¿ªÔÆÌåÓý

HP 5086-7906 YTO


 

¿ªÔÆÌåÓý

I forgot to send the mixer information.

?

Blue wire, PIN SW: band0= +9.81v and band 1=-9.56V

Green wire, BIAS: +3.9V

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 30 ao?t 2024 09:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

having received no feedback, I continue with a brief update on my attempts to bring my HP8595E back to its former glory 3
Summary and state of the art:
it all started with the failure of the main oscillator YTO (A3A7 on the block diagram); This component was replaced with the same model, purchased "as is" and installed after it was found to work and with good output across the entire range
Once installed, the analyzer returned to sweep (sometimes it gave the error LO UNLVL) but above all it revealed the problem that
no signal present in band 0 (9kHz -2.7GHz) is seen by the analyzer, (so not even the calibration signal which however turned out to be in specification and going to look for it in the analyzer I find it up to the low pass output - A3A3 on the block diagram)
I have not gone further yet
On the contrary, the signals in band 1 (2.7 - 6.5GHz) are visible and correct in frequency but attenuated
That of the attenuation of the signals in the 5- 6.5 GHz region is a fairly well-known problem with this analyzer and the diagnoses on the web have associated it with a misalignment of the yttrium spheres due to the countless thermal cycles that the SYTF undergoes over time
I opened my SYTF (particularly deaf starting from 6GHz) and I was able to see the misalignment in particular of a sphere (in the photo barely visible, under the gold mesh)
I await the arrival of a 2.5mm socket wrench after which I will try to realign
Any suggestions on how to deal with the problem of "blindness" in band 0 are welcome
adri

?

?

?


 

¿ªÔÆÌåÓý

Yves,
thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF

I must say that the operation done under the microscope is quite simple
The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done
Adri
?
?
?
?

----- Messaggio originale -----
Da: Yves Tardif via groups.io <yves_tardif@...>
Rispondi a: <[email protected]>
Inviato: 31/08/2024 03:07:53
Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

I forgot to send the mixer information.

?

Blue wire, PIN SW: band0= +9.81v and band 1=-9.56V

Green wire, BIAS: +3.9V

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 30 ao?t 2024 09:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

having received no feedback, I continue with a brief update on my attempts to bring my HP8595E back to its former glory 3
Summary and state of the art:
it all started with the failure of the main oscillator YTO (A3A7 on the block diagram); This component was replaced with the same model, purchased "as is" and installed after it was found to work and with good output across the entire range
Once installed, the analyzer returned to sweep (sometimes it gave the error LO UNLVL) but above all it revealed the problem that
no signal present in band 0 (9kHz -2.7GHz) is seen by the analyzer, (so not even the calibration signal which however turned out to be in specification and going to look for it in the analyzer I find it up to the low pass output - A3A3 on the block diagram)
I have not gone further yet
On the contrary, the signals in band 1 (2.7 - 6.5GHz) are visible and correct in frequency but attenuated
That of the attenuation of the signals in the 5- 6.5 GHz region is a fairly well-known problem with this analyzer and the diagnoses on the web have associated it with a misalignment of the yttrium spheres due to the countless thermal cycles that the SYTF undergoes over time
I opened my SYTF (particularly deaf starting from 6GHz) and I was able to see the misalignment in particular of a sphere (in the photo barely visible, under the gold mesh)
I await the arrival of a 2.5mm socket wrench after which I will try to realign
Any suggestions on how to deal with the problem of "blindness" in band 0 are welcome
adri

?

?

?


 

¿ªÔÆÌåÓý

Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!
After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm
The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal

Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input

On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification

The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V
?
This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works
Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output

In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days

Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable
?
Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?
Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)
?
This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly
thanks again and best regards
Adri
?
?
?

----- Messaggio originale -----
Da: Yves Tardif via groups.io <yves_tardif@...>
Rispondi a: <[email protected]>
Inviato: 06/09/2024 23:21:52
Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2
?
As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri
?

----- Messaggio originale -----
Da: Yves Tardif via groups.io <yves_tardif@...>
Rispondi a: <[email protected]>
Inviato: 08/09/2024 01:35:49
Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

Hi,

?

Good news.

?

I just reassembled my 8595E, and after some checks, everything seems normal.

?

I put a comparative image between our two Spectrum.

I also did a sweep in 100 kHz to better see the harmonics in high frequencies.

So, I pasted two images from 0Hz-3GHz and 3GHz to 6GHz.

Input attenuator at 0dB, Reference level at -20dBm.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

Adri, Here for information, is the DANL and flatness test that I did, at the end of April 2024, after the YTF change and recalibration.

When you have completed everything, you should have similar results.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

If another sample is useful I can coerce my HP8595E into providing one. Just let me know.

Wilko


 

¿ªÔÆÌåÓý

Hi Adri,

?

A friend of mine had a faulty 8595E, the problem comes from the option card 130.

But I took the opportunity to look at the second converter, so here are some details on this module.

?

Regards

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

Hello Yves,

This second converter assembly is also used in the 8563E and also ESA
E4407B spectrum analyzers. In fact I think it is common for all
6.5/13.5/26.5GHz versions.

Mine had failed in the E4407B and it looks like it is related to an
incorrectly manufactured transformer inside the assembly.

I attached the Agilent Service Note related to this issue.

Part number affected is 5086-7958.

Thanks for the nice PDF with all explanations.

Regards,
Razvan

On 13/09/2024 07:51, Yves Tardif via groups.io wrote:
Hi Adri,

A friend of mine had a faulty 8595E, the problem comes from the option
card 130.

But I took the opportunity to look at the second converter, so here are
some details on this module.

Regards

Yves


 

¿ªÔÆÌåÓý

Yves, thanks for the info about the 2nd converter? After having installed the one bought on the bay?I checked?the original one?on the bench and it?seems to work normally
?
I'm starting to suspect the blindness on the LOW BAND was due to a bad connection I did when the repair was suspended years ago ¡­ ouch !
?
Meanwhile I'm investigating?a certain deafness on the HIGH BAND (from 2.4 to 11 dB)
?
I measured the insertion loss of the SYTF I had realigned and found figures?from 5.2??to? 8.5 dB on the range? 3 to 6 GHz
?
These figures?seem not too bad?and?the AMP CAL routine should be able to recover them ...unless there is some hardware problem in the generation of the tuning current (I performed the insertion loss measurements by manually tuning? with an external precision power supply)
Will see ...
Adri
?
?
?
?

----- Messaggio originale -----
Da: Yves Tardif via groups.io <yves_tardif@...>
Rispondi a: <[email protected]>
Inviato: 13/09/2024 07:51:28
Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

Hi Adri,

?

A friend of mine had a faulty 8595E, the problem comes from the option card 130.

But I took the opportunity to look at the second converter, so here are some details on this module.

?

Regards

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

Hi Razvan,
thanks also for the service note. I have an E4407B, I had not
noticed that it was the same second converter.

Yves

-----Message d'origine-----
De?: [email protected]
<[email protected]> De la part de Razvan Popescu via
groups.io
·¡²Ô±¹´Ç²â¨¦?: 13 septembre 2024 03:49
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

Hello Yves,

This second converter assembly is also used in the 8563E and also ESA E4407B
spectrum analyzers. In fact I think it is common for all 6.5/13.5/26.5GHz
versions.

Mine had failed in the E4407B and it looks like it is related to an
incorrectly manufactured transformer inside the assembly.

I attached the Agilent Service Note related to this issue.

Part number affected is 5086-7958.

Thanks for the nice PDF with all explanations.

Regards,
Razvan

On 13/09/2024 07:51, Yves Tardif via groups.io wrote:
Hi Adri,

A friend of mine had a faulty 8595E, the problem comes from the option
card 130.

But I took the opportunity to look at the second converter, so here
are some details on this module.

Regards

Yves


 

¿ªÔÆÌåÓý

?

Hi Aldri,

?

?

For your second converter, when you do debugging it's easy to make a mistake, there is so much manipulation.

You have to keep the positive side, this part is set and you have a second spare converter.

?

For the YTF, it's not easy, centering the spheres is not enough, don't forget that it's a bandpass filter. So you have to align the filter for a certain reference frequency, but which one, we could assume that it's in the center, but not necessarily, it could be at the beginning for example (lowest frequency). The filter will deform according to the frequency, it's about finding a happy medium. The YTF calibration will find the coefficients of a polynomial trying to have the best for all frequencies. The level differences will be compensated during the frequency flatness calibration. However, don't forget that if you have too much loss to compensate, the DANL test could become out of spec.

?

The fact that you use a precise external power supply is not a problem for the alignment of your YTF. The important parameter of this power supply is its stability.

?

I had sent you a picture of my 8595E, here it is again, it represents a cumulative test of the frequencies:

yellow curve -> test at 3GHz

cyan curve -> test at 5GHz

magenta curve -> test at 6.5GHz

We can see that on the left, there could have been an improvement by re-adjusting one of the spheres, but nothing is certain. The insertion loss varied from 2 to 3 dB between the minimum and the maximum of these 3 frequencies. You have similar results, I think your YTF has good results...nothing is perfect :)

?

You say that you have problems in the high band, what are the frequencies in play. from 2.9GHz to 6.5GHz? (highband).

For the second converter, the path loss for the highband is -1.8dB. After this point two bands go through the same path, so if the YTF is correct, either there is a problem with the YTF control source, the first converter or the second converter (highband section). You should be able to easily find where the excess loss is.

?

You are at the end of your troubles, don't give up, you will find it.

?

?

--------------------?

Here is the verification of this second converter after reassembly.

The flatness is good, only 0.03dB for 5MHz width at output IF 321.4MHz (lowBand).

Multiplier x6 for 3600 MHz, adjusted for best match (lower loss at 321.4MHz), very sensitive adjustments.

Path for 321.4MHz highBand was verified too.

?

One sweep with 50 MHz span, another with 2950MHz span¡­to see the stop band.

The loss for IF highband include loss of filter and RF switch, around -1.8dB.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 13 septembre 2024 19:14
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves, thanks for the info about the 2nd converter? After having installed the one bought on the bay?I checked?the original one?on the bench and it?seems to work normally

?

I'm starting to suspect the blindness on the LOW BAND was due to a bad connection I did when the repair was suspended years ago ¡­ ouch !

?

Meanwhile I'm investigating?a certain deafness on the HIGH BAND (from 2.4 to 11 dB)

?

I measured the insertion loss of the SYTF I had realigned and found figures?from 5.2??to? 8.5 dB on the range? 3 to 6 GHz

?

These figures?seem not too bad?and?the AMP CAL routine should be able to recover them ...unless there is some hardware problem in the generation of the tuning current (I performed the insertion loss measurements by manually tuning? with an external precision power supply)

Will see ...

Adri

?

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 13/09/2024 07:51:28

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

A friend of mine had a faulty 8595E, the problem comes from the option card 130.

But I took the opportunity to look at the second converter, so here are some details on this module.

?

Regards

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

Hi Adri,

?

I had borrowed a faulty 8595E from a friend to study the second converter.

?

The problem with this 8595E was that the AMP CAL was not working and displayed an error message pointing to option 130 (narrow resolution bandwidth).

Looking at the calibration constants, I noticed the calibration values ??were very far apart for some filters.

After checking, I had no choice but to realign the filters.

I followed the procedures and the problem was solved.

This Spectrum had no hardware defects and option 130 had no problem.

So, I did the full calibration, and voila, this 8595E now has a second life.

?

Have you managed to get your 8595E working properly again?

?

Before filters adjustments at left, and after at right.

?

Yves

?

?

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 13 septembre 2024 19:14
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves, thanks for the info about the 2nd converter? After having installed the one bought on the bay?I checked?the original one?on the bench and it?seems to work normally

?

I'm starting to suspect the blindness on the LOW BAND was due to a bad connection I did when the repair was suspended years ago ¡­ ouch !

?

Meanwhile I'm investigating?a certain deafness on the HIGH BAND (from 2.4 to 11 dB)

?

I measured the insertion loss of the SYTF I had realigned and found figures?from 5.2??to? 8.5 dB on the range? 3 to 6 GHz

?

These figures?seem not too bad?and?the AMP CAL routine should be able to recover them ...unless there is some hardware problem in the generation of the tuning current (I performed the insertion loss measurements by manually tuning? with an external precision power supply)

Will see ...

Adri

?

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 13/09/2024 07:51:28

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

A friend of mine had a faulty 8595E, the problem comes from the option card 130.

But I took the opportunity to look at the second converter, so here are some details on this module.

?

Regards

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??:
[email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??:
[email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??:
[email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

hello Yves, great repair that of the opt 130
In my case the opt 130 has always worked well and I have never needed to study it in detail
Did you do the realignment by adjusting the trimmers on the board itself?
The adventure with my 8595E continues
To sum up, after my last email I focused on a more accurate and uniform alignment of the SYTF. With the help of the microscope, a strong light source, a precision power supply and a fairly reliable frequency source I obtained an insertion loss as per the attached excel file
After reassembling the RF front-end everything worked fine again, including the MXR CAL routine that?used to?give the?error? "insufficient harmonics"
When I moved from the "messy" version (pic)? reinserting everything into its space (which is very crowded, also due to the existence, in my case, of the tracking gen)?and connected?the SA input to CAL OUT I observed a type of intermittent noise, in the form of spikes of intensity about -60dBm that appear and disappear in the area of ??the highest harmonics (I already checked the?CAL OUT signal; it?is clean and fine)
Now I think I will have to re-disassemble the front-end (ouch) and try to understand from where does this noise originate :-(?
If you are interested I?can send you a short video of the problem (don't think I?can post videos here)
Thanks for asking, Yves
Adri
?

----- Messaggio originale -----
Da: Yves Tardif via groups.io <yves_tardif@...>
Rispondi a: <[email protected]>
Inviato: 18/09/2024 23:47:21
Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

Hi Adri,

?

I had borrowed a faulty 8595E from a friend to study the second converter.

?

The problem with this 8595E was that the AMP CAL was not working and displayed an error message pointing to option 130 (narrow resolution bandwidth).

Looking at the calibration constants, I noticed the calibration values ??were very far apart for some filters.

After checking, I had no choice but to realign the filters.

I followed the procedures and the problem was solved.

This Spectrum had no hardware defects and option 130 had no problem.

So, I did the full calibration, and voila, this 8595E now has a second life.

?

Have you managed to get your 8595E working properly again?

?

Before filters adjustments at left, and after at right.

?

Yves

?

?

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 13 septembre 2024 19:14
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves, thanks for the info about the 2nd converter? After having installed the one bought on the bay?I checked?the original one?on the bench and it?seems to work normally

?

I'm starting to suspect the blindness on the LOW BAND was due to a bad connection I did when the repair was suspended years ago ¡­ ouch !

?

Meanwhile I'm investigating?a certain deafness on the HIGH BAND (from 2.4 to 11 dB)

?

I measured the insertion loss of the SYTF I had realigned and found figures?from 5.2??to? 8.5 dB on the range? 3 to 6 GHz

?

These figures?seem not too bad?and?the AMP CAL routine should be able to recover them ...unless there is some hardware problem in the generation of the tuning current (I performed the insertion loss measurements by manually tuning? with an external precision power supply)

Will see ...

Adri

?

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 13/09/2024 07:51:28

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

A friend of mine had a faulty 8595E, the problem comes from the option card 130.

But I took the opportunity to look at the second converter, so here are some details on this module.

?

Regards

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 10 septembre 2024 16:26
??:
[email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves,
I received the 2nd converter and, after testing it on the bench (if you are interested I will give you the two test configurations) I installed it and voil¨¤, the machine is back to working like in the old days
It passes the amplitude and frequency calibrations and the confidence test without any problem
As it happened before the death of the YTO the analyzer still has a certain "deafness" after 5.5GHz
By virtue of the alignment of the spheres that I did a few weeks ago The defect is milder than it was then and I think it is due to a sub-optimal loss of insertion of the SYTF A3A2

?

As proof of the "deafness", while all the calibration routines, including the CAL YTF routine are performed without problems, the CAL MXR is not completed and gives the message CAL: HARMONICS INSUFFICIENT In fact I attach a screen of the 300MHz calibration signal from which it would seem that the very last harmonics are missing
Thank you in advance if can you send the analogous screen of your analyzer
In a few days I will receive a more reliable 6GHz RF generator than the current one and I will use it to try to perfect the alignment
Yves, I think that now you can reassemble and close your analyzer :-) Many thanks for all the measurements that you sent me and that were useful to trace the problem!
Adri

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 08/09/2024 01:35:49

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri

?

Here are the measurements connected to the first 2 converters as a reference.

?

For the second converter I get a voltage of -1.678V instead of -3.5V as written in the service manual, I would say that we have similar voltages for this particular point.

It seems far from the nominal point, but there may be an error in the manual.

?

You are close to the goal, and this module may be repairable if there is no micro circuit, I have never opened a second converter in this series, but it is a possibility.

?

When you open your second converter, send me pictures, I'm curious.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 6 septembre 2024 19:30
??:
[email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

dear Yves, I didn't realize that you had made the measurements expressly to help me! I owe you then a double (even triple) thanks! Especially considering the fact that the front end of this instrument is really very dense, complicated and delicate!

After my last email, in addition to doing the (essentially mechanical) intervention on the SYTF, I continued to check the input signal through the various sections by alternately injecting 300MHz @-20dBm and 5GHz @0 dBm

The signals were found to be in specification up to the A3A9 bandpass (where the two paths join to form the intermediate frequency of 321.4MHz). At this point I could see the 5GHz signal (converted), present at that point at -47dBm (the specification says -43/-56dBm).
Instead, no trace of the 300MHz input signal


Investigatine?the second converter A3A4 in more depth
I checked?the signals at its input


On J1 there is 3.9214GHz at -46dBm? --->OK
On J3 there is 321.4MHz @ -37.7dBm? --->OK
On J4 I did not check 600MHz @-1dBm but the fact that the converter works in band 1 tells me that the oscillator is present and is in specification


The power supply is quite OK (5V are actually 4.75; 10V are 9.98 and the DC that controls the internal switch that should be +15V? (band 0) and -3.5V?(band 1) is actually +14.60V and -1.35V

?

This last value is rather far from the nominal and would have worried me BUT band 1 is exactly the one that works

Conclusion: it would seem that the culprit is the second converter and in particular (probably) the internal switch that has the sole task of letting the signal? (which is already at 321.4MHz and does not need to be converted)?pass from its input to the output


In the next few days I will test the converter that meanwhile?has been?disassembled?; In the meantime I found a cheap one on the bay that should arrive in about ten days


Yves, I have a question: on the block diagram A3A4 is represented as a block that has 5 ports of which one (J5) would seem to be the connector that carries the power and the DC that controls the switch and that connects to A7 Analog Interface through the W35 cable

?

Actually,?this connector on the real component, does not have a name
Instead,?the name J5?, stamped on the case , indicates an SMB connector that has the writing LO TEST and in my analyzer?it is the one that?conveys the 600MHz LO signal?

Now,?on the block diagram the connector that conveys the 600MHz signal has the name J4 (in my analyzer, the J4 connector was disconnected)

?

This confusion of names and roles made me doubt that years ago, when I had disassembled this section I had not reassembled it correctly

thanks again and best regards

Adri

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 06/09/2024 23:21:52

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

I had sent you the measurements of the RF input-output mixer signals verification, I had disassembled the High Frequency section of my 8595E to be able to provide you with this information.

If you have no more questions relating to this section, I will close my 8595E and do a flatness check to see if everything is correct.

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 31 ao?t 2024 12:23
??:
[email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves,

thank you very much for all the suggestions and information with which I will continue my investigation
In the meantime I received the 2.5mm socket wrench and after?a little mod to adapt it?for the purpose (photo), I proceeded to align the spheres of the SYTF


I must say that the operation done under the microscope is quite simple

The nuts must be loosened (not too much) and with a small screwdriver that acts as a delicate lever, ?the supports of?the?spheres can be moved so that the coupling ring completely shadows them (i.e. the sphere is exactly in the center)
Removing the gold mesh was not necessary
The insertion loss has improved enormously:
at 5 GHz it is -6dBm at 6GHz it is -10dB (before the signal starting from this frequency was no longer visible) at 6.5 GHz it is only -3dB
I will install it when I have solved the problem of "blindness" in band 0 since the calibration constants that the instrument has in memory (in practice the tuning current data) are relative to the original SYTF and the "new" SYTW requires a? SYTF CAL is done

Adri

?

?


 

¿ªÔÆÌåÓý

Hi Adri,

?

Yes, I am curious, you can send me a video on the intermittent noise problem, by going through OneDrive for example, or Dropbox.

It could be that one of the PLLs is the cause.

?

For the 8595E on which I was trying to repair, the problem was not on the 130 option as the error message seemed to indicate, but rather an incorrect alignment on the basic filters.

The AMP CAL stopped because it was unable to find the two -3dB points of a filter.

?

Yes, I realigned all the filters following the procedure in the service manual.

There are 4 crystal filter sections and 4 LC filters.

Simple procedure but a bit long, and you must not make any mistakes, otherwise you have to start from the beginning.

I did the job slowly; everything went well and the original problem was solved.

I recalibrated the instrument and took the opportunity to improve my software for flatness alignment, now after the test, all data is entered automatically...which speeds up the work.

?

In my opinion, someone opened this 8595E to try to make a repair, the second converter had been opened, 2 screws were missing and 2 others were reset to 50%.

I had to realign the second converter after closing it correctly.

?

For the filter section, it is clear that one touched the adjustments, because half were well aligned, and for the others, the adjustments were very far from the optimal value.

?

In total, 8 screws were missing on this instrument.

Probably an interrupted repair¡­

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 19 septembre 2024 09:07
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves, great repair that of the opt 130
In my case the opt 130 has always worked well and I have never needed to study it in detail
Did you do the realignment by adjusting the trimmers on the board itself?
The adventure with my 8595E continues
To sum up, after my last email I focused on a more accurate and uniform alignment of the SYTF. With the help of the microscope, a strong light source, a precision power supply and a fairly reliable frequency source I obtained an insertion loss as per the attached excel file
After reassembling the RF front-end everything worked fine again, including the MXR CAL routine that?used to?give the?error? "insufficient harmonics"
When I moved from the "messy" version (pic)? reinserting everything into its space (which is very crowded, also due to the existence, in my case, of the tracking gen)?and connected?the SA input to CAL OUT I observed a type of intermittent noise, in the form of spikes of intensity about -60dBm that appear and disappear in the area of ??the highest harmonics (I already checked the?CAL OUT signal; it?is clean and fine)
Now I think I will have to re-disassemble the front-end (ouch) and try to understand from where does this noise originate :-(?

If you are interested I?can send you a short video of the problem (don't think I?can post videos here)
Thanks for asking, Yves
Adri

?


 

¿ªÔÆÌåÓý

Yves, your mention about the SW related to the flatness test is interesting. Is it software written by you or does it exist on the market?
As for my 8595E I suspected that the intermittent spikes could originate from the somewhat chaotic way in which I had restored the path of the various? front-end cables
After? extracting the entire front-end section?I reinserted it, paying attention to establishing cable paths as orderly as possible
Once the work was finished suspence:??the spikes were no longer there? ! ; I re-ran then the calibrations routines (AMP; FREQ; SYTF and MXR) and and all ended without problems
?fingers crossed
For fun I tested with the Tracking Gen some coaxial cables (with relative N connectors/adapters) of various qualities and lengths. I attach the images ( vertical scale is 1dBm/div, and?testing is done?sweeping the entire frequency range of the tracking gen i.e. 9kHz - 2.9GHz)
?
Adri
?
?
?
?

----- Messaggio originale -----
Da: Yves Tardif via groups.io <yves_tardif@...>
Rispondi a: <[email protected]>
Inviato: 20/09/2024 01:15:13
Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

Hi Adri,

?

Yes, I am curious, you can send me a video on the intermittent noise problem, by going through OneDrive for example, or Dropbox.

It could be that one of the PLLs is the cause.

?

For the 8595E on which I was trying to repair, the problem was not on the 130 option as the error message seemed to indicate, but rather an incorrect alignment on the basic filters.

The AMP CAL stopped because it was unable to find the two -3dB points of a filter.

?

Yes, I realigned all the filters following the procedure in the service manual.

There are 4 crystal filter sections and 4 LC filters.

Simple procedure but a bit long, and you must not make any mistakes, otherwise you have to start from the beginning.

I did the job slowly; everything went well and the original problem was solved.

I recalibrated the instrument and took the opportunity to improve my software for flatness alignment, now after the test, all data is entered automatically...which speeds up the work.

?

In my opinion, someone opened this 8595E to try to make a repair, the second converter had been opened, 2 screws were missing and 2 others were reset to 50%.

I had to realign the second converter after closing it correctly.

?

For the filter section, it is clear that one touched the adjustments, because half were well aligned, and for the others, the adjustments were very far from the optimal value.

?

In total, 8 screws were missing on this instrument.

Probably an interrupted repair¡­

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 19 septembre 2024 09:07
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves, great repair that of the opt 130
In my case the opt 130 has always worked well and I have never needed to study it in detail
Did you do the realignment by adjusting the trimmers on the board itself?
The adventure with my 8595E continues
To sum up, after my last email I focused on a more accurate and uniform alignment of the SYTF. With the help of the microscope, a strong light source, a precision power supply and a fairly reliable frequency source I obtained an insertion loss as per the attached excel file
After reassembling the RF front-end everything worked fine again, including the MXR CAL routine that?used to?give the?error? "insufficient harmonics"
When I moved from the "messy" version (pic)? reinserting everything into its space (which is very crowded, also due to the existence, in my case, of the tracking gen)?and connected?the SA input to CAL OUT I observed a type of intermittent noise, in the form of spikes of intensity about -60dBm that appear and disappear in the area of ??the highest harmonics (I already checked the?CAL OUT signal; it?is clean and fine)
Now I think I will have to re-disassemble the front-end (ouch) and try to understand from where does this noise originate :-(?

If you are interested I?can send you a short video of the problem (don't think I?can post videos here)
Thanks for asking, Yves
Adri

?


 

¿ªÔÆÌåÓý

Hi Adri,

?

Glad to hear you finally managed to get that 8595E back up and running.

?

As for the software, I wrote it in vb.net. It only does basic tests:

?

Flatness test and save calibration constants table on demand (12MHz-6.5GHz).

Low frequency flatness test only, no calibration constants (9kHz - 10MHz).

DANL resBW=1kHz test

DANL resBW=30Hz test for option 130

?

I will add more tests soon.

This utility saves me time, because everything can be done manually.

?

Yves

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 21 septembre 2024 12:15
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

Yves, your mention about the SW related to the flatness test is interesting. Is it software written by you or does it exist on the market?
As for my 8595E I suspected that the intermittent spikes could originate from the somewhat chaotic way in which I had restored the path of the various? front-end cables
After? extracting the entire front-end section?I reinserted it, paying attention to establishing cable paths as orderly as possible
Once the work was finished suspence:??the spikes were no longer there? ! ; I re-ran then the calibrations routines (AMP; FREQ; SYTF and MXR) and and all ended without problems

?fingers crossed
For fun I tested with the Tracking Gen some coaxial cables (with relative N connectors/adapters) of various qualities and lengths. I attach the images ( vertical scale is 1dBm/div, and?testing is done?sweeping the entire frequency range of the tracking gen i.e. 9kHz - 2.9GHz)
?

Adri

?

?

?

?

----- Messaggio originale -----

Da: Yves Tardif via groups.io <yves_tardif@...>

Rispondi a: <[email protected]>

Inviato: 20/09/2024 01:15:13

Oggetto: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO


Hi Adri,

?

Yes, I am curious, you can send me a video on the intermittent noise problem, by going through OneDrive for example, or Dropbox.

It could be that one of the PLLs is the cause.

?

For the 8595E on which I was trying to repair, the problem was not on the 130 option as the error message seemed to indicate, but rather an incorrect alignment on the basic filters.

The AMP CAL stopped because it was unable to find the two -3dB points of a filter.

?

Yes, I realigned all the filters following the procedure in the service manual.

There are 4 crystal filter sections and 4 LC filters.

Simple procedure but a bit long, and you must not make any mistakes, otherwise you have to start from the beginning.

I did the job slowly; everything went well and the original problem was solved.

I recalibrated the instrument and took the opportunity to improve my software for flatness alignment, now after the test, all data is entered automatically...which speeds up the work.

?

In my opinion, someone opened this 8595E to try to make a repair, the second converter had been opened, 2 screws were missing and 2 others were reset to 50%.

I had to realign the second converter after closing it correctly.

?

For the filter section, it is clear that one touched the adjustments, because half were well aligned, and for the others, the adjustments were very far from the optimal value.

?

In total, 8 screws were missing on this instrument.

Probably an interrupted repair¡­

?

Yves

?

De?: [email protected] <[email protected]> De la part de alfa beta
·¡²Ô±¹´Ç²â¨¦?: 19 septembre 2024 09:07
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO

?

hello Yves, great repair that of the opt 130
In my case the opt 130 has always worked well and I have never needed to study it in detail
Did you do the realignment by adjusting the trimmers on the board itself?
The adventure with my 8595E continues
To sum up, after my last email I focused on a more accurate and uniform alignment of the SYTF. With the help of the microscope, a strong light source, a precision power supply and a fairly reliable frequency source I obtained an insertion loss as per the attached excel file
After reassembling the RF front-end everything worked fine again, including the MXR CAL routine that?used to?give the?error? "insufficient harmonics"
When I moved from the "messy" version (pic)? reinserting everything into its space (which is very crowded, also due to the existence, in my case, of the tracking gen)?and connected?the SA input to CAL OUT I observed a type of intermittent noise, in the form of spikes of intensity about -60dBm that appear and disappear in the area of ??the highest harmonics (I already checked the?CAL OUT signal; it?is clean and fine)
Now I think I will have to re-disassemble the front-end (ouch) and try to understand from where does this noise originate :-(?

If you are interested I?can send you a short video of the problem (don't think I?can post videos here)
Thanks for asking, Yves
Adri

?