CG551AP Low SAC Uncal

 

I have 1 CG551AP which is having issues with its Low SAC that I
can't seem to understand. The output of the Low SAC is being driven
to -12.5V, after some stuffing around I have found that the output
of U1311 which provides offset adjustment for the Low SAC has is
output at about -13V despite the inverting input (roughly -13V)
sitting well below the non-inverting input (roughly -4.5V). I have
tried removing U1311 which allows the Low SAC to function normally
apart from there being an offset which is unable to be corrected
for. I have also tried replacing both U1311 (LM725) and U1211
(LF356) which did not help at all.

The big mystery for me is why the output of U1311 is not at positive
saturation rather than what appears to be negative saturation given
that the voltage of the inverting input is below that of the non
inverting input. The input voltages appear to be in the specified
range for the supply voltage being used i.e. +/-13.5V (worst case)
for a supply of +/-15V. I am even considering the possibility that
the circuit may be going unstable due to the frequency compensation
components going out of spec.

Has anyone experienced this problem before with the Low SAC in the
CG551AP or had similar issues with OP-Amps?


Re: [TekScopes2] Re:Replacing caps

Jonathan Thornton
 

Re: [TekScopes2] Re:Replacing caps


Hi everybody
Hurrah, I just got my TDS 544A working, usual acq fault, replaced all the
electrolytics on the acquisition PCB - still same fault, so I replaced all the
double and quad SOIC op.amps (TL 02/04 )--fault finding by attrition

I removed the caps by making a 2 pronged fork out of wire for the tip of the
soldering iron, to unsolder both sides of the cap simultaneously.
You can alternate the iron from one side of the cap to the other whilst lifting
the cap, but there a serious risk of pulling a pad off.
For the ICs I made a U shaped tip for the end of the iron out of brass sheet,
which with a bit of adjustment gripped the IC and unsoldered both sides at the
same time, very quick.

I then cleaned up all the pads with desoldering (the spellcheck tried to
change this to 'disordering' braid) and cleaned the board with isopropyl
alchohol.

For replacement i used solder paste and a very fine iron tip, all helped
enormously by a great purchase on XXXX,
a nearly new Leica StereoZoom 4 binocular microscope for under $180, it's one
of those things you don't know how you managed without
Finally a wash in the dishwasher.

Jonathan



--
I am using the free version of SPAMfighter for home users.
SPAMfighter has removed 12850 spam emails to date.
Paying users do not have this message in their emails.
Get the free SPAMfighter here:


Re: [TekScopes2] Re:Replacing caps

Jonathan Thornton
 

Hi everybody
Hurrah, I just got my TDS 544A working, usual acq fault, replaced all the electrolytics on the acquisition PCB - still same fault, so I replaced all the double and quad SOIC op.amps (TL 02/04 )--fault finding by attrition

I removed the caps by making a 2 pronged fork out of wire for the tip of the soldering iron, to unsolder both sides of the cap simultaneously.
You can alternate the iron from one side of the cap to the other whilst lifting the cap, but there a serious risk of pulling a pad off.
For the ICs I made a U shaped tip for the end of the iron out of brass sheet, which with a bit of adjustment gripped the IC and unsoldered both sides at the same time, very quick.

I then cleaned up all the pads with desoldering (the spellcheck tried to change this to 'disordering' braid) and cleaned the board with isopropyl alchohol.

For replacement i used solder paste and a very fine iron tip, all helped enormously by a great purchase on XXXX,
a nearly new Leica StereoZoom 4 binocular microscope for under $180, it's one of those things you don't know how you managed without
Finally a wash in the dishwasher.

Jonathan

--
I am using the free version of SPAMfighter for home users.
SPAMfighter has removed 12850 spam emails to date.
Paying users do not have this message in their emails.
Get the free SPAMfighter here:


TDS460A display intensity?

 

Just acquired a TDS460A.
Display intensity seems low (I have is set to 100)?and the contrast adjustment does nothing (on both internal and external monitor). Is this a common problem?
?
Is there an hours counter on it somewhere? wondering if the CRT is going out or some other problem. Seems to work perfectly otherwise.
?
Thanks,
?
David.
?




[Non-text portions of this message have been removed]


THM565 rear connector

 

Folks,
I bought two Tektronix THM565 portable scopes at a local surplus
dealer for a very cheap price.
However, the units came without batteries.
I could download the operating manual and found that there is an ac
adapter for this model, the THM5AC, but the only source I could find
for them was charging a modest value of $365.00 for one.
Since I do not even know if the units are operational, it makes no
sense trying to buy one.
Does anybody knows the pin assignment for the THM565 rear connector in
order to allow me to build a DIY power supply and test the units.
Thanks in advance for any help.
Roger


Re: Somewhat OT: HP 8553B/8443A system needs new home (Los Angeles area)

 

All of the items are now taken.
--John Gord

--- In TekScopes@..., "johncharlesgord" <johngord@...> wrote:

All,
I posted this on the HP/Agilent group and haven't had anyone jump
at the chance to pick this stuff up:
I have upgraded my spectrum analyzers and I no longer have room for
the old units. The following items are available free to anyone who
can use them and is willing pick them up. (I live in Venice,
California.)

140T/8552B/8553B system with a 8443A (late model, LED) tracking
generator.
2 each 8556A
8250A storage normalizer
Tek 604 monitor (use with 8250A)
HP 1331A variable persistence display

Let me know if you are interested.
-- John Gord ( johngord(at)verizon.net )


Re: Type 111 Pretrigger Pulse Generator not up to speed

Craig Sawyers
 

I have heard that
the frequency response of the recent digital scopes is not
guassian - that the rise time is
somewhat faster than bandwidth alone would indicate. That
makes me think that my
estimate of 111 risetime is actually erring on the high side.
That is one of the reasons that I prefer scopes from the steam age. What
you see is generally what is happening, rather than something that has been
digitised and processed using a process I have no control over.

Mostly I'm just glad it works at all, actually, I shouldn't
complain too hard. I suppose the
3S gear and the 111 can bootstrap themselves up to speed, one
step at a time.
Yeah - that sounds a good move. The 111 is actually a really nice pulser,
and should do nicely.

I also have a very odd avalanche pulser of more recent vintage that fits in
a TM5xx series power unit. The part number is not marked on the front, and
by all accounts there are several variants of this rather rare, and somewhat
prototypic beast - I've only heard of another two, and all three are
different. Over 4 years ago I put some pics on the tekscopes gallery,
including some sampler traces of the rise time, which measures an
aberration-free 250ps on an S6 sampling head (<30ps rise) - it is a fast
beastie. Gallery is called "Wierd TM500 plug in".

Craig


Re: [TekScopes2] Re:Replacing caps

davec
 

If I have a whole bunch to replace, I use solder paste, hot air, and a
quartz under board preheater. If I have only one or two to replace, I
use a fine tipped soldering iron and some #28AWG solder.
-=-=-=-

No hot air SMT apparatus or oven. Will be using fine tip iron.

Can you actually touch the tip to the caps' leads while the cap is resting in place? Or do you heat the pad while pressing down on the cap? (Flux paste is probably a good idea, too, I'd think...)

Thanks,
Dave


Re: Replacing caps (was: Help with Tek 2445)

davec
 

Chuck,
What about the small SM electro caps? Seems to me that these might be difficult to solder (being originally wave soldered) -- getting an iron in there might be difficult. I read that some folks use radial caps with very short, bent leads soldered to the pads.

Your experience?

Thanks,
Dave

-=-=-=-

Hi Dave,

It really doesn't matter, but because I order from Mouser most of the
time, I end up buying Nichicon. For the low voltage radial caps, I usually
use either UPW's or UHE's.

For higher voltage radial caps, I use UVZ's.

With axial caps, your choices become much more limited. Nichicon only makes
85C axial caps, but Xicon has a small series of 105C axial.

When I buy from Digikey, I tend to use the Panasonic FC, FM,EB,ED,EE...

Just about anything made today is an order of magnitude better than those
made when the 2445 was built.

-Chuck Harris

davec wrote:
Generally, I replace any bad electrolytic with 105C rated low ESR caps.
They are much longer lived than the 85C caps, and are good in switching
power supply filters.

-Chuck Harris
-=-=-=-

Chuck,
What manufacturer & series caps do you recommend?

Thanks,
Dave
------------------------------------

Yahoo! Groups Links



Re: Type 111 Pretrigger Pulse Generator not up to speed

 

--- In TekScopes@..., "Craig Sawyers" <c.sawyers@...> wrote:

Subject: [TekScopes] Type 111 Pretrigger Pulse Generator not
up to speed


I finally got around to replacing the caps in my Type 111 (s/n 490)
and find that it is not as fast as advertised; it's around 790pS,
where i had expected <=500pS.
Trying to deconvolve a fast rise time from a longer one is fraught with
problems. 500ps on a 1ns instrument response is a nominal 1.12ns - which is
only 14% different from your measured 1.275ns. Are you sure your timebase
is that accurate on the fastest sweep ranges (they often aren't).
I've got a chicken-and-egg problem here. I have a pile of 560-type sampling plugins that
I want to fix up, and need to get the 111 going to help do that :) Right now I'm borrowing
time on a modern Tek TDS4000 series. 350MHz digital phosphor unit. I have heard that
the frequency response of the recent digital scopes is not guassian - that the rise time is
somewhat faster than bandwidth alone would indicate. That makes me think that my
estimate of 111 risetime is actually erring on the high side.

Mostly I'm just glad it works at all, actually, I shouldn't complain too hard. I suppose the
3S gear and the 111 can bootstrap themselves up to speed, one step at a time.


281 pulser gets hot

Tim Phillips
 

From Tim P (UK)
I have a 281 TDR pulser on a 3S1.
After about 10 minutes the pulser
feels very hot. Should it?
I get a nice 1 volt pulse OK.
Tim.


Re: Tek 315D Oscilloscope Instruction Manual

 

--- On Fri, 9/19/08, Pierre F1BLF <pierrot8254@...> wrote:

From: Pierre F1BLF <pierrot8254@...>
Subject: [TekScopes] Tek 315D Oscilloscope Instruction Manual
To: TekScopes@...
Date: Friday, September 19, 2008, 9:29 PM






Hi Tekscopers,

I'm looking for a Tek 315D Instruction Manual,
preferably in a scanned form of PDF.

Many thanks in advance.

Pierre F1BLF


I could use one of those manuals too as I have one of those fine old relics in my collection. It is the only scope I have that I seriously thought of washing as it's the filthiest. Has broken fan mounts thanks to UPS but otherwise it is in very good physical condition. Complete with stacks of selenium rectifiers with a sticker that fell off stating that those had been replaced ages ago. Also contains plenty of "bumblebee" caps. Think I might be missing a cover plate over the HV section though. Occasionally I'll look at the scope but not planning on restoration any time soon, mainly due to the washing issue keeping me from even getting started. I?have?the scoped?wrapped in plastic so it doesn't get any dirtier!
?

?
Sorry to hijack the message. A few good pics in?there of the little beast if you care to look.
?
?
Dave


Somewhat OT: HP 8553B/8443A system needs new home (Los Angeles area)

 

All,
I posted this on the HP/Agilent group and haven't had anyone jump
at the chance to pick this stuff up:
I have upgraded my spectrum analyzers and I no longer have room for
the old units. The following items are available free to anyone who
can use them and is willing pick them up. (I live in Venice, California.)

140T/8552B/8553B system with a 8443A (late model, LED) tracking
generator.
2 each 8556A
8250A storage normalizer
Tek 604 monitor (use with 8250A)
HP 1331A variable persistence display

Let me know if you are interested.
-- John Gord ( johngord(at)verizon.net )


Tek 315D Oscilloscope Instruction Manual

 

Hi Tekscopers,

I'm looking for a Tek 315D Instruction Manual,
preferably in a scanned form of PDF.

Many thanks in advance.

Pierre F1BLF


part numbers for 1% resistors

jmassen418
 

Greetings,

Does someone have a list of the tektronix parts numbers for 1/8w 1%
metal film resistors? The prefix is 321. For example the part number
for a 1.47k resistor is 321-0209-00. I am trying to organize my part
bins.

jerry


Re: Type 111 Pretrigger Pulse Generator not up to speed

Craig Sawyers
 

Subject: [TekScopes] Type 111 Pretrigger Pulse Generator not
up to speed


I finally got around to replacing the caps in my Type 111 (s/n 490)
and find that it is not as fast as advertised; it's around 790pS,
where i had expected <=500pS.
Trying to deconvolve a fast rise time from a longer one is fraught with
problems. 500ps on a 1ns instrument response is a nominal 1.12ns - which is
only 14% different from your measured 1.275ns. Are you sure your timebase
is that accurate on the fastest sweep ranges (they often aren't).

The only sure way to measure it more accurately is to use something with a
response much faster than what you are trying to measure - ie a sampling set
up (which is precisely what the pretrigger of the 111 is for - getting the
sampling timebase up and running before the measurement event). To get a
good grip on the rise time and potential aberrations, you need a rise time
at least three times faster than what you are trying to measure (for a 10%
error). So at most at 170ps rise time. An S2, S4 or S6 sampling head will
do this, or a 1S2 from the earlier generation.

My main scope is set up with a spectrum analyser at the moment, or I would
get sampling up and running and measure my own type 111. It would just be
confirmation though - I've measured it before, and know for certain that it
gives less than 500ps rise.

Craig


Option 13 in an R7912?

Jose V. Gavila
 

Hello!

Today I have received my last Tek toy, an R7912 mainframe :-). It has a label in the plug-in compartment base which says 'Option 13'. I wonder if someone know what does it mean.

Thanks!

JOSE
-------------------------------------------------
73 EB5AGV - JOSE V. GAVILA
IM99sm La Canyada - Valencia (SPAIN)

Vintage Radio and Test Equipment

European Boatanchors List

Amateur Repairs List


Need 475 scope

Dave_s
 

Hi folks,

The repair of my sick 475 scope is not occurring EASILY despite SO AMNY HELPFUL suggestions from members of this group. It is possible that the repair may occur.

Just in case, I am now searching for a replacement usable 475, 475A, or 485 in good working condition. What is available? Never having used EBAY, I am also interested in other sources that refurbish, calibrate, offer good used scopes, etc.


Still optimistic that I will find the fault parts.


Thank you, 73, W6MIK, Dave


2N1516 & 2N964 needed

Tim Phillips
 

from Tim P. (UK)
I need some 2N1516 and 2N964 s (Ge)
anyone know of a supply? Sphere can't help.
many thanks
Tim


Re: 2215 Problem

tom jobe
 

Hello Gary,
For what it's worth...
That same capacitor, 285-1222-00, was also used in the military version of
the 2235, It was used directly across the incoming AC lines just before the
full wave bridge rectifier as part number C904, and it did not have a
resistor in series with it..
I have a few partially stripped military 2235 mainboards, and they are all
missing C904, so I would guess that it must have been a component that had
problems.
Other 22XX scopes I looked at use a .0022 uF (2200 pF) 250 V capacitor for
C904.
tom jobe...

----- Original Message -----
From: "Garey Barrell" <k4oah@...>
To: "Tekscopes" <TekScopes@...>
Sent: Thursday, September 18, 2008 7:55 PM
Subject: Re: [TekScopes] 2215 Problem


Stefan -

It's not right across the line. It's in series with a 470 ohm 1/2W CC
resistor (which now measures 19 ohms) and looks like a "snubber" across
the AC line "after" a series inductor. There's an MOV (250VAC) and a
.001 cap across the line just after the fuse on the "line" side of the
inductor. This unit is rated for 220 primary power, although it's never
been used on anything but 120VAC.

73, Garey - K4OAH
Glen Allen, VA



Stefan Trethan wrote:
I'm assuming this was the x-capacitor (across the power line)?

If so the very same thing happened to me and a fluke 8842a. It had just
come
from singapure, and worked just fine for a few hours beforehand.

The cap leaked a nasty brown liquid (and acrid smell). It stained my
fingers
for days. The cap was actually inside a Schaffner brand mains filter
module.

Anyway, if it was the x-capacitor, you should replace it with a rated x1
capacitor for mains filter purposes.

ST

On Thu, Sep 18, 2008 at 11:06 PM, Garey Barrell
<k4oah@...>wrote:


I was using my 2215 for an hour or so the other night, and moved over
to
the desk to look at something else. About 30 minutes later I heard a
"sizzle - pop" and smoke comes rolling out the back of the cabinet!
Both sides! I walked over and turned up the Brightness, and had a
trace. So I shut it off, took it out of the case and did a close
inspection. Nothing! I couldn't find my small TORX bits (don't ever
move.......) but everything looked fine. So I turned it back on,
measured the supply voltages and all were right on..... Scope seemed to
be operating normally!?!?!

Today, I found my little TORX bits, and removed the inductor (the
HUUUMMMMMM) from the back panel and lo and behold, on the main board
underneath the inductor cover was what remained of C926, fanned out
like
a Rolodex. P/N 285-1222-00, 0.068 uF @ 250 VDC, 20%. This is a tiny
rectangular 1/4" x 3/4" by 1/2" high epoxy thing. Of course none of
the
million and a half capacitors that I have will even come close to
fitting mechanically.

So. Anyone have a source for these things, or perhaps have an "extra"
one.

help.....

--
73, Garey - K4OAH
Glen Allen, VA

------------------------------------

Yahoo! Groups Links