Re: Novice question - 475, what mode is this?
The Tek 475 is analog, with a CRT screen, so the tools for LCD screens will probably not work. The intensity knob may be turned down too low (as Bob mentions), but on my 475, if I turn down the intensity until I can barely see the trace on just the A sweep, and then go to Mix, I can still see the A trace along with the B trace. But when I turn it down further, the B trace is still visible as the A disappears. There may be a dirty Mix switch or some other switch.
toggle quoted message
Show quoted text
On 6/22/2018 8:27 AM, Jeff Urban wrote:
On Thu, Jun 21, 2018 at 02:29 pm, Bob Albert wrote:Is the intensity set too low?Sounds like you are not seeing the left side of the trace. Couple possible reasons for this, one is that the brightness or pedestal on your monitor is set too low. There are monitor test programs for this, as it is hard to tell just where black is black but none of the grays are black. One such program is called ntest.exe, which is totally stand alone, would run off anything, like a CD or stick or whatever. One of the patterns is exactly for that and says adjust until you see some particular block in the pattern. That should eliminate any black clipping. |
Re: was capacitors now about switches
Just want to make it clear that the "buttons" I referred to in the OP are the small, self-contained switches that solder onto the board individually - not the rubbery sheet kind that carry carbon films that make contact on patterns on the board. I've had plenty of troubles with the rubbery kind, but never a problem until now, with the discrete snap-action switches, as long as the "snap" was still good. I have had plenty that had mechanical failure from wearout or breakage, where the plastic actuators were deformed or bent off-axis (or even the whole switch unit popped open) from too much force, or the detent action was lost and the contact would either short or open permanently.
Ed |
Re: was capacitors now about switches
Hello,
PCB buttons can and do fail all the time. A fun example was the Vernee Apollo lite, where too tight manufacturing tolerances allowed sweat grease to accumulate. I have summed it all up in a video, just in case anyone feels like taking a look: Tam -- With best regards Tam Hanna --- Enjoy electronics? Join 8300 other followers by visiting the Crazy Electronics Lab at |
Re: Novice question - 475, what mode is this?
On Thu, Jun 21, 2018 at 02:29 pm, Bob Albert wrote:
Sounds like you are not seeing the left side of the trace. Couple possible reasons for this, one is that the brightness or pedestal on your monitor is set too low. There are monitor test programs for this, as it is hard to tell just where black is black but none of the grays are black. One such program is called ntest.exe, which is totally stand alone, would run off anything, like a CD or stick or whatever. One of the patterns is exactly for that and says adjust until you see some particular block in the pattern. That should eliminate any black clipping. Another possibility is if you have an edge lit LCD monitor. As the viewing angle changes vertically in those, the brightness changes. This effect varies, some models are better/worse than others. Some are terrible and you only get a limited range or it does all kinds of funky crap like show a negative image or all the color disappears if you are too low or high, things like that. The one I got - ntest.exe, is from Nokia. If you can't find a suitable program I can upload it or you can hit me up with an email and I can send it as an attachment in a reply. |
Re: Novice question - 475, what mode is this?
That's "MIX" - mixed mode. What it does is instead of intensifying the part of the trace to be expanded by the B timebase it simply switches to B timebase at that point of the sweep, wherever you set the delay control.
On the left you are going at at the slower A sweep on the right it actually is intensified but since the sweep is faster it needs that. Push the intensify button and you'll get the whole trace of the slower sweep where the sine waves look compressed but right where the B sweep starts in mixed mode it will be intensified. Looking at the mixed mode tell me that the intensified portion will only be a few cycles - a small version of what you see on the right. Then of course when you go to B sweep the whole screen gets the delayed part - which is currently on the right in mixed mode. I hope you got that, sometimes I am not the best explainer in the world. The manual might do better, not sure. I am not sure of the usefulness of mixed mode, I can't think of a case where I would use it. Well effectively, I like playing around with such things so it is useful for that...:-) |
was capacitors now about switches
Greetings,
Allow me to add to the comments by Ed and Shailendra regarding push button switches. I had the remote for an auto fail and drain the batteries. Exploring the problem I first determined that the remote was not transmitting. From there I discovered that the conductive coating on the contact areas of the molded buttons had dislodged and made conductive paths (shorts) on the switch areas of the PCB. After cleaning this residue away I was able to return the unit to proper function with a very neat little product from ¡°ButtonWorx¡±. They make different size peel and stick conductive buttons which stick to the PCB over the switch circuit area and below the molded button. Depressing the molded button in turn depresses the ¡°new¡± button. I have no affiliation with ¡°ButtonWorx¡±, I¡¯m simply happy I discovered their product. Regards, Ken |
Re: Speaking of capacitor problems - it's not always their fault
On Thu, Jun 21, 2018 at 09:36 am, Ed Breya wrote:
Hello Ed, I too have experienced the same kind of switch failure on several devices, including a LCD TV, a PC monitor, and music systems. The tiny tactile switches do fail after some time. It is worth checking them when the controls appear locked up, or a certain function is continuously active without pressing any button. Replacement is the best solution, but if you can't find a replacement, you can use a contact cleaner to soak the button for some time, and then exercise it to restore operation. Although the buttons appear sealed, some amount of liquid seeps in and helps make the switch work for some time. Eventually, you must replace the switch to get ling term fix. Regards Shailendra |
Re: Tek 465 no display
Michael A. Terrell
toggle quoted message
Show quoted text
-----Original Message-----
From: Jon Batters <jonbatters32@...>There is only the meter module, and a current transformer. It is connected across the AC line, and it uses a current transformer. This is shown in a photo in the listing. I haven't tested it, yet. None of my True RMS meters have a current reading function. They are Fluke 8920/8921/8922 series meters. |
Re: - Early Telequipment 'Scope
Hi Adrian,
toggle quoted message
Show quoted text
Thanks for your reply. So far, you are the only group member who has expressed interest. I attach photograph of the S31. You may be able to see that the input connector has been changed (by the previous owner) to the more conventional BNC type. Otherwise, I don't think there are any changes. It is not obvious how the case can be removed, but it seems to be constructed from 6 sections. I think that the rear panel must clip on somehow and hold the sides in place which, in turn, hold the top and bottom vented panels in place. Hope the photo is of interest and you could certainly have the scope if you don't mind travelling up to Yorkshire sometime. No rush, since the scope has just been stored in the loft for the past 40 years! Just let me know if you decide not to have it and I will also let you know if there is any other interest. Ken. On 19:42, Thu, 21 Jun 2018 Adrian, <Adrian@...> wrote:
Hi Ken, |
Re: Novice question - 475, what mode is this?
Bob Albert
Is the intensity set too low?
toggle quoted message
Show quoted text
Bob On Thursday, June 21, 2018, 1:45:58 PM PDT, Dmitri Shuev <dshuev@...> wrote:
Hello, everybody: I am about to test a 475 that I have acquired for my radio alignment projects, and since I am relatively new to the oscilloscopes, I wonder if I am missing something... I am trying to replicate what I think is a Delayed Sweep combination trace, but for some reason rotating the Time/Delay Time control, and trying to move the switch point with the Delay Time Position pot does not result in the combination curve ( sample here - /g/TekScopes/photo/60634/0?p=Name,,,20,1,0,0) - all I am getting is a blank on the left before the trace begins... Am I understanding this mode correctly? Something is wrong with timebase control? Or am I missing something fundamental? Thank you, Dmitri |
Novice question - 475, what mode is this?
Hello, everybody:
I am about to test a 475 that I have acquired for my radio alignment projects, and since I am relatively new to the oscilloscopes, I wonder if I am missing something... I am trying to replicate what I think is a Delayed Sweep combination trace, but for some reason rotating the Time/Delay Time control, and trying to move the switch point with the Delay Time Position pot does not result in the combination curve ( sample here - /g/TekScopes/photo/60634/0?p=Name,,,20,1,0,0) - all I am getting is a blank on the left before the trace begins... Am I understanding this mode correctly? Something is wrong with timebase control? Or am I missing something fundamental? Thank you, Dmitri |
Re: Tek 465 no display
On 2018-06-21 12:51 PM, musicamex@... wrote:
Albert and LolPol,I've scanned the 465 manual with all foldouts in one piece, not stitched. I uploaded it to Tekwiki: --Toby
|
Re: 465 bad rise time
This the the rise time now at 5mV. I could probably get it better but but my bnc cables and 50ohm terminator are not great quality. Should I keep working at it?
toggle quoted message
Show quoted text
/g/TekScopes/photo/59361/4?p=Name,,,20,1,0,0 On Thu, Jun 21, 2018 at 12:00 pm, Fabio Trevisan wrote:
|
Re: Up to date capacitor list for Tek 2465A and 2465B scopes (2018)
Hello,
In principle, Menachems approach works - looks like this: However, be careful of vibration and inductance - SMD parts tend to be better in both of these regards. -- With best regards Tam HANNA (emailing on a keyboard-less handset, sorry for spelling mistakes and brevity) |
Re: 465 bad rise time
Hello lop pol,
toggle quoted message
Show quoted text
I've followed up this topic from the start, without having actually jumped in, hoping since the beginning that someone would ask or make remarks about two or three important things (to check, or to be aware of). 1. That's a question... What type of Vertical Output board does your 465 have. the one discrete or the one with the custom Tek IC? It's important because the High Freq. compensation networks are different from one type of board to the other and since your problem manifests in both channels, there's a great likelihood the problem is located on the Vertical Output Amplifier board, instead of on the pre-amplifier. Without that information is difficult to advise further. 2. If your vertical amplifier is of the "Teck IC" kind, look also to the BIAS trimpot. If the wiper is open, the amplifier may be operating on the wrong bias point and that is supposed to impact the high-frequency performance. Since the procedure for calibrating the bias calls for applying a 100Mhz signal and adjusting the bias for the highest p-p display, from there we deduct it has a marked effect on the low nanosecond time constant, which is exactly where your issue lies. 3. Besides what Raymond already mentioned, that is not uncommon for some of those trimmer caps fail open (I had 3 of those failed on my 464), I want to bring to your attention that ALL the trimpots that are used on the H.F. compensation networks are connected as "rheostats" so, a false contact or open wiper will make any of them appear open, or in the best case, the full trimpot's resistance, which will be just as bad for the step response. Since many of them are in series with trimmer caps, it's easy to check them in circuit, without needing to lift any leg, or remove any component. Last but not least, despite your problem clearly doesn't seem to be on the pre-amplifiers, when you find out the culprit and eventually replace one faulty trimmer cap or faulty trimmer resistor, you will eventually need to go through the Vertical High Frequency step response calibration, and for that, it's important that you follow the procedure, and from your description, I think you're already doing wrong... While choosing for the right input level / input attenuator (pass-through) and vertical attenuator settings, it's important to do all the High-Frequency calibration with the oscilloscope set to the most sensitive setting... i.e.: 5mV/div (not 10mV, not 100mV... and not 200mV/div). In other words, you need to source 25mVpp to the oscilloscope's input (for a 5 division display). This is what will assure you that none of the high impedance input attenuators are upsetting the subtle high frequency adjustments and artifacts. While - in theory - the high-impedance input attenuators shouldn't degrade the step response of the osciloscope, they're not perfect and they do cause some minor disturbances and aberrations to the step response (and they can also be misadjusted, something that you will only correct at a later stage) so, to make a "clean" H.F. calibration, you need to make sure you got them (the high impedance input attenuators) out of the way. The only compromise to that rule is the input probe itself... If you intend to use the oscilloscope with a particular probe (let's say, a 10:1), then you need to connect that probe to the input, use the scope set to 5mV/div, and apply, accordingly, a 250mVpp signal to the probe (if it's a 10:1 probe). This will assure you will compensate in the scope, for the probe's H.F. inaccuracies or artifacts. Yet on this topic, you need to be sure to "source" that signal (regardless if directly to the input or through your preferred probe), from a low impedance source (50Ohm). The latter is important because you don't want the frequency response of the incoming signal to be affected by variations caused by the 'scope's input capacitance, or your probe's input capacitance. The lower the impedance of the source signal, the less the input capacitance will matter and, of course, you will also want to assure the interconnect cable is properly terminated. Good luck with that, Krgrds, Fabio On Sun, Jun 17, 2018 at 06:47 pm, lop pol wrote:
|
Re: - Early Telequipment 'Scope
Hi Ken,
toggle quoted message
Show quoted text
I would be interested if nobody nearer to you pops up - bit of a drive from Eastern Anglia. Does it have its case by any chance? Mine does not and I'm trying to re-create one based on a very tiny, badly scanned image from a copy of a 1957 Wireless World, and struggling! All I can deduce at the moment is that it may have been constructed more like the single '5-sided box' design of the later S51 than the three piece 'clam shell' design of the face-lifted S31 because it is missing all the fixing holes it would need to hold the base part of that on. I would really appreciate a pic! All the best, Adrian On 6/21/2018 1:35 PM, Ken wrote:
Hi Adrian and group, |
Re: Tek 465 no display
Albert and LolPol,
Thank you for your input. Somehow i missed this download. I found LV test points and values in section 6 of this manual. I'll report my findings this weekend when i have some time to get back to testing....maybe sooner. The monsoon season here has started and my 4 hectares eats up allot of my free time in the lab, but green is beautiful compared to dry brown winters at our 1200 m elevation. I also found allot of info in the charts and schematics in the last part of the 465 late .pdf. The display troubleshooting will be of help hopefully once i verify my LV test points. I dont have a HV probe for my VOM and don't have the source here for the appropriate resistors to build an appropriate voltage divider, so I'm hoping the solution is something in the LV circuit. Is there a source for pdf or other format files of the foldout pages in unchopped original layout? It would make reading them much easier. I guess that I could print/splice/tape the ones i need.... Thanks again! I feel the cloud lifting! Russ |
Re: Up to date capacitor list for Tek 2465A and 2465B scopes (2018)
On Thu, Jun 21, 2018 at 09:37 am, machineguy59 wrote:
OUCH!! Yahoo email to groups.io lost context. So here is the message again. Menahem, Thanks for the information. I was not clear in expressing my concern. I will clarify my problem. The A5 boards I have suffered severe electrolyte damage. One of them (my current project) lost three pads and two traces to the etching of spilled electrolyte. I replace these pads with pads cut from foil and glued to the board with CW2500 epoxy. You can see an example of this process here: Although my hand cut pads are not as perfect as his machine cut pads. I get pretty close. Also, my hand cut traces are not as narrow as original. I will try to post pictures of my work next time I have the scope opened up. I did not take "before" pictures. Its tedious but it works and saves the board without using jumper wires or gluing parts to the board. My concern is that CW2500 epoxy is great stuff but still not as good as original for strength. The weight of a leaded part leveraged by the leads would very likely wrench my replacement pads from the board. |