Re: 8340A Fault - ADC
The ADC is in a wide DIP (don't know the dimensions from heart, but it has the same width as an ordinary EPROM).
By
Tobias Pluess
·
#86403
·
|
Re: 8340A Fault - ADC
I also googled but a minute or two.
What is the housing of that ADC IS? Worst case, one can IMHO grab an SMD part and make a fourlayer adapter at the usual PCB making suspects
By
Tam Hanna
·
#86402
·
|
Re: 8340A Fault - ADC
Hi Calvin
I checked the A27 parts list. Following chips seem to be used:
NE5018F DAC - can be found on eBay
1826-0881 ADC - this is actually an ADC856C. I am unsure whether this can be replaced
By
Tobias Pluess
·
#86401
·
|
Re: 8340A Fault - ADC
Hi Calvin
I have the complete 8340A service manuals in paper on my desk. I try to find someone who can scan these plus size pages. My scanner at home or in the office can't do that.
Unfortunately,
By
Tobias Pluess
·
#86400
·
|
Re: 8340A Fault - ADC
Hi Lex,
Yes, it’s ADC fault. The instrument has Fault annunciator lit. I asked the guy to press [shift][manual], and the ADC lights are blinking the the Entry Display. The other faults that would
By
Calvin Guan
·
#86399
·
|
Re: 8340A Fault - ADC
Hello,
It depends on who you're buying from. Personally I would go for it if I knew the basic diagnostic self tests were working, the seller isn't selling a physically damaged unit and if I had the
By
Lexter Negron
·
#86398
·
|
Re: 8340A Fault - ADC
Fault could be either the ADC chip or the signals controlling it. I
believe the 8340 permits direct control of the peripheral boards.
Diagnostic procedure would be to measure value at ADC input and
By
Bruce
·
#86397
·
|
8340A Fault - ADC
Hi there,
I am looking into buying a 8340A with known problem — Fault light on, fault code ADC. From the service manual, it says ADC fault indicates CPU detected ADC circuit problem on A27 level
By
Calvin Guan
·
#86396
·
|
Re: HP 5335A counter without fancy options, has HPIB -- fail 2.1 U25 RAM
John
I could use a parts unit.
What sort of compensation do you need?
Dennis
--------------------------------------------
By
czelusniakd
·
#86395
·
|
HP 5335A counter without fancy options, has HPIB -- fail 2.1 U25 RAM
Anyone want it? Parts of it? No oven crystal in there, no 1.2GHz input C.
By
John Griessen
·
#86394
·
|
Re: OT Omni Spectra Semi-Rigid Tools
Not mentioned in this catalog, but there are contentlists for the individual parts.
JimMc
By
J Mcvein
·
#86393
·
|
Re: OT Omni Spectra Semi-Rigid Tools
Jim,
Any mention of a T-200/500? That’s what I have here.
Thanks,
Steve
WB0DBS
By
Steve - Home
·
#86392
·
|
Re: OT Omni Spectra Semi-Rigid Tools
I have a 1991 OS catalog here, and there is a toolingsection.? T-250 is listed as SSMA-(OSSM), T-450A is TNC.T-400 is for Type N.
I have never been fortunate enough to find a T-200 in goodshape in
By
J Mcvein
·
#86391
·
|
1740A Horizontal setup
Hi everyone,
I'm restoring a 1740A back to health again and its up and running but some form of calibration is now required.
I have a Heathkit IG 4505 calibrator and the volts / div is pretty
By
graham.l.n.butcher@...
·
#86390
·
|
Re: HP 8341A sucessfully FIXED!
Hi Tony
wow, cool, thanks. These files should work for the 8341A as well I think. I will give it a try ;-)
Since I cannot run HP Basic stuff, I will translate the programs to Python which runs nicely
By
Tobias Pluess
·
#86389
·
|
Re: HP 8341A sucessfully FIXED!
I have the one for the 8340B if that helps.
Original files here:
https://1drv.ms/f/s!Amqar8_XQ9UzkKVReIlocjFDGipQfw https://1drv.ms/f/s!Amqar8_XQ9UzkKVReIlocjFDGipQfw
Unlocked files
By
Tony Goodhew
·
#86388
·
|
Re: HP 8341A sucessfully FIXED!
Hi,
I saw that there was an automated test program which calibrates the attenuator and the ALC using HPIB commands.
However, I didn't find that program on the net.
Does anyone have a copy of
By
Tobias Pluess
·
#86387
·
|
Re: 5342A: blown sampler?
My counter counts reliably up to 20 GHz. I could not yet test higher frequencies.
I wonder whether somebody has a ROM dump of it?
It would be nice if I could add the DAC option, but this requires a
By
Tobias Pluess
·
#86386
·
|
Re: KNOBS HP1701 SCOPE
Hi? Keith;......
Thanks for the reply...
My scope the HP 1701, has three (3) knobs on the shafts
that control 'Main TIME/DEV, Delay TIME/Dev, and Sweep
Select'.? I beleive they were used on
By
Irvin Schwartz <sl45000@...>
·
#86385
·
|
Re: 8566B PL1 Unlock, PL2 Unlock and Ref Unlock
The failed cap has an ESR of 56.32k and slowly dropping. The capacitance is 48.78nF and rising. This was measured with an Agilent 1733C. I have an HP bench unit down in the lab so I'll walk down
By
Jerry Hancock
·
#86384
·
|