¿ªÔÆÌåÓý

Re: Troubleshooting a 8591e with opt 130


 

Hey Yves,?
?
Thanks once more for all your fast and helpful replies! Many (paid) customersupport isn't as helpfull as you are.
?
There is no jumper on a14W1, option 130 is installed and the jumper was missing on the option module, to get it working I took the jumper from A14w1 and installed it on the opt 130 card.?
?
I did quite some measurements with the instrument both in factory default cal and after it had done it's selfcal.?
Factory default cal:
I used the TinySa generator function to generate 300MHz @-71dBm
8591E set to 300MHz, 0Hz Span, Ref level 0dBm
Levels measured:
Ref level Mrkr
-50 -65
-40 -66,5
-30 -68,2
-20 -69,6
-10 -70,7
0 -71,5
?
There seems to be more than 50dB of range but I might misinterpret.?
I also did the 1dB steps, seemed to show 10 steps with the measured level changing about 1dB.?
?
Next I connected the 8591E Cal signal directly to the input, span 0 Hz, freq 300MHz, measured level -22.5dBm. Connected TinySA to the Aux IF output which measured 21.4MHz, -49.8dBm. Quite a difference from your -38dBm.
I also connected the TinySA to the 2nd converter LO Out, measured 1.799GHz, -17.3dBm
And the 600MHz out: 600MHz, 0dBm.
?
After selfcal:
Then I did a full Selfcal, did again notice strange behaviour during 3dB BW Cal, checked a vid online of a full selfcal, there the levels seem to be properly on the screen, during this 8591E selfcal it looks like this:
?
After selfcal I redid the previous measurements
?
Ref level mrkr
-50 -55
-40 -56
-30 -57
-20 -60
-10 -70
0 -70
?
There is a jump with the -20dB level as can be seen in the pictures too.
I also checked the IF out after cal
8591E Cal signal directly to the input, span 0 Hz, freq 300MHz, measured level -5dBm. Connected TinySA to the Aux IF output which measured 21.4MHz, -39.2dBm. A lot closer to your -38dBm. Is the too low IF output causing an overcompensation during calibration?
?
Btw, i bought a nullmodem cable and a usb to serial adapter. I can now receive the plotter/printer output from the 8591E in serial terminal on my pc. I'll see if I can make a plot out of it with some of the online hpgl plot software.
?