Keyboard Shortcuts
Likes
Search
Cleaning Tek475A Attenuator Switches
¿ªÔÆÌåÓýThis post describes a procedure that I used to remove and clean the Ch1 and Ch 2 input assemblies for a Tek 475A scope.? My scope was showing intermittent and erratic traces even after switching ranges and tapping on the range switch.? What I discovered, and repaired, made the vertical display like new again. I followed the procedure shown in the 475A Service Manual, pages 4-24, 4-25.? This manual is available on line.? The dis-assembly involves removing the Vertical Preamp Board and the Vertical Mode ¨C Gain Switch Board and the associated Ch 1 and Ch 2 attenuator assemblies.? It sounds formidable but it can be done.? Take pictures with a digital camera or draw pictures in your own style to help you retrace your steps for re-assembly. The input attenuators are mechanically operated by the range knobs (Volts/Div) which have cams that operate very small ¡°switches¡± on the Gain Switch circuit board.? The attenuators are the little rectangular plastic assemblies which plug into the Vertical Mode ¨C Gain Switch Boards (Ch 1 and Ch 2).? At first I tried using contact cleaner on the Attenuator module pins, sockets and board-mounted switch contacts (see photos) but that had almost no effect on the quality of the vertical displays. Looking at the bottom of one of the Attenuator modules I noticed that some of the solder joints looked quite bad and had signs of corrosion.? The Service Manual calls for replacement of the Attenuator Modules.? They are available on line, but I decided to look further.? This is where you will want to have a somewhat steady hand, a good work light and an eye loupe or a head band magnifier.? I use one made by Donegan Optical, the OptiVisor model with 5 diopter optics.? I bought one at a local model / hobby store.? They are also available on line, new or used. Before removing any of the Attenuator modules I marked them by Channel, and by location.? Then I removed the modules one at a time, and did a visual inspection of all of the solder joints on the bottom of the attenuator modules. I even opened up some of the modules being careful not to bump any of the variable elements (resistors or capacitors).? Where a solder joint looked bad I applied a small dot of liquid flux and carefully applied the soldering iron tip and a very slight amount of new solder if needed.? I did not use any flux remover (or alcohol). When I reassembled everything I was very happy
with the results.? In
summary, I believe that the main cause of the erratic displays was
the vibration sensitivity of the attenuator module cold solder
joints.? You could pay for
new / used attenuator modules but I found that it is possible to
correct the problem first hand.
Photo 1: Vertical Preamp Board and Vertical
Mode¨CGain Switch Board.? Channel 1 attenuator cover removed to
show attenuators.
Photo 2: Vertical Mode ¨C Gain Switch Board with all attenuator modules removed to show switch contacts. (Mark the modules before removing them!)
Photo 3: Vertical Mode ¨C Gain
Switch Board close up showing board-mounted switch contacts.
Carefully clean the little ¡°forks¡± with good contact cleaner or
alcohol and a small strip of paper.
Photo 4: Attaching Vertical Mode ¨C Gain Switch Board to front panel.? Each input attenuator module attaches to the front panel with two nuts.? I used an Allen wrench to hold the nut and a cut up wooden stick to turn the nut onto the screw.
Photo 5: Use a nut driver to tighten the nuts to the front panel screws. |