¿ªÔÆÌåÓý

Date
Re: HP 3456A Repair help needed
As an observation: Any time the presence or absence of a measurement probe (scope, meter, etc) makes a difference, you're dealing with a marginal situation. Your probe represents at least a
By Harvey White · #126531 ·
Re: HP 3456A Repair help needed
-------- Hi John, Great news and congratulations for the repair. You already debugged it 99% of the way by identifying 1M-ohm from gate to ground makes a difference. That 100k resistor is not carrying
By Ozan · #126530 ·
Re: HP 3456A Repair help needed
Ozan, THANK YOU! The 100k resistor had failed open. I can¡¯t believe it. I'm slightly embarrassed, because I was sure I had already checked these. It is always a great idea to have a second set of
By John Quinn · #126529 ·
Re: HP 8640B stiff main tuning
Roy is right. In many greases, one of the major heavy components is soap. Look at what happens when soap residue hardens. But, be careful; not all soaps dissolve in water. Meths can sometimes help. 73
By Brian · #126528 ·
Re: H/P 8753B disassembly
Thankyou Peter - just what the Doctor ordered............................Don C.
By donald collie · #126527 ·
Re: HP 3456A Repair help needed
----- Are resistors R122 and R143 approximately correct and can you trace connectivity from TP104 to gate of Q103? The resistors don't need to be accurate but voltage at gate of Q103 should follow
By Ozan · #126526 ·
HP 3456A Repair help needed
Hello, I'm looking for help repairing my beloved HP3456A DMM. It is a later made model (serial #2201A13361) with most of the IC's date codes from 1984. When I received the unit several years ago, I
By John Quinn · #126525 ·
Re: HP 8640B stiff main tuning
The issue here is the same thing that causes fans to not work well when they age. The original lubricant is made up of a number of different fractions, and over long periods of time the lighter ones
By Roy J. Tellason, Sr. · #126524 ·
Re: H/P 8753B disassembly
Try the manual 08753-90156 [image: image.png] [image: image.png] etc, there are pictures of the assemblies. Be very careful in the HV area as static can blow components (the DAC I believe). Make
By peter bunge · #126523 ·
H/P 8753B disassembly
In damp weather my 8753B EHT plays up causing the display to disappear, and reappear rapidly, as though something is tracking. I have decided to clean the soot off all the EHT components with
By donald collie · #126522 ·
Re: 4395A software
Rico, Excellent work. Wondering if you found out how the calibration constants are stored?
By ROMDUMP · #126521 ·
Re: 4395A software
Hi Sergey, I found a way to upgrade the options to 1D6, 010, 001 This is the reference: How to upgrade: 1. Change this line to this values: 2. Calculate the Checksum over this area: 3. Add +1 =
By Rico Sonderegger · #126520 ·
Re: HP 8640B stiff main tuning
I have just put 8640B Service Manual Volumes 1 and 2 inc CCT Pn 08640-90215 May 1991 .pdf Onto this site in the files also other info including 8640 note for sticking tuning Paul
By Paul Bicknell · #126519 ·
Re: HP 8640B stiff main tuning
Avoid dissassembly. Get service manual. Try some gentle solvent then lube on the shaft , perhaps Iso Alcohol? Jon
By Froggie the Gremlin · #126518 ·
Re: HP 8640B stiff main tuning
No, its been sitting on the bench in operating position. Electrically, the oscillator works fine with no problems, so I don't think its a problem with the internal grease migration. Its just that the
By mondial · #126517 ·
Re: HP 8640B stiff main tuning
You didn't store it on its back, did you? If so, that could be the problem. You might try storing it face-down for a few days/weeks before taking the cavity apart. There's some discussion of
By John Miles · #126516 ·
Re: Shipping / Moving HP Equipment HP16702B - and other 19" rack equipment like a 3325B
Bingo............. H.
By Harvey White · #126515 ·
Re: Shipping / Moving HP Equipment HP16702B - and other 19" rack equipment like a 3325B
Avoid moving into hi-rise buildings in Florida... especially those close to salty waters! cheers Martn
By Martin · #126514 ·
Re: Thoughts on unusual snubber failure mode AC Source HP6813B ?
That's perfect, thanks a heap Askild :-o) I aided keysight first head office in USA for that but, instead the 'applications' engineer reminded me it was out of service and as it was upgraded to 6813B
By nichevlct@... · #126513 ·
Re: Shipping / Moving HP Equipment HP16702B - and other 19" rack equipment like a 3325B
Understand where you want to live and what's there for you. Not everyone would wish to live where you do, and how you do. Dave's opinion is 100% valid for Dave, Mine is 100% valid for me. That's