¿ªÔÆÌåÓý

ctrl + shift + ? for shortcuts
© 2025 Groups.io

Re: 7000 series on screen graphics question 7704A versus 7633


Jim Popwell Jr
 

HI CHRIS,
I would do the astig and refocus¡­.astig¡­and refocus for sharpest beam definition. use the calibrator and adjust for best looking sq wave¡­then kill the vert deflection and the horz deflection¡­turn the intensity down and adjust both again for the sharpest spot¡­.increase the intensity and readjust both again¡­then go back to normal sweep¡­¡­.

NOW DO THE CAL procedure on the display bd.

there are a couple of adjustments on the bd that will straighten out the "jumpy double" characters.
won't take much time but will make a big difference.
if i remember right these characters were created by different currents at different times¡­ so a little too much or too little current misshaped the characters.
i often find the right controls, then adjust them by eye for best display, by finding the sweet spot of the adjustment. i.e. find the ends of the range where the adjustment makes a total mess with the adjustment one way then going back thru the best look¡­until you get a mess again¡­¡­...then splitting the difference.
jim

On Jul 13, 2012, at 3:30 AM, Chris Wilson wrote:



13/07/2012 11:27

Is the astigmatism adjustment internal on these David? If I look
closely it almost appears to be trying to show two numerals / letters
at the same time, possibly the figure 1 (one)? As it's otherwise 100%
and new old stock I'd quite like to get this facet of its operation
fixed. Thanks.

--
Best Regards,
Chris Wilson.
mailto: chris@...



------------------------------------

Yahoo! Groups Links


Join [email protected] to automatically receive all group messages.