Mark Wendt
 

Unforch, I don't have a HV spark generator. Though I could think of a few nefarious uses I could put it too... ;-)

Mark

On 01/01/2013 05:28 PM, Tom Miller wrote:


Can you put a circle around where you think it is open?
The strap just appears to be a support for the glass rod.
Do you have any kind of a HV spark generator that you could connect across the heater pins so as to show off the gap?
I am thinking open heater element.
Tom

----- Original Message -----
*From:* wendtmk <mailto:mark.wendt@...>
*To:* TekScopes@... <mailto:TekScopes@...>
*Sent:* Tuesday, January 01, 2013 5:11 PM
*Subject:* [TekScopes] Re: Tek 154-0812-00 Storage CRT

Before I go and do all that, would you guys look at the photos I
just put up of the pin and strap, and see if you think it's even
possible? To me, the gap between the pin and the strap is a little
bit too large. The two photos are in the folder called SC503 CRT.

Mark