Re: 465B Oscilloscope No -8V


Rob
 

¿ªÔÆÌåÓý

Given the initial problem you described, I believe you¡¯re looking for capacitors that check a consistent relatively low ohm with a standard ohm meter (i.e. do not charge to show open or build towards open). Open or bad caps (high esr wise) would cause you to have high ripple but not lack of DC.

?

Unless I missed something along the way. Hopefully helpful

Rob


--- In TekScopes@..., "rwdplz1" wrote:
>
> OK, I checked C4419, it is a 5000uF 35V capacitor. ESR is 0.08, which should be about right. However, after desoldering it from the board (which is a real pain with the two large mounting legs), and measuring capacitance, one of my meters couldn't measure it, and the other read 0.185 nano Farads. Both of these meters read the capacitance of other small capacitors I have on hand perfectly, so the meters are fine. ESR of the other large capacitors is about the same, but I'd rather not desolder them from the board if I don't have to, to prevent damage to the fragile traces.
>
> --- In TekScopes@..., "Tom Miller" wrote:
> >
> > OK, good. I would think the next thing to check is C-4419, a 5000 uF can capacitor.
> >
> > Regards,
> > Tom
> >
> > ----- Original Message -----
> > From: rwdplz1
> > To: TekScopes@...
> > Sent: Tuesday, January 15, 2013 2:40 AM
> > Subject: [TekScopes] Re: 465B Oscilloscope No -8V
> >
> >
> >
> > The resistance between the -8V test point and ground (either a test point labeled ground on the board or the scope case) is 49.7 ohms.
> >
> > I replaced one tantalum, C4203, 33uF, with no effect.
> >
> > --- In TekScopes@..., "pdxareaid" wrote:
> > >
> > > you might want to check for shorted tantalum caps in -8 circuits.
> > > see:
> > >
> > > phil
> > >
> > >
> > >
> > > --- In TekScopes@..., "rwdplz1" wrote:
> > > >
> > > > Hello, I am trying to fix my Tektronix 465B Oscilloscope. I have the service manual and have been using it to try and figure out what's wrong. When turned on, there is nothing on the screen, even when pressing and holding beam finder, and the power light doesn't come on. The only control that seems to work is scale illumination, which makes the screen glow orange when turned up.
> > > >
> > > > I checked the voltages at the test points on the main board A4 as described on page 4-24, and found all of them to be within spec: +55, +15, +5, and +110, EXCEPT for the -8, which measures 0.2mV. The power supply wiring diagram at the back of the manual shows two blue wires from the transformer, going through a fuse F4501 (desoldered and removed from board, it's good), then through a rectifier bridge CR4411, which is also good.
> > > >
> > > > So where do I go from here? What voltage should I be seeing at the two blue wires, because I'm starting to suspect that part of the transformer? Thank you
> > > >
> > >
> >
>

Join [email protected] to automatically receive all group messages.