Re: N channel fet 151-1121-00 -- V10206
Craig Sawyers
Not precisely sure what I'm looking at. Is that the said FET on a 7CT1N in your 'scope? Because if it
toggle quoted message
Show quoted text
is, I would say that was an ex-FET. It has gone to join the choir eternal. -----Original Message----- |
Re: Tek 465 "B" Sweep
Chuck Harris
I don't remember it that way. The entire 475 can be calibrated
toggle quoted message
Show quoted text
sitting out of its case on the bench. The entire 465 family can be calibrated sitting out of its case on the bench. The adjustment holes in the case are an indication that some of the more sensitive circuits are not very stable, and require occasional touch-up by the operator. -Chuck Harris ¸é±ð²Ô¨¦±ð wrote: check the side of the case there may be "holes" for some of the sensitive |
Re: Tek 465 "B" Sweep
On Fri, Nov 2, 2018 at 05:50 PM, ¸é±ð²Ô¨¦±ð wrote:
These holes provide access to e.g. vertical position, attenuator balance and DC gain settings. They are there as a convenience to allow some adjustments "in the field" without opening the case. These adjustments are not critical or sensitive in that they can be done without access to special equipment; just the eye and the inbuilt calibrator. Adjustments with the case removed are not part of the Adjustments Procedure as per the Service Manual. Raymond |
Re: Tek 465 "B" Sweep
check the side of the case there may be "holes" for some of the sensitive adjustments. there are on a 475 and the case needs to be installed for those adjustments.
toggle quoted message
Show quoted text
¸é±ð²Ô¨¦±ð On 2018-11-02 5:34 a.m., Chuck Harris wrote:
Definitely not normal! If tektronix needed you to put a special |
Re: Tek 465 "B" Sweep
Chuck Harris
That's a very good point. When I was using CFL illumination
toggle quoted message
Show quoted text
in a desk lamp, at my bench, I used to see a lot of excess noise signals in hi gain, hi-Z circuitry. In addition, carrying a smart phone everywhere I go has shown up some spurious signals as the phone calls home, sends responses to wifi beacons, and transfers data... Not to mention the very strong magnet in the speaker erasing magstripes on keycards. Perhaps a little spelunking with a loop of wire and the scope in a high gain setting? -Chuck Harris Raymond Domp Frank wrote: On Fri, Nov 2, 2018 at 11:59 AM, Jeff Urban wrote:No, it is not normal although the area is sensitive to external signals, like hum; you're covering and grounding a high-impedance part of the vertical input amp, directly after the attenuator block with attenuation set at 200 mV/div. |
Re: Tek 465 "B" Sweep
On Fri, Nov 2, 2018 at 11:59 AM, Jeff Urban wrote:
No, it is not normal although the area is sensitive to external signals, like hum; you're covering and grounding a high-impedance part of the vertical input amp, directly after the attenuator block with attenuation set at 200 mV/div. Are you sure there's no external source producing something like 50 kHz, if I'm reading the horizontal settings correctly? Raymond |
Re: Tek 465 "B" Sweep
Chuck Harris
Definitely not normal! If tektronix needed you to put a special
toggle quoted message
Show quoted text
shield there, there would be a shield with a part number specified in the list of things required for calibration. What you have is probably a point grounded by a screw that has corroded, a bypass cap that is sub-optimal, or possibly an active part that has gotten wild.. look for a replacement part that has too much gain to be stable in the circuit, too long leads, needs a ferrite bead... Also look for socket pins on active parts that are marginal. -Chuck Harris Jeff Urban wrote: AHA ! |
Re: Tek 465 "B" Sweep
AHA !
/g/TekScopes/photo/77251/1?p=Name,,,20,1,0,0 That is a piece of blank PC board on the left there grounded by a clip lead. The trace is smaller than it looks in the picture, I probably should've turned the intensity down but it's pretty clean from here. If that is normal for having the cover off, how do you work on the vertical front end in these things ? How do you even begin to calibrate it ? Enquiring and hopeless minds want to know. It is a day. |
Re: Tek 465 "B" Sweep
Well well well, Q4566 open. A depletion mode FET, and I couldn't find one. I look at the print, I think about those 2N7000s I got. I look at the print again. I think about the 2N7000s again.
All they are doing is turning the thing on and off, not even sure why they used an FET, a bipolar would have worked. (I almost bet it would if I just stuck it in there) So I think about this and go ahead and throw in the 2N7000 and it works. I suppose I should get a 2SK30 or something for it, but it DOES work. If nothing else it proves the fault, that part of the troubleshooting is over. /g/TekScopes/photo/77251/1?p=Name,,,20,1,0,0 As you can see, as Murphy's law and a few other things are strictly enforced around here, it has come up with another problem. The wrong FET, the most I can figure it doing is throwing off the calibration of the B sweep. So what for now. The oscillation in the trace appears to be about 50KHz. It varies randomly at frequency and by the AC line at amplitude. (I thin) Its amplitude is lowest when the V/div. is set to any range with a "5". The problem wasn't there initially but it popped up before I did anything to it so wasn't me. I cooked the unit watching Clint Walker shooting bad guys (one hour) and when I came back the problem was the same. It came on after a few minutes of running the first time and has stuck around. As such I doubt it is filter caps. I think it is a ground somewhere going to the front end. The only time the oscillation goes away is either when the channel is disabled by the switch to the right of the input section, or in the 5mV/div setting with the AD/DC switch in the ground position. It is affected by the variable gain. (uncal.) So it appears I have my work cut out for me now. The main thing I hope for it that it doesn't fix itself. I don't like that, and I don't like just soldering things up and hoping for the best. One place I worked at insisted, if you soldered and it fixed it they wanted to know WHICH connection it was. I now agree with that. So anyway, any ideas on this new problem I would appreciate. In a few minutes I am going to call it a day, it is 6:45 AM and I like to work, but I also like to eat and sleep. If I happen to fix it in five minutes you'll be the second to know. All in all though, this has increased my suspicion of old FETs, I've had a few bad in audio equipment as well, they almost always open up, I haven't seen a shorted one yet. So if you're doing that oldie but goodie stuff keep an eye on those FETs. |
Re: 425 Mil vertical module needed....seeking
Bert Haskins......
Calling Bert Haskins...calling Bert Haskins ....CALLING BERT HASKINS Confusion reigns....I'm only new here....I don't know whether this reply goes to you or Fabio. I'm a bit of a 'headless chook' at present, Bert... Are you in Brazil? Bert Can you contact me on goldmort@... with a price for the vertical? Yes I am interested. -- Jack |
2465BCT CRT date code ?
I have recently acquired a spare CRT for my 2465BCT and am trying to understand the markings as follows:
P/N?: 154-0850-00 S/N?: 149403 Other?: 41-91 Is this possibly a date code, perhaps corresponding to week 41 of year 1991? If not, is there another meaning? No other markings on the label. Thanks, Rick, K8EZB |
Re: Tek 2465B Bezel Removal
" Or, you could just slide the blue filter up, and out at the bottom
toggle quoted message
Show quoted text
and remove it. The gorp might be just there. " Thank you Chuck. I did not know you could do that. It was rather easy and it cleaned up nicely!! The blue filter re-installed just as easy as it came out. Dave On Thu, Nov 1, 2018 at 1:38 PM Chuck Harris <cfharris@...> wrote:
Or, you could just slide the blue filter up, and out at the bottom |
Re: 425 Mil vertical module needed....seeking
Ok Fabio...finally I see it ...well some of it....Is that Brazilian site related to Bert? There's no contact detail for him.
toggle quoted message
Show quoted text
Can you assist there?...The vertical on the Brazilian site is "identical" in appearance to the 425 but I recall, and must re-check his letter, Dinos advising they are not related CRO's. I'm a bit like Alice in Wonderland at the moment. My Regards -----Original Message-----
From: Fabio Trevisan Sent: Thursday, November 1, 2018 10:49 PM To: [email protected] Subject: Re: [TekScopes] 425 Mil vertical module needed....seeking Hi Jack, Bert Haskins... he's replied to your topic, see msg reference below. /g/TekScopes/message/152031 Most of his message is quote from previous messages... And his reply is a two liner, at the end. Rgrds, Fabio -- Jack |
Re: 425 Mil vertical module needed....seeking
Hi again for your interesting information. Amongst it you wrote "But Bert already chimed in that he has a NOS 465M module... so I think you already have a good lead"
toggle quoted message
Show quoted text
I didn't receive that....was it related to my inquiry> and where will I find it. My regards -----Original Message-----
From: Fabio Trevisan Sent: Thursday, November 1, 2018 10:59 AM To: [email protected] Subject: Re: [TekScopes] 425 Mil vertical module needed....seeking Hi Jack, The google translations are perfect, both in Portuguese and Spanish. Only the expression "mileage" in Portuguese and in Spanish are not usually used for anything else than vehicles. For the strict meaning of "usage", we just say "uso" (which is the exact translation of usage). In view of the answer from Dinos (thanks Dinos for correcting my assumptions), the contact with this seller in Brazil may not yield anything fruitful, as this guy seem to be a lonely wolf (he doesn't seem to have anything else than this 455 parts... and since the parts differ from the AN/USM 425 that you have, those parts probably aren't interchangeable with yours. But Bert already chimed in that he has a NOS 465M module... so I think you already have a good lead. About Tektronix, vs anything else... I also have mixed feelings about it... In my first job in the 80s, most of the scopes were Philips and I loved them... and we had one Tek 465 which I was never comfortable to work with. I found back then that arrangement of the time base knobs clumsy and awkward to use, and that the lack of an ALT delayed time base was really a big weakness... and I still think it was... I really don't know how Tek got away with selling so well the 46x line with that useless "Mix" mode, when they had earlier designs which already had the "ALT" time base. To some extent, some of that awkwardness in using the Tek scopes have disappeared as I became used to its "User Interface", and I learnt also to appreciate that many of those features that I took for granted on the Philips scopes were actually invented and appeared first on some Tek scope many years before (which I didn't know back then). Today I own a Tek 7623A with 3 plugins, and a jap scope, a 60MHz triple trace Kenwood scope... And I like both very much. The Kenwood is very powerful for its relatively simple design (there are no double sided PCBs!). Wish you luck with your 465M Rgrds, Fabio On Wed, Oct 31, 2018 at 06:22 PM, Jack wrote: -- Jack |
Re: Tek 547 A Sweep display during retrace
I found one more fast part. If you happen to have an MPSH55 in your junk box, or you can get one more easily than a BC556 or KSA992 (Netty Electronics has a few dozen left), it will do great with 100pF to 220pF added to C370.
With 100pF added across C370 MPSH55 55ns 35ns 80V Motorola, old With 220pF added across C370 MPSH55 50ns 25ns 80V Motorola, old I have not found a modern part that does not require C370 augmentation, and I'm giving up the search. |
Re: Tek 2465B Bezel Removal
Chuck Harris
Or, you could just slide the blue filter up, and out at the bottom
toggle quoted message
Show quoted text
and remove it. The gorp might be just there. -Chuck Harris David Kuhn wrote: Craig, |
Re: 577 retrace problem
Hi Fred,
toggle quoted message
Show quoted text
Make a note to yourself - Craig is much smarter than me for several reasons: He knew the answer right away; He was able to describe it in one paragraph; He had a better way to test this with a cascode combination. Albert is also smarter - He knew to look this up in the archives which probably took a minute instead of reinventing the wheel like I did. Before you go to all the trouble of trying a flip flop there is an easier way which Craig first suggested 6 years ago. I just tried it and it does work. Switch to the FAST step rate, and then vary the NUMBER OF STEPS until you see an even number of traces on the screen. No loops! When you do this the curve tracer is applying a step as the collector sinewave voltage is rising and right at the top of the sine wave it applies the next step and the beam displays the next trace as the sine wave is falling. So each step lasts half as long. If you look very carefully where the loops were, every other trace will have a slight bump up (where the top half of the loop would have been) and the steps in between will have a slight dip down. With an odd NUMBER OF STEPS, the traces will have loops and they will be flickering which is the worst combination I probably did read Craig's explanation 6 years ago and I'm sure it sailed right over my head at the time. I know a lot more now than I did then. I still had to figure this out my way so I would learn for once and for all why I get those loops sometimes and other times they are not there. The first time I really thought about these loops was when I was testing my Vacuum Tube Curve Tracer Adapter on my 575, 576, and 577. At the time I knew certain combinations of STEP RATE and NUMBER OF STEPS would make them go away, but I never put two and two together like Craig did. Dennis Tillman W7PF -----Original Message----- -- Dennis Tillman W7PF TekScopes Moderator |
Re: Tek 2465B Bezel Removal
On Thu, 1 Nov 2018 12:14:50 -0400, you wrote:
Okay, I think the analog would be more useful possibly in trouble-shooting,Any digital scope is a sampling system, so it looks at a waveform, does an average during the sample period, and then ignores the signal to the next sampling period. Pretty much as long as you sample fast enough, you'll get most of everything. However, when you get to the upper limit of the scope's frequency response, the analog scope fails gracefully, at least giving you an idea of something there. The digital scope has a relatively sharp cutoff to keep aliasing down, and has a different kind of response. Where the digital scope is superior, IMHO, has to do with the storage and computation functions. A 7000 series scope can get pretty close to a lot of the things a digital scope can do, within limits,of course. I've got analog scopes that go to 1 Ghz, digital ones that go to 500 Mhz (and the main analog scope is a 7904, so 500 Mhz). I have sampling units that can go to about 10 Ghz, in case I need to look there (haven't so far). Can't answer there, you may want to consider a private sale to someone on the list rather than ebay. Best is selling in person so that both people get an idea of what they're getting. I really don't have a lot of selling experience and with me having aIf the business bought it, then likely yes, if you bought it, then it's a private sale, I'd think. Definitely not an expert here. oh, and Hamfests are a good possibility depending on where you are. Anywho, back to work. Thanks for the response.Sure. Definitely Harvey
|