looking around for quite sometime I found that I am only getting a little more than 10VAC p-p off one side of T1235. Shouldn't I be getting more than twice that? Or am I misinterpreting that? I am getting about 300V p-p at TP1238, TP1231 and TP1234. is that too low? We already discussed the two large power supply caps C1216 and C1217- they have checked out OK AFAIK. I won't be using my isolation transformer again until I can replace it with a better one, so I may be limited in what I can safely do. Here are some photos: One side of T1235 (note timebase): TP1238 (bottom of C1229, one side of VR1239): TP1231 (same as TP1234): So the main question is to find out why TP1625 is so low, and this is my attempt to trace the problem back to its origin. I'm confused about how to calculate the expected voltages- if TP1238 is 300VAC p-p, and T1235 turns ratio is 1:55, should the output of T1235 be 300/55 = ~ 5.5 VAC p-p? But then how could we get near 55V at TP1625? BTW- using my DMM, U1275 pin 6 is about 7.06V (should be 7.5) and pin 7 is -1.92 (should be -2). Voltage across the 20V zener VR1279 is only about 9.75V. Thanks for any insights you can give, Keith Otstertag |
Re: Tek 602 Display Unit
My 602 fan is a square boxer, roughly 3.5in x 3.5in x 1.5in.
toggle quoted message
Show quoted text
Curly --- In TekScopes@..., "Dave Brown" <tractorb@...> wrote:
|
Re: 7B92A with tunnel diodes vs. version without
raymonddompfrank
Of course, where I wrote "...the maximum signal that...", I meant to write "...the maximum *frequency* that..."
toggle quoted message
Show quoted text
Sorry for the inaccuracy. Raymond --- In TekScopes@..., "raymonddompfrank" <r.domp.frank@...> wrote:
|
Re: 7704A Readout Capability
Hi,
toggle quoted message
Show quoted text
Try running the readout intensity control throughout its range to make sure it is not sitting on a bad spot. Press your beam finder and see it dots appear above and below the reduced trace. Both plugins support readout. If you have not done so, get a manual and get familiar with your fine machine. Jerry Massengale
-----Original Message----- From: dabyram To: TekScopes Sent: Sun, Jul 1, 2012 5:00 pm Subject: [TekScopes] 7704A Readout Capability
?
I recently acquired a Tek 7704A with 7A26 and 7B53A plugins. The 'scope seems to operate fine except I have no readout functions on the CRT. I am not sure if these plugins have the capability or if there is a problem with the readout circuitry in the mainframe.
Any thoughts would be appreciated. Tom |
7B92A with tunnel diodes vs. version without
raymonddompfrank
Recently, I acquired a 7B92A that in contrast with the unit that I had, turns out to have no (discrete?) tunnel diodes in the trigger circuits but instead, uses hybrids packaged like the ones in e.g. a 7904A's vertical amp. I have seen references to this newer version of the 7B92A but haven't been able to find any description of the reasons why and any possible consequences. I would guess (and could appreciate) lower cost, higher reliability?
I wonder whether the circuits in these hybrids still contain circuits with tunnel diodes or do they contain more traditional circuits and if so, are there any functional differences? i could imagine that the maximum signal that the unit will trigger on would be lower than the tunnel diode-equipped version. I would much appreciate if any of you experts could shed some light on this. Raymond |
Re: 7704A Readout Capability
7000 series Plug-ins that lack readout capability have an N suffix.
You could check if the 7704A has option 1 which removes the readout, look at the third of four potentiometers from the top which should be the readout intensity control, or look to see if the readout PC board is even inside. On Sun, 01 Jul 2012 22:00:24 -0000, "dabyram" <tom.byram@...> wrote: I recently acquired a Tek 7704A with 7A26 and 7B53A plugins. The 'scope seems to operate fine except I have no readout functions on the CRT. I am not sure if these plugins have the capability or if there is a problem with the readout circuitry in the mainframe. |
7704A Readout Capability
I recently acquired a Tek 7704A with 7A26 and 7B53A plugins. The 'scope seems to operate fine except I have no readout functions on the CRT. I am not sure if these plugins have the capability or if there is a problem with the readout circuitry in the mainframe.
Any thoughts would be appreciated. Tom |
Error code list for Tektronics TDS scopes
John C
I was wondering if anyone knew where a list of error codes was for the TDS6xxx series oscilloscopes.
Specifically, I have a TDS6604 with an error code of 531 on the calibration test, which appears to be a memory storage error from the materials I could fine. Thanks all! :D |
File - Posting Rules
Feel free to check out a site I run:
It focuses more on modern scopes of all brands, as well as electronics and test in general. I hope you'll join, and if you're interested in writing for the site - either blog posts or regular commentage - please get in touch. Please edit any posts/replies to the list to minimize quoted material to that required for continuity. Please do not send personal replies to the list. Replies currently go to the list by default. Those congesting the list with with either of the above may have posting rights revoked. Only by following these simple rules will we have a "clean" archive. They also reduce inbox clutter, make digests much nicer to read, improve searches, and reduce the chance of having old messages deleted. To post to the group, mail to: TekScopes@... It is now possible to send attachments with postings. In the message, Yahoo replaces the file with a link, and they store the file on-site for a limited time. I occasionally get emails meant for the group mistakenly sent to TekScopes-owner@...! I just delete these, so please watch where you're sending... Please use the archive groups listed on the home page for file uploads, as our own area is pretty much full! If you really want to put a file in our directory, contact me directly at TekScopes-owner@.... If you do not wish to belong to TekScopes, you may unsubscribe by sending an email to TekScopes-unsubscribe@... You may also visit the Yahoo web site to modify your subscription: Check out the files, links, and photos sections, and remember, you can search the group's archives from its home page, or when reading messages. The group's home is at: Regards, Michael Dunn Listowner, TekScopes |