Re: Tek 7704a Incorrect Readout Position


 

BTW, the reason I asked about that is if you swap them and the readout is centered horizontally, then the problem is likely in your vertical amplifier board. If it exhibits the same offset horizontally, then the problem is likely in the readout board. It's easy enough to try.

Thanks,
Barry - N4BUQ

If you swap the cables coming from the readout to the vertical and horizontal
boards, what does that show?

Thanks,
Barry

On Nov 17, 2024, at 12:46 PM, talynprior via groups.io
<talynprior@...> wrote:

?top is now at center, bottom is below the bottom of the screen but still there.
?when I use the beam finder button I can see that all is displaying well, and
?readout value changes when i change volts per division, it"s just in the wrong
?location






Join [email protected] to automatically receive all group messages.