Troubleshooting a 8591e with opt 130


 

Hello all, I recently bought a 8591e spectrum analyzer. This analyzer was originally produced with option 130 narrow bandwidth if.
The opt 130 card came in a separate shielded bag when I bought the analyzer because it would give errors with the option installed according to previous owner.
?
I did some initial tests and now wondering what the best way to proceed is.?
?
I started withouts the opt 130 installed. Basic functions seem ok but there is noise around ref level -9 / -10db over the full bandwidth. Not there at other ref levels and gone with cal>correct off. I'm assuming there's crappy cal data.
So did several freq and ampt self call attempts. At first I got a cal span sens fail. Later self call succeeded but conf check failed on Res Bw Shape fail 30000 & 10000.
?
Next I installed the opt 130. Now there is zero signal on the analyzer, it's a flatline at -140dBm.?
?
Should I try to find the error in the opt 130 and do a new self call with it installed or should I focus on the Res Bw shape fail first?
?
Unit looks fine overall so would be a great addition to my hobby shack when I can get it functional again.
?
Thanks!


 

I continued my testing and debugging with some success.
The opt 130 is showing signal now, there are supposed to be 2 jumpers on that board and only 1 was present, the other was probably used for A14W1 by the previous owner, when he removed the option. With that jumper in place i'm no longer just looking at the nosiefloor.
With the option installed the cal fails on NBW 200Hz Notch failed.?
Doubt that i'lll be able to fix that without schematics or adjustment options on the opt 130.
?
Interestingly the unit does show a -22dB cal signal with the factory default cal at 350MHz in full span and 300MHz at 100MHz span.?
During ampt cal the signal is way above the screen during 3db BW steps, i'm guessing something is going wrong there.
After running freq and amp cal the 300MHz is back at -5dB instead of -20dB, even though the cal failed and correction off doesn't change it either.?
?
Starting to wonder how feasible this repair will be.
?


 

Continuing on the cal signal I checked the attenuator steps.
After factor default and with correction off I found the following:
At 300MHz:
Att? |? Level
10? -22
20? -12
30? -21
40? -26
50? -19
60? -21
?
at 1500 MHz
Att? |? Level
10? -28
20? -32
30? -27
40? -32
50? -30
60? -30
?
I'm guessing my 20db attenuator is toast but I would have expected to have seen the issue coming back at the higher attenuation levels. Could this behavior be caused by something else??
?


 

¿ªÔÆÌåÓý

?

Hi,

?

It would be easier to check the attenuator between the RF input of the 8591E and the output of the attenuator.

This way no other circuits would be involved. You will have a better idea of ??the state of the attenuator.

?

Yves

?

De?: [email protected] <[email protected]> De la part de jansen.marco90@...
·¡²Ô±¹´Ç²â¨¦?: 3 novembre 2024 14:09
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

Continuing on the cal signal I checked the attenuator steps.

After factor default and with correction off I found the following:

At 300MHz:

Att? |? Level

10? -22

20? -12

30? -21

40? -26

50? -19

60? -21

?

at 1500 MHz

Att? |? Level

10? -28

20? -32

30? -27

40? -32

50? -30

60? -30

?

I'm guessing my 20db attenuator is toast but I would have expected to have seen the issue coming back at the higher attenuation levels. Could this behavior be caused by something else??

?


 

¿ªÔÆÌåÓý

Hi,

I did the same tests as you, results in the tables.

?

I attached a picture showing the diodes of the first mixer, I did the repair of two Spectrum this year with damaged diodes, so the one with which I did the attenuator tests today.

Maybe you should do this check. Ideally, if you had another measuring instrument that would allow you to know the RF level at the J3 connector of the first converter, you could know if the first converter is OK.

?

correction at OFF

SPAN 100kHz

resBW 10kHz

vidBW 100Hz

MKR peak

REFlevel at -10dBm

?

300MHz

-20dBm

ATT

Level measured

0

-19.4

10

-19.3

20

-19.7

30

-19.7

40

-19.8

50

-19.9

60

-20.0

1500MHz

-20dBm

ATT

Level measured

0

-21.2

10

-21.2

20

-21.5

30

-21.5

40

-21.8

50

-21.7

60

-21.8

?

Yves

?

De?: [email protected] <[email protected]> De la part de jansen.marco90@...
·¡²Ô±¹´Ç²â¨¦?: 3 novembre 2024 14:09
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

Continuing on the cal signal I checked the attenuator steps.

After factor default and with correction off I found the following:

At 300MHz:

Att? |? Level

10? -22

20? -12

30? -21

40? -26

50? -19

60? -21

?

at 1500 MHz

Att? |? Level

10? -28

20? -32

30? -27

40? -32

50? -30

60? -30

?

I'm guessing my 20db attenuator is toast but I would have expected to have seen the issue coming back at the higher attenuation levels. Could this behavior be caused by something else??

?


 

Hey Yves,
?
Thanks for your reply! Those measurements look a lot more like what I expected. Took me a few days but got my hands on a nanoVNA and hooked it up to only the attenuator today. The noisefloor of the VNA at high frequencies is too high to really say something about the attenuator but at 100 and 400MHz the measurements are usable. It looks like the 10db is tilted and the 20dB is attenuating less than the 10dB. The 30dB looks ok.?
i'd say this attenuator is toast, might still be more issues along the path though.?
? Levels ? ?
? 100MHz 300MHz 600MHz
cable through 1.5dB 0.0dB -0.1dB
analyzer off -66dB -53db -64dB
10db att -21dB -15dB -12dB
20db att -15dB -15dB -17dB
30db att -34db -30 dB ?-30dB
40db att -51dB -43dB ?-44dB
50db att -45dB -44dB -49dB
60db att ?-66dB -53dB -66dB
Wanted to backup the frequency response correction constants of the unit today but found out they are all gone already. So even if the attenuator and other issues could be fixed it would still need some form of calibration. Starting to feel more and more like flogging a dead horse.
?
Marco


 

I went ahead and took the attenuator out. It's a JFW 50p-1120 and it was definitely abused by a previous owner. Googled the part number hoping to find other teardown pictures to find out replacement values but no luck so far


 

¿ªÔÆÌåÓý

Hi,

?

If you have lost your calibration constants, it will be possible to rebuild them.

For the moment there is the attenuator to repair... it is easily achievable according to the image.

?

But I fear for your mixer, there is a good chance that it is damaged.

But that too is repairable, it is only two SOT-23 components to replace (diodes).

?

Attenuator values-based of attenuation.

?

?

?

Attenuator from RF input to first converter input. At left, absolute loss vs frequency, at right, VNA normalized to 0dB attenuator trace.

?

?

Yves

?

De?: [email protected] <[email protected]> De la part de jansen.marco90@...
·¡²Ô±¹´Ç²â¨¦?: 6 novembre 2024 14:35
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

Hey Yves,

?

Thanks for your reply! Those measurements look a lot more like what I expected. Took me a few days but got my hands on a nanoVNA and hooked it up to only the attenuator today. The noisefloor of the VNA at high frequencies is too high to really say something about the attenuator but at 100 and 400MHz the measurements are usable. It looks like the 10db is tilted and the 20dB is attenuating less than the 10dB. The 30dB looks ok.?

i'd say this attenuator is toast, might still be more issues along the path though.?

?

Levels

?

?

?

100MHz

300MHz

600MHz

cable through

1.5dB

0.0dB

-0.1dB

analyzer off

-66dB

-53db

-64dB

10db att

-21dB

-15dB

-12dB

20db att

-15dB

-15dB

-17dB

30db att

-34db

-30 dB

-30dB

40db att

-51dB

-43dB

-44dB

50db att

-45dB

-44dB

-49dB

60db att

-66dB

-53dB

-66dB

Wanted to backup the frequency response correction constants of the unit today but found out they are all gone already. So even if the attenuator and other issues could be fixed it would still need some form of calibration. Starting to feel more and more like flogging a dead horse.

?

Marco


 

That image was very helpful! I ordered replacement components according to your values. Both for the through hole and the smd resistors. I went with thin film ones because couldn't get any specific RF resistors in the right values.
They should be in early next week.??
?
I also got a set of BNC attenuators (Hameg HZ24 rated to 1GHz), so I hooked up the cal signal straight into the mixer, with the external attenuators.?
?
Measured the 300MHz with the same settings you did before:

correction at OFF

SPAN 100kHz

resBW 10kHz

vidBW 100Hz

MKR peak

REFlevel at -10dBm

?

found the following:

Attenuators?? 300MHz 1500MHz
10dB -14,4 -30,9
20dB -25,3 -41,3
10+20dB -36,8 -55
So the 300MHz looks quite okay I'd say and the 1500MHz is not too bad considering the 1GHz rated attenuators.
?
Though at 1500MHz I had to use a bigger span because I couldn't get the peak centered in small span. Peak>CF would not work and manually scrolling, the frequency would change but the peak would stay in it's place. This was without proper warmup of the SA btw.?
?
Would that mean that the mixer diodes have to be replaced asweel or would other measurements be needed to answer that??
?
Marco


 

¿ªÔÆÌåÓý

Hi Marco,

?

I don't know what value your Spectrum's attenuator was set to.

Even if it wasn't connected, the IF gain will change according to that value.

So, I did a test at 0dB and 10dB, even if the Spectrum's internal attenuator wasn't connected.

I inserted a precise level at -20dBm at 300 or 1500 MHz.

?

Now looking at the results, it may seem surprising that the values ??at 300 and 1500 MHz are very similar.

In fact, this is normal in my case since, in the calibration table (Flatness) there is a gap of only 1dB between the level at 300 MHz and 1500 MHz, at 1500MHz the signal is attenuated by 1dB, and this 1dB is mainly due to the cables connecting the Spectrum input to the attenuator and from the attenuator to the first converter.

Also, the attenuator does not have a flat frequency response.

Since I connected directly to the first converter, these losses are not included in the test measurement.

?

300 MHz

Spectrum ATT

EXT ATT

0 dB

10 dB

0 dB att

-19.4

-9.4

10db att

-29.4

-19.4

20db att

-39.4

-29.4

30db att

-49.5

-39.4

1500 MHz

Spectrum ATT

EXT ATT

0 dB

10 dB

0 dB att

-19.5

-9.4

10db att

-29.7

-19.5

20db att

-39.8

-29.7

30db att

-49.9

-39.8

?

?

De?: [email protected] <[email protected]> De la part de jansen.marco90@...
·¡²Ô±¹´Ç²â¨¦?: 8 novembre 2024 07:53
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

That image was very helpful! I ordered replacement components according to your values. Both for the through hole and the smd resistors. I went with thin film ones because couldn't get any specific RF resistors in the right values.

They should be in early next week.??

?

I also got a set of BNC attenuators (Hameg HZ24 rated to 1GHz), so I hooked up the cal signal straight into the mixer, with the external attenuators.?

?

Measured the 300MHz with the same settings you did before:

correction at OFF

SPAN 100kHz

resBW 10kHz

vidBW 100Hz

MKR peak

REFlevel at -10dBm

?

found the following:

Attenuators??

300MHz

1500MHz

10dB

-14,4

-30,9

20dB

-25,3

-41,3

10+20dB

-36,8

-55

So the 300MHz looks quite okay I'd say and the 1500MHz is not too bad considering the 1GHz rated attenuators.

?

Though at 1500MHz I had to use a bigger span because I couldn't get the peak centered in small span. Peak>CF would not work and manually scrolling, the frequency would change but the peak would stay in it's place. This was without proper warmup of the SA btw.?

?

Would that mean that the mixer diodes have to be replaced asweel or would other measurements be needed to answer that??

?

Marco


 

¿ªÔÆÌåÓý

Marco, I forgot to include these remarks

?

Here are two signs of a damaged mixer that I have already noticed (blue curves):

- At 0 Hz, the carrier exceeds the reference level

- PRESET: ATTENUATOR = 0 dB (whether the attenuator is installed or not does not change anything). Between 0Hz and 10 MHz, the noise increases in the lower frequencies instead of remaining relatively flat.

?

Yves

?


 

Hi Yves,
?
I did the measurements all with the SA's attenuator set to 10dBm. I redid the measurements with both SA attenuator settings, see table below.
I'm still using the cal out of the SA because I don't have a more trustworthy source yet.
?
300 MHz Spectrum ATT ?
EXT ATT 0 dB 10 dB ?
0 dB att -15 -5 ref level 0dBm
10db att -25,9 -15,7 ?
20db att -37,5 -26,3 ?
30db att -49,5 -37,9 ?
? ? ? ?
1500 MHz Spectrum ATT ?
EXT ATT 0 dB 10 dB ?
0 dB att -30,8 -19,8 ?
10db att -45,7 -34,2 ?
20db att -56,5 -44,9 ?
30db att -68,4 -57 ?
?
On the other images, I tried to reproduce them, I'm just taking pictures of the screen though. Are you using a GPIB interface to make those? Or can it also be done with the serial output?
The noise increase at lower frequencies doesn't seem to be there. There is an offset in the frequency though, the 0Hz peak seems to be around 580kHz
?
in a broader span with the cal signal connected (through 10dB external attenuator) The 0Hz peak is near the reference level, the 300MHz signal is also quite high.?
?
So i'll try to fix the attenuator and move on with another try of the selfcall from there. I'm thinking of getting a TinySA ultra to be able to do some verification.?
?
Marco


 

¿ªÔÆÌåÓý

Hi Marco,

?

Here are your results and mine tabulated.

That the source signal is not exactly -20.0dBm, will not change the conclusion.

Just like you, I used my 300 MHz and harmonics calibrator as a signal source.

I had done the tests for the Spectrum attenuator at 0dB or 10dB, not knowing what this condition was in your previous email.

Here we have the 2 conditions, so I used them.

?

Using the fact that there is a test where there is no external attenuator, allows us to evaluate the external attenuators of 10dB and 20dB as well as their sum.

We were able to assume that the values ??of the external attenuators (10dB and 20dB) are quite close to their nominal values.

This allows us to see the linearity over a 30dB range.

Calculating the difference between, without attenuator and 10dB, I would expect -10dB difference, if the attenuator was precisely 10dB.

And whether the Spectrum is with its attenuator configuration of 0dB or 10dB, should not change this result, even for jumps of 20dB, or 30dB.

?

So, columns E and F should show values ??of -10, -20 and -30 if the external attenuators were perfect and the mixer was perfectly linear.

Column G should show the value of 0dB everywhere.

?

If you look at the results of my Spectrum 8591A, the results follow the logic of results -10, -20 and -30, then 0 in column G

?

But your Spectrum 8591E has non-negligible deviations, and considering the images, I would say that your first converter is damaged¡­ but it is easily repairable.

I will send you the details.

?

-----------

?

For screenshots, it's a software I made for a GPIB interface, but I can certainly use the serial link if that can help you.

?

Having lost your calibration constants, it is probably normal to have this offset.

But it would be preferable for now to do the attenuator repair before doing a CAL FREQ.

?

If you manage to get a TinySA ultra, you will be able to directly measure the output of the first converter at 2.1214GHz

?

Yves

?

De?: [email protected] <[email protected]> De la part de jansen.marco90 via groups.io
·¡²Ô±¹´Ç²â¨¦?: 9 novembre 2024 08:57
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

Hi Yves,

?

I did the measurements all with the SA's attenuator set to 10dBm. I redid the measurements with both SA attenuator settings, see table below.

I'm still using the cal out of the SA because I don't have a more trustworthy source yet.

?

300 MHz

Spectrum ATT

?

EXT ATT

0 dB

10 dB

?

0 dB att

-15

-5

ref level 0dBm

10db att

-25,9

-15,7

?

20db att

-37,5

-26,3

?

30db att

-49,5

-37,9

?

?

?

?

?

1500 MHz

Spectrum ATT

?

EXT ATT

0 dB

10 dB

?

0 dB att

-30,8

-19,8

?

10db att

-45,7

-34,2

?

20db att

-56,5

-44,9

?

30db att

-68,4

-57

?

?

On the other images, I tried to reproduce them, I'm just taking pictures of the screen though. Are you using a GPIB interface to make those? Or can it also be done with the serial output?

The noise increase at lower frequencies doesn't seem to be there. There is an offset in the frequency though, the 0Hz peak seems to be around 580kHz

?

?

in a broader span with the cal signal connected (through 10dB external attenuator) The 0Hz peak is near the reference level, the 300MHz signal is also quite high.?

?

?

So, I¡¯ll try to fix the attenuator and move on with another try of the self-call from there. I'm thinking of getting a TinySA ultra to be able to do some verification.?

?

Marco


 

Hi Yves,
?
That's a clear overview, I agree it's better to repair the first converter.
I googled for the diodes for a bit, found the HSMS-2812 to be obsolete. Did find the Macom MA4E1338B1 series to be pretty close on RF specs but it has a breakdown voltage of only 8v compared to the 15v of the HSMS-2812. Not too sure wether that would be a problem though. They aren't sold in matched pairs so i'd have to try to match them myself I guess.
So would be good to know where you got the replacement diodes and which you got.?
?
My SA doesn't have the GPIB port but it does have the serial and parrallel port, I found some other projects online of people making printer/plotter emulators to be able to get screencaptures but none look as good as yours do.
?
I should be able to start fixing the attenuator next week.
?
Marco
?
?


 

¿ªÔÆÌåÓý

Hi Marco,

?

The diode Macom MA4E1338B1 would be OK, I had also found this equivalent.

There was also TPM2840-03, a little bit more performing than the original diode, but really very close...only 5V protection however.

?

I found a picture showing the section about the first converter of the 8591E which is different from the 8591A.

?

The method I used:

  • I replaced cr1 and cr2
  • I close the lid and I test, I see if the signal to the monitor at 0Hz, is lower than ideally, at the first division (upper left, picture 8591 LO leakage).
  • If the diodes are not matched well enough, the carrier at 0Hz will exceed the first division at the top of the left screen.
  • In such a case, change CR2 only, because easier to replace and redo the test.
  • It may work from the first attempt as in my case, but I had still tried a second attempt out of curiosity, and got slightly better, so I did not continue with further attempts.
  • I successfully closed, tested and calibrated everything because there was no other problem.

?

I will do some tests with the screen capture program controlling the rs-232 port.

?

Yves

?

?

De?: [email protected] <[email protected]> De la part de jansen.marco90 via groups.io
·¡²Ô±¹´Ç²â¨¦?: 10 novembre 2024 04:40
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

Hi Yves,

?

That's a clear overview, I agree it's better to repair the first converter.

I googled for the diodes for a bit, found the HSMS-2812 to be obsolete. Did find the Macom MA4E1338B1 series to be pretty close on RF specs but it has a breakdown voltage of only 8v compared to the 15v of the HSMS-2812. Not too sure wether that would be a problem though. They aren't sold in matched pairs so i'd have to try to match them myself I guess.

So would be good to know where you got the replacement diodes and which you got.?

?

My SA doesn't have the GPIB port but it does have the serial and parrallel port, I found some other projects online of people making printer/plotter emulators to be able to get screencaptures but none look as good as yours do.

?

I should be able to start fixing the attenuator next week.

?

Marco

?

?


 

Hi Yves,
(and the rest of the group ;) )
?
Yesterday I soldered new resistors in the attenuator, it needed some tweaking to get the response flat but I finished it today.
I checked all the settings with a LiteVNA. The 10dB isn't perfect but close enough for now.?
Att Setting 300MHz 600MHz 900MHz 1,2GHz 1,5GHz 1,8GHz
10dB -10,5 -10,8 -11 -11 -10,7 -10
20dB -21 -21,2 -21,2 -21,3 -21,5 -21,6
30dB -30,8 -30,7 -30,4 -30,2 -30 -30
10+20 -30,6 -30,9 -31 -31 -30,6 -29,6
10+30 -40,5 -40,4 -40,2 -40 -39,2 -38
20+30 -50,9 -50,6 -50,4 -50 -49,8 -49,7
10+20+30 -60,7 -61,2 -61,5 -62 -62,1 -61,2
?
I build the attenuator back into the analyzer and checked the function.
The resulting levels are looking better, not perfect though. I'm expecting this will be fixed with new mixer diodes which should be in soon.
Before Fix ? ? Now ?
?300MHz:
?
?
?300MHz:
?
Att
Level
?
Att
Level
10
-22
?
10
-15,7
20
-12
?
20
-16,3
30
-21
?
30
-17,3
40
-26
?
40
-18,7
50
-19
?
50
-20,9
60
-21
?
60
-21,8
?
I did a full self call which successfully passed and a confidence test which also passed.
The resulting Cal signal on screen though is still up at -5dBm instead of the expected -20dBm (I double checked the Cal output of the SA with another SA and it is around -20dBm. )
I'm starting to think there must be a problem with the calibration attenuator in A12, causing this level mismatch. Any experience with that? Or thoughts on how to verify it? Otherwise I guess i'm just going to take the A12 module out to have a look.
?
Marco


 

¿ªÔÆÌåÓý

Hi Marco,

?

The A12 module acts on a range determined by the Reference Level, between 0dBm and -50dBm.

With the following configuration, I was able to verify that all the attenuation jumps are functional.

?

I connect the reference to the Spectrum input, but put an external 30dB attenuator there.

Thus, the source at 300 MHz which normally is -20dBm, will have been read as -50dBm.

?

PRESET

CENTER FREQ 300 MHz

SPAN 0

REF LEVEL 0DBM.

?

By changing the reference level from 0dBm to -50dBm, by 10dB jump with the UP/DOWN arrows, you will see if all the attenuators are functional.

And at your choice for a Reference Level between 0dBm and -40dBm, make 1dB jumps to check the calibration attenuator.

?

But having as RF level read at 300 MHz of -5dBm instead of -20dBm, add another 15dB of external attenuation and put the resBW at 100kHz.

The goal is to check the 50dB range of the A12 module. If A12 has 50dB of range then, the problem is maybe on A16.

?

But, a measurement on AUX IF could be revealing.

Without any external attenuator, enter the 300 MHz reference to the Spectrum and put the reference level at 0dBm.

On my side I read, -38.2dBm at 21.4MHz.

The AUX IF OUTPUT on the rear panel is connected to A13, the module that follows A12.

?

Yves

?

?

?

De?: [email protected] <[email protected]> De la part de jansen.marco90 via groups.io
·¡²Ô±¹´Ç²â¨¦?: 13 novembre 2024 15:50
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

Hi Yves,

(and the rest of the group ;) )

?

Yesterday I soldered new resistors in the attenuator, it needed some tweaking to get the response flat but I finished it today.

I checked all the settings with a LiteVNA. The 10dB isn't perfect but close enough for now.?

Att Setting

300MHz

600MHz

900MHz

1,2GHz

1,5GHz

1,8GHz

10dB

-10,5

-10,8

-11

-11

-10,7

-10

20dB

-21

-21,2

-21,2

-21,3

-21,5

-21,6

30dB

-30,8

-30,7

-30,4

-30,2

-30

-30

10+20

-30,6

-30,9

-31

-31

-30,6

-29,6

10+30

-40,5

-40,4

-40,2

-40

-39,2

-38

20+30

-50,9

-50,6

-50,4

-50

-49,8

-49,7

10+20+30

-60,7

-61,2

-61,5

-62

-62,1

-61,2

?

I build the attenuator back into the analyzer and checked the function.

The resulting levels are looking better, not perfect though. I'm expecting this will be fixed with new mixer diodes which should be in soon.

Before Fix

?

?

Now

?

300MHz:

?

?300MHz:

?

Att

Level

?

Att

Level

10

-22

?

10

-15,7

20

-12

?

20

-16,3

30

-21

?

30

-17,3

40

-26

?

40

-18,7

50

-19

?

50

-20,9

60

-21

?

60

-21,8

?

I did a full self call which successfully passed and a confidence test which also passed.

The resulting Cal signal on screen though is still up at -5dBm instead of the expected -20dBm (I double checked the Cal output of the SA with another SA and it is around -20dBm. )

I'm starting to think there must be a problem with the calibration attenuator in A12, causing this level mismatch. Any experience with that? Or thoughts on how to verify it? Otherwise I guess i'm just going to take the A12 module out to have a look.

?

Marco


 

¿ªÔÆÌåÓý

Hi Marco,

?

Just to be sure, is the A14W1 jumper there or not.

Because, if there is a jumper installed at that position, it must be removed according to the documentation on that option 130.

?

Yves

?

?

De?: [email protected] <[email protected]> De la part de jansen.marco90@...
·¡²Ô±¹´Ç²â¨¦?: 1 novembre 2024 16:38
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Troubleshooting a 8591e with opt 130

?

I continued my testing and debugging with some success.

The opt 130 is showing signal now, there are supposed to be 2 jumpers on that board and only 1 was present, the other was probably used for A14W1 by the previous owner, when he removed the option. With that jumper in place i'm no longer just looking at the nosiefloor.

With the option installed the cal fails on NBW 200Hz Notch failed.?

Doubt that i'lll be able to fix that without schematics or adjustment options on the opt 130.

?

Interestingly the unit does show a -22dB cal signal with the factory default cal at 350MHz in full span and 300MHz at 100MHz span.?

During ampt cal the signal is way above the screen during 3db BW steps, i'm guessing something is going wrong there.

After running freq and amp cal the 300MHz is back at -5dB instead of -20dB, even though the cal failed and correction off doesn't change it either.?

?

Starting to wonder how feasible this repair will be.

?


 

Hey Yves,?
?
Thanks once more for all your fast and helpful replies! Many (paid) customersupport isn't as helpfull as you are.
?
There is no jumper on a14W1, option 130 is installed and the jumper was missing on the option module, to get it working I took the jumper from A14w1 and installed it on the opt 130 card.?
?
I did quite some measurements with the instrument both in factory default cal and after it had done it's selfcal.?
Factory default cal:
I used the TinySa generator function to generate 300MHz @-71dBm
8591E set to 300MHz, 0Hz Span, Ref level 0dBm
Levels measured:
Ref level Mrkr
-50 -65
-40 -66,5
-30 -68,2
-20 -69,6
-10 -70,7
0 -71,5
?
There seems to be more than 50dB of range but I might misinterpret.?
I also did the 1dB steps, seemed to show 10 steps with the measured level changing about 1dB.?
?
Next I connected the 8591E Cal signal directly to the input, span 0 Hz, freq 300MHz, measured level -22.5dBm. Connected TinySA to the Aux IF output which measured 21.4MHz, -49.8dBm. Quite a difference from your -38dBm.
I also connected the TinySA to the 2nd converter LO Out, measured 1.799GHz, -17.3dBm
And the 600MHz out: 600MHz, 0dBm.
?
After selfcal:
Then I did a full Selfcal, did again notice strange behaviour during 3dB BW Cal, checked a vid online of a full selfcal, there the levels seem to be properly on the screen, during this 8591E selfcal it looks like this:
?
After selfcal I redid the previous measurements
?
Ref level mrkr
-50 -55
-40 -56
-30 -57
-20 -60
-10 -70
0 -70
?
There is a jump with the -20dB level as can be seen in the pictures too.
I also checked the IF out after cal
8591E Cal signal directly to the input, span 0 Hz, freq 300MHz, measured level -5dBm. Connected TinySA to the Aux IF output which measured 21.4MHz, -39.2dBm. A lot closer to your -38dBm. Is the too low IF output causing an overcompensation during calibration?
?
Btw, i bought a nullmodem cable and a usb to serial adapter. I can now receive the plotter/printer output from the 8591E in serial terminal on my pc. I'll see if I can make a plot out of it with some of the online hpgl plot software.
?


 

¿ªÔÆÌåÓý

Hi Marco, I'll look into a way to test the A12 module, but here are a few notes...

?

The test you did injecting the 300 MHz reference directly to the Spectrum, the measurement on the screen was -5dBm, and at the AUX IF output being -39.2dBm means that the overall gain of the Spectrum input was normal, if the REF LEVEL was 0dBm.

This would mean that the compensation was added after the A12 stage. But it is probably during the SELFCAL that the constants received incorrect values.

?

To try to find a bad stage for the gain, it will probably be necessary to work in CAL CORRECT mode at OFF

?

It would be a good idea to measure the LOG detector by doing the following test (CAL CORRECT at OFF):

- Ref Level at 0dBm, 300 MHz, zero span, resBW 1kHz, sweep time 1sec

- Put a signal of about -70dBm, and adjusted so that the trace arrives on the -70dBm line.

- Raise the level in 10dB steps, from -70 to -10. The trace should move in 10dB steps as in the picture.

?

Filters check (resBW):

- CAL CORRECT to OFF

- REF 300 MHz connected to input 8591E

- Verify each resBW include option 130 (resBW 300Hz, 200Hz, 100Hz and 30 Hz).

- Here I have superimposed the images to show you that there will be a little difference in levels and frequency, which is corrected during SELFCAL, but no big gaps between the filters.

?

?

?

?

?

-----------------------------------------------

?

?

Here are the three images in comparison: My 8591A, your 8591E after SELFCAL and your 8591EE with default CAL.

- I also injected -70dBm, starting from this level it will be easier to evaluate the value of the excess gain.

- For my 8591A, when I change the REF LEVEL by -10dB, the trace goes up by +10dB, and this, for the 50dB of range involving the A12 module.

- For your 8591E, some jumps have a little more than 10dB but mainly the last one, for a total of +5dB with default CAL and nearly +15dB after SELFCAL.

?

?

Send me a text file captured by the rs-232 serial link with your hpgl data, I will look at the behavior with my program.

?

?