Hi Barry,
toggle quoted message
Show quoted text
By removing transistors and seeing +0.6V we confirmed power transistor is not responsible for the fault (at least not the -33V output issue). Voltage across R9041 is 0.672V, this turns on the power transistor. However, in this state (output exceeding target voltage), Q7042 should be off and voltage across R9041 should be zero. This means there is current through Q7042. It could either be faulty Q7042 or faulty Q8051. Transistor Q8051 has a good base drive, it should be able to short across Q7042 and turn it off. Is C-E terminals of Q7042 correct, i.e. not swapped? If you short across C-E of Q8051 (with or without Q8051 in place is OK, you can do it at the socket) does output go to ~ +0.6V? If it does Q8051 is bad. If it doesn't Q7042 is bad or C-E swapped. Is Q7042 base voltage of -29.3V wrt to its emitter or ground? Vbe of -29.3V is unlikely and would damage the transistor. Ozan On Sun, Dec 22, 2024 at 12:20 PM, Barry Breaux wrote:
|