Displayed signal becomes compressed when position is moved to the left SOLVED!


 

Found it! Magnifier registration R749 had the center lead broken. And was not suppling the 30V to U760.

--- In TekScopes@..., "david_s5y" <david_s5@...> wrote:

Hi,

On a tek 2230 in non-store mode, the displayed signal becomes compressed when position is moved to the left.

So if I have a 5K square wave displayed at .1 Ms the first 1/2 of the wave will be 60% the width if I move it to the right. It also looks like brighter.

+==+==+==+==+==+==+==+==+==+==+
---___---___--- <moved to right>

+--- Short and brighter.
V
==___---___---

What needs adjusting