Date

Re: 2465B Display issues..................

David Garrido
 

Is there any reason that sticks out as to why my 1.36 ref would vary?

David


Re: 2465B Display issues..................

David Garrido
 

Tom,

I did go through the R2010 procedure. But now that I think about it, isn't there something that dictates a follow thru on something (?) to make sure the procedure is written to RAM?

David


Re: 2465B Display issues..................

Tom Miller
 

¿ªÔÆÌåÓý

Hi David,
?
Your measurements look ok to me. Did you go through the DAC adjustment procedure?
?
See attached. You want the voltage difference between the +1.25 and - 1.25 to be exactly 2.500 volts.
?
So, at first, set the -1.36 to -1.2504 then check them both. Allow a 20 minute warm-up with a fan blowing
on the main board hybrids.
?
HTH,
Tom
?
?

----- Original Message -----
Sent: Friday, December 10, 2010 10:28 PM
Subject: [TekScopes] Re: 2465B Display issues..................

?



Hello All,

I am hoping someone with either experience on this A5 Board problem can help or someone with a 2465B can give me actual readings in circuit that might be able to shed some light on the topic.

My 2465B Ser No. B053XXX with an SMD A5 Board info is as follows with all info pulled in circuit:

-1.36 varies between 1.0431 and 1.34 ish
+1.25 is steady at 1.2496

Top of R2016(1.0k) is 9.999V and the bottom is 8.981V. Resistance is 0.977k
Top of R2013(10.0k) is 8.981V and the bottom is 156.95mV. Resistance is 7.12k
Top of R2012(10.0k) is 154.41mV and the bottom is 0.0V. Resistance is 7.76k
Top of R2011(4.75k) is 8.981V and the bottom is 8.336V. Resistance is 4.438k
Top of R2014(221k) is 8.336V and the bottom is 0.3555V. Resistance is 14.11k
Left side of R2015(82.5) is 8.336V and the right side is 156.94mV. Resistance is 10.6k

I am clearly not an engineer and could use any help offered.

Cheers,

David


No virus found in this message.
Checked by AVG -
Version: 10.0.1170 / Virus Database: 426/3307 - Release Date: 12/10/10


Re: 2465B Display issues..................

David Garrido
 

Hello All,

I am hoping someone with either experience on this A5 Board problem can help or someone with a 2465B can give me actual readings in circuit that might be able to shed some light on the topic.

My 2465B Ser No. B053XXX with an SMD A5 Board info is as follows with all info pulled in circuit:

-1.36 varies between 1.0431 and 1.34 ish
+1.25 is steady at 1.2496

Top of R2016(1.0k) is 9.999V and the bottom is 8.981V. Resistance is 0.977k
Top of R2013(10.0k) is 8.981V and the bottom is 156.95mV. Resistance is 7.12k
Top of R2012(10.0k) is 154.41mV and the bottom is 0.0V. Resistance is 7.76k
Top of R2011(4.75k) is 8.981V and the bottom is 8.336V. Resistance is 4.438k
Top of R2014(221k) is 8.336V and the bottom is 0.3555V. Resistance is 14.11k
Left side of R2015(82.5) is 8.336V and the right side is 156.94mV. Resistance is 10.6k

I am clearly not an engineer and could use any help offered.

Cheers,

David


New 7xxx readout board

EB4APL
 

I think that somebody recently mentioned a "new type" readout board, microprocessor based, even I think he attached a picture or a link to it. I think that he mentioned that his manual has not the info about his board.
I tried searching the archives but I was unable to find anything. Was I dreaming or was it real?

Ignacio


Re: Taps and setscrews

 

Whatever you get, make sure to get plug or bottoming taps too, not just starter taps as you may see in some sets. For knobs and blind holes you won't have much clearance behind the threads.

Way OT - I vowed never to buy another Craftsman power tool. Thirty years ago I bought a cheap Black & Decker circular saw at Kmart for $30 on sale, used it successfully on lots of lumber projects, and then put an abrasive blade on it to cut steel stock and car body panels, etc., but the plastic blade guard started to melt through. I lined it with copper sheeting and continued to use it that way for many years. A few years later I treated myself to a nice new $100 Craftsman saw, but didn't get around to using it seriously until a year later, at which time it blew its bearings and seized up after only a few lumber cuts - of course just past warranty. Sears would not exchange it, and I was out of luck. I continued to use the old B&D for many years, building several decks, various remodeling projects, and cutting countless pieces of steel. About ten years ago I got a new Makita, and retired the B&D to to its final duty as a concrete saw - sans blade guard, which finally fell apart. It has since then sat outside many months in the rain, and cut through about a hundred linear feet of rock and concrete. It just keeps going and going, even though its bearings have long ago disintegrated internally, it's covered with rust, and full of dirt and grit.

I have had no problems with Craftsman hand tools, but I haven't bought any for quite a few years, so don't know what they're like now. I still exchange worn out screwdrivers there - at least they honor that. I figure they still owe me a lot of screwdrivers to compensate for that saw.

Ed

--- In TekScopes@..., Artekmedia <manuals@...> wrote:

With All due respect ( and USED TO B E A HUGE craftsman fan myslef)
Craftsman 20 years ago was good quality stuff ...Craftsman today is
made from recycled automobiles and I'll leave it up to the reader to
figure out where most of their stuff is manufactured ...not the same
......
Hi,
For knob repair I sometimes need to retap the threads on the knobs.
The last tap and die set I bought was a cheap disaster. Any
suggestions for buying good taps? setscrews?
jerry


Re: RPR Conversion from Microfiche to OCR'ed PDF...............

David Garrido
 

Hakan,

The way things are going we will need to know if you will contribute your fiche by morning in the US. ?But seriously, this group of people NEVER ceases to amaze me. ?Truly astounding response folks. ?We are half way there and just need to double the contributions that have already come in. ?Lets try and get this done sooner than later. ?This is going to be the most useful conversion yet.

Thank you for all your support. ?As soon as we get all the contribution commitments I will get this off to the conversion folks.

Cheers,

David






?
?
???

Hello Everyone,

I have taken the liberty to rename the "CONVERTING FICHE TO PDF" thread.

We are about 20% to the finish line and I will continue to update the topic about once a week to let everyone know about the progress of the project.

Again, so far I have received commitments of about $200.00 USD. ?We are well on our way to getting this done folks. ?Lets keep it going. ?I will not ask for the deposit into my account until we have commitments enough to cover the project cost.......about $1000.00 USD if we are able to use Hakan's latest copy of December 01, 1997.

Please email me offline and I will put you on the list for your commitment and keep you updated privately as well as the list here.

Cheers,

David

?
???



--
David Garrido



Home of the "Classic" TEK Project and dedicated to the "not for profit" preservation, restoration, maintenance, repair, and archival of all "Classic" Tektronix Test and Measurement Equipment.


Re: Taps and setscrews

 

jmassen418a@... wrote:


Hi,

For knob repair I sometimes need to retap the threads on the knobs.
The last tap and die set I bought was a cheap disaster. Any
suggestions for buying good taps? setscrews?

A 6-32 & a 4-40 tap should cover most of your needs. If you already
bought a set you should have a T-handle wrench. Plug & bottoming
style taps work where you don't have much length beyond the end
of the threads. Taper style will need more length than you can get
in a knob. McMaster, as mentioned before, is a good source. So is
Enco if you stick with US-made tooling. I've bought lots of stuff
from Enco. Hardened cup-point setscrews are pretty cheap by the box,
same sources.

HTH,
-ls-


Re: 485-Intensity control

 

¿ªÔÆÌåÓý

Goran-
?
Again thanks for the nice response. One of the great challenges one has with this equipment is sourcing replacement parts, and in the case of high voltage components it's a big challenge. This morning I was searching the internet and came across otelectronics.com. They have both diodes and high voltage capacitors at very attractive prices and the shipping is cheap. If you're interested I'll let you know if they perform as advertised; if so they're one that other members of the forum may wish to note. The bottom line is I'll change all the voltage doubling components and see if that solves it.
?
Thanks again,
Phil
kd6rm

----- Original Message -----
Sent: Friday, December 10, 2010 4:05 AM
Subject: Re: [TekScopes] Re: 485-Intensity control

?

Phil,
?
This is a difficult part. I had a problem with U1624 (741) in my 7904 scope, check that both pins 2 and 3 are at ground level, they should be. I am not sure if you can say at this point that the regulation circuitry is good just because you can adjust TP1614 to +95V while having the -3kV jumping up and down. The regulator circuitry can not operate stand alone, there is a tremendous gain here, U1624 runs open loop.
?
You probably have a problem with the doubling circuitry. I am not familiar with testing high voltage caps. You could have a problem with arcing and/or discharging here, one way to find out is of course to replace the caps all together.
?
G?ran
?


Re: Taps and setscrews

 

WOW!

I should have been able to guess that suits inside Sears would corrupt the Craftsman brand and quality. ?I am sorry to hear that.

Here in Colorado Springs, CO we have an industrial supply company - Shop Tools - which offers quality tooling to machinists. ?Additionally, like I previously indicated McMaster Carr provides good quality items that are reasonably priced and delivered swiftly.

Not to start an off topic rant - but I sure would like to see American ingenuity and quality return on a grand scale to the manufacturing realm. ?Right now we're driving down the wrong road into oblivion.

Try McMaster Carr - their web site is very good.

Ken

On 10Dec, 2010, at 4:47 PM, Artekmedia wrote:

?

With All due respect ( and? USED TO B E A HUGE craftsman fan myslef)? Craftsman? 20 years ago was good quality stuff ...Craftsman today is made from recycled automobiles and I'll leave it up to the reader to figure out where most of their stuff is manufactured ...not the same stuff anymore

Go visit your local tool and die shop and ask them where they by quality taps these days ...and for that matter what to avoid

Dave

On 12/10/2010 5:37 PM, Ken Chalfant wrote:

?

Hi,


Craftsman tools are reasonable. ?I've had a couple sets for going on twenty years and they serve me well.

For a wide variety of individual taps and setscrews check out . ?That's the website for McMaster Carr, a very well established industrial supply company in Chicago. ?They have been around since the dawn of time and are a fantastic resource.

Ken

On 10Dec, 2010, at 3:59 PM, jmassen418a@... wrote:

?


Hi,
?
For knob repair I sometimes need to retap the threads on the knobs. The last tap and die set I bought was a cheap disaster. Any suggestions for buying good taps? setscrews?
?
jerry



--
Dave & Lynn Henderson
Manuals@...

PO Box 175
Welch,MN 55089



Re: Taps and setscrews

tom jobe
 

Hi Jerry,
Could I suggest that you buy a few thread forming taps ( aka "roll taps")
for the common sizes found on Tektronix equipment?
A normal tap cuts metal away to straighten up a deformed thread, and a roll
tap pushes the existing metal back into its proper position. The main
difference being that you have not lost additional material out of the
threaded area. Also the material has now been work hardened by the thread
forming action which is probably good.
The common small threads on these older Tektronix scopes seem to be either
4-40, 5-50 or 6-32 with an occasional 8-32. I keep roll taps in 4-40, 5-40,
and 6-32.
Any machine shop supply, McMaster-Carr and the like, or eBay will have them.
Try your local hobby shop or eBay for the set screws, they sell them in
small quantities. If you buy set screws from a machine shop supply you have
to buy a box of them which is usually 100, but the price will be very good
per set screw if you can use that many.
When using any roll tap, be sure to use some good lubricant so the base
metal and the tap don't weld themselves together.
tom jobe...
PS If you use a roll tap to make a new thread, the hole needs to be the
correct size or there will be no place for the metal to move to if the hole
is too small, and you will break the tap.

----- Original Message -----
From: <jmassen418a@...>
To: <tekscopes@...>
Sent: Friday, December 10, 2010 2:59 PM
Subject: [TekScopes] Taps and setscrews




Hi,

For knob repair I sometimes need to retap the threads on the knobs. The last
tap and die set I bought was a cheap disaster. Any suggestions for buying
good taps? setscrews?

jerry


Re: Taps and setscrews

Artekmedia
 

¿ªÔÆÌåÓý

With All due respect ( and? USED TO B E A HUGE craftsman fan myslef)? Craftsman? 20 years ago was good quality stuff ...Craftsman today is made from recycled automobiles and I'll leave it up to the reader to figure out where most of their stuff is manufactured ...not the same stuff anymore

Go visit your local tool and die shop and ask them where they by quality taps these days ...and for that matter what to avoid

Dave

On 12/10/2010 5:37 PM, Ken Chalfant wrote:
?

Hi,


Craftsman tools are reasonable. ?I've had a couple sets for going on twenty years and they serve me well.

For a wide variety of individual taps and setscrews check out . ?That's the website for McMaster Carr, a very well established industrial supply company in Chicago. ?They have been around since the dawn of time and are a fantastic resource.

Ken

On 10Dec, 2010, at 3:59 PM, jmassen418a@... wrote:

?


Hi,
?
For knob repair I sometimes need to retap the threads on the knobs. The last tap and die set I bought was a cheap disaster. Any suggestions for buying good taps? setscrews?
?
jerry



-- 
Dave & Lynn Henderson
Manuals@...

PO Box 175
Welch,MN 55089


Re: Taps and setscrews

Don Patterson
 

Jerry,

If you live near a larger metropolitan area, look for an industrial tool supply store. ?In the Detroit area, Production Tool, and some others provide quality taps and hardware to local companies. ?Your area should have a similar business.

Don Patterson


On Dec 10, 2010, at 5:59 PM, jmassen418a@... wrote:

?


Hi,
?
For knob repair I sometimes need to retap the threads on the knobs. The last tap and die set I bought was a cheap disaster. Any suggestions for buying good taps? setscrews?
?
jerry



Re: Taps and setscrews

 

Hi to Group......We use stainless set screws, you can get them on line, at MicroFasteners, or at ACE. McMaster Carr carries a wide range of Taps and Dies, excellent quality .?
????

?
?
Thank You
Kiss-Electronics
Ms Ashley Hall
183 N 5th Avenue
Cornelius, Oregon
97113
?
?
W7DUZ
?
?


-----Original Message-----
From: jmassen418a@...
To: tekscopes@...
Sent: Fri, Dec 10, 2010 2:59 pm
Subject: [TekScopes] Taps and setscrews




Hi,
?
For knob repair I sometimes need to retap the threads on the knobs. The last tap and die set I bought was a cheap disaster. Any suggestions for buying good taps? setscrews?
?
jerry



Re: Taps and setscrews

 

Hi,

Craftsman tools are reasonable. ?I've had a couple sets for going on twenty years and they serve me well.

For a wide variety of individual taps and setscrews check out . ?That's the website for McMaster Carr, a very well established industrial supply company in Chicago. ?They have been around since the dawn of time and are a fantastic resource.

Ken

On 10Dec, 2010, at 3:59 PM, jmassen418a@... wrote:

?


Hi,
?
For knob repair I sometimes need to retap the threads on the knobs. The last tap and die set I bought was a cheap disaster. Any suggestions for buying good taps? setscrews?
?
jerry



Re: Semi OT - Using Scope calibration signal as test signal

 

There was a suggestion of using a 1K series resistor which I will try.
Bob


--- On Fri, 12/10/10, lilacbarn wrote:

From: lilacbarn
Subject: [TekScopes] Re: Semi OT - Using Scope calibration signal as test signal
To: TekScopes@...
Date: Friday, December 10, 2010, 2:52 PM

?


--- In , "Robert" wrote:
>
> Can a 7603 calibration signal be used to test a PC LCD monitor inverter transformer? I was thinking of feeding it into the primary side and see what comes out.
> Bob
>
The 7603 spec (1984 ed) for the calibrator says that the 4V output is approx 450 Ohm output resistance, while the 40mV and 400mV are approx 50 Ohm.
The circuit schematic indicates that the 4V is driven directly from the collector of Q1072 which is a PNP device part#151-0220-00 which Sphere.bc.ca says is a 2N3906 rated at 40V Vceo/Vcbo and ??? joules of energy. The 450 Ohms is the load of the 40mV and 400mV attenuator on the collector pin.
Because of the direct connection I would be cautious about connecting anything inductive to this 4V calibrator output.
The other two outputs have quite a bit of protection in the resistive attenuator network, but excessive dissipation in those resistors may necessitate recalibration afterwords.

My 2 cents,
Geoff.



Re: RPR Conversion from Microfiche to OCR'ed PDF...............

David Garrido
 

Ok Folks...........we are 30% of the way home!!

The response in the last three hours has been awesome!!!!!!

Lets keep it going. ?We need more help for this to work out well for all involved.

Cheers,

David





?
?
???

Hello Everyone,

I have taken the liberty to rename the "CONVERTING FICHE TO PDF" thread.

We are about 20% to the finish line and I will continue to update the topic about once a week to let everyone know about the progress of the project.

Again, so far I have received commitments of about $200.00 USD. ?We are well on our way to getting this done folks. ?Lets keep it going. ?I will not ask for the deposit into my account until we have commitments enough to cover the project cost.......about $1000.00 USD if we are able to use Hakan's latest copy of December 01, 1997.

Please email me offline and I will put you on the list for your commitment and keep you updated privately as well as the list here.

Cheers,

David

?
???



--
David Garrido



Home of the "Classic" TEK Project and dedicated to the "not for profit" preservation, restoration, maintenance, repair, and archival of all "Classic" Tektronix Test and Measurement Equipment.


Taps and setscrews

 


Hi,
?
For knob repair I sometimes need to retap the threads on the knobs. The last tap and die set I bought was a cheap disaster. Any suggestions for buying good taps? setscrews?
?
jerry


Re: Semi OT - Using Scope calibration signal as test signal

 

--- In TekScopes@..., "Robert" <go_boating_fast@...> wrote:

Can a 7603 calibration signal be used to test a PC LCD monitor inverter transformer? I was thinking of feeding it into the primary side and see what comes out.
Bob
The 7603 spec (1984 ed) for the calibrator says that the 4V output is approx 450 Ohm output resistance, while the 40mV and 400mV are approx 50 Ohm.
The circuit schematic indicates that the 4V is driven directly from the collector of Q1072 which is a PNP device part#151-0220-00 which Sphere.bc.ca says is a 2N3906 rated at 40V Vceo/Vcbo and ??? joules of energy. The 450 Ohms is the load of the 40mV and 400mV attenuator on the collector pin.
Because of the direct connection I would be cautious about connecting anything inductive to this 4V calibrator output.
The other two outputs have quite a bit of protection in the resistive attenuator network, but excessive dissipation in those resistors may necessitate recalibration afterwords.

My 2 cents,
Geoff.


RPR Conversion from Microfiche to OCR'ed PDF...............

David Garrido
 

Hello Everyone,

I have taken the liberty to rename the "CONVERTING FICHE TO PDF" thread.

We are about 20% to the finish line and I will continue to update the topic about once a week to let everyone know about the progress of the project.

Again, so far I have received commitments of about $200.00 USD. We are well on our way to getting this done folks. Lets keep it going. I will not ask for the deposit into my account until we have commitments enough to cover the project cost.......about $1000.00 USD if we are able to use Hakan's latest copy of December 01, 1997.

Please email me offline and I will put you on the list for your commitment and keep you updated privately as well as the list here.

Cheers,

David