Re: 2220 POST displays a large X on the screen
Question for the OP.......does your scope have either of the side options? RS232, GPIB? If so disconnect them. The battery backup is on the side option on these so if there is no option installed the scope shouldn't really be booting into any weird state.
---In TekScopes@..., <edbreya@...> wrote : I looked at the 2230 service manual. In section 5 it describes adjusting the store/nonstore registration using a box generator accessible from the diagnostics menu. So the X isn't normally seen or user accessible. I think it was probably a factory setup mode for coarse checking and alignment during production, to make sure the digital display basically worked and was on-screen - if it was too far out of whack, you may not be able to read the menus. Also, in production, you wouldn't want to be fooling around waiting for boot-up and navigating menus while firing up a unit. A hardware override can force the mode, and it's ready to go, and you can see the status. I don't recall how much automation was applied, but they likely checked the fine details via GPIB once the units were proofed and operational past a certain point. If you look at the screen shot, you can imagine that if the X was tweaked slightly up and left, the center would be on the CRT origin, and the lines would intersect the graticule grids almost perfectly, which would be correct. I'm still guessing a jumper or DIP switch setting activates it, but there likely won't be anything said about it in the manual. |
Re: 2465B U400 Availability?
I've only played around with it briefly, but I'm pretty sure both cursors move smoothly. Everything seems to be pointing to U400 but have not checked voltages yet.
toggle quoted message
Show quoted text
Previous owner had taken it to a cal lab and the paperwork came back from there stating U400 is bad. I don't know if a cal lab would have checked voltages, though. Will know more soon. Thanks, Barry - N4BUQ ----- Original Message -----
From: "very_fuzzy_logic@... [TekScopes]" <TekScopes@...> |
Re: 2220 POST displays a large X on the screen
I looked at the 2230 service manual. In section 5 it describes adjusting the store/nonstore registration using a box generator accessible from the diagnostics menu.
So the X isn't normally seen or user accessible. I think it was probably a factory setup mode for coarse checking and alignment during production, to make sure the digital display basically worked and was on-screen - if it was too far out of whack, you may not be able to read the menus. Also, in production, you wouldn't want to be fooling around waiting for boot-up and navigating menus while firing up a unit. A hardware override can force the mode, and it's ready to go, and you can see the status. I don't recall how much automation was applied, but they likely checked the fine details via GPIB once the units were proofed and operational past a certain point. If you look at the screen shot, you can imagine that if the X was tweaked slightly up and left, the center would be on the CRT origin, and the lines would intersect the graticule grids almost perfectly, which would be correct. I'm still guessing a jumper or DIP switch setting activates it, but there likely won't be anything said about it in the manual. |
Re: 2465B U400 Availability?
You are right, it could be U400.? But I would try the other things first, lol.? In Barry's case, he has a source for an alternate U400.? I would try that AFTER?verifying all voltages are in spec.? No need to damage the alternative U400 with bad supply voltages.?
I am waiting to hear that your "new" U400 corrects your problem. From: "very_fuzzy_logic@... [TekScopes]" <TekScopes@...> To: TekScopes@... Sent: Wednesday, March 30, 2016 3:21 PM Subject: Re: [TekScopes] Re: 2465B U400 Availability? ? Two points: Barry's 2465B and mine have very similar issues with non-linearity on ch1 and ch2 but ch3 and ch4 OK. At least on my scope the power rails are ok for voltage and ripple. I also checked the power supplies where they have common decoupling to U100 and U200 and so far have not found anything amiss. There are threads here and elsewhere suggesting U400 as a possible cause and it's price is consistent with it failing frequently! I haven't yet checked U400 bias current or the voltages on the channel select inputs. Secondly, Barry has several noisy analogue controls, DAC stability could be a common cause. Do the y and t cursors move smoothly? If they don't then it would be well worth checking the DAC output, there is setup procedure near the beginning of the calibration section of the manual that puts the DAC output on one of the DIP header pins and it responds to cursor movement - should be a good test. Meanwhile my replacement U400 has been stuck in transit for over a week en route to the UK! Roger #yiv2028227971 #yiv2028227971 -- #yiv2028227971ygrp-mkp {border:1px solid #d8d8d8;font-family:Arial;margin:10px 0;padding:0 10px;}#yiv2028227971 #yiv2028227971ygrp-mkp hr {border:1px solid #d8d8d8;}#yiv2028227971 #yiv2028227971ygrp-mkp #yiv2028227971hd {color:#628c2a;font-size:85%;font-weight:700;line-height:122%;margin:10px 0;}#yiv2028227971 #yiv2028227971ygrp-mkp #yiv2028227971ads {margin-bottom:10px;}#yiv2028227971 #yiv2028227971ygrp-mkp .yiv2028227971ad {padding:0 0;}#yiv2028227971 #yiv2028227971ygrp-mkp .yiv2028227971ad p {margin:0;}#yiv2028227971 #yiv2028227971ygrp-mkp .yiv2028227971ad a {color:#0000ff;text-decoration:none;}#yiv2028227971 #yiv2028227971ygrp-sponsor #yiv2028227971ygrp-lc {font-family:Arial;}#yiv2028227971 #yiv2028227971ygrp-sponsor #yiv2028227971ygrp-lc #yiv2028227971hd {margin:10px 0px;font-weight:700;font-size:78%;line-height:122%;}#yiv2028227971 #yiv2028227971ygrp-sponsor #yiv2028227971ygrp-lc .yiv2028227971ad {margin-bottom:10px;padding:0 0;}#yiv2028227971 #yiv2028227971actions {font-family:Verdana;font-size:11px;padding:10px 0;}#yiv2028227971 #yiv2028227971activity {background-color:#e0ecee;float:left;font-family:Verdana;font-size:10px;padding:10px;}#yiv2028227971 #yiv2028227971activity span {font-weight:700;}#yiv2028227971 #yiv2028227971activity span:first-child {text-transform:uppercase;}#yiv2028227971 #yiv2028227971activity span a {color:#5085b6;text-decoration:none;}#yiv2028227971 #yiv2028227971activity span span {color:#ff7900;}#yiv2028227971 #yiv2028227971activity span .yiv2028227971underline {text-decoration:underline;}#yiv2028227971 .yiv2028227971attach {clear:both;display:table;font-family:Arial;font-size:12px;padding:10px 0;width:400px;}#yiv2028227971 .yiv2028227971attach div a {text-decoration:none;}#yiv2028227971 .yiv2028227971attach img {border:none;padding-right:5px;}#yiv2028227971 .yiv2028227971attach label {display:block;margin-bottom:5px;}#yiv2028227971 .yiv2028227971attach label a {text-decoration:none;}#yiv2028227971 blockquote {margin:0 0 0 4px;}#yiv2028227971 .yiv2028227971bold {font-family:Arial;font-size:13px;font-weight:700;}#yiv2028227971 .yiv2028227971bold a {text-decoration:none;}#yiv2028227971 dd.yiv2028227971last p a {font-family:Verdana;font-weight:700;}#yiv2028227971 dd.yiv2028227971last p span {margin-right:10px;font-family:Verdana;font-weight:700;}#yiv2028227971 dd.yiv2028227971last p span.yiv2028227971yshortcuts {margin-right:0;}#yiv2028227971 div.yiv2028227971attach-table div div a {text-decoration:none;}#yiv2028227971 div.yiv2028227971attach-table {width:400px;}#yiv2028227971 div.yiv2028227971file-title a, #yiv2028227971 div.yiv2028227971file-title a:active, #yiv2028227971 div.yiv2028227971file-title a:hover, #yiv2028227971 div.yiv2028227971file-title a:visited {text-decoration:none;}#yiv2028227971 div.yiv2028227971photo-title a, #yiv2028227971 div.yiv2028227971photo-title a:active, #yiv2028227971 div.yiv2028227971photo-title a:hover, #yiv2028227971 div.yiv2028227971photo-title a:visited {text-decoration:none;}#yiv2028227971 div#yiv2028227971ygrp-mlmsg #yiv2028227971ygrp-msg p a span.yiv2028227971yshortcuts {font-family:Verdana;font-size:10px;font-weight:normal;}#yiv2028227971 .yiv2028227971green {color:#628c2a;}#yiv2028227971 .yiv2028227971MsoNormal {margin:0 0 0 0;}#yiv2028227971 o {font-size:0;}#yiv2028227971 #yiv2028227971photos div {float:left;width:72px;}#yiv2028227971 #yiv2028227971photos div div {border:1px solid #666666;height:62px;overflow:hidden;width:62px;}#yiv2028227971 #yiv2028227971photos div label {color:#666666;font-size:10px;overflow:hidden;text-align:center;white-space:nowrap;width:64px;}#yiv2028227971 #yiv2028227971reco-category {font-size:77%;}#yiv2028227971 #yiv2028227971reco-desc {font-size:77%;}#yiv2028227971 .yiv2028227971replbq {margin:4px;}#yiv2028227971 #yiv2028227971ygrp-actbar div a:first-child {margin-right:2px;padding-right:5px;}#yiv2028227971 #yiv2028227971ygrp-mlmsg {font-size:13px;font-family:Arial, helvetica, clean, sans-serif;}#yiv2028227971 #yiv2028227971ygrp-mlmsg table {font-size:inherit;font:100%;}#yiv2028227971 #yiv2028227971ygrp-mlmsg select, #yiv2028227971 input, #yiv2028227971 textarea {font:99% Arial, Helvetica, clean, sans-serif;}#yiv2028227971 #yiv2028227971ygrp-mlmsg pre, #yiv2028227971 code {font:115% monospace;}#yiv2028227971 #yiv2028227971ygrp-mlmsg * {line-height:1.22em;}#yiv2028227971 #yiv2028227971ygrp-mlmsg #yiv2028227971logo {padding-bottom:10px;}#yiv2028227971 #yiv2028227971ygrp-msg p a {font-family:Verdana;}#yiv2028227971 #yiv2028227971ygrp-msg p#yiv2028227971attach-count span {color:#1E66AE;font-weight:700;}#yiv2028227971 #yiv2028227971ygrp-reco #yiv2028227971reco-head {color:#ff7900;font-weight:700;}#yiv2028227971 #yiv2028227971ygrp-reco {margin-bottom:20px;padding:0px;}#yiv2028227971 #yiv2028227971ygrp-sponsor #yiv2028227971ov li a {font-size:130%;text-decoration:none;}#yiv2028227971 #yiv2028227971ygrp-sponsor #yiv2028227971ov li {font-size:77%;list-style-type:square;padding:6px 0;}#yiv2028227971 #yiv2028227971ygrp-sponsor #yiv2028227971ov ul {margin:0;padding:0 0 0 8px;}#yiv2028227971 #yiv2028227971ygrp-text {font-family:Georgia;}#yiv2028227971 #yiv2028227971ygrp-text p {margin:0 0 1em 0;}#yiv2028227971 #yiv2028227971ygrp-text tt {font-size:120%;}#yiv2028227971 #yiv2028227971ygrp-vital ul li:last-child {border-right:none !important;}#yiv2028227971 |
Re: 2465B U400 Availability?
Two points:
Barry's 2465B and mine have very similar issues with non-linearity on ch1 and ch2 but ch3 and ch4 OK. At least on my scope the power rails are ok for voltage and ripple. I also checked the power supplies where they have common decoupling to U100 and U200 and so far have not found anything amiss. There are threads here and elsewhere suggesting U400 as a possible cause and it's price is consistent with it failing frequently! I haven't yet checked U400 bias current or the voltages on the channel select inputs. Secondly, Barry has several noisy analogue controls, DAC stability could be a common cause. Do the y and t cursors move smoothly? If they don't then it would be well worth checking the DAC output, there is setup procedure near the beginning of the calibration section of the manual that puts the DAC output on one of the DIP header pins and it responds to cursor movement - should be a good test. Meanwhile my replacement U400 has been stuck in transit for over a week en route to the UK! Roger |
Re: 2465B U400 Availability?
Only Channels 1 and 2 exhibit vertical linearity issues.? This is re-assuring for the CRT.? Since its only Channels 1 & 2 but its both channels in the same way, I suspect its a circuit problem and not likely the U400.? Keep in mind that bad supply rails can cause vertical distortion AND less sharp traces.? Its unusual for these scopes to have good supply rails after 20+ years.? Nearly all require re-capping the power supply and the A5 board as a minimum.This is the reason for my concern about being dropped.? This case is very sturdy, meant to be a portable scope.? I cant imagine how to distort the case with a normal blow.? I have an older rack mounted 2465 that was dropped hard, broke the CRT (at the neck, near the rear pins) and bent the screws holding it to the rack.? But the frame is still straight. As for the broken fanI have never owned?a 2400 series scope that didn't come with some kind of fan issue.? Yours seems minor compared to others.? From: "Barry n4buq@... [TekScopes]" <TekScopes@...> To: TekScopes@... Sent: Wednesday, March 30, 2016 8:11 AM Subject: Re: [TekScopes] Re: 2465B U400 Availability? ? Only Channels 1 and 2 exhibit vertical linearity issues. Channels 3 and 4 appear to be okay. It won't be a big deal to swap U400 with my 2445 to see if it fixes the problem. I do notice the trace is not as sharp as I'd like/expect (my 2445 shows a bit crisper and I think they use the same CRT), so not sure if there's internal damage or just needs adjusting. There's a slight dent in the bottom rear of the cabinet and I think I can tell a small distortion in one of the frame pieces that meets up with the rear panel but the direction of that distortion seems odd (e.g. it would indicate an outward force rather than an inward force such as a strike to the rear panel, etc.). No telling just what/how it was hit. Overall, the frame seems twisted about 1/8" or more (doesn't sit flat on the bench). It's probably not too difficult to straighten this out. As for the broken fan, the way the fan "shroud" is designed, it presses against the center of the fan motor and it wouldn't really take very much pressure (with the cabinet removed) to break the small plastic supports that center the motor in the frame. I plan to CA glue those back in and see how that holds up. If not, a replacement appears pretty easy to find. The fan is slightly noisier than the other style fan in my 2445. I wish it had the same style fan in this model. BTW, as I mentioned, this appears to be a 1994 manufacture date and a 65k serial number. Was this pretty close to the end of production? I've looked for serial number correlation but not finding too much on that for this model. TekWiki just says "Produced from 1984 to ?". Thanks, Barry - N4BUQ ----- Original Message ----- From: "machine guy machineguy59@... [TekScopes]" <TekScopes@...>#yiv1268545622 -- #yiv1268545622ygrp-mkp {border:1px solid #d8d8d8;font-family:Arial;margin:10px 0;padding:0 10px;}#yiv1268545622 #yiv1268545622ygrp-mkp hr {border:1px solid #d8d8d8;}#yiv1268545622 #yiv1268545622ygrp-mkp #yiv1268545622hd {color:#628c2a;font-size:85%;font-weight:700;line-height:122%;margin:10px 0;}#yiv1268545622 #yiv1268545622ygrp-mkp #yiv1268545622ads {margin-bottom:10px;}#yiv1268545622 #yiv1268545622ygrp-mkp .yiv1268545622ad {padding:0 0;}#yiv1268545622 #yiv1268545622ygrp-mkp .yiv1268545622ad p {margin:0;}#yiv1268545622 #yiv1268545622ygrp-mkp .yiv1268545622ad a {color:#0000ff;text-decoration:none;}#yiv1268545622 #yiv1268545622ygrp-sponsor #yiv1268545622ygrp-lc {font-family:Arial;}#yiv1268545622 #yiv1268545622ygrp-sponsor #yiv1268545622ygrp-lc #yiv1268545622hd {margin:10px 0px;font-weight:700;font-size:78%;line-height:122%;}#yiv1268545622 #yiv1268545622ygrp-sponsor #yiv1268545622ygrp-lc .yiv1268545622ad {margin-bottom:10px;padding:0 0;}#yiv1268545622 #yiv1268545622actions {font-family:Verdana;font-size:11px;padding:10px 0;}#yiv1268545622 #yiv1268545622activity {background-color:#e0ecee;float:left;font-family:Verdana;font-size:10px;padding:10px;}#yiv1268545622 #yiv1268545622activity span {font-weight:700;}#yiv1268545622 #yiv1268545622activity span:first-child {text-transform:uppercase;}#yiv1268545622 #yiv1268545622activity span a {color:#5085b6;text-decoration:none;}#yiv1268545622 #yiv1268545622activity span span {color:#ff7900;}#yiv1268545622 #yiv1268545622activity span .yiv1268545622underline {text-decoration:underline;}#yiv1268545622 .yiv1268545622attach {clear:both;display:table;font-family:Arial;font-size:12px;padding:10px 0;width:400px;}#yiv1268545622 .yiv1268545622attach div a {text-decoration:none;}#yiv1268545622 .yiv1268545622attach img {border:none;padding-right:5px;}#yiv1268545622 .yiv1268545622attach label {display:block;margin-bottom:5px;}#yiv1268545622 .yiv1268545622attach label a {text-decoration:none;}#yiv1268545622 blockquote {margin:0 0 0 4px;}#yiv1268545622 .yiv1268545622bold {font-family:Arial;font-size:13px;font-weight:700;}#yiv1268545622 .yiv1268545622bold a {text-decoration:none;}#yiv1268545622 dd.yiv1268545622last p a {font-family:Verdana;font-weight:700;}#yiv1268545622 dd.yiv1268545622last p span {margin-right:10px;font-family:Verdana;font-weight:700;}#yiv1268545622 dd.yiv1268545622last p span.yiv1268545622yshortcuts {margin-right:0;}#yiv1268545622 div.yiv1268545622attach-table div div a {text-decoration:none;}#yiv1268545622 div.yiv1268545622attach-table {width:400px;}#yiv1268545622 div.yiv1268545622file-title a, #yiv1268545622 div.yiv1268545622file-title a:active, #yiv1268545622 div.yiv1268545622file-title a:hover, #yiv1268545622 div.yiv1268545622file-title a:visited {text-decoration:none;}#yiv1268545622 div.yiv1268545622photo-title a, #yiv1268545622 div.yiv1268545622photo-title a:active, #yiv1268545622 div.yiv1268545622photo-title a:hover, #yiv1268545622 div.yiv1268545622photo-title a:visited {text-decoration:none;}#yiv1268545622 div#yiv1268545622ygrp-mlmsg #yiv1268545622ygrp-msg p a span.yiv1268545622yshortcuts {font-family:Verdana;font-size:10px;font-weight:normal;}#yiv1268545622 .yiv1268545622green {color:#628c2a;}#yiv1268545622 .yiv1268545622MsoNormal {margin:0 0 0 0;}#yiv1268545622 o {font-size:0;}#yiv1268545622 #yiv1268545622photos div {float:left;width:72px;}#yiv1268545622 #yiv1268545622photos div div {border:1px solid #666666;height:62px;overflow:hidden;width:62px;}#yiv1268545622 #yiv1268545622photos div label {color:#666666;font-size:10px;overflow:hidden;text-align:center;white-space:nowrap;width:64px;}#yiv1268545622 #yiv1268545622reco-category {font-size:77%;}#yiv1268545622 #yiv1268545622reco-desc {font-size:77%;}#yiv1268545622 .yiv1268545622replbq {margin:4px;}#yiv1268545622 #yiv1268545622ygrp-actbar div a:first-child {margin-right:2px;padding-right:5px;}#yiv1268545622 #yiv1268545622ygrp-mlmsg {font-size:13px;font-family:Arial, helvetica, clean, sans-serif;}#yiv1268545622 #yiv1268545622ygrp-mlmsg table {font-size:inherit;font:100%;}#yiv1268545622 #yiv1268545622ygrp-mlmsg select, #yiv1268545622 input, #yiv1268545622 textarea {font:99% Arial, Helvetica, clean, sans-serif;}#yiv1268545622 #yiv1268545622ygrp-mlmsg pre, #yiv1268545622 code {font:115% monospace;}#yiv1268545622 #yiv1268545622ygrp-mlmsg * {line-height:1.22em;}#yiv1268545622 #yiv1268545622ygrp-mlmsg #yiv1268545622logo {padding-bottom:10px;}#yiv1268545622 #yiv1268545622ygrp-msg p a {font-family:Verdana;}#yiv1268545622 #yiv1268545622ygrp-msg p#yiv1268545622attach-count span {color:#1E66AE;font-weight:700;}#yiv1268545622 #yiv1268545622ygrp-reco #yiv1268545622reco-head {color:#ff7900;font-weight:700;}#yiv1268545622 #yiv1268545622ygrp-reco {margin-bottom:20px;padding:0px;}#yiv1268545622 #yiv1268545622ygrp-sponsor #yiv1268545622ov li a {font-size:130%;text-decoration:none;}#yiv1268545622 #yiv1268545622ygrp-sponsor #yiv1268545622ov li {font-size:77%;list-style-type:square;padding:6px 0;}#yiv1268545622 #yiv1268545622ygrp-sponsor #yiv1268545622ov ul {margin:0;padding:0 0 0 8px;}#yiv1268545622 #yiv1268545622ygrp-text {font-family:Georgia;}#yiv1268545622 #yiv1268545622ygrp-text p {margin:0 0 1em 0;}#yiv1268545622 #yiv1268545622ygrp-text tt {font-size:120%;}#yiv1268545622 #yiv1268545622ygrp-vital ul li:last-child {border-right:none !important;}#yiv1268545622 |
Re: 2220 POST displays a large X on the screen
Oh well, so much for that. Another possibility is to duplicate 2230 ROMs and try in a 2220.
As I mentioned, the purpose of the big X could be to align the storage display with the CRT. The lines should cross at the center, and the slopes should be to confirm linearity and set them to cross the CRT graticule lines at the intersections all the way to the corners. These adjustments are in the vector generator, as I recall. There would be offset and gain adjustment for vertical and horizontal. Maybe it's a factory setting not normally accessible to the user. If the X isn't for that, then there may be something in the diagnostic menu to draw some kind of pattern for this purpose. There should be something in the service manual about this alignment process. Ed |
Re: 2220 POST displays a large X on the screen
The ROMS in the 2220 are 1603976-04 and 160-3975-04, different to the 2230 and a quick comparison between the boards doesn't appear to have much difference apart from the obvious lack of cursor control and missing switches.
I don't think the X is a cal feature, at least I've never seen it and don't see what use it would be. Very strange. ---In TekScopes@..., <edbreya@...> wrote : I don't recall exactly how the 2220 was differentiated from the 2230 for product positioning, but I do know the guts are mostly the same. I think the 2220 was reduced in performance by limiting the BW to 60 MHz (with parts on the main board) vs 100 for the 2230, and maybe lower sample rate and record length. I think the brain identified the model by jumpers (0 ohm resistor jumpers) on the digital board, although they may instead have different ROMs. If anyone with these models can compare, it would be evident. I also thought the 2220 had on-screen readout, but fewer features within the operating menus. If the ROMs are the same between models, this could be a case where a 2220 can be mostly converted to a 2230, except for any major items missing. I saw the picture of the big X - it certainly is big, and appears to be generated by the digital display system, like a calibration signal to align the storage mode to the CRT face. Maybe there's a jumper or DIP switch set to the wrong position, or loose, forcing a cal mode. Another possibility is that it's a fault indicator showing X-out - that there's a major problem such as the brain being confused about what model it should be, so it just sits there. Perhaps there's an intermittent fault in the ID process, or maybe someone hacked it part way and got it all confused. Obviously, the brain must be working in order to present the X on screen. I'd recommend studying the manuals and circuits and boards and ROMs and jumpers and DIP switches to see if anything looks different from normal. Especially check in the adjustments section, to see what's done for aligning the store/nonstore registration. I'd guess it's a jumper or DIP switch setting, rather than a diagnostic menu item, to make the X, and pots on the vector board to tweak it to the CRT. Ed |
Introducing myself and my Tek 465 Project
Hello all, new to the group and happy to be here.
I just wrote a long intro about me and my new Tek 465 into the Yahoo groups 'new post' textbox, but it seems to have vanished ... so I'll briefly recap (pardon the pun). I am a web and database programmer during the day but have dabbled in electronics since I was 10. I'm in my 30s now and getting back into modding my audio tube amps. I've always been building kits and fixing through parts swapping in TV power supply boards, digital Arduino projects, RC planes etc. But I've never really learned the math or theory/design side of the analog world. As I'm looking to build some personal audio gear (active crossovers, maybe some tube mono-blocks) I'm teaching myself more theory. And I'm a very visual person and like learning through gear so thought I'd get an analog o-scope. Well I stumbed upon an as-is Tek 465 on Craiglist that I picked up Monday near Chicago. I'm blown away by the detail in the service manual I found online and the serviceability, and am starting with the power supply diag in the manual to try to get things running. I don't have high expectations for accuracy - I know a detailed calibration is a profession in itself. But getting even rough readings would be a big help for me as I learn. Current condition of the scope is a single dot in the center with no sweep. Fan and lights work, knobs in good shape. Very complete scope but the sweep IC gets super hot. I figured I'd have to swap a lot of caps and transistors and maybe some regulators, but hoping I don't have to source the custom ICs (but found a source so I guess there's hope). So far I've confirmed a solid +55v rail, +15v rail and a drooping +5v and no -8v with my DMM. The +5v droop goes away if I isolate the horizontal amp board by pulling the connector. I posted on the EEVBlog forum and got a response saying to check the ripple as well; seems I need another (working) scope? Any hope that a cheap USB scope (opposite end of spectrum from this beast!) would suffice? Thanks in advance for your knowledge; I'll ask more detailed questions and post pics as I get into things. - Neil |
Re: 2220 POST displays a large X on the screen
I don't recall exactly how the 2220 was differentiated from the 2230 for product positioning, but I do know the guts are mostly the same. I think the 2220 was reduced in performance by limiting the BW to 60 MHz (with parts on the main board) vs 100 for the 2230, and maybe lower sample rate and record length. I think the brain identified the model by jumpers (0 ohm resistor jumpers) on the digital board, although they may instead have different ROMs. If anyone with these models can compare, it would be evident. I also thought the 2220 had on-screen readout, but fewer features within the operating menus.
If the ROMs are the same between models, this could be a case where a 2220 can be mostly converted to a 2230, except for any major items missing. I saw the picture of the big X - it certainly is big, and appears to be generated by the digital display system, like a calibration signal to align the storage mode to the CRT face. Maybe there's a jumper or DIP switch set to the wrong position, or loose, forcing a cal mode. Another possibility is that it's a fault indicator showing X-out - that there's a major problem such as the brain being confused about what model it should be, so it just sits there. Perhaps there's an intermittent fault in the ID process, or maybe someone hacked it part way and got it all confused. Obviously, the brain must be working in order to present the X on screen. I'd recommend studying the manuals and circuits and boards and ROMs and jumpers and DIP switches to see if anything looks different from normal. Especially check in the adjustments section, to see what's done for aligning the store/nonstore registration. I'd guess it's a jumper or DIP switch setting, rather than a diagnostic menu item, to make the X, and pots on the vector board to tweak it to the CRT. Ed |
New to Group, looking for help with Tek 465 Repair
Hi all, just introducing myself and my current project to the group. I am a web programmer during the day but have always been interested in electronics. Always have been a 'part swapper' level of a electronic repair tech but know a bit of theory and how to read schematics etc. Also am a fan of audio tube amplifiers and looking to build my own circuits for home audio projects (preamp, active crossovers). Also need to fix a TV power supply for my brother at the moment; and have a tape deck I need to diagnose deeper than I can do with a DMM. So now I'm wanting to learn more; taking a class and just now getting into scopes and the deeper math side of things. But sticking with low Mhz needs and analog for starters.
Picked up a Tek 465 off Craiglist on Monday - it has a single non-sweeping dot in the center and a toasty hot sweep IC. I've been following the service manual diag steps and seeing low voltage on the +5v rail until the horizontal amp board connector is removed, then it stabilizes to 5v per my DMM. No -8v to be seen but solid +55 and +15. I have been told on the EEBlog forum to check for ripple; seems I need to buy another (working) scope to get that confirmed which I think I'll do. Any chance the less expensive USB O-Scopes can do ripple? I see the custom Tek Sweep IC can be had from some specialty web sites so I may order one of those based on mine running too hot to touch. Otherwise I'm planning to test and swap all caps and transistors as needed. I also have a quality Tek o-probe on the way and a good Weller soldering station and pretty well equipped bench. Thanks in advance; I'm hoping to ask more detailed questions as I have them but am starting testing from the power supply end. - Neil |
Re: New file uploaded to TekScopes
Besides, the links don't work. In the "New file..." messages the folder name is either truncated (too long?), or changed afterwards in the Files section (is that possible?). Very strange, the text -fixed is removed from the folder name and replaced by a space.
toggle quoted message
Show quoted text
Albert ---In TekScopes@..., <malcolm.r.hunter@...> wrote : Hi Darren, If you upload photos to the Photos section, people can view them in their web browser rather than having to download them first. Malcolm On 30 March 2016 at 14:42, <TekScopes@... mailto:TekScopes@...> wrote:
> > > > Hello, > > This email message is a notification to let you know that > a file has been uploaded to the Files area of the TekScopes > group. > > File : /2247A-PSU-BurnMark-Near-A18R2254 /Before-1of1.jpg > Uploaded by : darren_ting_hc <darren_ting_hc@... mailto:darren_ting_hc@...> > Description : Before Recap and burn mark fix > > You can access this file at the URL: > > > > To learn more about file sharing for your group, please visit: > > > > Regards, > > darren_ting_hc <darren_ting_hc@... mailto:darren_ting_hc@...> |
Re: Question Regarding Power Supply Ripple Measurement - 2465B
Barry, it's what you suggest, it will be ripple after double rectification. But the repetition frequency is still 60 Hz; only in case of ideal symmetry the frequency appears as 120 Hz.
Hence set the trigger mode at LINE. If you trigger on the signal then the trigger circuitry will normally recognize 120 positive (at + slope) level crossings per second. Depending on sweep speed and hold-off time you have two alternating waveforms superimposed, or "at random" one of these two (maybe stable, maybe now and then jumping to the other waveform). Albert ---In TekScopes@..., <n4buq@...> wrote : In the 2400 series manuals, the table that defines the measurements for voltage and ripple state (for ripple measurement): p-p Ripple at Two Times Line Frequency How is this supposed to be done? With a 60Hz line frequency (U.S.), is this supposed to be performed with the scope measuring the ripple triggered at 120Hz or am I misinterpreting what's being stated? Is it simply assumed that any ripple will present itself at twice the line frequency (full-wave rectification) and the scope being used to measure this will simply trigger at that frequency (assumedly using a fairly slow sweep speed to better see the waveform)? Thanks, Barry - N4BUQ |
Re: Question Regarding Power Supply Ripple Measurement - 2465B
They are referring to ripple produced by the 120 Hz variation from the
toggle quoted message
Show quoted text
full wave rectifier at the input. Set the measuring oscilloscope to line triggering; that is what it is there for. On Wed, 30 Mar 2016 11:31:48 -0400 (EDT), you wrote:
In the 2400 series manuals, the table that defines the measurements for voltage and ripple state (for ripple measurement): |
Re: 2220 POST displays a large X on the screen
Data on the 2220 and 2221 is pretty slim; I have never see schematics
toggle quoted message
Show quoted text
or the service manual for them. Similarities with the 2230 which make it the closest match and 2235 series in general include: - Production dates coinciding with the 2230. - 20pF input capacitance and 2mV/div sensitivity. - Maximum 50ns/div unmagnified sweep. - 20 MS/s digitizer and 100ns peak detect. Will the 2230 display an X for diagnostic purposes? I do not remember. On Wed, 30 Mar 2016 11:37:44 +0000 (UTC), you wrote:
HiSorry, I thought I had posted an image. I've added one to the albums: |
Question Regarding Power Supply Ripple Measurement - 2465B
In the 2400 series manuals, the table that defines the measurements for voltage and ripple state (for ripple measurement):
p-p Ripple at Two Times Line Frequency How is this supposed to be done? With a 60Hz line frequency (U.S.), is this supposed to be performed with the scope measuring the ripple triggered at 120Hz or am I misinterpreting what's being stated? Is it simply assumed that any ripple will present itself at twice the line frequency (full-wave rectification) and the scope being used to measure this will simply trigger at that frequency (assumedly using a fairly slow sweep speed to better see the waveform)? Thanks, Barry - N4BUQ |
Tek 7K plugins available - spring garage cleaning
Hello to all again!
Jim Reese from Dayton, Ohio?back in action. I am going through the piles of stuff and have many 7K plugins available at low prices. All have had basic function testing in a 7704A unless otherwise noted. Add shipping from zip 45344. 7A16A, $207A18, $307A26, $407B50A, $307B51A, $357B92A, $50 7B92A non-working (short on power supply line, probably bad capacitor), $25 I have many others available that I have not gotten too yet that I will post later.? Let me know of anything you are looking for. Regards, Jim |
to navigate to use esc to dismiss