Re: Wanted 067-0681-01 Tunnel Diode Pulser by new member


 

On Mon, May 21, 2018 at 05:44 pm, Siggi wrote:


Hey Lol,

Your nice old 485s come equipped with graticule markers specifically for
rise time measurement. See this video <> for
how to use them to get repeatable and comparable measurements between
scopes and signal sources.

Siggi
On Mon, May 21, 2018 at 19:46 lop pol via Groups.Io <the_infinite_penguin=
[email protected]> wrote:

On Mon, May 21, 2018 at 03:57 pm, Raymond Domp Frank wrote:


Hi lop pol,
Thanks for doing the tests. I don't understand the outcome, however.
Just looking at your images:
In the TD (tunneldiode pulser) image, assuming a TD rt of 100ps (<=125ps
specified), I see a 600ps rise time (rt), indicating about 590 MHz bw of
the
485 against a spec. of 350 MHz. Even for that 'scope, it seems very high
bw to
me. 30% faster than spec'ed would be "normal", but 70%? Even with a
theoretical rt of 0ps for the TD, the 485's bw would be about 580 MHz,
about
65% faster than specified for this model! It *could* be caused by a slow
1
ns/div. time base speed.

A second strange thing is the suggested rt of the PG506 (pg). I see a rt
of
1.6 ns. Assuming the specified rt for the 485 @ <= 1 ns (350 MHz), the
pg's rt
would be 1.24 ns, against a spec of < 1 ns. Not so good but possible,
since
the later (towards the top) part of the edge shows a speed reduction
(shoulder
rounding) that is not typical for this instrument. Also, the amplitude
on the
screen is a bit higher than used for the 485's specification, which could
result in it performing a bit slower. Also, using the pg's variable
attenuator
would reduce its rt. BTW, the "slower shoulder" could be a side effect of
using the attenuator pot to reduce the output signal!

So, two things I don't quite understand:
- Exceptionally fast (high bw) 485
- Exceptionally slow PG506.

Unless I'm mistaken in my observations, I don't see how to reconcile
them.

Raymond
I have added another 485 for our assessment.
067-0681-01 /g/TekScopes/photo/50828/2?p=Name,,,20,1,0,0
and pg506 fast rise
/g/TekScopes/photo/50828/3?p=Name,,,20,1,0,0



Hey thanks for that. Makes it a ton easier.